Nicholas Watson Nicholas Watson
Reply | Threaded
Open this post in threaded view
♦
♦
|

10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

To enable Zemax simulations for a customized two photon microscope (university research application), I am looking for the presciption for the Olympus 10x 0.6 NA objective, model: XLPLN10XSVMP 3mm WD. The presciption can usually be found in the patent submitted by Olympus Corporation. Has anyone been able to find this presciption/patent? Thank you.
Zdenek Svindrych-2 Zdenek Svindrych-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,
I have a similar request - a prescription of Zeiss LD Achroplan 20x 0.40
Korr. It's quite an old lens and I don't know where to start looking for
details.
The idea is to use it as a water-immersion lens for multiphoton
microscopy (so the chromatic aberration is not an issue). The
issue actually is the very long working distance (11 mm in air, some 7 mm
in water...) making in a bit impractical for some applications.
Thanks!
Best, zdenek

On Thu, Mar 11, 2021 at 9:21 AM Nicholas Watson <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> To enable Zemax simulations for a customized two photon microscope
> (university research application), I am looking for the presciption for the
> Olympus 10x 0.6 NA objective, model: XLPLN10XSVMP 3mm WD. The presciption
> can usually be found in the patent submitted by Olympus Corporation. Has
> anyone been able to find this presciption/patent? Thank you.
>


--
--
Zdenek Svindrych, Ph.D.
Research Scientist - Microscopy Imaging Specialist
Department of Biochemistry and Cell Biology
Geisel School of Medicine at Dartmouth
Lingbo Kong Lingbo Kong
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi All,

You can search this web
https://sites.google.com/site/danreiley/MicroscopeObjectives
to find the same or similar objective lens.

Good luck!

Bob

On Thu, Mar 11, 2021, 7:01 AM Zdenek Svindrych <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi All,
> I have a similar request - a prescription of Zeiss LD Achroplan 20x 0.40
> Korr. It's quite an old lens and I don't know where to start looking for
> details.
> The idea is to use it as a water-immersion lens for multiphoton
> microscopy (so the chromatic aberration is not an issue). The
> issue actually is the very long working distance (11 mm in air, some 7 mm
> in water...) making in a bit impractical for some applications.
> Thanks!
> Best, zdenek
>
> On Thu, Mar 11, 2021 at 9:21 AM Nicholas Watson <
> [hidden email]>
> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > To enable Zemax simulations for a customized two photon microscope
> > (university research application), I am looking for the presciption for
> the
> > Olympus 10x 0.6 NA objective, model: XLPLN10XSVMP 3mm WD. The presciption
> > can usually be found in the patent submitted by Olympus Corporation. Has
> > anyone been able to find this presciption/patent? Thank you.
> >
>
>
> --
> --
> Zdenek Svindrych, Ph.D.
> Research Scientist - Microscopy Imaging Specialist
> Department of Biochemistry and Cell Biology
> Geisel School of Medicine at Dartmouth
>
Benjamin Smith Benjamin Smith
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I've also used the models that Bob recommended, and they're pretty good.
That said, I've found that objectives are corrected enough that a simple
paraxial lens with a radius matching the backaperture of the objective is
sufficient in most cases, unless you are trying to do non-sequential
modelling.

One thing I would love, though, would be some models of immersion
objectives.

On Thu, Mar 11, 2021 at 10:28 AM Lingbo Kong <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi All,
>
> You can search this web
> https://sites.google.com/site/danreiley/MicroscopeObjectives
> to find the same or similar objective lens.
>
> Good luck!
>
> Bob
>
> On Thu, Mar 11, 2021, 7:01 AM Zdenek Svindrych <[hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi All,
> > I have a similar request - a prescription of Zeiss LD Achroplan 20x 0.40
> > Korr. It's quite an old lens and I don't know where to start looking for
> > details.
> > The idea is to use it as a water-immersion lens for multiphoton
> > microscopy (so the chromatic aberration is not an issue). The
> > issue actually is the very long working distance (11 mm in air, some 7 mm
> > in water...) making in a bit impractical for some applications.
> > Thanks!
> > Best, zdenek
> >
> > On Thu, Mar 11, 2021 at 9:21 AM Nicholas Watson <
> > [hidden email]>
> > wrote:
> >
> > > *****
> > > To join, leave or search the confocal microscopy listserv, go to:
> > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > > Post images on http://www.imgur.com and include the link in your
> > posting.
> > > *****
> > >
> > > To enable Zemax simulations for a customized two photon microscope
> > > (university research application), I am looking for the presciption for
> > the
> > > Olympus 10x 0.6 NA objective, model: XLPLN10XSVMP 3mm WD. The
> presciption
> > > can usually be found in the patent submitted by Olympus Corporation.
> Has
> > > anyone been able to find this presciption/patent? Thank you.
> > >
> >
> >
> > --
> > --
> > Zdenek Svindrych, Ph.D.
> > Research Scientist - Microscopy Imaging Specialist
> > Department of Biochemistry and Cell Biology
> > Geisel School of Medicine at Dartmouth
> >
>


--
Benjamin E. Smith, Ph. D.
Imaging Specialist, Vision Science
University of California, Berkeley
195 Weill Hall
Berkeley, CA  94720-3200
Tel  (510) 642-9712
Fax (510) 643-6791
e-mail: [hidden email]
https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/
Stan Schwartz-2 Stan Schwartz-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

In reply to this post by Nicholas Watson
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

For people looking for microscope objective “prescriptions” or design criteria these following links are an incredible resource.
https://www.degruyter.com/document/doi/10.1515/aot-2019-0002/html
https://www.degruyter.com/document/doi/10.1515/aot-2019-0013/html
https://www.degruyter.com/document/doi/10.1515/aot-2019-0014/html

Best Regards,
Stan
Martin Wessendorf-5 Martin Wessendorf-5
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 10x objective for two photon microscopy (XLPLN10XSVMP 3mm WD)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Wow.  Thank you, Stan!

Martin Wessendorf

On Sat, Mar 13, 2021 at 1:34 PM Stan Schwartz <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> For people looking for microscope objective “prescriptions” or design
> criteria these following links are an incredible resource.
> https://www.degruyter.com/document/doi/10.1515/aot-2019-0002/html
> https://www.degruyter.com/document/doi/10.1515/aot-2019-0013/html
> https://www.degruyter.com/document/doi/10.1515/aot-2019-0014/html
>
> Best Regards,
> Stan
>