20X objective for 2P work

classic Classic list List threaded Threaded
♦
♦
13 messages Options
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear listers,
We're currently on the search for a high NA water immersion 20x objective to
use for intravital 2P imaging.  As such, we need something that does it all,
short wavelength transmission for SHG and long wavelength transmission for
our Ti:Saph excitation source.  Any thoughts on what folks use/have
seen/would like to have/have heard about would be appreciated.

Thanks,
c

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
PattyJansma PattyJansma
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=646

I think it may only be sold with the Flo-view1000MPE but does what you
want.

Patty

On 10/18/2010 11:20 AM, Carl Boswell wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear listers,
> We're currently on the search for a high NA water immersion 20x
> objective to use for intravital 2P imaging.  As such, we need
> something that does it all, short wavelength transmission for SHG and
> long wavelength transmission for our Ti:Saph excitation source.  Any
> thoughts on what folks use/have seen/would like to have/have heard
> about would be appreciated.
>
> Thanks,
> c
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> University of Arizona
> 520-954-7053
> FAX 520-621-3709
Stephen Firth (Med) Stephen Firth (Med)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We use the Leica 20x 1.0NA dipping on our 2P (Leica CFS w. MaiTai), it's
worked well for all our deep/intravital imaging.
Cheers,
Stephen


>
> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear listers,
>> We're currently on the search for a high NA water immersion 20x objective
>> to use for intravital 2P imaging.  As such, we need something that does it
>> all, short wavelength transmission for SHG and long wavelength transmission
>> for our Ti:Saph excitation source.  Any thoughts on what folks use/have
>> seen/would like to have/have heard about would be appreciated.
>>
>> Thanks,
>> c
>>
>> Carl A. Boswell, Ph.D.
>> Molecular and Cellular Biology
>> University of Arizona
>> 520-954-7053
>> FAX 520-621-3709
>>
>


--
Stephen Firth
Manager of Advanced Optical Microscopy
Monash Micro Imaging
Monash University
Room G58 Building 75
Phone 990 55612
Mob 0400 695 425
Cameron Nowell Cameron Nowell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

In reply to this post by PattyJansma
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi List,

Unfortunately the Olympus 25x is only sold as part of a complete system
(we have one and it has been great). But the older 20x XLumPlanFL works
pretty well too for visible and IR wavelengths and you should be able to
buy one of those separately.


Cheers

Cam


Cameron J. Nowell
Microscopy Manager
Centre for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008
Royal Melbourne Hospital
Victoria, 3050
AUSTRALIA
Office: +61 3 9341 3155
Mobile: +61422882700
Fax: +61 3 9341 3104
Facility Website



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Patty Jansma
Sent: Tuesday, 19 October 2010 6:07 AM
To: [hidden email]
Subject: Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
46

I think it may only be sold with the Flo-view1000MPE but does what you
want.

Patty

On 10/18/2010 11:20 AM, Carl Boswell wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear listers,
> We're currently on the search for a high NA water immersion 20x
> objective to use for intravital 2P imaging.  As such, we need
> something that does it all, short wavelength transmission for SHG and
> long wavelength transmission for our Ti:Saph excitation source.  Any
> thoughts on what folks use/have seen/would like to have/have heard
> about would be appreciated.
>
> Thanks,
> c
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> University of Arizona
> 520-954-7053
> FAX 520-621-3709


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
Adrian Smith-6 Adrian Smith-6
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

Nikon also has a new 25x objective for multiphoton work - specifications look very similar to the Olympus  and it available without buying a complete system... however I've not had any feedback from actual users yet.

I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some stage but haven't had any success getting that organised yet.

We've had good results with the Olympus 20x NA 0.95 (standard and special IR versions).

Regards,

Adrian Smith
Centenary Institute, Sydney, Australia

On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi List,
>
> Unfortunately the Olympus 25x is only sold as part of a complete system
> (we have one and it has been great). But the older 20x XLumPlanFL works
> pretty well too for visible and IR wavelengths and you should be able to
> buy one of those separately.
>
>
> Cheers
>
> Cam
>
>
> Cameron J. Nowell
> Microscopy Manager
> Centre for Advanced Microscopy
> Ludwig Institute for Cancer Research
> PO Box 2008
> Royal Melbourne Hospital
> Victoria, 3050
> AUSTRALIA
> Office: +61 3 9341 3155
> Mobile: +61422882700
> Fax: +61 3 9341 3104
> Facility Website
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Patty Jansma
> Sent: Tuesday, 19 October 2010 6:07 AM
> To: [hidden email]
> Subject: Re: 20X objective for 2P work
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
> 46
>
> I think it may only be sold with the Flo-view1000MPE but does what you
> want.
>
> Patty
>
> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear listers,
>> We're currently on the search for a high NA water immersion 20x
>> objective to use for intravital 2P imaging.  As such, we need
>> something that does it all, short wavelength transmission for SHG and
>> long wavelength transmission for our Ti:Saph excitation source.  Any
>> thoughts on what folks use/have seen/would like to have/have heard
>> about would be appreciated.
>>
>> Thanks,
>> c
>>
>> Carl A. Boswell, Ph.D.
>> Molecular and Cellular Biology
>> University of Arizona
>> 520-954-7053
>> FAX 520-621-3709
>
>
> This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error.
> The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi,

We use the 20x and its really a workhorse!

I was told recently that the 20x from olympus has better response in red/infrared range than de 25x despite of the lower NA. Can anyone confirm this?

All the best,
Nuno Moreno, PhD
Head of Equipment Management
Instituto Gulbenkian de Ciência
Fundação Calouste Gulbenkian
Tel . +351 4464538
Fax. +351 4407970





