Alexa 750

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Olivier Bardot Olivier Bardot
Reply | Threaded
Open this post in threaded view
♦
♦
|

Alexa 750

Hello
We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use a
fifth label and have found that Alexa750 seems to be a good possibility.
We are working on Zeiss LSM 510 Meta.
Does anyone have been using this dye?
Is it really possible to discriminate Cy3 from Alexa750?
I will appreciate any advices.
Thanks
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

=
Hi Olivier, 

I doubt the META detector can read emissions past 720-750 nm. Also, if you look at the excitation/emission spectra of Alexa 750, such as here:


you will see that with standard confocal lasers (488, 543/561/633), you will achieve an excitation efficiency of at most 10%. Finally, if you want to use a conventional detector, you would need an appropriate emission filter on your confocal, such as a longpass 730 or higher, which is probably not standard, and you probably would have rather poor detection efficiency, unless you have special IR PMTs...

One suggestion would be to use a dye in the red range that you can spectrally separate from Cy3, or maybe use something with a large stokes shift, such as a Qdot 605 or similar, that you can excite at around 450 or 488, and detect in the red channel, and therefore can be separated from Cy3 based on the different excitation efficiency. Something like Cy5-PE might work too (Excitation at 488, emission at 680)


--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024





On Nov 5, 2008, at 12:32 AM, Olivier Bardot wrote:

Hello
We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use a
fifth label and have found that Alexa750 seems to be a good possibility.
We are working on Zeiss LSM 510 Meta.
Does anyone have been using this dye?
Is it really possible to discriminate Cy3 from Alexa750?
I will appreciate any advices.
Thanks

Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

Hi Olivier,

I agree with Julio that the Alexa 750 may not be the best choice.  One
possibility is to add in another red dye that could be separated from
the Cy3 using the META detector - either Alexa 568 or Rhodamine Red-X
should work fine, in our experience.  Another is to add a second green
colour and separate that from the FITC using the META - or maybe swap
from FITC to AF 488 if possible? - we have separated AF 488 and AF 514
nicely using the META, and the META detector is anyway most sensitive in
the green emission region.  You would need to tweak the labelling
conditions so that the colours you are using with the META detector are
pretty well balanced, and we also tend to prefer collecting the regular
channels in one scan and the fluorochromes to be unmixed by the META in
a separate scan.- e.g. if you were using two reds, set up a
DAPI/FITC/Cy5 regular scan, followed by a lambda stack just imaging the
reds.  Then combine the results after unmixing.  If you try to collect
everything on one lambda scan, you will have the problem of having to
uncheck the window before and after each of the additional laser lines
in the lambda stack when you do the unmixing.

I hope this makes sense, if not please feel free to contact me offline.

Best,
Alison

Julio Vazquez wrote:

> =
> Hi Olivier,
>
> I doubt the META detector can read emissions past 720-750 nm. Also, if
> you look at the excitation/emission spectra of Alexa 750, such as here:
>
> http://www.bdbiosciences.com/spectra/
>
> you will see that with standard confocal lasers (488, 543/561/633),
> you will achieve an excitation efficiency of at most 10%. Finally, if
> you want to use a conventional detector, you would need an appropriate
> emission filter on your confocal, such as a longpass 730 or higher,
> which is probably not standard, and you probably would have rather
> poor detection efficiency, unless you have special IR PMTs...
>
> One suggestion would be to use a dye in the red range that you can
> spectrally separate from Cy3, or maybe use something with a large
> stokes shift, such as a Qdot 605 or similar, that you can excite at
> around 450 or 488, and detect in the red channel, and therefore can be
> separated from Cy3 based on the different excitation efficiency.
> Something like Cy5-PE might work too (Excitation at 488, emission at 680)
>
>
> --
> Julio Vazquez
> Fred Hutchinson Cancer Research Center
> Seattle, WA 98109-1024
>
>
> http://www.fhcrc.org/
>
>
>
> On Nov 5, 2008, at 12:32 AM, Olivier Bardot wrote:
>
>> Hello
>> We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to
>> use a
>> fifth label and have found that Alexa750 seems to be a good possibility.
>> We are working on Zeiss LSM 510 Meta.
>> Does anyone have been using this dye?
>> Is it really possible to discriminate Cy3 from Alexa750?
>> I will appreciate any advices.
>> Thanks
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
rjpalmer rjpalmer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

One could also try a long-Stokes-shift fluor such as Chromeo 494 that
is excited together with AF 488 or FITC, but that emits in the mid
600s.  No unmixing required when doing sequential scans.

>Hi Olivier,
>
>I agree with Julio that the Alexa 750 may not be the best choice.
>One possibility is to add in another red dye that could be separated
>from the Cy3 using the META detector - either Alexa 568 or Rhodamine
>Red-X should work fine, in our experience.  Another is to add a
>second green colour and separate that from the FITC using the META -
>or maybe swap from FITC to AF 488 if possible? - we have separated
>AF 488 and AF 514 nicely using the META, and the META detector is
>anyway most sensitive in the green emission region.  You would need
>to tweak the labelling conditions so that the colours you are using
>with the META detector are pretty well balanced, and we also tend to
>prefer collecting the regular channels in one scan and the
>fluorochromes to be unmixed by the META in a separate scan.- e.g. if
>you were using two reds, set up a DAPI/FITC/Cy5 regular scan,
>followed by a lambda stack just imaging the reds.  Then combine the
>results after unmixing.  If you try to collect everything on one
>lambda scan, you will have the problem of having to uncheck the
>window before and after each of the additional laser lines in the
>lambda stack when you do the unmixing.
>
>I hope this makes sense, if not please feel free to contact me offline.
>
>Best,
>Alison
>
>Julio Vazquez wrote:
>>=
>>Hi Olivier,
>>I doubt the META detector can read emissions past 720-750 nm. Also,
>>if you look at the excitation/emission spectra of Alexa 750, such
>>as here:
>>
>>http://www.bdbiosciences.com/spectra/
>>
>>you will see that with standard confocal lasers (488, 543/561/633),
>>you will achieve an excitation efficiency of at most 10%. Finally,
>>if you want to use a conventional detector, you would need an
>>appropriate emission filter on your confocal, such as a longpass
>>730 or higher, which is probably not standard, and you probably
>>would have rather poor detection efficiency, unless you have
>>special IR PMTs...
>>
>>One suggestion would be to use a dye in the red range that you can
>>spectrally separate from Cy3, or maybe use something with a large
>>stokes shift, such as a Qdot 605 or similar, that you can excite at
>>around 450 or 488, and detect in the red channel, and therefore can
>>be separated from Cy3 based on the different excitation efficiency.
>>Something like Cy5-PE might work too (Excitation at 488, emission
>>at 680)
>>
>>
>>--
>>Julio Vazquez
>>Fred Hutchinson Cancer Research Center
>>Seattle, WA 98109-1024
>>
>>
>>http://www.fhcrc.org/
>>
>>
>>
>>On Nov 5, 2008, at 12:32 AM, Olivier Bardot wrote:
>>
>>>Hello
>>>We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use a
>>>fifth label and have found that Alexa750 seems to be a good possibility.
>>>We are working on Zeiss LSM 510 Meta.
>>>Does anyone have been using this dye?
>>>Is it really possible to discriminate Cy3 from Alexa750?
>>>I will appreciate any advices.
>>>Thanks
>>
>
>--
>Alison J. North, Ph.D.,
>Research Assistant Professor and Director of the Bio-Imaging Resource Center,
>The Rockefeller University,
>1230 York Avenue,
>New York,
>NY 10065.
>Tel: office ++ 212 327 7488
>Tel: lab   ++ 212 327 7486
>Fax:       ++ 212 327 7489


--
Robert J. Palmer Jr., Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-594-0025
fax 301-402-0396
Guillermo Palchik Guillermo Palchik
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

Another possibility, if you already are thinking of having 5 colors,  
you should start considering using quantum dots...

On Nov 5, 2008, at 1:20 PM, Robert J. Palmer Jr. wrote:

> One could also try a long-Stokes-shift fluor such as Chromeo 494  
> that is excited together with AF 488 or FITC, but that emits in the  
> mid 600s.  No unmixing required when doing sequential scans.
>
>> Hi Olivier,
>>
>> I agree with Julio that the Alexa 750 may not be the best choice.  
>> One possibility is to add in another red dye that could be  
>> separated from the Cy3 using the META detector - either Alexa 568  
>> or Rhodamine Red-X should work fine, in our experience.  Another is  
>> to add a second green colour and separate that from the FITC using  
>> the META - or maybe swap from FITC to AF 488 if possible? - we have  
>> separated AF 488 and AF 514 nicely using the META, and the META  
>> detector is anyway most sensitive in the green emission region.  
>> You would need to tweak the labelling conditions so that the  
>> colours you are using with the META detector are pretty well  
>> balanced, and we also tend to prefer collecting the regular  
>> channels in one scan and the fluorochromes to be unmixed by the  
>> META in a separate scan.- e.g. if you were using two reds, set up a  
>> DAPI/FITC/Cy5 regular scan, followed by a lambda stack just imaging  
>> the reds.  Then combine the results after unmixing.  If you try to  
>> collect everything on one lambda scan, you will have the problem of  
>> having to uncheck the window before and after each of the  
>> additional laser lines in the lambda stack when you do the unmixing.
>>
>> I hope this makes sense, if not please feel free to contact me  
>> offline.
>>
>> Best,
>> Alison
>>
>> Julio Vazquez wrote:
>>> =
>>> Hi Olivier,
>>> I doubt the META detector can read emissions past 720-750 nm.  
>>> Also, if you look at the excitation/emission spectra of Alexa 750,  
>>> such as here:
>>>
>>> http://www.bdbiosciences.com/spectra/
>>>
>>> you will see that with standard confocal lasers (488,  
>>> 543/561/633), you will achieve an excitation efficiency of at most  
>>> 10%. Finally, if you want to use a conventional detector, you  
>>> would need an appropriate emission filter on your confocal, such  
>>> as a longpass 730 or higher, which is probably not standard, and  
>>> you probably would have rather poor detection efficiency, unless  
>>> you have special IR PMTs...
>>>
>>> One suggestion would be to use a dye in the red range that you can  
>>> spectrally separate from Cy3, or maybe use something with a large  
>>> stokes shift, such as a Qdot 605 or similar, that you can excite  
>>> at around 450 or 488, and detect in the red channel, and therefore  
>>> can be separated from Cy3 based on the different excitation  
>>> efficiency. Something like Cy5-PE might work too (Excitation at  
>>> 488, emission at 680)
>>>
>>>
>>> --
>>> Julio Vazquez
>>> Fred Hutchinson Cancer Research Center
>>> Seattle, WA 98109-1024
>>>
>>>
>>> http://www.fhcrc.org/
>>>
>>>
>>>
>>> On Nov 5, 2008, at 12:32 AM, Olivier Bardot wrote:
>>>
>>>> Hello
>>>> We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want  
>>>> to use a
>>>> fifth label and have found that Alexa750 seems to be a good  
>>>> possibility.
>>>> We are working on Zeiss LSM 510 Meta.
>>>> Does anyone have been using this dye?
>>>> Is it really possible to discriminate Cy3 from Alexa750?
>>>> I will appreciate any advices.
>>>> Thanks
>>>
>>
>> --
>> Alison J. North, Ph.D.,
>> Research Assistant Professor and Director of the Bio-Imaging  
>> Resource Center,
>> The Rockefeller University,
>> 1230 York Avenue,
>> New York,
>> NY 10065.
>> Tel: office ++ 212 327 7488
>> Tel: lab   ++ 212 327 7486
>> Fax:       ++ 212 327 7489
>
>
> --
> Robert J. Palmer Jr., Ph.D.
> Natl Inst Dental Craniofacial Res - Natl Insts Health
> Oral Infection and Immunity Branch
> Bldg 30, Room 310
> 30 Convent Drive
> Bethesda MD 20892
> ph 301-594-0025
> fax 301-402-0396

Guillermo Palchik
[hidden email]
Fredrik Wermeling Fredrik Wermeling
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

One word of caution regarding the Qdots is that since they are excited by all lasers you can´t
do sequential scanning (or at least you won´t benefit from it)  and thus need to really trust
your filter settings. I have however used Qdot605 successfully in combination with alexa488
and alexa647. Nowadays I prefer to use alexa555 instead of the Qdot605, to be able to
record individual channels and minimize crosstalk. For four color staining I´ve included a
Pacific Blue conjugated antibody, recorded in the dapi channel, to the above combination
with ok result.

Good luck!
/Fredrik    

--------------------------------------
Fredrik Wermeling, PhD-student
Karolinska Institutet
Department of Medicine
Unit of Clinical Allergy Research
L2:04 Karolinska Hospital
SE-171 76 Stockholm
Sweden

phone: +46-8-51776696
mail: [hidden email]
fax: +46-8-335724
--------------------------------------

----- Original Message -----
From: Guillermo Palchik <[hidden email]>
Date: Wednesday, November 5, 2008 8:11 pm
Subject: Re: Alexa 750
To: [hidden email]

> Another possibility, if you already are thinking of having 5
> colors,  
> you should start considering using quantum dots...
>
> On Nov 5, 2008, at 1:20 PM, Robert J. Palmer Jr. wrote:
>
> > One could also try a long-Stokes-shift fluor such as Chromeo 494  
> > that is excited together with AF 488 or FITC, but that emits in
> the  
> > mid 600s.  No unmixing required when doing sequential scans.
> >
> >> Hi Olivier,
> >>
> >> I agree with Julio that the Alexa 750 may not be the best
> choice.  
> >> One possibility is to add in another red dye that could be  
> >> separated from the Cy3 using the META detector - either Alexa
> 568  
> >> or Rhodamine Red-X should work fine, in our experience.  Another
> is  
> >> to add a second green colour and separate that from the FITC
> using  
> >> the META - or maybe swap from FITC to AF 488 if possible? - we
> have  
> >> separated AF 488 and AF 514 nicely using the META, and the META  
> >> detector is anyway most sensitive in the green emission region.  
>
> >> You would need to tweak the labelling conditions so that the  
> >> colours you are using with the META detector are pretty well  
> >> balanced, and we also tend to prefer collecting the regular  
> >> channels in one scan and the fluorochromes to be unmixed by the  
> >> META in a separate scan.- e.g. if you were using two reds, set
> up a  
> >> DAPI/FITC/Cy5 regular scan, followed by a lambda stack just
> imaging  
> >> the reds.  Then combine the results after unmixing.  If you try
> to  
> >> collect everything on one lambda scan, you will have the problem
> of  
> >> having to uncheck the window before and after each of the  
> >> additional laser lines in the lambda stack when you do the
> unmixing.>>
> >> I hope this makes sense, if not please feel free to contact me  
> >> offline.
> >>
> >> Best,
> >> Alison
> >>
> >> Julio Vazquez wrote:
> >>> =
> >>> Hi Olivier,
> >>> I doubt the META detector can read emissions past 720-750 nm.  
> >>> Also, if you look at the excitation/emission spectra of Alexa
> 750,  
> >>> such as here:
> >>>
> >>> http://www.bdbiosciences.com/spectra/
> >>>
> >>> you will see that with standard confocal lasers (488,  
> >>> 543/561/633), you will achieve an excitation efficiency of at
> most  
> >>> 10%. Finally, if you want to use a conventional detector, you  
> >>> would need an appropriate emission filter on your confocal,
> such  
> >>> as a longpass 730 or higher, which is probably not standard,
> and  
> >>> you probably would have rather poor detection efficiency,
> unless  
> >>> you have special IR PMTs...
> >>>
> >>> One suggestion would be to use a dye in the red range that you
> can  
> >>> spectrally separate from Cy3, or maybe use something with a
> large  
> >>> stokes shift, such as a Qdot 605 or similar, that you can
> excite  
> >>> at around 450 or 488, and detect in the red channel, and
> therefore  
> >>> can be separated from Cy3 based on the different excitation  
> >>> efficiency. Something like Cy5-PE might work too (Excitation at
>
> >>> 488, emission at 680)
> >>>
> >>>
> >>> --
> >>> Julio Vazquez
> >>> Fred Hutchinson Cancer Research Center
> >>> Seattle, WA 98109-1024
> >>>
> >>>
> >>> http://www.fhcrc.org/
> >>>
> >>>
> >>>
> >>> On Nov 5, 2008, at 12:32 AM, Olivier Bardot wrote:
> >>>
> >>>> Hello
> >>>> We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We
> want  
> >>>> to use a
> >>>> fifth label and have found that Alexa750 seems to be a good  
> >>>> possibility.
> >>>> We are working on Zeiss LSM 510 Meta.
> >>>> Does anyone have been using this dye?
> >>>> Is it really possible to discriminate Cy3 from Alexa750?
> >>>> I will appreciate any advices.
> >>>> Thanks
> >>>
> >>
> >> --
> >> Alison J. North, Ph.D.,
> >> Research Assistant Professor and Director of the Bio-Imaging  
> >> Resource Center,
> >> The Rockefeller University,
> >> 1230 York Avenue,
> >> New York,
> >> NY 10065.
> >> Tel: office ++ 212 327 7488
> >> Tel: lab   ++ 212 327 7486
> >> Fax:       ++ 212 327 7489
> >
> >
> > --
> > Robert J. Palmer Jr., Ph.D.
> > Natl Inst Dental Craniofacial Res - Natl Insts Health
> > Oral Infection and Immunity Branch
> > Bldg 30, Room 310
> > 30 Convent Drive
> > Bethesda MD 20892
> > ph 301-594-0025
> > fax 301-402-0396
>
> Guillermo Palchik
> [hidden email]
>
rjpalmer rjpalmer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

In reply to this post by Olivier Bardot
Re the five-color stuff, one could also try a long-Stokes-shift fluor
such as Chromeo 494 that is excited together with AF 488 or FITC, but
that emits in the mid 600s.  No unmixing required but presumes your
sample can survive sequential scans.

