405, 488, 561 and 647nm lasers

classic Classic list List threaded Threaded
♦
♦
6 messages Options
Mayandi Sivaguru Mayandi Sivaguru
Reply | Threaded
Open this post in threaded view
♦
♦
|

405, 488, 561 and 647nm lasers

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 405, 488, 561 and 647nm lasers

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Diode lasers can mostly meet those power requirements these days; the problem is the 'economical' part.
There are so-called 'white' lasers that can reach all those wavelength ranges, but after you slice out the part of the spectrum you want to use you probably won't have 100mW left.
It's a tricky problem!  Really the only way I know of to meet these parameters off-hand is to just buy the individual lasers, but you're looking at ~$10k each on average for these lines, possibly more.

Craig


On Fri, Sep 12, 2008 at 12:33 PM, Mayandi Sivaguru <[hidden email]> wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu


Kevin Braeckmans Kevin Braeckmans
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 405, 488, 561 and 647nm lasers

In reply to this post by Mayandi Sivaguru
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

 

You could check out the solid state lasers from Cobolt (www.cobolt.se): up to 100 mW for 491 nm, up to 75 mW for 561 nm. They are much less expensive compared to other manufacturers (e.g. CrystaLaser), at least in Europe two years ago when we purchased two of them. We use them in a widefield set-up for single particle tracking.

 

I have never heard of a 647 nm diode laser, though. There are 636 – 640 nm diode lasers, but as far as I know they are limited to about 30 mW. I would be interested if someone knows of more powerful solid state lasers in this wavelength range.

 

Best regards,

 

Kevin

 

 

 

Van: Confocal Microscopy List [mailto:[hidden email]] Namens Mayandi Sivaguru
Verzonden: vrijdag 12 september 2008 20:34
Aan: [hidden email]
Onderwerp: 405, 488, 561 and 647nm lasers

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal


Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

dc-4 dc-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 405, 488, 561 and 647nm lasers

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear All,

 

No commercial interest at all, but in response to the discussion on diode lasers, I have just noticed that there is a 642nm diode from Opnext, the Opnext HL6385DG.

 

The blurb from optics.org follows:

 

The Opnext HL6385DG is a single longitudinal mode, low aspect ratio laser diode with optical output power of 150mW CW (typ), at a lasing wavelength of 642nm, operating current of 280mA, threshold current of 110mA, operating voltage of 2.6V and LD reverse voltage of 2V. The operating temperature range is -10 degrees C to +40 degrees C.

Supplied in a 5.6mm package with LD-only pin configuration, this is a single longitudinal mode laser diode with TE mode oscillation. The 5.6mm package size is particularly suited to small, energy-efficient displays and low-power, high-quality optical applications.

Download specifications from Photonic Products’ website:
http://www.photonic-products.com/news/opnext-hl6385dg-laser-diode.html (this is the UK supplier)

How well this would suit I have no idea.

 

Opnext website is www.opnext.com

 

All the best

 

Darran Clements

www.babworth.com

Babworth: Insight and Analysis for the Internet, Communications and IT Industries

 

The information contained in this message is the confidential property of Babworth and may be privileged and protected from disclosure. If you are not the intended recipient, any use, dissemination, distribution or copying is strictly prohibited. If you have received this message in error, please e-mail the sender. 

Whilst we have taken reasonable precautions to ensure that this e-mail and any attachment has been checked for viruses, we cannot guarantee that they are virus free and we cannot accept liability for any damage sustained as a result of software viruses. We would advise that you carry out your own virus checks, especially before opening an attachment.

Babworth Ltd. Registered office: 35 Firs Avenue, London N11 3NE, UK. Registered in England and Wales. Registration No. 6620167.


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kevin Braeckmans
Sent: 16 September 2008 07:33
To: [hidden email]
Subject: Re: 405, 488, 561 and 647nm lasers

 

Hi,

 

You could check out the solid state lasers from Cobolt (www.cobolt.se): up to 100 mW for 491 nm, up to 75 mW for 561 nm. They are much less expensive compared to other manufacturers (e.g. CrystaLaser), at least in Europe two years ago when we purchased two of them. We use them in a widefield set-up for single particle tracking.

 

I have never heard of a 647 nm diode laser, though. There are 636 – 640 nm diode lasers, but as far as I know they are limited to about 30 mW. I would be interested if someone knows of more powerful solid state lasers in this wavelength range.

 

Best regards,

 

Kevin

 

 

 

Van: Confocal Microscopy List [mailto:[hidden email]] Namens Mayandi Sivaguru
Verzonden: vrijdag 12 september 2008 20:34
Aan: [hidden email]
Onderwerp: 405, 488, 561 and 647nm lasers

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal


Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

Stephen Cody Stephen Cody
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 405, 488, 561 and 647nm lasers

In reply to this post by Kevin Braeckmans
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear List,

 

Shiv just mentioned Cobalt lasers. So I thought I would use this opportunity to give this company some well deserved praise.