On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>
> Nikon also has a new 25x objective for multiphoton work - specifications look very similar to the Olympus  and it available without buying a complete system... however I've not had any feedback from actual users yet.
>
> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some stage but haven't had any success getting that organised yet.
>
> We've had good results with the Olympus 20x NA 0.95 (standard and special IR versions).
>
> Regards,
>
> Adrian Smith
> Centenary Institute, Sydney, Australia
>
> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi List,
>>
>> Unfortunately the Olympus 25x is only sold as part of a complete system
>> (we have one and it has been great). But the older 20x XLumPlanFL works
>> pretty well too for visible and IR wavelengths and you should be able to
>> buy one of those separately.
>>
>>
>> Cheers
>>
>> Cam
>>
>>
>> Cameron J. Nowell
>> Microscopy Manager
>> Centre for Advanced Microscopy
>> Ludwig Institute for Cancer Research
>> PO Box 2008
>> Royal Melbourne Hospital
>> Victoria, 3050
>> AUSTRALIA
>> Office: +61 3 9341 3155
>> Mobile: +61422882700
>> Fax: +61 3 9341 3104
>> Facility Website
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]]
>> On Behalf Of Patty Jansma
>> Sent: Tuesday, 19 October 2010 6:07 AM
>> To: [hidden email]
>> Subject: Re: 20X objective for 2P work
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
>> 46
>>
>> I think it may only be sold with the Flo-view1000MPE but does what you
>> want.
>>
>> Patty
>>
>> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Dear listers,
>>> We're currently on the search for a high NA water immersion 20x
>>> objective to use for intravital 2P imaging.  As such, we need
>>> something that does it all, short wavelength transmission for SHG and
>>> long wavelength transmission for our Ti:Saph excitation source.  Any
>>> thoughts on what folks use/have seen/would like to have/have heard
>>> about would be appreciated.
>>>
>>> Thanks,
>>> c
>>>
>>> Carl A. Boswell, Ph.D.
>>> Molecular and Cellular Biology
>>> University of Arizona
>>> 520-954-7053
>>> FAX 520-621-3709
>>
>>
>> This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error.
>> The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


 
>I was told recently that the 20x from olympus has better response in
>red/infrared range than de 25x

 the 25x has a higher transmission in the red/IR. To achieve the same laser power out of the objective e.g you don't want to kill your tissue,  you need to attenuate more. If you are limited in laser power, e.g. at he fringes of the spectrum and for deep imaging you get more through with the 25x. That is waht is was designed for.

best wishes

Andreas


 

 

-----Original Message-----
From: Nuno Moreno <[hidden email]>
To: [hidden email]
Sent: Tue, 19 Oct 2010 9:31
Subject: Re: 20X objective for 2P work


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Hi,



We use the 20x and its really a workhorse!



I was told recently that the 20x from olympus has better response in

red/infrared range than de 25x despite of the lower NA. Can anyone confirm this?



All the best,

Nuno Moreno, PhD

Head of Equipment Management

Instituto Gulbenkian de Ciência

Fundação Calouste Gulbenkian

Tel . +351 4464538

Fax. +351 4407970











On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:



> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

> Hi all,

>

> Nikon also has a new 25x objective for multiphoton work - specifications look

very similar to the Olympus  and it available without buying a complete

system... however I've not had any feedback from actual users yet.

>

> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some

stage but haven't had any success getting that organised yet.

>

> We've had good results with the Olympus 20x NA 0.95 (standard and special IR

versions).

>

> Regards,

>

> Adrian Smith

> Centenary Institute, Sydney, Australia

>

> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:

>

>> *****

>> To join, leave or search the confocal microscopy listserv, go to:

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>> *****

>>

>> Hi List,

>>

>> Unfortunately the Olympus 25x is only sold as part of a complete system

>> (we have one and it has been great). But the older 20x XLumPlanFL works

>> pretty well too for visible and IR wavelengths and you should be able to

>> buy one of those separately.

>>

>>

>> Cheers

>>

>> Cam

>>

>>

>> Cameron J. Nowell

>> Microscopy Manager

>> Centre for Advanced Microscopy

>> Ludwig Institute for Cancer Research

>> PO Box 2008

>> Royal Melbourne Hospital

>> Victoria, 3050

>> AUSTRALIA

>> Office: +61 3 9341 3155

>> Mobile: +61422882700

>> Fax: +61 3 9341 3104

>> Facility Website

>>

>>

>>

>> -----Original Message-----

>> From: Confocal Microscopy List [mailto:[hidden email]]

>> On Behalf Of Patty Jansma

>> Sent: Tuesday, 19 October 2010 6:07 AM

>> To: [hidden email]

>> Subject: Re: 20X objective for 2P work

>>

>> *****

>> To join, leave or search the confocal microscopy listserv, go to:

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>> *****

>>

>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6

>> 46

>>

>> I think it may only be sold with the Flo-view1000MPE but does what you

>> want.

>>

>> Patty

>>

>> On 10/18/2010 11:20 AM, Carl Boswell wrote:

>>> *****

>>> To join, leave or search the confocal microscopy listserv, go to:

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>>> *****

>>>

>>> Dear listers,

>>> We're currently on the search for a high NA water immersion 20x

>>> objective to use for intravital 2P imaging.  As such, we need

>>> something that does it all, short wavelength transmission for SHG and

>>> long wavelength transmission for our Ti:Saph excitation source.  Any

>>> thoughts on what folks use/have seen/would like to have/have heard

>>> about would be appreciated.

>>>

>>> Thanks,

>>> c

>>>

>>> Carl A. Boswell, Ph.D.

>>> Molecular and Cellular Biology

>>> University of Arizona

>>> 520-954-7053

>>> FAX 520-621-3709

>>

>>

>> This communication is intended only for the named recipient and may contain

information that is confidential, legally privileged or subject to copyright;

the Ludwig Institute for Cancer Research Ltd does not waive any rights if you

have received this communication in error.

>> The views expressed in this communication are those of the sender and do not

necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.


 
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

is there any numbers? Opinions I already have.





On Oct 19, 2010, at 10:02 AM, Andreas Bruckbauer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
>
>
>> I was told recently that the 20x from olympus has better response in
>> red/infrared range than de 25x
>
> the 25x has a higher transmission in the red/IR. To achieve the same laser power out of the objective e.g you don't want to kill your tissue,  you need to attenuate more. If you are limited in laser power, e.g. at he fringes of the spectrum and for deep imaging you get more through with the 25x. That is waht is was designed for.
>
> best wishes
>
> Andreas
>
>
>
>
>
>
> -----Original Message-----
> From: Nuno Moreno <[hidden email]>
> To: [hidden email]
> Sent: Tue, 19 Oct 2010 9:31
> Subject: Re: 20X objective for 2P work
>
>
> *****
>
> To join, leave or search the confocal microscopy listserv, go to:
>
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
> *****
>
>
>
> Hi,
>
>
>
> We use the 20x and its really a workhorse!
>
>
>
> I was told recently that the 20x from olympus has better response in
>
> red/infrared range than de 25x despite of the lower NA. Can anyone confirm this?
>
>
>
> All the best,
>
> Nuno Moreno, PhD
>
> Head of Equipment Management
>
> Instituto Gulbenkian de Ciência
>
> Fundação Calouste Gulbenkian
>
> Tel . +351 4464538
>
> Fax. +351 4407970
>
>
>
>
>
>
>
>
>
>
>
> On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:
>
>
>
>> *****
>
>> To join, leave or search the confocal microscopy listserv, go to:
>
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>> *****
>
>>
>
>> Hi all,
>
>>
>
>> Nikon also has a new 25x objective for multiphoton work - specifications look
>
> very similar to the Olympus  and it available without buying a complete
>
> system... however I've not had any feedback from actual users yet.
>
>>
>
>> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some
>
> stage but haven't had any success getting that organised yet.
>
>>
>
>> We've had good results with the Olympus 20x NA 0.95 (standard and special IR
>
> versions).
>
>>
>
>> Regards,
>
>>
>
>> Adrian Smith
>
>> Centenary Institute, Sydney, Australia
>
>>
>
>> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:
>
>>
>
>>> *****
>
>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>> *****
>
>>>
>
>>> Hi List,
>
>>>
>
>>> Unfortunately the Olympus 25x is only sold as part of a complete system
>
>>> (we have one and it has been great). But the older 20x XLumPlanFL works
>
>>> pretty well too for visible and IR wavelengths and you should be able to
>
>>> buy one of those separately.
>
>>>
>
>>>
>
>>> Cheers
>
>>>
>
>>> Cam
>
>>>
>
>>>
>
>>> Cameron J. Nowell
>
>>> Microscopy Manager
>
>>> Centre for Advanced Microscopy
>
>>> Ludwig Institute for Cancer Research
>
>>> PO Box 2008
>
>>> Royal Melbourne Hospital
>
>>> Victoria, 3050
>
>>> AUSTRALIA
>
>>> Office: +61 3 9341 3155
>
>>> Mobile: +61422882700
>
>>> Fax: +61 3 9341 3104
>
>>> Facility Website
>
>>>
>
>>>
>
>>>
>
>>> -----Original Message-----
>
>>> From: Confocal Microscopy List [mailto:[hidden email]]
>
>>> On Behalf Of Patty Jansma
>
>>> Sent: Tuesday, 19 October 2010 6:07 AM
>
>>> To: [hidden email]
>
>>> Subject: Re: 20X objective for 2P work
>
>>>
>
>>> *****
>
>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>> *****
>
>>>
>
>>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
>
>>> 46
>
>>>
>
>>> I think it may only be sold with the Flo-view1000MPE but does what you
>
>>> want.
>
>>>
>
>>> Patty
>
>>>
>
>>> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>
>>>> *****
>
>>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>>> *****
>
>>>>
>
>>>> Dear listers,
>
>>>> We're currently on the search for a high NA water immersion 20x
>
>>>> objective to use for intravital 2P imaging.  As such, we need
>
>>>> something that does it all, short wavelength transmission for SHG and
>
>>>> long wavelength transmission for our Ti:Saph excitation source.  Any
>
>>>> thoughts on what folks use/have seen/would like to have/have heard
>
>>>> about would be appreciated.
>
>>>>
>
>>>> Thanks,
>
>>>> c
>
>>>>
>
>>>> Carl A. Boswell, Ph.D.
>
>>>> Molecular and Cellular Biology
>
>>>> University of Arizona
>
>>>> 520-954-7053
>
>>>> FAX 520-621-3709
>
>>>
>
>>>
>
>>> This communication is intended only for the named recipient and may contain
>
> information that is confidential, legally privileged or subject to copyright;
>
> the Ludwig Institute for Cancer Research Ltd does not waive any rights if you
>
> have received this communication in error.
>
>>> The views expressed in this communication are those of the sender and do not
>
> necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
>
>
>
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

>is there any numbers? Opinions I already have.



at 800 nm, 2.5% laser power setting (AOM transmission)
25x: 10.4 mW, 20x: 7.9 mW

at 950 nm:
25x: 3.2 mW, 20x: 2.85

at 1020 mW 6%
25x: 2.61, 20x 2.39 mW

these were measured on the Olympus system which fills the back aperture
of the 25x objective, but underfills the 20x objective, when i set the beam expander to get the same fill factor for the 25x as for the 20x, i measure
11.8 mW (800nm), 3.5mW (950), 2.91 mW (1020)

so i conclude about 20-50% more transmission in the IR for the 25x
These values were measured in air, so there is some total reflection expected for the 25x NA1.05 objective. The measured power levels also depend on
the output characteristics of the laser. Furthermore there are some uncertainties depending on the type of power meter used, i think this was done with
a photodiode based sensor (was a while ago).

best wishes

Andreas







On Oct 19, 2010, at 10:02 AM, Andreas Bruckbauer wrote:



> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

>

>

>> I was told recently that the 20x from olympus has better response in

>> red/infrared range than de 25x

>

> the 25x has a higher transmission in the red/IR. To achieve the same laser

power out of the objective e.g you don't want to kill your tissue,  you need to

attenuate more. If you are limited in laser power, e.g. at he fringes of the

spectrum and for deep imaging you get more through with the 25x. That is waht is

was designed for.

>

> best wishes

>

> Andreas

>

>

>

>

>

>

> -----Original Message-----

> From: Nuno Moreno <[hidden email]>

> To: [hidden email]

> Sent: Tue, 19 Oct 2010 9:31

> Subject: Re: 20X objective for 2P work

>

>

> *****

>

> To join, leave or search the confocal microscopy listserv, go to:

>

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

> *****

>

>

>

> Hi,

>

>

>

> We use the 20x and its really a workhorse!

>

>

>

> I was told recently that the 20x from olympus has better response in

>

> red/infrared range than de 25x despite of the lower NA. Can anyone confirm

this?

>

>

>

> All the best,

>

> Nuno Moreno, PhD

>

> Head of Equipment Management

>

> Instituto Gulbenkian de Ciência

>

> Fundação Calouste Gulbenkian

>

> Tel . +351 4464538

>

> Fax. +351 4407970

>

>

>

>

>

>

>

>

>

>

>

> On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:

>

>

>

>> *****

>

>> To join, leave or search the confocal microscopy listserv, go to:

>

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>> *****

>

>>

>

>> Hi all,

>

>>

>

>> Nikon also has a new 25x objective for multiphoton work - specifications look



>

> very similar to the Olympus  and it available without buying a complete

>

> system... however I've not had any feedback from actual users yet.

>

>>

>

>> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some

>

> stage but haven't had any success getting that organised yet.

>

>>

>

>> We've had good results with the Olympus 20x NA 0.95 (standard and special IR

>

> versions).

>

>>

>

>> Regards,

>

>>

>

>> Adrian Smith

>

>> Centenary Institute, Sydney, Australia

>

>>

>

>> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:

>

>>

>

>>> *****

>

>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>> *****

>

>>>

>

>>> Hi List,

>

>>>

>

>>> Unfortunately the Olympus 25x is only sold as part of a complete system

>

>>> (we have one and it has been great). But the older 20x XLumPlanFL works

>

>>> pretty well too for visible and IR wavelengths and you should be able to

>

>>> buy one of those separately.