--
Robert J. Palmer Jr., Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-594-0025
fax 301-402-0396
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

In reply to this post by Olivier Bardot
AF750 has been tried in my lab on a 510 META with no success (633 excitation).  As has been mentioned by others, the standard laser lines and emission filters are not suitable.  We ended up going on a widefield 'scope with an AF750 filter from Chroma and a metal halide lamp.  Even though the output of metal halide and the QE of the camera are both poor at ≥750 nm, we got good images with long integration times (> 1 sec.) on a cooled CCD camera.
 
-Esteban

On Wed, Nov 5, 2008 at 2:32 AM, Olivier Bardot <[hidden email]> wrote:
Hello
We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use a
fifth label and have found that Alexa750 seems to be a good possibility.
We are working on Zeiss LSM 510 Meta.
Does anyone have been using this dye?
Is it really possible to discriminate Cy3 from Alexa750?
I will appreciate any advices.
Thanks



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax


M. van de corput M. van de corput
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750

Another option is using combinatorial labeling. I don't know if that is
possible for your targets but it will avoid an increase in scan time and
you can use the labels you have used before and are familiar with.
Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput,
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Wo, 5 november, 2008 11:58 pm, schreef G. Esteban Fernandez:

> AF750 has been tried in my lab on a 510 META with no success (633
> excitation).  As has been mentioned by others, the standard laser lines
> and
> emission filters are not suitable.  We ended up going on a widefield
> 'scope
> with an AF750 filter from Chroma and a metal halide lamp.  Even though the
> output of metal halide and the QE of the camera are both poor at ™750 nm,
> we
> got good images with long integration times (> 1 sec.) on a cooled CCD
> camera.
>
> -Esteban
>
> On Wed, Nov 5, 2008 at 2:32 AM, Olivier Bardot <[hidden email]> wrote:
>
>> Hello
>> We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use
>> a
>> fifth label and have found that Alexa750 seems to be a good possibility.
>> We are working on Zeiss LSM 510 Meta.
>> Does anyone have been using this dye?
>> Is it really possible to discriminate Cy3 from Alexa750?
>> I will appreciate any advices.
>> Thanks
>>
>
>
>
> --
> G. Esteban Fernandez, Ph.D.
> Associate Director
> Molecular Cytology Core Facility
> University of Missouri
> 120 Bond Life Sciences Center
> Columbia, MO  65211
>
> http://www.biotech.missouri.edu/mcc/
>
> 573-882-4895
> 573-884-9395 fax
>
Kilgore, Jason Kilgore, Jason
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Alexa 750 - Vendor Response

In reply to this post by G. Esteban Fernandez
 
**Vendor Response**
 
Hi, all,
 
I should note that we have a 510 META here at Molecular Probes.  Though I don't think I have tried using Alexa Fluor 750 on it, Qdot 800 is not able to be visualized on that system.  Qdot 705, or anything lower, is able to be imaged, though.
 
Of course, widefield microscopes work fine with Alexa Fluor 750 or Qdot 800 (with correct filter sets and detectors).
 
Cheers,
 
Jason
 
Jason A. Kilgore
Invitrogen Corporation
Molecular Probes Labeling and Detection Technologies
Technical Service Scientist
Eugene, OR  97402-9132
phone: (541) 335-0353
fax:      (541) 335-0238
Please visit our website:  www.probes.invitrogen.com
 
 P Save trees.  Please don't print this e-mail unless you really need to.


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of G. Esteban Fernandez
Sent: Wednesday, November 05, 2008 2:58 PM
To: [hidden email]
Subject: Re: Alexa 750

AF750 has been tried in my lab on a 510 META with no success (633 excitation).  As has been mentioned by others, the standard laser lines and emission filters are not suitable.  We ended up going on a widefield 'scope with an AF750 filter from Chroma and a metal halide lamp.  Even though the output of metal halide and the QE of the camera are both poor at ≥750 nm, we got good images with long integration times (> 1 sec.) on a cooled CCD camera.
 
-Esteban

On Wed, Nov 5, 2008 at 2:32 AM, Olivier Bardot <[hidden email]> wrote:
Hello
We are using DAPI, FITC, Cy3 and Cy5 for immunolabelling. We want to use a
fifth label and have found that Alexa750 seems to be a good possibility.
We are working on Zeiss LSM 510 Meta.
Does anyone have been using this dye?
Is it really possible to discriminate Cy3 from Alexa750?
I will appreciate any advices.
Thanks



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax