Hi, could anyone be kind enough to provide some advice about photonconversion of Dendra2 by two photon laser? We tried the 810nm laser according to a paper (http://cshprotocols.cshlp.org/content/2011/10/pdb.top065904.full), but still have not succeeded yet. Anything tricky about it? Thanks.
Aget
Institute of Physiology II,University of Tübingen
Keplerstraße 15,
72074, Tübingen, Germany
|
|
Manja Schubert |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Aget, I was very surprised by the findings of the mentioned paper because I have also tried to switch dendra2 with a Ti:Saph laser a couple years ago, but I never managed to do so in the whole range of the laser (690 - 1040nm). After contacting the developer of Dendra2 (http://www.nature.com/nprot/journal/v2/n8/full/nprot.2007.291.html) and discussing this issue we concluded it is maybe not possible to do so with an IR laser. They suggested to us a 405 nm laser if available or exposing it to violet light for 5 min. Because two photon microscopy is a non-linear process the excitation spectra can show some unexpected shifts, f.e. Alexa 488 = 750nm 2P or eGFP = 950nm 2P. I found the following home pages always helpful for a good starting point if you use a fluorescence dye at 2P. (http://www.drbio.cornell.edu/cross_sections.html ; http://www.spectra.arizona.edu/) . If I right recall I think Kaede can be converted with a Ti:Saph laser at 750nm (http://adsabs.harvard.edu/abs/2008SPIE.6860E..26W). You could try this wavelength for Dendra2 too. Good luck. Manja On 16/07/2012 18:54, aget wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, could anyone be kind enough to provide some advice about photonconversion > of Dendra2 by two photon laser? We tried the 810nm laser according to a > paper (http://cshprotocols.cshlp.org/content/2011/10/pdb.top065904.full), > but still have not succeeded yet. Anything tricky about it? Thanks. > Aget > > -- > View this message in context: http://confocal-microscopy-list.588098.n2.nabble.com/2photon-conversion-of-Dendra2-tp7578658.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. -- Dr. Manja Schubert University of Bergen Department of Clinical Medicine Paraneoplastic Neurological Syndromes Group Haukeland University Hospital 5020 Bergen Norway Tel:+47-55 58 67 15 Fax: + 47-55 58 63 60 |
|
|
Thank you so much Manja, the websites are really helpful! I also wrote to the developer of Dendra2 who politely replied that they dont have experience with 2p conversion. Then i wrote to the authors of the paper describing 2p conversion by 810nm but got no reply so far (more than a month). It seems IR laser is not good at doing this job. Maybe i can try Kaede which can also be used to track cells.
Institute of Physiology II,University of Tübingen
Keplerstraße 15,
72074, Tübingen, Germany
|
|
|
Good news, the author of the paper i provided the link, Dr.John Condeelis replied, he said 'We have found the best multiphoton photoconversion to be at 30mW of 815nm for 300s using a parked beam'. Now i can try this, though i wonder how can i park the beam to the whole cell(?).
Institute of Physiology II,University of Tübingen
Keplerstraße 15,
72074, Tübingen, Germany
|
|
|
In reply to this post by aget
Hi Aget,
In our hands PAGFP and PSCFP2 are by far the best at two photon photoswitching. We have struggled to get Dendra, KikGR, and mEOS to work well, though given the right conditions (low power, long photoactivation times) perhaps things would work better. I haven't tried Kaede, but it appears that the chromophore sequence is similar to the others I mentioned. Jay Unruh Research Advisor Stowers Institute for Medical Research -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of aget Sent: Tuesday, July 17, 2012 4:46 AM To: [hidden email] Subject: Re: 2photon conversion of Dendra2 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Thank you so much Manja, the websites are really helpful! I also wrote to the developer of Dendra2 who politely replied that they dont have experience with 2p conversion. Then i wrote to the authors of the paper describing 2p conversion by 810nm but got no reply so far (more than a month). It seems IR laser is not good at doing this job. Maybe i can try Kaede which can also be used to track cells. ----- Institute of Physiology II,University of Tübingen Keplerstraße 15, 72074, Tübingen, Germany -- View this message in context: http://confocal-microscopy-list.588098.n2.nabble.com/2photon-conversion-of-Dendra2-tp7578658p7578668.html Sent from the Confocal Microscopy List mailing list archive at Nabble.com. |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** If you really need to, you can frequency-double a Ti:Saph to get 405. I've done this for some spectrophotometry to get a highly tunable visible source. You fire the Ti:Saph beam through a crystal like BBO, then rotate the crystal until you see conversion. The resulting light will be half the wavelength your laser is tuned to. For instance to get 405, tune the Ti:Saph to 810nm. It's surprisingly simple to do. Craig On Tue, Jul 17, 2012 at 10:20 AM, Unruh, Jay <[hidden email]> wrote: > Hi Aget, > > In our hands PAGFP and PSCFP2 are by far the best at two photon > photoswitching. We have struggled to get Dendra, KikGR, and mEOS to work > well, though given the right conditions (low power, long photoactivation > times) perhaps things would work better. I haven't tried Kaede, but it > appears that the chromophore sequence is similar to the others I mentioned. > > Jay Unruh > Research Advisor > Stowers Institute for Medical Research > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of aget > Sent: Tuesday, July 17, 2012 4:46 AM > To: [hidden email] > Subject: Re: 2photon conversion of Dendra2 > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Thank you so much Manja, the websites are really helpful! I also wrote to > the developer of Dendra2 who politely replied that they dont have > experience with 2p conversion. Then i wrote to the authors of the paper > describing 2p conversion by 810nm but got no reply so far (more than a > month). It seems IR laser is not good at doing this job. Maybe i can try > Kaede which can also be used to track cells. > > ----- > Institute of Physiology II,University of Tübingen Keplerstraße 15, 72074, > Tübingen, Germany > -- > View this message in context: > http://confocal-microscopy-list.588098.n2.nabble.com/2photon-conversion-of-Dendra2-tp7578658p7578668.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. > |
|
|
In reply to this post by Unruh, Jay
Thanks Jay. i checked Kaede and realized that it might be an issue for its diffusion in host cells. Dendra2 still seems ideal mammalian cell tracker.
Institute of Physiology II,University of Tübingen
Keplerstraße 15,
72074, Tübingen, Germany
|
|
|
In reply to this post by Craig Brideau
That sounds a great idea, i will try it. Thanks Craig.
Institute of Physiology II,University of Tübingen
Keplerstraße 15,
72074, Tübingen, Germany
|
|
Jerry (Gerald) Sedgewick |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** July 30, 12:00pm (noon) Central Daylight Time (U.S.). For Photoshop Extended versions CS3 (version 10) - CS6 (version 13). To sign up, go to: www.Imagingandanalysis.com/web072012cm.html. You will learn to post-process movie files and things like putting movies side-by-side, adding arrows and other symbols, adding captions, and so on. Here is a list of what will be shown: - Open Image stacks as image sequences; or as a z- or time-series saved in the uncompressed AVI format - Maintain a log of what you did to movie files - Open DICOM series as movies - Add text, arrows, captions, etc. to movies - Filter movies all at once: De-Interlace, Sharpen, etc. - Place movies side-by-side; overlay movies to merge - Adjust tonal range for optimizing to output devices - Export completed movies for frame-by-frame "scientific" replay Cheers, Jerry -- Jerry (Gerald) Sedgewick Scientific Imaging & Image Analysis Consultant Technical Writer / Regulatory Consultant 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 http://www.imagingandanalysis.com http://www.quickphotoshop.com Author: “Scientific Imaging with Photoshop: Methods, Measurement and Output” |
|
Craig Brideau |
|
|
In reply to this post by aget
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Glad I could help! If you don't have a crystal, most laser crystal dealers can direct you to the best type and cut of crystal for what you want to do. I have a piece of BBO that works quite well and is easy to tune. Other crystals may be more efficient for 405nm production though, so ask around. Craig On Thu, Jul 19, 2012 at 8:11 AM, aget <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > That sounds a great idea, i will try it. Thanks Craig. > > ----- > Institute of Physiology II,University of Tübingen > Keplerstraße 15, > 72074, Tübingen, Germany > -- > View this message in context: > http://confocal-microscopy-list.588098.n2.nabble.com/2photon-conversion-of-Dendra2-tp7578658p7578695.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. > |
| Free forum by Nabble | Edit this page |