 

Awhile ago I decided to replace our aging Argon ion laser on our MRC-1000/1024 system. I ordered a Cobalt “Dual Calypso” Solid state laser that gives two lines 491nm and 532nm each line 20mW. After receiving the laser I needed to build a heat sink and order new dichroics etc. So the process took me awhile. When finally I had finished incorporating the laser into our home built laser combiner, I noticed there were some fluctuations in the laser intensity.

 

After a couple of quick emails to the company describing the problem, they sent out a data cable by post. When this arrived it was a simple matter to plug the laser power supply into a computer with network access. They were then able to gain remote access to the laser and an engineer conducted a full diagnostic and service remotely, and solved the problem!

 

It appeared that our heat sink was too efficient, and the laser could not reach it set temperature points. They increased the power output of the laser to 25 + 25mW and adjusted three separate temperature controller set points and the laser has performed beautifully ever since. That was really impressive service, and I certainly would recommend Cobalt.

 

No commercial interest in Cobalt.

 

Cheers Steve

Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research

PO Box 2008 Royal Melbourne Hospital
Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kevin Braeckmans
Sent: Tuesday, 16 September 2008 4:33 PM
To: [hidden email]
Subject: Re: 405, 488, 561 and 647nm lasers

 

Hi,

 

You could check out the solid state lasers from Cobolt (www.cobolt.se): up to 100 mW for 491 nm, up to 75 mW for 561 nm. They are much less expensive compared to other manufacturers (e.g. CrystaLaser), at least in Europe two years ago when we purchased two of them. We use them in a widefield set-up for single particle tracking.

 

I have never heard of a 647 nm diode laser, though. There are 636 – 640 nm diode lasers, but as far as I know they are limited to about 30 mW. I would be interested if someone knows of more powerful solid state lasers in this wavelength range.

 

Best regards,

 

Kevin

 

 

 

Van: Confocal Microscopy List [mailto:[hidden email]] Namens Mayandi Sivaguru
Verzonden: vrijdag 12 september 2008 20:34
Aan: [hidden email]
Onderwerp: 405, 488, 561 and 647nm lasers

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal


Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research.

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 405, 488, 561 and 647nm lasers

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Wow!  That is impressive service!  Thanks for the tip!

Craig


On Tue, Sep 16, 2008 at 6:24 PM, Stephen Cody <[hidden email]> wrote:

Dear List,

 

Shiv just mentioned Cobalt lasers. So I thought I would use this opportunity to give this company some well deserved praise.

 

Awhile ago I decided to replace our aging Argon ion laser on our MRC-1000/1024 system. I ordered a Cobalt "Dual Calypso" Solid state laser that gives two lines 491nm and 532nm each line 20mW. After receiving the laser I needed to build a heat sink and order new dichroics etc. So the process took me awhile. When finally I had finished incorporating the laser into our home built laser combiner, I noticed there were some fluctuations in the laser intensity.

 

After a couple of quick emails to the company describing the problem, they sent out a data cable by post. When this arrived it was a simple matter to plug the laser power supply into a computer with network access. They were then able to gain remote access to the laser and an engineer conducted a full diagnostic and service remotely, and solved the problem!

 

It appeared that our heat sink was too efficient, and the laser could not reach it set temperature points. They increased the power output of the laser to 25 + 25mW and adjusted three separate temperature controller set points and the laser has performed beautifully ever since. That was really impressive service, and I certainly would recommend Cobalt.

 

No commercial interest in Cobalt.

 

Cheers Steve

Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research

PO Box 2008 Royal Melbourne Hospital
Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kevin Braeckmans
Sent: Tuesday, 16 September 2008 4:33 PM


To: [hidden email]
Subject: Re: 405, 488, 561 and 647nm lasers

 

Hi,

 

You could check out the solid state lasers from Cobolt (www.cobolt.se): up to 100 mW for 491 nm, up to 75 mW for 561 nm. They are much less expensive compared to other manufacturers (e.g. CrystaLaser), at least in Europe two years ago when we purchased two of them. We use them in a widefield set-up for single particle tracking.

 

I have never heard of a 647 nm diode laser, though. There are 636 – 640 nm diode lasers, but as far as I know they are limited to about 30 mW. I would be interested if someone knows of more powerful solid state lasers in this wavelength range.

 

Best regards,

 

Kevin

 

 

 

Van: Confocal Microscopy List [mailto:[hidden email]] Namens Mayandi Sivaguru
Verzonden: vrijdag 12 september 2008 20:34
Aan: [hidden email]
Onderwerp: 405, 488, 561 and 647nm lasers

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal


Dear Listers, is there a single or multiple Economical source/s for the diode laser lines 405, 488, 561 and 647 nm lasers with higher total output power 100-1000 mW.
Thanking you in advance
Shiv
 

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu