Blue beads for SIM calibration.

classic Classic list List threaded Threaded
♦
♦
2 messages Options
Ian Dobbie Ian Dobbie
Reply | Threaded
Open this post in threaded view
♦
♦
|

Blue beads for SIM calibration.

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Everyone,

I am wondering if anyone has any good source of Blue (405 excitation,
460 emission) beads for calibrating SIM systems. We have some of the
PSF spec 170 nm beads from invitrogen but really struggling to get
enough signal from the beads. The system has an air objective so the 170
nm beads are still sub resolution.

We previously had a brilliant sample of 100 nm beads fro Duke scientific
but they stopped producing them at least a decade ago.

Thanks

Ian
simon walker (BI)-2 simon walker (BI)-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Blue beads for SIM calibration.

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Our flow team were asked to beta test CPNs  https://www.streambio.co.uk/ which I also tested for SIM calibration.
They worked very well, especially for the 405 nm laser where we'd previously struggled to find anything bright enough.
Simon
(no commercial interest).


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ian Dobbie
Sent: 14 May 2021 16:14
To: [hidden email]
Subject: Blue beads for SIM calibration.

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Everyone,

I am wondering if anyone has any good source of Blue (405 excitation,
460 emission) beads for calibrating SIM systems. We have some of the PSF spec 170 nm beads from invitrogen but really struggling to get enough signal from the beads. The system has an air objective so the 170 nm beads are still sub resolution.

We previously had a brilliant sample of 100 nm beads fro Duke scientific but they stopped producing them at least a decade ago.

Thanks

Ian