						Emma King-2 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We would like to take serial sections (20 microns thick) of rat spinal cord (up to 1cm long), label them with a bright and specific fluorescent label, acquire images (we can use either a widefield or confocal microscope) and then reconstruct the data in to one, 3D structure. We have the sectioning, staining and acquisition parameters/options under control, but don't have any software capable of reconstructing the resultant images. Does anyone have any suggestions as to what software to use and where to get it from? Any help much appreciated. Cheers, Emma Advanced Microscopy Unit School of Biomedical Sciences Univeristy of Nottingham  | 
			
 
	
	 | 
						Louis Villeneuve | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, In fact you want to add mutltiple Z-series together in order to form one huge serie. First, you will need a good computer and 2nd, I have done it using Huygens software (SVI). Also, I am pretty sure that there is a plugin In" Image J" that can do that. Do not forget to do that with a good computer!!!! Best regards, Louis Louis Villeneuve Associé de recherche - Microscopie confocale Institut de Cardiologie de Montreal- Centre de recherche 5000 est Bélanger Montreal (Qc), Canada H1T 3C8 514-376-3330 ext 3511 514-376-1355 (fax) Emma King <[hidden email]>@LISTS.UMN.EDU Envoyé par : Confocal Microscopy List <[hidden email]> 2011-11-18 09:39 Veuillez répondre à Confocal Microscopy List <[hidden email]> A [hidden email] cc Objet 3D reconstruction of multiple, fluorescently labelled, sections ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We would like to take serial sections (20 microns thick) of rat spinal cord (up to 1cm long), label them with a bright and specific fluorescent label, acquire images (we can use either a widefield or confocal microscope) and then reconstruct the data in to one, 3D structure. We have the sectioning, staining and acquisition parameters/options under control, but don't have any software capable of reconstructing the resultant images. Does anyone have any suggestions as to what software to use and where to get it from? Any help much appreciated. Cheers, Emma Advanced Microscopy Unit School of Biomedical Sciences Univeristy of Nottingham  | 
			
 
	
	 | 
						Louis Kerr | 
| 
					
		 | 
				
					
	 
				In reply to this post by Emma King-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, Have you seen the work from Stephen Smith's Lab at Stanford with array tomography? This is a link to a Cold Spring Harbor Protocol. http :// cshprotocols . cshlp .org/content/2010/11/ pdb .top89.full Good luck, Louie ----- Original Message ----- From: "Emma King" < emma .[hidden email]> To: CONFOCALMICROSCOPY @LISTS. UMN . EDU Sent: Friday, November 18, 2011 9:39:57 AM Subject: 3D reconstruction of multiple, fluorescently labelled , sections ***** To join, leave or search the confocal microscopy listserv , go to: http ://lists. umn . edu /cgi-bin/ wa ?A0= confocalmicroscopy ***** We would like to take serial sections (20 microns thick) of rat spinal cord (up to 1cm long), label them with a bright and specific fluorescent label, acquire images (we can use either a widefield or confocal microscope) and then reconstruct the data in to one, 3D structure. We have the sectioning, staining and acquisition parameters/options under control, but don't have any software capable of reconstructing the resultant images. Does anyone have any suggestions as to what software to use and where to get it from? Any help much appreciated. Cheers, Emma Advanced Microscopy Unit School of Biomedical Sciences Univeristy of Nottingham -- Louis Kerr lkerr @ mbl . edu Research and Education Support Coordinator Marine Biological Laboratory 7 MBL Street Woods Hole, MA 02543 508-289-7273 508-540-6902 (FAX) 508-292-0289 (Cell phone) VISIT OUR WEB SITES: www . mbl . edu www .courses. mbl . edu  | 
			
 
	
	 | 
						tineke vendrig | 
| 
					
		 | 
				
					
	 
				In reply to this post by Emma King-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We use the Andor software for our 3D movies, works nice and good! Tineke Vendrig -- Tineke Vendrig, ing technical engineer optical microscopy Delft University of Technology Bionano Science Kavli Institute of Nanoscience Lorentzweg 1 2628LJ Delft room F185 Tel: +31 15 2789299 Fax:+31 15 2781202 email: [hidden email] mobile phone: +31 624341412 2011/11/18 Emma King <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > We would like to take serial sections (20 microns thick) of rat spinal > cord (up to 1cm long), label them with a bright and specific fluorescent > label, acquire images (we can use either a widefield or confocal > microscope) and then reconstruct the data in to one, 3D structure. > > We have the sectioning, staining and acquisition parameters/options > under control, but don't have any software capable of reconstructing > the resultant images. Does anyone have any suggestions as to what > software to use and where to get it from? > > Any help much appreciated. > > Cheers, > Emma > > Advanced Microscopy Unit > School of Biomedical Sciences > Univeristy of Nottingham >  | 
			
