Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Andrew, Have you looked at the Nikon Perfect Focus system or the Olympus ZDC? These are both laser based systems that reflect an IR laser off the cover glass of your sample and maintain focus not based on image analysis but based on reading the reflected laser beam. For the system you've described, I'd want to couple one of these with a software or hardware solution for finding the best focused plane based on image analysis and then trigger the laser system to maintain that position.
Chris Tully On 8/28/07, Andrew Resnick <[hidden email]> wrote:
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
As far as autofocus capabilities, there are some manufacturers that claim
to have it- Nikon's setup for TIRF also comes to mind. I've been on
a personal mission at trade shows to (gently) berate the manufacturers to
come up with a solution for focus drift due to thermal fluctuations- the
hardware exists, the missing link is a decent software feedback
loop. None have expressed interest, the typical excuse being that
thermal fluctuations have nothing to do with the optical performance of
the microscope, and thus is not their problem.
Another capability I lobby for is the ability to acquire fast z-stacks-
the camera should be the limiting rate. Again, it's primarily a
software issue and it's seen as a niche application.
As for pre-setting the alignment of the optics, my point wasn't that it
is/was not possible, my point was that we didn't have the time to solve
that particular issue in addition to the thousands of other issues.
Like any large project, what was needed was to recognize that not every
problem could be solved, and to prioritize the issues. That wasn't done,
so no problem got solved.
Andy
At 07:52 AM 8/26/2007, you wrote:
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
This is really fascinating, and it's a crying shame that the
project didn't get off the
ground. It's also a classic example of how NASA is brilliant in
spending huge sums
of money just to stuff things up. (At the time of writing this I
don't know whether the
latest Shuttle crew has safely returned to Earth or not, knowing that
their heat shield
tiles have been holed by foam from the fuel tanks)
In terms of the design features, widefield autofocus has been available
for years but
I can't vouch for the situation back in 1999. I would have thought
that for such a
tightly specified sample, Koehler illumination could have been preset,
and likewise
I'd have thought that a suitably rugged phase contrast condenser would
not require
re-aligning in orbit. But 2 litres of immersion oil still
worries me!
Andrew, this is a most extraorinarily fascinating story, and I do hope
that even
if the microscope didn't fly some useful land-based improvements to
automated
microscopy flowed on from it.
Guy
From: Confocal Microscopy List on
behalf of Robert J. Palmer Jr.
Sent: Sat 25/08/2007 4:54 AM
To: [hidden email]
Subject: Re: A microscope on Mars?
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
This is all quite interesting from the standpoint of what NASA (and DOE
and DOD) can dream up as tasks, and how they and the associated
contractors approach them. This particular example looks like a
materials science track (?). The thread originally dealt with a
biological application. I heard a talk from a person at Carnegie on
his experience with designing an automated epifluorescence microscope
intended for rover deployment. This thing was used to photograph
lichens in the Atacama desert. There are no lichens (or fungi or
algae) on the surface of Mars - MAYBE some fossilized bacteria or, if one
is really really lucky, some bacterial spores (viability undetermined)
near the poles. So, the only way to hunt for these things is to do
geological thin sections that can be examined using high resolution light
microscopy (EM doesn't work, but Raman has!), then have an AI system (or
real-time interpretation by a living scientist) to recognize things as
bacteria - things that are not terribly different than abundant
abiological geological "artifacts".
I would not be surprised to find out that a system like this has at least
been proposed to NASA, if not gone to feasibility tests. Anyway,
that's the view from a professional pessimist......
PS - for those interested, a long-standing research route at NASA and ESA
has been to look for bacteria in space (collection of particles and
subsequent analysis/growth epxts.
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Ok, here
goes. I apologize in advance for the length; it's just a sign of
how committed I was to the project.
Let me first say that this project (the Light Microscopy Module, LMM) was
an incredible experience for me, and I learned an amazing amount of stuff
from some truly world-class engineers and technicians. In my
opinion, the failure of this project was due to the fear of NASA and
contractor managers to make timely and unpopular decisions. At it's
peak, the LMM contractor team numbered about 50, there was another 10
NASA technical folks and the Principle Investigator (PI) teams added
another 5 or 10 post-docs and grad students.