>

>>>

>

>>>

>

>>> Cheers

>

>>>

>

>>> Cam

>

>>>

>

>>>

>

>>> Cameron J. Nowell

>

>>> Microscopy Manager

>

>>> Centre for Advanced Microscopy

>

>>> Ludwig Institute for Cancer Research

>

>>> PO Box 2008

>

>>> Royal Melbourne Hospital

>

>>> Victoria, 3050

>

>>> AUSTRALIA

>

>>> Office: +61 3 9341 3155

>

>>> Mobile: +61422882700

>

>>> Fax: +61 3 9341 3104

>

>>> Facility Website

>

>>>

>

>>>

>

>>>

>

>>> -----Original Message-----

>

>>> From: Confocal Microscopy List [mailto:[hidden email]]

>

>>> On Behalf Of Patty Jansma

>

>>> Sent: Tuesday, 19 October 2010 6:07 AM

>

>>> To: [hidden email]

>

>>> Subject: Re: 20X objective for 2P work

>

>>>

>

>>> *****

>

>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>> *****

>

>>>

>

>>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6

>

>>> 46

>

>>>

>

>>> I think it may only be sold with the Flo-view1000MPE but does what you

>

>>> want.

>

>>>

>

>>> Patty

>

>>>

>

>>> On 10/18/2010 11:20 AM, Carl Boswell wrote:

>

>>>> *****

>

>>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>>> *****

>

>>>>

>

>>>> Dear listers,

>

>>>> We're currently on the search for a high NA water immersion 20x

>

>>>> objective to use for intravital 2P imaging.  As such, we need

>

>>>> something that does it all, short wavelength transmission for SHG and

>

>>>> long wavelength transmission for our Ti:Saph excitation source.  Any

>

>>>> thoughts on what folks use/have seen/would like to have/have heard

>

>>>> about would be appreciated.

>

>>>>

>

>>>> Thanks,

>

>>>> c

>

>>>>

>

>>>> Carl A. Boswell, Ph.D.

>

>>>> Molecular and Cellular Biology

>

>>>> University of Arizona

>

>>>> 520-954-7053

>

>>>> FAX 520-621-3709

>

>>>

>

>>>

>

>>> This communication is intended only for the named recipient and may contain

>

> information that is confidential, legally privileged or subject to copyright;

>

> the Ludwig Institute for Cancer Research Ltd does not waive any rights if you

>

> have received this communication in error.

>

>>> The views expressed in this communication are those of the sender and do not



>

> necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.

>

>

>


 
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi,

Now I think I understand why on Nimejen they have the 20x. They have an OPO and use >1100 nm for mcherry. It might be that there is a turning point in performance because at 1020 the difference is only 9%.

Thank you for those numbers. Even with technical limitations better those than nothing!
NM