 
	
	 | 
						Johannes Schindelin | 
| 
					
		 | 
				
					
	 
				In reply to this post by Louis Villeneuve
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On Fri, 18 Nov 2011, [hidden email] wrote: > Emma King <[hidden email]>@LISTS.UMN.EDU > Envoyé par : Confocal Microscopy List <[hidden email]> > 2011-11-18 09:39 > > > We would like to take serial sections (20 microns thick) of rat spinal > > cord (up to 1cm long), label them with a bright and specific > > fluorescent label, acquire images (we can use either a widefield or > > confocal microscope) and then reconstruct the data in to one, 3D > > structure. > > > We have the sectioning, staining and acquisition parameters/options > > under control, but don't have any software capable of reconstructing > > the resultant images. Does anyone have any suggestions as to what > > software to use and where to get it from? > > In fact you want to add mutltiple Z-series together in order to form one > huge serie. First, you will need a good computer and 2nd, I have done > it using Huygens software (SVI). > > Also, I am pretty sure that there is a plugin In" Image J" that can do > that. Do not forget to do that with a good computer!!!! plugins to do the job: Volume Viewer (non-accelerated) and 3D Viewer (this one uses your graphics adapter's accelerated 3D rendering). However, if you need to have surface models, you might want to look into the TrakEM2 plugin. It is very powerful, therefore it takes some training before one can exploit its capabilities in full. If you do not want to bother to find all the bits and pieces required for these plugins to run, please feel free to download Fiji (http://fiji.sc/) which bundles ImageJ with a lot of plugins (including the 3 mentioned above). Ciao, Johannes  | 
			
 
	
	 | 
						Armstrong, Brian | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You could try Reconstruct from NIH Human Brain project (http://synapses.clm.utexas.edu/tools/reconstruct/reconstruct.stm ). In my opinion, Amira, Imaris, and Volocity, are probably best at 3D visualization tools. Cheers, Brian D Armstrong PhD Assistant Research Professor Light Microscopy Core Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Johannes Schindelin Sent: Friday, November 18, 2011 10:19 AM To: [hidden email] Subject: Re: RE 3D reconstruction of multiple, fluorescently labelled, sections ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On Fri, 18 Nov 2011, [hidden email] wrote: > Emma King <[hidden email]>@LISTS.UMN.EDU > Envoyé par : Confocal Microscopy List <[hidden email]> > 2011-11-18 09:39 > > > We would like to take serial sections (20 microns thick) of rat spinal > > cord (up to 1cm long), label them with a bright and specific > > fluorescent label, acquire images (we can use either a widefield or > > confocal microscope) and then reconstruct the data in to one, 3D > > structure. > > > We have the sectioning, staining and acquisition parameters/options > > under control, but don't have any software capable of reconstructing > > the resultant images. Does anyone have any suggestions as to what > > software to use and where to get it from? > > In fact you want to add mutltiple Z-series together in order to form one > huge serie. First, you will need a good computer and 2nd, I have done > it using Huygens software (SVI). > > Also, I am pretty sure that there is a plugin In" Image J" that can do > that. Do not forget to do that with a good computer!!!! If all you want to do is a volume rendering, there are indeed 2 ImageJ plugins to do the job: Volume Viewer (non-accelerated) and 3D Viewer (this one uses your graphics adapter's accelerated 3D rendering). However, if you need to have surface models, you might want to look into the TrakEM2 plugin. It is very powerful, therefore it takes some training before one can exploit its capabilities in full. If you do not want to bother to find all the bits and pieces required for these plugins to run, please feel free to download Fiji (http://fiji.sc/) which bundles ImageJ with a lot of plugins (including the 3 mentioned above). Ciao, Johannes --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. ---------------------------------------------------------------------  | 
			
 
	
	 | 
						Graham Wright | 
| 
					
		 | 
				
					
	 
				In reply to this post by Emma King-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, Sorry for contributing a bit late to this discussion or if somebody has already mentioned this (I didn't see it if so). If you make the 'huge series' using ImageJ/Fiji (in Fiji through Image>Stacks>Manipulation) you can then try to use the StackReg plugin from EPFL (http://bigwww.epfl.ch/thevenaz/stackreg/) which requires the turboreg plugin, linked from the same site, to align the stack (these are both built into Fiji in Plugins>Registration). Follow this with some cropping and it might work out for further use in any of the 3D visualisation programs discussed. If each image shares enough overlap/similarity with the previous it should work. Good luck, Graham -- IMB Microscopy Unit Institute of Medical Biology Singapore http://www.imb.a-star.edu.sg/imu/index.html On 18 November 2011 22:39, Emma King <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > We would like to take serial sections (20 microns thick) of rat spinal > cord (up to 1cm long), label them with a bright and specific fluorescent > label, acquire images (we can use either a widefield or confocal > microscope) and then reconstruct the data in to one, 3D structure. > > We have the sectioning, staining and acquisition parameters/options > under control, but don't have any software capable of reconstructing > the resultant images. Does anyone have any suggestions as to what > software to use and where to get it from? > > Any help much appreciated. > > Cheers, > Emma > > Advanced Microscopy Unit > School of Biomedical Sciences > Univeristy of Nottingham >  | 
			
 
	
	 | 
						phil laissue | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma I am very happy with Nikon's NIS-Elements software, but just wanted to second Graham Wright's & Louis Kerr's input - the stackreg algorithm works a charm. I often use its modified version MultiStackReg, aligning one channel and applying the same translation to the other(s). Very useful in case of moving cells, stage drift etc. http://www.stanford.edu/~bbusse/work/downloads.html Speaking of which: Many thanks to Thierry Thévenaz & Michael Unser and Brad Busse & Steven Smith for these fabulous open-source tools! Philippe Philippe Laissue, PhD Bioimaging Facility, University of Essex, UK  | 
			
 
	
	 | 
						Dmitry Sokolov | 
| 
					
		 | 
				
					
	 
				In reply to this post by Graham Wright
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, StackReg plugin did not work for me on a very similar task a couple of years ago. The slices were found not only rotated on the slides but also deformed at cutting. Automatic alignment without absolute reference points or sites can be problematic if not impossible due to the distortions especially. My experience on the subject is summarized here: http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Alignment I would appreciate to hear a better solution if available. Please feel free to modify the MIAWiki and contact me offline when required. With best wishes, Dmitry MIAWiki for Mass Collaboration ------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------ Dr. Dmitry Sokolov *LYNX Analytics: * Image Analysis - Data Processing - Knowledge Transfer /Scanning Confocal / Atomic Force Microscopy & Spectroscopy/ Mob: +64(21)063-5382 ------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------ On 21/11/2011 10:58 p.m., Graham Wright wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Emma, > > Sorry for contributing a bit late to this discussion or if somebody has > already mentioned this (I didn't see it if so). > > If you make the 'huge series' using ImageJ/Fiji (in Fiji through > Image>Stacks>Manipulation) you can then try to use the StackReg plugin from > EPFL (http://bigwww.epfl.ch/thevenaz/stackreg/) which requires the turboreg > plugin, linked from the same site, to align the stack (these are both built > into Fiji in Plugins>Registration). Follow this with some cropping and it > might work out for further use in any of the 3D visualisation programs > discussed. > > If each image shares enough overlap/similarity with the previous it should > work. > > Good luck, > Graham > > -- > IMB Microscopy Unit > Institute of Medical Biology > Singapore > http://www.imb.a-star.edu.sg/imu/index.html > > On 18 November 2011 22:39, Emma King<[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> We would like to take serial sections (20 microns thick) of rat spinal >> cord (up to 1cm long), label them with a bright and specific fluorescent >> label, acquire images (we can use either a widefield or confocal >> microscope) and then reconstruct the data in to one, 3D structure. >> >> We have the sectioning, staining and acquisition parameters/options >> under control, but don't have any software capable of reconstructing >> the resultant images. Does anyone have any suggestions as to what >> software to use and where to get it from? >> >> Any help much appreciated. >> >> Cheers, >> Emma >> >> Advanced Microscopy Unit >> School of Biomedical Sciences >> Univeristy of Nottingham >>  | 
			
 
	
	 | 
						Daniel James White | 
| 
					
		 | 
				
					
	 
