![]() ![]() |
Santiago Correa |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Everyone, We recently acquired a 1999 micro radiance AG2 confocal lasers scanning microscope. We have been working on getting it to run but have hit a block. when we initialize the lasersharp 2000 program to begin using the laser, we get several error messages. These are the error popups, -Path not found -Failed to initialize 1394 channel -Initializing firewire (SVIS) failed -Error 1168 hardware problem in channel initiation We are using the same desktop computer, and programs that were used it originally. We do not have or know what the firewire is supposed to connect to, I have tried looking online for troubleshooting but there is nothing on the error readouts. If anyone has had experience with this system please let me know. We are very excited to have this equipment and eager to use it. Thank you all for your time. -Santiago |
![]() ![]() |
Guy Cox-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Firewire is a high-speed bus - an Apple product - that the computer uses to connect to the scan head. There will be a firewire board plugged in to the control computer, and a firewire cable connecting the board to the scan head. Check that these are all present and correctly plugged in. If there is a problem with these the system cannot run. Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Santiago Correa Sent: Tuesday, 4 November 2014 6:02 AM To: [hidden email] Subject: Biorad Micro radiance AG2 Lasersharp 2000 Error ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Everyone, We recently acquired a 1999 micro radiance AG2 confocal lasers scanning microscope. We have been working on getting it to run but have hit a block. when we initialize the lasersharp 2000 program to begin using the laser, we get several error messages. These are the error popups, -Path not found -Failed to initialize 1394 channel -Initializing firewire (SVIS) failed -Error 1168 hardware problem in channel initiation We are using the same desktop computer, and programs that were used it originally. We do not have or know what the firewire is supposed to connect to, I have tried looking online for troubleshooting but there is nothing on the error readouts. If anyone has had experience with this system please let me know. We are very excited to have this equipment and eager to use it. Thank you all for your time. -Santiago |
![]() ![]() |
Santiago Correa |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thank you for the reply. It seems that the FireWire controller board is missing. We are now looking for potential boards. Do anyone have any recommendations in particular for the board. Perhaps brand or specs? Thank you again. Santiago On Tuesday, November 4, 2014, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Firewire is a high-speed bus - an Apple product - that the computer uses > to connect to the scan head. There will be a firewire board plugged in to > the control computer, and a firewire cable connecting the board to the scan > head. Check that these are all present and correctly plugged in. If there > is a problem with these the system cannot run. > > Guy > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email] > <javascript:;>] On Behalf Of Santiago Correa > Sent: Tuesday, 4 November 2014 6:02 AM > To: [hidden email] <javascript:;> > Subject: Biorad Micro radiance AG2 Lasersharp 2000 Error > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello Everyone, > > We recently acquired a 1999 micro radiance AG2 confocal lasers scanning > microscope. We have been working on getting it to run but have hit a block. > when we initialize the lasersharp 2000 program to begin using the laser, > we get several error messages. > These are the error popups, > > -Path not found > > -Failed to initialize 1394 channel > > -Initializing firewire (SVIS) failed > > -Error 1168 hardware problem in channel initiation > > We are using the same desktop computer, and programs that were used it > originally. > We do not have or know what the firewire is supposed to connect to, I have > tried looking online for troubleshooting but there is nothing on the error > readouts. If anyone has had experience with this system please let me know. > We are very excited to have this equipment and eager to use it. Thank you > all for your time. > > -Santiago > ... [show rest of quote]
|
Free forum by Nabble | Edit this page |