On Oct 19, 2010, at 4:28 PM, Andreas Bruckbauer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
>> is there any numbers? Opinions I already have.
>
>
>
> at 800 nm, 2.5% laser power setting (AOM transmission)
> 25x: 10.4 mW, 20x: 7.9 mW
>
> at 950 nm:
> 25x: 3.2 mW, 20x: 2.85
>
> at 1020 mW 6%
> 25x: 2.61, 20x 2.39 mW
>
> these were measured on the Olympus system which fills the back aperture
> of the 25x objective, but underfills the 20x objective, when i set the beam expander to get the same fill factor for the 25x as for the 20x, i measure
> 11.8 mW (800nm), 3.5mW (950), 2.91 mW (1020)
>
> so i conclude about 20-50% more transmission in the IR for the 25x
> These values were measured in air, so there is some total reflection expected for the 25x NA1.05 objective. The measured power levels also depend on
> the output characteristics of the laser. Furthermore there are some uncertainties depending on the type of power meter used, i think this was done with
> a photodiode based sensor (was a while ago).
>
> best wishes
>
> Andreas
>
>
>
>
>
>
>
> On Oct 19, 2010, at 10:02 AM, Andreas Bruckbauer wrote:
>
>
>
>> *****
>
>> To join, leave or search the confocal microscopy listserv, go to:
>
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>> *****
>
>>
>
>>
>
>>
>
>>> I was told recently that the 20x from olympus has better response in
>
>>> red/infrared range than de 25x
>
>>
>
>> the 25x has a higher transmission in the red/IR. To achieve the same laser
>
> power out of the objective e.g you don't want to kill your tissue,  you need to
>
> attenuate more. If you are limited in laser power, e.g. at he fringes of the
>
> spectrum and for deep imaging you get more through with the 25x. That is waht is
>
> was designed for.
>
>>
>
>> best wishes
>
>>
>
>> Andreas
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>> -----Original Message-----
>
>> From: Nuno Moreno <[hidden email]>
>
>> To: [hidden email]
>
>> Sent: Tue, 19 Oct 2010 9:31
>
>> Subject: Re: 20X objective for 2P work
>
>>
>
>>
>
>> *****
>
>>
>
>> To join, leave or search the confocal microscopy listserv, go to:
>
>>
>
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>
>
>> *****
>
>>
>
>>
>
>>
>
>> Hi,
>
>>
>
>>
>
>>
>
>> We use the 20x and its really a workhorse!
>
>>
>
>>
>
>>
>
>> I was told recently that the 20x from olympus has better response in
>
>>
>
>> red/infrared range than de 25x despite of the lower NA. Can anyone confirm
>
> this?
>
>>
>
>>
>
>>
>
>> All the best,
>
>>
>
>> Nuno Moreno, PhD
>
>>
>
>> Head of Equipment Management
>
>>
>
>> Instituto Gulbenkian de Ciência
>
>>
>
>> Fundação Calouste Gulbenkian
>
>>
>
>> Tel . +351 4464538
>
>>
>
>> Fax. +351 4407970
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>> On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:
>
>>
>
>>
>
>>
>
>>> *****
>
>>
>
>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>
>
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>
>
>>> *****
>
>>
>
>>>
>
>>
>
>>> Hi all,
>
>>
>
>>>
>
>>
>
>>> Nikon also has a new 25x objective for multiphoton work - specifications look
>
>
>
>>
>
>> very similar to the Olympus  and it available without buying a complete
>
>>
>
>> system... however I've not had any feedback from actual users yet.
>
>>
>
>>>
>
>>
>
>>> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some
>
>>
>
>> stage but haven't had any success getting that organised yet.
>
>>
>
>>>
>
>>
>
>>> We've had good results with the Olympus 20x NA 0.95 (standard and special IR
>
>>
>
>> versions).
>
>>
>
>>>
>
>>
>
>>> Regards,
>
>>
>
>>>
>
>>
>
>>> Adrian Smith
>
>>
>
>>> Centenary Institute, Sydney, Australia
>
>>
>
>>>
>
>>
>
>>> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:
>
>>
>
>>>
>
>>
>
>>>> *****
>
>>
>
>>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>
>
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>
>
>>>> *****
>
>>
>
>>>>
>
>>
>
>>>> Hi List,
>
>>
>
>>>>
>
>>
>
>>>> Unfortunately the Olympus 25x is only sold as part of a complete system
>
>>
>
>>>> (we have one and it has been great). But the older 20x XLumPlanFL works
>
>>
>
>>>> pretty well too for visible and IR wavelengths and you should be able to
>
>>
>
>>>> buy one of those separately.
>
>>
>
>>>>
>
>>
>
>>>>
>
>>
>
>>>> Cheers
>
>>
>
>>>>
>
>>
>
>>>> Cam
>
>>
>
>>>>
>
>>
>
>>>>
>
>>
>
>>>> Cameron J. Nowell
>
>>
>
>>>> Microscopy Manager
>
>>
>
>>>> Centre for Advanced Microscopy
>
>>
>
>>>> Ludwig Institute for Cancer Research
>
>>
>
>>>> PO Box 2008
>
>>
>
>>>> Royal Melbourne Hospital
>
>>
>
>>>> Victoria, 3050
>
>>
>
>>>> AUSTRALIA
>
>>
>
>>>> Office: +61 3 9341 3155
>
>>
>
>>>> Mobile: +61422882700
>
>>
>
>>>> Fax: +61 3 9341 3104
>
>>
>
>>>> Facility Website
>
>>
>
>>>>
>
>>
>
>>>>
>
>>
>
>>>>
>
>>
>
>>>> -----Original Message-----
>
>>
>
>>>> From: Confocal Microscopy List [mailto:[hidden email]]
>
>>
>
>>>> On Behalf Of Patty Jansma
>
>>
>
>>>> Sent: Tuesday, 19 October 2010 6:07 AM
>
>>
>
>>>> To: [hidden email]
>
>>
>
>>>> Subject: Re: 20X objective for 2P work
>
>>
>
>>>>
>
>>
>
>>>> *****
>
>>
>
>>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>
>
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>
>
>>>> *****
>
>>
>
>>>>
>
>>
>
>>>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
>
>>
>
>>>> 46
>
>>
>
>>>>
>
>>
>
>>>> I think it may only be sold with the Flo-view1000MPE but does what you
>
>>
>
>>>> want.
>
>>
>
>>>>
>
>>
>
>>>> Patty
>
>>
>
>>>>
>
>>
>
>>>> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>
>>
>
>>>>> *****
>
>>
>
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>
>>
>
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>>
>
>>>>> *****
>
>>
>
>>>>>
>
>>
>
>>>>> Dear listers,
>
>>
>
>>>>> We're currently on the search for a high NA water immersion 20x
>
>>
>
>>>>> objective to use for intravital 2P imaging.  As such, we need
>
>>
>
>>>>> something that does it all, short wavelength transmission for SHG and
>
>>
>
>>>>> long wavelength transmission for our Ti:Saph excitation source.  Any
>
>>
>
>>>>> thoughts on what folks use/have seen/would like to have/have heard
>
>>
>
>>>>> about would be appreciated.
>
>>
>
>>>>>
>
>>
>
>>>>> Thanks,
>
>>
>
>>>>> c
>
>>
>
>>>>>
>
>>
>
>>>>> Carl A. Boswell, Ph.D.
>
>>
>
>>>>> Molecular and Cellular Biology
>
>>
>
>>>>> University of Arizona
>
>>
>
>>>>> 520-954-7053
>
>>
>
>>>>> FAX 520-621-3709
>
>>
>
>>>>
>
>>
>
>>>>
>
>>
>
>>>> This communication is intended only for the named recipient and may contain
>
>>
>
>> information that is confidential, legally privileged or subject to copyright;
>
>>
>
>> the Ludwig Institute for Cancer Research Ltd does not waive any rights if you
>
>>
>
>> have received this communication in error.
>
>>
>
>>>> The views expressed in this communication are those of the sender and do not
>
>
>
>>
>
>> necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
>
>>
>
>>
>
>>
>
>
>
Adrian Smith-6 Adrian Smith-6
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi,

If you are referring to the LaVision BioTec systems at Nijmegen then the other reason they will have the 20x is that Olympus won't sell them the 25x as it is not an Olympus system :)

We are in the same boat here with our LaVision BioTec systems with OPO.

We actually have two versions of the 20x Olympus objective - one "normal" and one "special IR" which has much better transmission in the longer wavelengths. I don't have any rigorous numbers or 25x to compare with so I won't post anything here.

We have also heard that the current 20x Olympus objectives (might be NA 1.0?) are not as good for the OPO as the older NA0.95 special IR version... so it certainly seems that not all Olympus 20x objectives are created equal!

Regards,

Adrian Smith
Centenary Institute, Sydney, Australia



On 20/10/2010, at 3:05 AM, Nuno Moreno wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi,
>
> Now I think I understand why on Nimejen they have the 20x. They have an OPO and use >1100 nm for mcherry. It might be that there is a turning point in performance because at 1020 the difference is only 9%.
>
> Thank you for those numbers. Even with technical limitations better those than nothing!
> NM
>
>
>
>
> On Oct 19, 2010, at 4:28 PM, Andreas Bruckbauer wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>>> is there any numbers? Opinions I already have.
>>
>>
>>
>> at 800 nm, 2.5% laser power setting (AOM transmission)
>> 25x: 10.4 mW, 20x: 7.9 mW
>>
>> at 950 nm:
>> 25x: 3.2 mW, 20x: 2.85
>>
>> at 1020 mW 6%
>> 25x: 2.61, 20x 2.39 mW
>>
>> these were measured on the Olympus system which fills the back aperture
>> of the 25x objective, but underfills the 20x objective, when i set the beam expander to get the same fill factor for the 25x as for the 20x, i measure
>> 11.8 mW (800nm), 3.5mW (950), 2.91 mW (1020)
>>
>> so i conclude about 20-50% more transmission in the IR for the 25x
>> These values were measured in air, so there is some total reflection expected for the 25x NA1.05 objective. The measured power levels also depend on
>> the output characteristics of the laser. Furthermore there are some uncertainties depending on the type of power meter used, i think this was done with
>> a photodiode based sensor (was a while ago).
>>
>> best wishes
>>
>> Andreas
>>
>>
>>
>>
>>
>>
>>
>> On Oct 19, 2010, at 10:02 AM, Andreas Bruckbauer wrote:
>>
>>
>>
>>> *****
>>
>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>> *****
>>
>>>
>>
>>>
>>
>>>
>>
>>>> I was told recently that the 20x from olympus has better response in
>>
>>>> red/infrared range than de 25x
>>
>>>
>>
>>> the 25x has a higher transmission in the red/IR. To achieve the same laser
>>
>> power out of the objective e.g you don't want to kill your tissue,  you need to
>>
>> attenuate more. If you are limited in laser power, e.g. at he fringes of the
>>
>> spectrum and for deep imaging you get more through with the 25x. That is waht is
>>
>> was designed for.
>>
>>>
>>
>>> best wishes
>>
>>>
>>
>>> Andreas
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>> -----Original Message-----
>>
>>> From: Nuno Moreno <[hidden email]>
>>
>>> To: [hidden email]
>>
>>> Sent: Tue, 19 Oct 2010 9:31
>>
>>> Subject: Re: 20X objective for 2P work
>>
>>>
>>
>>>
>>
>>> *****
>>
>>>
>>
>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>>
>>
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>>
>>
>>> *****
>>
>>>
>>
>>>
>>
>>>
>>
>>> Hi,
>>
>>>
>>
>>>
>>
>>>
>>
>>> We use the 20x and its really a workhorse!
>>
>>>
>>
>>>
>>
>>>
>>
>>> I was told recently that the 20x from olympus has better response in
>>
>>>
>>
>>> red/infrared range than de 25x despite of the lower NA. Can anyone confirm
>>
>> this?
>>
>>>
>>
>>>
>>
>>>
>>
>>> All the best,
>>
>>>
>>
>>> Nuno Moreno, PhD
>>
>>>
>>
>>> Head of Equipment Management
>>
>>>
>>
>>> Instituto Gulbenkian de Ciência
>>
>>>
>>
>>> Fundação Calouste Gulbenkian
>>
>>>
>>
>>> Tel . +351 4464538
>>
>>>
>>
>>> Fax. +351 4407970
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>>
>>
>>> On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:
>>
>>>
>>
>>>
>>
>>>
>>
>>>> *****
>>
>>>
>>
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>>
>>
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>>
>>
>>>> *****
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> Hi all,
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> Nikon also has a new 25x objective for multiphoton work - specifications look
>>
>>
>>
>>>
>>
>>> very similar to the Olympus  and it available without buying a complete
>>
>>>
>>
>>> system... however I've not had any feedback from actual users yet.
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some
>>
>>>
>>
>>> stage but haven't had any success getting that organised yet.
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> We've had good results with the Olympus 20x NA 0.95 (standard and special IR
>>
>>>
>>
>>> versions).
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> Regards,
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> Adrian Smith
>>
>>>
>>
>>>> Centenary Institute, Sydney, Australia
>>
>>>
>>
>>>>
>>
>>>
>>
>>>> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:
>>
>>>
>>
>>>>
>>
>>>
>>
>>>>> *****
>>
>>>
>>
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>>
>>
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>>
>>
>>>>> *****
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Hi List,
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Unfortunately the Olympus 25x is only sold as part of a complete system
>>
>>>
>>
>>>>> (we have one and it has been great). But the older 20x XLumPlanFL works
>>
>>>
>>
>>>>> pretty well too for visible and IR wavelengths and you should be able to
>>
>>>
>>
>>>>> buy one of those separately.
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Cheers
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Cam
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Cameron J. Nowell
>>
>>>
>>
>>>>> Microscopy Manager
>>
>>>
>>
>>>>> Centre for Advanced Microscopy
>>
>>>
>>
>>>>> Ludwig Institute for Cancer Research
>>
>>>
>>
>>>>> PO Box 2008
>>
>>>
>>
>>>>> Royal Melbourne Hospital
>>
>>>
>>
>>>>> Victoria, 3050
>>
>>>
>>
>>>>> AUSTRALIA
>>
>>>
>>
>>>>> Office: +61 3 9341 3155
>>
>>>
>>
>>>>> Mobile: +61422882700
>>
>>>
>>
>>>>> Fax: +61 3 9341 3104
>>
>>>
>>
>>>>> Facility Website
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> -----Original Message-----
>>
>>>
>>
>>>>> From: Confocal Microscopy List [mailto:[hidden email]]
>>
>>>
>>
>>>>> On Behalf Of Patty Jansma
>>
>>>
>>
>>>>> Sent: Tuesday, 19 October 2010 6:07 AM
>>
>>>
>>
>>>>> To: [hidden email]
>>
>>>
>>
>>>>> Subject: Re: 20X objective for 2P work
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> *****
>>
>>>
>>
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>>
>>
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>>
>>
>>>>> *****
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6
>>
>>>
>>
>>>>> 46
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> I think it may only be sold with the Flo-view1000MPE but does what you
>>
>>>
>>
>>>>> want.
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> Patty
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> On 10/18/2010 11:20 AM, Carl Boswell wrote:
>>
>>>
>>
>>>>>> *****
>>
>>>
>>
>>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>
>>>
>>
>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>
>>>
>>
>>>>>> *****
>>
>>>
>>
>>>>>>
>>
>>>
>>
>>>>>> Dear listers,
>>
>>>
>>
>>>>>> We're currently on the search for a high NA water immersion 20x
>>
>>>
>>
>>>>>> objective to use for intravital 2P imaging.  As such, we need
>>
>>>
>>
>>>>>> something that does it all, short wavelength transmission for SHG and
>>
>>>
>>
>>>>>> long wavelength transmission for our Ti:Saph excitation source.  Any
>>
>>>
>>
>>>>>> thoughts on what folks use/have seen/would like to have/have heard
>>
>>>
>>
>>>>>> about would be appreciated.
>>
>>>
>>
>>>>>>
>>
>>>
>>
>>>>>> Thanks,
>>
>>>
>>
>>>>>> c
>>
>>>
>>
>>>>>>
>>
>>>
>>
>>>>>> Carl A. Boswell, Ph.D.
>>
>>>
>>
>>>>>> Molecular and Cellular Biology
>>
>>>
>>
>>>>>> University of Arizona
>>
>>>
>>
>>>>>> 520-954-7053
>>
>>>
>>
>>>>>> FAX 520-621-3709
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>>
>>
>>>
>>
>>>>> This communication is intended only for the named recipient and may contain
>>
>>>
>>
>>> information that is confidential, legally privileged or subject to copyright;
>>
>>>
>>
>>> the Ludwig Institute for Cancer Research Ltd does not waive any rights if you
>>
>>>
>>
>>> have received this communication in error.
>>
>>>
>>
>>>>> The views expressed in this communication are those of the sender and do not
>>
>>
>>
>>>
>>
>>> necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
>>
>>>
>>
>>>
>>
>>>
>>
>>
>>
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

In reply to this post by Nuno Moreno
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


 OLympus does not have the transmission data for the 25x NA 1.05 objective (its called XLPLN25xW-MP) online. When i remember correctly from prensentation by Eiji Yokoi  at the FOM 2008 where they launched this lens, the transmission was improved over the whole range of wavelengths. My measurement was with the old 20x, XLUMPLFL 20XW (NA0.95) and now this is redesigned and called XLUMPLFLN 20XW (NA 1.0), the transmission curve up to 1000 nm is available on the Olympus website
http://microscope.olympus-global.com/uis2/en/xlumplfln20xw/
Maybe Olympus can comment on the transmission > 1100 nm for all these objectives?