				In reply to this post by Emma King-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, > Date: Mon, 21 Nov 2011 09:18:27 +0000 > From: Emma King <[hidden email]> > Subject: Re: CONFOCALMICROSCOPY Digest - 17 Nov 2011 to 18 Nov 2011 (#2011-88) > Hi All, > > Thanks for all your messages. Louis described much better than I did what w= > e plan to do "In fact you want to add multiple Z-series together in order t= > o form one huge series"! OK, then we also have tools for that kind of thing in Fiji. > > We have Volocity (and obviously imageJ) to visualise 3D data and I believe= > we can add z-series together, on top of each other. My worry is we will ha= > ve issues aligning structures in one z-stack, from one tissue section, with= > the same structures in the next z-stack/tissue section. Your worries are well founded, and possibly already solved. See http://fiji.sc/TrakEM2_align_sections_tutorial http://bigwww.epfl.ch/thevenaz/stackreg/ and other tools in Fiji. There is nothing that will just work out of the box, but rather it wil take some figuring out, and possibly the writing of at least a macro to automate the many mouse clicks. If there is overlapping information in the top optical section of one stack and the bottom ptical section of the next z stack from the next physical section, then it might be possible to get the 3D stitching plugin to glue the z stacks together correctly. http://fiji.sc/Stitching_2D/3D http://fiji.sc/wiki/index.php/Stitching_2D/3D#2D_Stitching_and_3D_Stitching (might need to "rotate" the z stacks first so that the z becomes x for this plugin to work... or maybe not) cheers Dan > > I'll give it a go using the suggestions made and come back to you if I get = > stuck! > > Thanks again, > Em Dr. Daniel James White BSc. (Hons.) PhD Leader - Image Processing Facility, Senior Microscopist, Light Microscopy Facility. Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstrasse 108 01307 DRESDEN Germany +49 (0)15114966933 (German Mobile) +49 (0)351 210 2627 (Work phone at MPI-CBG) +49 (0)351 210 1078 (Fax MPI-CBG LMF) chalkie666 Skype http://www.bioimagexd.net BioImageXD http://fiji.sc Fiji - is just ImageJ (Batteries Included) http://www.chalkie.org.uk Dan's Homepages https://ifn.mpi-cbg.de Biopolis Dresden Imaging Platform (BioDIP) dan (at) chalkie.org.uk ( white (at) mpi-cbg.de )  | 
			
| 
					
		 | 
				
					
	
	 
		Hi
 
				Just something to add: Personally I use Fiji for this and registration> 'register virtual stack slices' is better that stackreg for large stacks and can align one colour by another etc. But I was going to say that with multiple sections like yours if you are not after quantitative flourescence then it is really good to float the image around a mean and have a common standard deviation of intensities. If you want to do this you can write a macro (or ask me) in FIJI or use IMOD software (which is definitely better on files too large to hold in the ram). If you are having trouble aligning but want a quantitative result then float the images align them but use the output alignment files to then align the original. Regards Kenton  | 
			
 
	
	 | 
						phil laissue | 
| 
					
		 | 
				
					
	 
				In reply to this post by Dmitry Sokolov
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Tissue deformations due to cutting are tricky, and usually require pretty heavy affine transformation (unless you photograph the block face, which can't be used in this case). Derek Magee presented an elegant solution a few years ago: Magee D., Treanor D., and Quirke P., A New Image Registration algorithm with application to 3D Histopathology *Proc. Microscopic Image Analysis with Applications in Biology (MICCAI Workshop)*, 2008 I'm not sure about the availability of the algorithm, but you could contact him directly via http://www.comp.leeds.ac.uk/drm/ (where you'll also find the pdf of said paper). There's many other approaches (e.g. Eric Bardinet et al.'s block matching or Amelia Cifor/Tony Pridmore/Alain Pitiot's work), but again, I don't know about algorithm availability - certainly no ImageJ plugins.  | 
			
 
	
	 | 
						Daniel James White | 
| 
					
		 | 
				
					
	 