Here's some background: around 1998, NASA approved a group of experiments
involving colloids and fluids for spaceflight on the International Space
Station. The first step in the process is for NASA to write a set
of "requirements" for each experiment. A
"requirement" is some capability that the flight instrument
must have in order to produce valid and useful scientific results.
For example, a requirement may be "obtain 30 images per
second". How these requirements come about is an interesting
story in itself, but in the end, NASA provides a list of requirements as
part of a contract bid- contractors look at the requirements list and
come up with a proposal to build *something that meets the requirements*
at a certain cost. NASA then awards the contract to whomever it
chooses.
I was hired by the contractor in late 1999, when a contract was being
generated to fly a microscope. That is to say, the contractor management
convinced NASA management that a microscope was *something that met the
requirements* for 4 flight experiments. Three of the experiments
involved imaging and manipulating colloidal emulsions, while the fourth
was a fluid heat transfer experiment, and the need was to image the
wetting boundary. There were approximately 300 pages (single
spaced) of requirements. At the time, none of us knew anything
about microscopes specifically.
To summarize the early planning stage, what was needed was a completely
automated, tricked-out microscope. Not motorized, but
*automated*. Once we invented this particular instrument, quite a
bit of excitement was generated throughout NASA, for obvious
reasons. It's cool, for one thing. Really sci-fi. The project
rapidly blew up , attracting the NASA bio community as well. Lots
of color drawings were made and distributed around. Websites were
created.
At the time, we knew of no imaging system that will *automatically* focus
on an unknown target within any sort of reasonable timeframe, and
maintain focus as the sample is moved around. We knew of no
microscope system that will optimize (and automatically adjust) DIC shear
for contrast. We knew of no microscope system that will
automatically align the phase rings in a phase contrast
system. We knew of no imaging system that would automatically
establish Kohler illumination. I still am not aware of any
microscope that will do any of this. If I am wrong, I would certainly
like to know about it. When we did the project, we had to motorize
the condenser turret and components, the epi- DIC prism, polarizer, and
waveplate, the tube lens turret and Bertrand lens, the viewing prisms,
all kinds of stuff. And any existing motors that did not meet NASA
specs had to be removed and replaced, the gears de-greased and re-greased
with NASA approved grease, the chassis de-painted, bolts changed, wiring
re-done, integrated circuits had to be replaced with radiation-hardened
components. I've disassembled and reassembled irises, stage slider
bearings, and entire microscopes. Irises are the worst, by far.
We obtained from the manufacturer complete blueprints for the microscope
and lenses. Complete lens specifications and drawings. Let me
tell you- those high NA immersion apochromats are absolute
marvels.
Here's how I would characterize the design phase: Imagine you work for
GM, and GM wants a spiffy new car. You head the team that is in charge of
the engine. LMM is the engine- the whole car is a larger structure that
supplies the LMM with electricity, coolant, saves data, etc. etc.
So you are to design an engine to certain specifications- torque,
horsepower, compression ratio, whatever. Problem: you don't know how much
room you have in the hood. You don't know where the transmission or
exhaust will connect. You don't know where the electrical
connections are. You don't know anything except the engine
requirements. Your first step may be to decide on the piston
arrangement- how many? inline? V? rotary? In any case, off you
go. And as problems arise, you work to solve them. There will be
meetings with the electrical team to decide where the alternator belt
will go. Meetings with the fuel supply group to determine how much gas is
needed. This goes along reasonably well until someone claims that
the car will not meet "car specifications"- maybe not enough
miles per gallon. Who's fault is that- the transmission team or the
engine team? Maybe it's the body design- too much drag or weight? Design
teams now compete to show the other team is at fault. Money and time is
wasted trying to meet unrealistic and arbitrary requirements.
Now we come to the safety concerns- astronauts are rare and delicate
flowers, after all- touch temperatures (no convection to draw off heat)
have to be below a certain value, the equipment must withstand "kick
loads"- astronauts may kick off the microscope to propel themselves
(!), anything breakable (including the samples and arc bulbs) has to be
enclosed 3x.... chemicals, etc. There were "leak paths"
that had to be plugged- the objective turret head was re-designed, we
thought about enclosing the immersion objectives with latex at some
point, it's insane what has to be done in the name of 'safety'.