 
Another issue when using these lenses on other manufacturers microscopes and these include Olympus microscope stands sold to other manufacturers such as Lavision, that they might have different scanning lenses or tube lenses leading to different results.

best wishes

Andreas

 

-----Original Message-----
From: Nuno Moreno <[hidden email]>
To: [hidden email]
Sent: Tue, 19 Oct 2010 17:05
Subject: Re: 20X objective for 2P work


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Hi,



Now I think I understand why on Nimejen they have the 20x. They have an OPO and

use >1100 nm for mcherry. It might be that there is a turning point in

performance because at 1020 the difference is only 9%.



Thank you for those numbers. Even with technical limitations better those than

nothing!

NM









On Oct 19, 2010, at 4:28 PM, Andreas Bruckbauer wrote:



> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

>> is there any numbers? Opinions I already have.

>

>

>

> at 800 nm, 2.5% laser power setting (AOM transmission)

> 25x: 10.4 mW, 20x: 7.9 mW

>

> at 950 nm:

> 25x: 3.2 mW, 20x: 2.85

>

> at 1020 mW 6%

> 25x: 2.61, 20x 2.39 mW

>

> these were measured on the Olympus system which fills the back aperture

> of the 25x objective, but underfills the 20x objective, when i set the beam

expander to get the same fill factor for the 25x as for the 20x, i measure

> 11.8 mW (800nm), 3.5mW (950), 2.91 mW (1020)

>

> so i conclude about 20-50% more transmission in the IR for the 25x

> These values were measured in air, so there is some total reflection expected

for the 25x NA1.05 objective. The measured power levels also depend on

> the output characteristics of the laser. Furthermore there are some

uncertainties depending on the type of power meter used, i think this was done

with

> a photodiode based sensor (was a while ago).

>

> best wishes

>

> Andreas

>

>

>

>

>

>

>

> On Oct 19, 2010, at 10:02 AM, Andreas Bruckbauer wrote:

>

>

>

>> *****

>

>> To join, leave or search the confocal microscopy listserv, go to:

>

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>> *****

>

>>

>

>>

>

>>

>

>>> I was told recently that the 20x from olympus has better response in

>

>>> red/infrared range than de 25x

>

>>

>

>> the 25x has a higher transmission in the red/IR. To achieve the same laser

>

> power out of the objective e.g you don't want to kill your tissue,  you need

to

>

> attenuate more. If you are limited in laser power, e.g. at he fringes of the

>

> spectrum and for deep imaging you get more through with the 25x. That is waht

is

>

> was designed for.

>

>>

>

>> best wishes

>

>>

>

>> Andreas

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>> -----Original Message-----

>

>> From: Nuno Moreno <[hidden email]>

>

>> To: [hidden email]

>

>> Sent: Tue, 19 Oct 2010 9:31

>

>> Subject: Re: 20X objective for 2P work

>

>>

>

>>

>

>> *****

>

>>

>

>> To join, leave or search the confocal microscopy listserv, go to:

>

>>

>

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>

>

>> *****

>

>>

>

>>

>

>>

>

>> Hi,

>

>>

>

>>

>

>>

>

>> We use the 20x and its really a workhorse!

>

>>

>

>>

>

>>

>

>> I was told recently that the 20x from olympus has better response in

>

>>

>

>> red/infrared range than de 25x despite of the lower NA. Can anyone confirm

>

> this?

>

>>

>

>>

>

>>

>

>> All the best,

>

>>

>

>> Nuno Moreno, PhD

>

>>

>

>> Head of Equipment Management

>

>>

>

>> Instituto Gulbenkian de Ciência

>

>>

>

>> Fundação Calouste Gulbenkian

>

>>

>

>> Tel . +351 4464538

>

>>

>

>> Fax. +351 4407970

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>>

>

>> On Oct 19, 2010, at 5:33 AM, Adrian Smith wrote:

>

>>

>

>>

>

>>

>

>>> *****

>

>>

>

>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>

>

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>

>

>>> *****

>

>>

>

>>>

>

>>

>

>>> Hi all,

>

>>

>

>>>

>

>>

>

>>> Nikon also has a new 25x objective for multiphoton work - specifications

look

>

>

>

>>

>

>> very similar to the Olympus  and it available without buying a complete

>

>>

>

>> system... however I've not had any feedback from actual users yet.

>

>>

>

>>>

>

>>

>

>>> I'm hoping to be able to test one on our LaVision BioTec TriMScopes at some

>

>>

>

>> stage but haven't had any success getting that organised yet.

>

>>

>

>>>

>

>>

>

>>> We've had good results with the Olympus 20x NA 0.95 (standard and special IR



>

>>

>

>> versions).

>

>>

>

>>>

>

>>

>

>>> Regards,

>

>>

>

>>>

>

>>

>

>>> Adrian Smith

>

>>

>

>>> Centenary Institute, Sydney, Australia

>

>>

>

>>>

>

>>

>

>>> On 19/10/2010, at 10:16 AM, Cameron Nowell wrote:

>

>>

>

>>>

>

>>

>

>>>> *****

>

>>

>

>>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>

>

>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>

>

>>>> *****

>

>>

>

>>>>

>

>>

>

>>>> Hi List,

>

>>

>

>>>>

>

>>

>

>>>> Unfortunately the Olympus 25x is only sold as part of a complete system

>

>>

>

>>>> (we have one and it has been great). But the older 20x XLumPlanFL works

>

>>

>

>>>> pretty well too for visible and IR wavelengths and you should be able to

>

>>

>

>>>> buy one of those separately.

>

>>

>

>>>>

>

>>

>

>>>>

>

>>

>

>>>> Cheers

>

>>

>

>>>>

>

>>

>

>>>> Cam

>

>>

>

>>>>

>

>>

>

>>>>

>

>>

>

>>>> Cameron J. Nowell

>

>>

>

>>>> Microscopy Manager

>

>>

>

>>>> Centre for Advanced Microscopy

>

>>

>

>>>> Ludwig Institute for Cancer Research

>

>>

>

>>>> PO Box 2008

>

>>

>

>>>> Royal Melbourne Hospital

>

>>

>

>>>> Victoria, 3050

>

>>

>

>>>> AUSTRALIA

>

>>

>

>>>> Office: +61 3 9341 3155

>

>>

>

>>>> Mobile: +61422882700

>

>>

>

>>>> Fax: +61 3 9341 3104

>

>>

>

>>>> Facility Website

>

>>

>

>>>>

>

>>

>

>>>>

>

>>

>

>>>>

>

>>

>

>>>> -----Original Message-----

>

>>

>

>>>> From: Confocal Microscopy List [mailto:[hidden email]]

>

>>

>

>>>> On Behalf Of Patty Jansma

>

>>

>

>>>> Sent: Tuesday, 19 October 2010 6:07 AM

>

>>

>

>>>> To: [hidden email]

>

>>

>

>>>> Subject: Re: 20X objective for 2P work

>

>>

>

>>>>

>

>>

>

>>>> *****

>

>>

>

>>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>

>

>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>

>

>>>> *****

>

>>

>

>>>>

>

>>

>

>>>> http://www.olympusamerica.com/seg_section/seg_pressDetails.asp?pressNo=6

>

>>

>

>>>> 46

>

>>

>

>>>>

>

>>

>

>>>> I think it may only be sold with the Flo-view1000MPE but does what you

>

>>

>

>>>> want.

>

>>

>

>>>>

>

>>

>

>>>> Patty

>

>>

>

>>>>

>

>>

>

>>>> On 10/18/2010 11:20 AM, Carl Boswell wrote:

>

>>

>

>>>>> *****

>

>>

>

>>>>> To join, leave or search the confocal microscopy listserv, go to:

>

>>

>

>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>

>>

>

>>>>> *****

>

>>

>

>>>>>

>

>>

>

>>>>> Dear listers,

>

>>

>

>>>>> We're currently on the search for a high NA water immersion 20x

>

>>

>

>>>>> objective to use for intravital 2P imaging.  As such, we need

>

>>

>

>>>>> something that does it all, short wavelength transmission for SHG and

>

>>

>

>>>>> long wavelength transmission for our Ti:Saph excitation source.  Any

>

>>

>

>>>>> thoughts on what folks use/have seen/would like to have/have heard

>

>>

>

>>>>> about would be appreciated.

>

>>

>

>>>>>

>

>>

>

>>>>> Thanks,

>

>>

>

>>>>> c

>

>>

>

>>>>>

>

>>

>

>>>>> Carl A. Boswell, Ph.D.

>

>>

>

>>>>> Molecular and Cellular Biology

>

>>

>

>>>>> University of Arizona

>

>>

>

>>>>> 520-954-7053

>

>>

>

>>>>> FAX 520-621-3709

>

>>

>

>>>>

>

>>

>

>>>>

>

>>

>

>>>> This communication is intended only for the named recipient and may contain



>

>>

>

>> information that is confidential, legally privileged or subject to copyright;



>

>>

>

>> the Ludwig Institute for Cancer Research Ltd does not waive any rights if you



>

>>

>

>> have received this communication in error.

>

>>

>

>>>> The views expressed in this communication are those of the sender and do

not

>

>

>

>>

>

>> necessarily reflect the views of the Ludwig Institute for Cancer Research

Ltd.

>

>>

>

>>

>

>>

>

>

>


 
Steffen Dietzel Steffen Dietzel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 20X objective for 2P work

In reply to this post by Adrian Smith-6
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Adrian is right, there was a change in manufacturing for the Olympus
20x, I believe in 2009. Before, there was the XLUMPLFL20x NA 0.95 and
the specially coated IR Version IR-XLUMPLFL20x, which is still used by
us and others for multi-photon microscopy (although originally designed
for IR-DIC).

Instead of those then discontinued lenses, the XLUMPLFLN20x (note the N)
was then offered. I once got the following transmission numbers from an
Olympus representative, didn't verify them by measuring myself:
                       500 nm   650nm  800nm      1200 nm
XLUMPLFL20x           0.73     0.66    0.60       0.50
XLUMPLFL20x-IR        0.42     0.61    0.74       0.69
XLUMPLFLN20x          0.77     0.77    0.68       0.52

So, for a two-photon process with backward-detected 500 nm emission, if
you do the math the new one should be better than both old versions,
essentially because you loose less of the emitted photons: (exc^2 * em.)
                  800/500        1200/500         1200/650
XLUMPLFL20x-IR    0.23            0,2              0.29
XLUMPLFLN20x      0.35            0,21             0.21

This predicted outcome for 500 nm emission was used by Olympus to argue
that an IR-coated version of the new lens is not needed.

However, if you are interessted in three-photon processes and/or forward
detection (eg. THG), then you should keep a close eye on any still
available old IR 20x. Same if you go for two-photon fluorescence with
red dyes. As far as I know, there is no IR-coated version of the new
objective (yet), hopefully I am wrong. Maybe if we push this a little.


While we are on the objectives: Did any one of you bothered to measure
the chromatic aberration for Ti:Sa lasers and OPOs? For our system
(LaVision BioTech) with the XLUMPLFL20x-IR, I measured a z-shift of the
focal plane of about 9 µm between 725 and 1050 and about 7 µm between
1175 and 1300 nm. I am wondering if this is in the normal range.

Steffen


On 20.10.2010 01:11, Adrian Smith wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi,
>
> If you are referring to the LaVision BioTec systems at Nijmegen then the other reason they will have the 20x is that Olympus won't sell them the 25x as it is not an Olympus system :)
>
> We are in the same boat here with our LaVision BioTec systems with OPO.
>
> We actually have two versions of the 20x Olympus objective - one "normal" and one "special IR"

which has much better transmission in the longer wavelengths. I don't
have any rigorous numbers

or 25x to compare with so I won't post anything here.
>
> We have also heard that the current 20x Olympus objectives (might be NA 1.0?) are not as good

for the OPO as the older NA0.95 special IR version... so it certainly
seems that not all Olympus

20x objectives are created equal!
>
> Regards,
>
> Adrian Smith
> Centenary Institute, Sydney, Australia
>
>
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Walter-Brendel-Zentrum für experimentelle Medizin (WBex)
Head of light microscopy