				In reply to this post by Emma King-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emma, Hi Dimitry, On Nov 25, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system wrote: > Date: Tue, 22 Nov 2011 07:57:48 +1300 > From: Dmitry Sokolov <[hidden email]> > Subject: Re: 3D reconstruction of multiple, fluorescently labelled, sections > > Hi Emma, > > StackReg plugin did not work for me on a very similar task a couple of ye= > ars ago. The slices were found not only rotated on the slides but also de= > formed at cutting. Automatic alignment without absolute reference points = > or sites can be problematic if not impossible due to the distortions espe= > cially. > My experience on the subject is summarized here: > http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Ali= > gnment > > I would appreciate to hear a better solution if available. This problem has been solved for electron microscopy, and the solution is appicable to fluorescence microscopy. The software solutions are available in Fiji already, and are the work of the Tomancak lab here at MPI-CBG. see info here: http://fiji.sc/wiki/index.php/Elastic_Alignment_and_Montage what you get is : Elastic registration of the image slices, so in the reconstruction objects dont joggle and jitter as you go through the z stack. here we have multiple z stack that need to be aligned, but the solutions in this software can be adapted to deal with that. So long as there is overlapping information between the ends of the stacks, all is well. cheers Dan > > Please feel free to modify the MIAWiki and contact me offline when requir= > ed. > > With best wishes, > Dmitry > MIAWiki for Mass Collaboration Dr. Daniel James White BSc. (Hons.) PhD Leader - Image Processing Facility, Senior Microscopist, Light Microscopy Facility. Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstrasse 108 01307 DRESDEN Germany +49 (0)15114966933 (German Mobile) +49 (0)351 210 2627 (Work phone at MPI-CBG) +49 (0)351 210 1078 (Fax MPI-CBG LMF) chalkie666 Skype http://www.bioimagexd.net BioImageXD http://fiji.sc Fiji - is just ImageJ (Batteries Included) http://www.chalkie.org.uk Dan's Homepages https://ifn.mpi-cbg.de Biopolis Dresden Imaging Platform (BioDIP) dan (at) chalkie.org.uk ( white (at) mpi-cbg.de )  | 
			
 
	
	 | 
						Dmitry Sokolov | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Dan, thanks a lot. The procedure looks very good for trial. I regret now that information on the elastic alignment was added to MIAWiki a couple of months after it's actual need and is still of interest to the Community. I hope finding the topic will be quicker from now on through the MIAWiki interactive search bar and via the permalink to the page: http://confocal-manawatu.pbworks.com/w/page/40316645/Elastic%20Alignment%20of%20Images%20ImageJ The page is cross-linked with "Manual Stack Alignment" http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Alignment#view=page and a couple of other knowledge nodes in accordance to principles of Contextual Integration: http://confocal-manawatu.pbworks.com/w/page/47947288/Contextual%20Integration Thanks again, Dmitry MIAWiki for Mass Collaboration: Not Published = Lost Forever On 25/11/2011 10:04 p.m., Daniel James White wrote: > Hi Emma, Hi Dimitry, > > On Nov 25, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system wrote: > >> Date: Tue, 22 Nov 2011 07:57:48 +1300 >> From: Dmitry Sokolov<[hidden email]> >> Subject: Re: 3D reconstruction of multiple, fluorescently labelled, sections >> >> Hi Emma, >> >> StackReg plugin did not work for me on a very similar task a couple of ye= >> ars ago. The slices were found not only rotated on the slides but also de= >> formed at cutting. Automatic alignment without absolute reference points = >> or sites can be problematic if not impossible due to the distortions espe= >> cially. >> My experience on the subject is summarized here: >> http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Ali= >> gnment >> >> I would appreciate to hear a better solution if available. > This problem has been solved for electron microscopy, and the solution is appicable to fluorescence microscopy. > The software solutions are available in Fiji already, > and are the work of the Tomancak lab here at MPI-CBG. > > see info here: > > http://fiji.sc/wiki/index.php/Elastic_Alignment_and_Montage > > what you get is : > Elastic registration of the image slices, > so in the reconstruction objects dont joggle and jitter as you go through the z stack. > > here we have multiple z stack that need to be aligned, > but the solutions in this software can be adapted to deal with that. > > So long as there is overlapping information between the ends of the stacks, all is well. > > cheers > > Dan > > > > > >> Please feel free to modify the MIAWiki and contact me offline when requir= >> ed. >> >> With best wishes, >> Dmitry >> MIAWiki for Mass Collaboration > Dr. Daniel James White BSc. (Hons.) PhD > > Leader - Image Processing Facility, > Senior Microscopist, > Light Microscopy Facility. > > Max Planck Institute of Molecular Cell Biology and Genetics > Pfotenhauerstrasse 108 > 01307 DRESDEN > Germany > > +49 (0)15114966933 (German Mobile) > +49 (0)351 210 2627 (Work phone at MPI-CBG) > +49 (0)351 210 1078 (Fax MPI-CBG LMF) > chalkie666 Skype > http://www.bioimagexd.net BioImageXD > http://fiji.sc Fiji - is just ImageJ (Batteries Included) > http://www.chalkie.org.uk Dan's Homepages > https://ifn.mpi-cbg.de Biopolis Dresden Imaging Platform (BioDIP) > dan (at) chalkie.org.uk > ( white (at) mpi-cbg.de ) > > > > > > > > > > > > > > > > > > >  | 
			
 
	
	 | 
						Dmitry Sokolov | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, sorry for the typo: It's great to see that the topic is of interest to the Community, of course! :0) Have a nice weekend, Dmitry MIAWiki for Mass Collaboration: Not Published = Lost Forever On 26/11/2011 8:14 a.m., Dmitry Sokolov wrote: > Hi Dan, > > thanks a lot. The procedure looks very good for trial. I regret now that information on the elastic alignment was added to MIAWiki a couple of months after it's actual need and is still of interest to the Community. > I hope finding the topic will be quicker from now on through the MIAWiki interactive search bar and via the permalink to the page: > http://confocal-manawatu.pbworks.com/w/page/40316645/Elastic%20Alignment%20of%20Images%20ImageJ > The page is cross-linked with "Manual Stack Alignment" > http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Alignment#view=page > and a couple of other knowledge nodes in accordance to principles of Contextual Integration: > http://confocal-manawatu.pbworks.com/w/page/47947288/Contextual%20Integration > > Thanks again, > Dmitry > MIAWiki for Mass Collaboration: > Not Published = Lost Forever > > On 25/11/2011 10:04 p.m., Daniel James White wrote: >> Hi Emma, Hi Dimitry, >> >> On Nov 25, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system wrote: >> >>> Date: Tue, 22 Nov 2011 07:57:48 +1300 >>> From: Dmitry Sokolov<[hidden email]> >>> Subject: Re: 3D reconstruction of multiple, fluorescently labelled, sections >>> >>> Hi Emma, >>> >>> StackReg plugin did not work for me on a very similar task a couple of ye= >>> ars ago. The slices were found not only rotated on the slides but also de= >>> formed at cutting. Automatic alignment without absolute reference points = >>> or sites can be problematic if not impossible due to the distortions espe= >>> cially. >>> My experience on the subject is summarized here: >>> http://confocal-manawatu.pbworks.com/w/page/38604941/Manual%20Stack%20Ali= >>> gnment >>> >>> I would appreciate to hear a better solution if available. >> This problem has been solved for electron microscopy, and the solution is appicable to fluorescence microscopy. >> The software solutions are available in Fiji already, >> and are the work of the Tomancak lab here at MPI-CBG. >> >> see info here: >> >> http://fiji.sc/wiki/index.php/Elastic_Alignment_and_Montage >> >> what you get is : >> Elastic registration of the image slices, >> so in the reconstruction objects dont joggle and jitter as you go through the z stack. >> >> here we have multiple z stack that need to be aligned, >> but the solutions in this software can be adapted to deal with that. >> >> So long as there is overlapping information between the ends of the stacks, all is well. >> >> cheers >> >> Dan >> >> >> >> >> >>> Please feel free to modify the MIAWiki and contact me offline when requir= >>> ed. >>> >>> With best wishes, >>> Dmitry >>> MIAWiki for Mass Collaboration >> Dr. Daniel James White BSc. (Hons.) PhD >> >> Leader - Image Processing Facility, >> Senior Microscopist, >> Light Microscopy Facility. >> >> Max Planck Institute of Molecular Cell Biology and Genetics >> Pfotenhauerstrasse 108 >> 01307 DRESDEN >> Germany >> >> +49 (0)15114966933 (German Mobile) >> +49 (0)351 210 2627 (Work phone at MPI-CBG) >> +49 (0)351 210 1078 (Fax MPI-CBG LMF) >> chalkie666 Skype >> http://www.bioimagexd.net BioImageXD >> http://fiji.sc Fiji - is just ImageJ (Batteries Included) >> http://www.chalkie.org.uk Dan's Homepages >> https://ifn.mpi-cbg.de Biopolis Dresden Imaging Platform (BioDIP) >> dan (at) chalkie.org.uk >> ( white (at) mpi-cbg.de ) >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> > >  | 
			