There was a 200+ page document with a list of safety concerns: for
example, we calculated at most 2000 ml of immersion oil was required over
the life of the 3 colloid experiments- so one question was "what
happens if all 2000 ml is dispensed at once?" Each concern
required at least 3 pages of reply- a plan, a backup plan, a backup to
the backup, and then a test, validation and verification procedure to
ensure that one of the three plans would work.
Then the microscope is launched (up to 9 g's) in pieces and assembled
on-orbit, without gravity loading things or keeping them in place.
This, after sitting in Florida in a non-environmentally controlled crate
for several months. And the space station is incredibly
(mechanically) noisy. And there are huge thermal swings, some from
the microscope and light sources- and the fans we needed to move the heat
off add to the vibrational environment. This is why the microscope
needed to be able to align the optical elements.
Remember, there is no user interface- it's all done by software. We
had no access to "diagnostic" images to determine if there is a
problem. There was no astronaut interface- no eyepiece, no computer
screen to view what was being acquired. Again, imagine writing down
instructions on how to set up a microscope from the box, giving the
instructions to an undergraduate who has never done that before, and
having the expectation that everything will work perfectly the first time
(because you only get one chance to run the experiment- once the
experiment is over, all you get is a stack of data tapes with all your
images when the shuttle is able to bring them down). In
reality, it's even worse than that- you don't get to write the
instructions, someone unfamiliar with how microscopes work writes down
the instructions, occasionally asking for your input.
Confocal was (surprisingly) the easy one- we used a Yokogawa spinning
disk- incredibly rugged, and the issue of "finding focus" is
eliminated. Confocal imaging would have met the overwhelming
majority of imaging requirements for the colloid experiments.
Oil immersion is do-able, actually- the oil preferentially clings to the
glass and objective, and we found some non-wetting coatings we could
apply to control where the oil went. The trick is dispensing the
oil without bubbles. We proved this out by flying a microscope on
the KC-135 "vomit comet". That was awesome.
And the data- we estimated 1 TB of information was going to be generated
*per day* (the requirements were 30 frames/sec, 1k x 1k 12-bit images)
for 12 months. Where does the data go? Downlinking provided about 5
MB at best per day, IIRC.
Here's an example of what is right about NASA technical folks- one of the
PIs wanted to fly about 1500 samples. How on earth could we
accommodate this? The solution was wafer-scale integration of sample
cells- a glass disk 1 mm thick, just like a huge microscope slide, was
bonded to a silicon wafer, which was then ground down to 100 microns
thickness, so we could image the whole thickness. The silicon was
etched in the pattern we needed: rows and columns of holes shaped like
'=0=', each hole/cell holding about 50 microliters of colloid. Each hole
got a small magnetic stir bar, a fraction of a mm long, so each sample
could be freshly mixed prior to imaging. Then, a glass plate was
bonded to the top and ground down to 150 microns thick: a # 1 1/2
coverslip. The silicon provided an excellent reflective surface- i.e.
something to focus on, and we could pack in about 300 cells per sample
tray. We knew the exact position of everything, and the exact
thickness of everything, so we solved quite a few problems in one fell
swoop. Of course, there is then the problem of filling and sealing
1500 50-microliter cells (each with different compositions- and the
composition had to be verified somehow) that have to maintain their
integrity for months, but we put that back onto the PI.
Anything that goes up in space is using 10-year old technology at
best. For gears and fans, that's not a big deal. But it is
for cameras. And data interlinks. And software. Again, going back
to the car analogy, once the piston layout is designed, a lot of things
are constrained. And as the project moves along, it's harder and harder
to change the piston layout. Once a particular microscope was
selected, we could not go back and even update to a newer model- we
bought 5 sets of everything to ensure we had spare parts.
In the end, the NASA expectations (as sold to the PIs) were totally
unrealistic, and the budget was likewise ridiculous. NASA
management honestly believed that we could simply buy what we needed,
that there were no problems to be solved. There was no money for
testing and evaluation, among other things. I really thought, until
I left 5 years later, that we could deliver something that would meet
about 80% of the requirements. Unfortunately, management's only
input was to periodically yell "failure is not an option!" and
so we spent 80% of our time on the least important 20%.
At 10:42 AM 8/24/2007, you wrote:
Search the CONFOCAL archive
at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
I agree, this is an excellent thread…
Alice Rodriguez, Ph.D.
Duke University
Rm 4331 French Family Science Center
Science Drive
Dept of Biology/DCMB Box 90338
Durham NC 27708
Mobile: 919 451-4682
From: Confocal Microscopy List [
[hidden email]] On Behalf Of Evelyn
Ralston
Sent: Friday, August 24, 2007 11:30 AM
To: [hidden email]
Subject: Re: A microscope on Mars?
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Please don't. It may not be what we think about every day but IMHO it's a
very interesting thread.
Evelyn Ralston, Ph.D.
Head, Light imaging Section,
Office of Science and Technology, NIAMS, NIH
Rm 1535, Bldg 50
Bethesda MD 20892-8023
tel 301-496-6164; FAX 301-402-2724
On Aug 24, 2007, at 11:54 AM, Andrew Resnick wrote:
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I don't want
to bore everyone on this list, so I'll take this offline.
Andy
At 08:50 AM 8/24/2007, you wrote:
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Well, automating oil immersion in zero gravity doesn't really
bear thinking about! But automated microscopes (non-immersion)
in which you simply 'post' in a slide and look at the images on
a screen are available from several manufacturers - wouldn't
one of those have been a good starting point? So far as I know
none are confocal but implementing confocal and spectrophotometry
would not be hard - laser tweezers might be more tricky. DIC,
phase
darkfield etc would be trivial.
Guy
From: Confocal Microscopy List on behalf of Andrew Resnick
Sent: Fri 24/08/2007 11:29 PM
To:
[hidden email]
Subject: Re: A microscope on Mars?
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal My prior job
was with a NASA contractor, our project was to put a full-featured
microscope (brightfield, darkfield, phase contrast, DIC, confocal, laser
tweezers, spectrophotometry, interferometry, oil immersion) onto the
space station for (initially) 4 condensed-matter experiments, and later
for biology. Because of safety regulations, the microscope could
not be operated by astronauts- they were to simply load a sample, close
the door, and the machine would do its thing. You cannot imagine how
difficult it is to automate a microscope- the oil immersion alone was a
nightmare. In the end, we were unable to pull it off.
There are some compact space-based microscopes (I think ESA has a phase
contrast scope), but AFAIK, landers typically have chemical-based
detectors as they are more sensitive and can be made more robust to the
rigors of launch and landing.
At 09:03 PM 8/23/2007, you wrote:
Search the CONFOCAL archive at
<a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal A thought
just occurred to me while reading a CNN news article about the planet
Mars: Why not send an automated microscope with a robotic lander to look
for life there in the soil? Surely more sophisticated instruments have
been sent to Mars. Has NASA or any other space agency ever considered
this kind of approach? Just curious.
John Oreopoulos, BSc,
PhD Candidate
University of Toronto
Institute For Biomaterials and Biomedical Engineering
Centre For Studies in Molecular Imaging
Tel: W:416-946-5022
Andrew Resnick, Ph. D.
Instructor
Department of Physiology and Biophysics
Case Western Reserve University
216-368-6899 (V)
216-368-4223 (F)
Andrew Resnick, Ph. D.
Instructor
Department of Physiology and Biophysics
Case Western Reserve University
216-368-6899 (V)
216-368-4223 (F)
Andrew Resnick, Ph. D.
Instructor
Department of Physiology and Biophysics
Case Western Reserve University
216-368-6899 (V)
216-368-4223 (F)
--
Robert J. Palmer Jr.,
Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-594-0025
fax 301-402-0396
Andrew Resnick, Ph. D.
Instructor
Department of Physiology and Biophysics
Case Western Reserve University
216-368-6899 (V)
216-368-4223 (F)
|