|    | Mel Symeonides | 
|   | 
		Hello,
 I am using a DeltaVision microscope with stock filters to image fixed cells labeled with the lipophilic tracer DiI and the cytoplasmic label Cell Tracker Green CMFDA. There are two co-cultured cell populations previously labeled with either dye and I am looking for double-labeled cells as a result of exchange of membrane or cytoplasm between cells. I believe I am experiencing bleed through of DiI into the FITC channel and want to see whether anyone else has experienced this or could predict whether there would be bleed through based on its spectral profile. The DeltaVision FITC/green filter has the following properties: Excitation: 490/20nm (laser at 488nm) Emission: 528/38nm If you load DiI up on Invitrogen's Fluorescence SpectraViewer and plug in the filter spectra above, you should notice two things: DiI will excite to around 20%, and around 5-10% of DiI emission will be collected by this filter. I think I am interpreting this correctly. What I actually experience during imaging is that in locations of very high levels of DiI fluorescence, there is low level signal in the FITC channel where there shouldn't be any of the green CMFDA dye. This signal precisely co-localizes with the strongest areas of DiI fluorescence, which leads me to believe that it is bleed-through. Moreover, I have quantified the level of the bled-through signal to around 0.5-1% of the corresponding level in the T-RED channel, which I use to excite and collect DiI fluorescence. This proportion is in line with what I would expect if the FITC channel excites DiI by 20% and collects 5% of that, if Fluorescence SpectraViewer is accurate. (20% * 5% = 1%) At 1%, it should not be an issue to simply scale it out, however there are two compounding issues: Firstly, DiI is an extremely strong label, which increases the level of bleed through, and secondly, Cell Tracker Green labels very weakly, to the extent that the DiI bleed through is sometimes stronger than the real green fluorescence! I am in the process of optimizing the green label in order to boost the signal. All this makes sense to me, but I wanted to see if I am getting this right. I have been offered the alternative explanation that the areas of "bleed through" are actually autofluorescence due to the cells taking up material from the environment, or the cells being unhappy due to manipulation. However, the co-localization of regions of intense DiI and non-specific green signal, as well as the spectral profile of DiI, suggest bleed-through to me. Thanks for reading! Mel | 
|    | Tamara Howard | 
|   | 
		I may have missed something, but have you actually looked 
 at cells labeled ONLY with DiI and checked the other channel under the imaging conditions you are using with the CMFDA? That would tell you right away if you are seeing bleed-through...it should be standard practice to always run a set of single labels when trying to do doubles/co-localizations - that way you'll know if you have any issues, instead of just guessing/assuming! Tamara On Mon, 26 Apr 2010 15:10:39 -0700 Mel Symeonides <[hidden email]> wrote: > Hello, > > I am using a DeltaVision microscope with stock filters >to image fixed cells > labeled with the lipophilic tracer DiI and the >cytoplasmic label Cell > Tracker Green CMFDA. There are two co-cultured cell >populations previously > labeled with either dye and I am looking for >double-labeled cells as a > result of exchange of membrane or cytoplasm between >cells. > > I believe I am experiencing bleed through of DiI into >the FITC channel and > want to see whether anyone else has experienced this or >could predict > whether there would be bleed through based on its >spectral profile. > > The DeltaVision FITC/green filter has the following >properties: > Excitation: 490/20nm (laser at 488nm) > Emission: 528/38nm > > If you load DiI up on Invitrogen's Fluorescence >SpectraViewer and plug in > the filter spectra above, you should notice two things: >DiI will excite to > around 20%, and around 5-10% of DiI emission will be >collected by this > filter. I think I am interpreting this correctly. > > > What I actually experience during imaging is that in >locations of very high > levels of DiI fluorescence, there is low level signal in >the FITC channel > where there shouldn't be any of the green CMFDA dye. >This signal precisely > co-localizes with the strongest areas of DiI >fluorescence, which leads me to > believe that it is bleed-through. Moreover, I have >quantified the level of > the bled-through signal to around 0.5-1% of the >corresponding level in the > T-RED channel, which I use to excite and collect DiI >fluorescence. This > proportion is in line with what I would expect if the >FITC channel excites > DiI by 20% and collects 5% of that, if Fluorescence >SpectraViewer is > accurate. (20% * 5% = 1%) > > At 1%, it should not be an issue to simply scale it out, >however there are > two compounding issues: Firstly, DiI is an extremely >strong label, which > increases the level of bleed through, and secondly, Cell >Tracker Green > labels very weakly, to the extent that the DiI bleed >through is sometimes > stronger than the real green fluorescence! I am in the >process of optimizing > the green label in order to boost the signal. > > > All this makes sense to me, but I wanted to see if I am >getting this right. > I have been offered the alternative explanation that the >areas of "bleed > through" are actually autofluorescence due to the cells >taking up material > from the environment, or the cells being unhappy due to >manipulation. > However, the co-localization of regions of intense DiI >and non-specific > green signal, as well as the spectral profile of DiI, >suggest bleed-through > to me. > > > Thanks for reading! > Mel > -- > View this message in context: >http://confocal-microscopy-list.588098.n2.nabble.com/Bleed-through-of-membrane-label-DiI-into-green-channel-tp4965362p4965362.html > Sent from the Confocal Microscopy List mailing list >archive at Nabble.com. *************************** Tamara Howard Cell Biology & Physiology UNM-HSC Albuquerque, NM *************************** | 
|    | John Oreopoulos | 
|   | 
		Based on the original comments about quantifying bleed-through signals, it sounds like Mel has made some attempt to look at singly labeled control samples of Cell Tracker Green and DiI separately. To further Tamara's suggestions, if you want to check for autofluorescence exclusively, try imaging a completely unlabelled sample with the same acquisition settings for both channels (preferably on the same day with identical preparation procedures minus the labelling step for the unlabelled sample). One way you could verify/confirm if the cause is really signal bleed-through into the opposite channel due to DiI fluorescence would be to reverse the colour of the probes by using spectral variants of the same labels. For DiI, you could switch to the green emitting variant DiO, and I believe there is a probe called Cell Tracker Red (no commercial interest for wherever you get these probes). Check and see if your colocalization coefficients change under these conditions. There are also longer wavelength versions of DiI, DiD and DiR, which might be compatible with your current Texas Red filters. If signal bleed through is the culprit, short of purchasing better filters optimized for these two dyes or switching probes, you could try linear unmixing or spectral imaging if your department has access to a confocal microscope with spectrally resolved detection. Alternatively, your current microscope system might have a shorter wavelength laser which can still sufficiently excite Cell Tracker Green with lesser excitation of DiI. John Oreopoulos, BSc, PhD Candidate University of Toronto Institute For Biomaterials and Biomedical Engineering Centre For Studies in Molecular Imaging On 2010-04-26, at 6:54 PM, Tamara A Howard wrote: 
 | 
|    | Mel Symeonides | 
|   | 
		Thanks Tamara & John. I have not imaged samples where there are no Cell Tracker Green-labeled cells around, but the phenomenon I am looking for occurs at extremely low frequency so in fact the vast majority of cells are singly labeled anyway. Also the cells are at very low density so that they are well spaced apart unless they are actually interacting as part of the phenomenon. I think at this point however I should really image singly labeled cells as well as unlabeled cells to look for autofluorescence, just to be 100% certain of what I am seeing. If there is no way around the issue I will have to look into the alternative labels you suggested.
 Just to clarify though - does my interpretation of the spectral profiles make sense? i.e. should I in fact expect about 1% bleed-through into the green channel? I am fairly new to this so I want to be sure I am thinking along the right tracks. Mel | 
|    | Daniel James White | 
|   | 
				In reply to this post by Mel Symeonides
			 
		Hi Mel,
 Begin forwarded message: > > Date: Mon, 26 Apr 2010 20:25:38 -0700 > From: Mel Symeonides <[hidden email]> > Subject: Re: Bleed through of membrane label DiI into green channel > > Thanks Tamara & John. I have not imaged samples where there are no Cell > Tracker Green-labeled cells around, but the phenomenon I am looking for > occurs at extremely low frequency so in fact the vast majority of cells are > singly labeled anyway. Also the cells are at very low density so that they > are well spaced apart unless they are actually interacting as part of the > phenomenon. I think at this point however I should really image singly > labeled cells as well as unlabeled cells to look for autofluorescence, just > to be 100% certain of what I am seeing. thats exactly what you should do. You must always do the single labelled control to check for bleed through, especially when you expect to get it by looking at the spectra. You have interpreted those correctly. > If there is no way around the issue > I will have to look into the alternative labels you suggested. this is by far the best choice. You can keep the Dil dye and use a a red or even far red dye (far red might be harder for your camera to detect) so as to minimise the possibility of wrong excitation and wrong emission. (crosstalk and bleed through...) Or do it the other way around. Watch out for "blue" dyes leaking into green detection, and "red" emitting dyes being excited by "blue" excitation light. Always check the spectra first. You can use image processing methods to separate these signals, but thats tricky, and error prone unless you really know what you are doing in a well characterized system. But why bother with all that? Its better to get spectrally separated data from the get go by using dyes that are better separated. > > Just to clarify though - does my interpretation of the spectral profiles > make sense? i.e. should I in fact expect about 1% bleed-through into the > green channel? Yes and Yes. > I am fairly new to this so I want to be sure I am thinking > along the right tracks. you are already doing much better than many others, by even recognizing the problem. I've seen nature and cell papers where this problem is clearly there but ignored. cheers Dan Dr. Daniel James White BSc. (Hons.) PhD Senior Microscopist / Image Visualisation, Processing and Analysis Light Microscopy and Image Processing Facilities Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstrasse 108 01307 DRESDEN Germany +49 (0)15114966933 (German Mobile) +49 (0)351 210 2627 (Work phone at MPI-CBG) +49 (0)351 210 1078 (Fax MPI-CBG LMF) http://www.bioimagexd.net BioImageXD http://pacific.mpi-cbg.de Fiji - is just ImageJ (Batteries Included) http://www.chalkie.org.uk Dan's Homepages https://ifn.mpi-cbg.de Dresden Imaging Facility Network dan (at) chalkie.org.uk ( white (at) mpi-cbg.de ) | 
|    | George McNamara | 
|   | 
				In reply to this post by Mel Symeonides
			 
		
Hi Mel, Yes, DiI excites nicely at 488 nm on our confocal scopes (see Li ... Wen et al 2008 Nature Protocols for our major use). Suggestions: a) switch to DiD or DiR (or DiA)? b) DeltaVision is quantitative - bleedthrough should be linear, subtract it. *** Instead of estimating from a web graph, the spectra data for most fluorophores is at the pubspectra zip file at http://www.sylvester.org/AICF/pubspectra/ The UA graphing site organized by Carl Boswell, http://www.mcb.arizona.edu/ipc/fret/index.html has a calculations option (for those fluorophores with full details). If anyone cannot find their favorite fluorophore spectra or filter, etc, in the Pubspectra.xlsx, I am happy to add more data but am about done with digitizing graphs - please send me data files or web links to data. George At 06:10 PM 4/26/2010, you wrote: Hello, George McNamara, Ph.D. Image Core Manager Analytical Imaging Core Facility University of Miami, Miller School of Medicine Miami, FL 33136 [hidden email] [hidden email] 305-243-8436 office http://www.sylvester.org/AICF (Analytical Imaging Core Facility) http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx file is in the zip file) http://home.earthlink.net/~geomcnamara | 