Oshel, Philip Eugene |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** All, I had this question put to me by a new faculty member, and don't have a ready answer: "Is there a ballpark percentage for how much less bright an LED vs a standard mercury lamp light?" This is for regular epifluorescence, not confocal. This is in the realm of arm-waving over a picture of beer (a good, dark stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part of the spectrum is used, and all that. Personally, I'd think the answer is more like, "Doesn't matter, the dimmer system is still too bright to use all the available light and not damage the specimen." But ... ? Phil -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
|
Armstrong, Brian |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Phil I would say that the new LED sources and Light Engines will be brighter than mercury sources. For example I compared an Exfo to a SOLA light engine and the light engine is brighter. The new Lumen Dynamics XLED should provide much better illumination than a mercury source. Hopefully, mercury sources will soon be a thing of the past. Cheers, Brian D Armstrong PhD Associate Research Professor Director, Light Microscopy Core Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philip Oshel Sent: Tuesday, November 05, 2013 9:57 AM To: [hidden email] Subject: Brightness difference Hg vs LED ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** All, I had this question put to me by a new faculty member, and don't have a ready answer: "Is there a ballpark percentage for how much less bright an LED vs a standard mercury lamp light?" This is for regular epifluorescence, not confocal. This is in the realm of arm-waving over a picture of beer (a good, dark stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part of the spectrum is used, and all that. Personally, I'd think the answer is more like, "Doesn't matter, the dimmer system is still too bright to use all the available light and not damage the specimen." But ... ? Phil -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |
|
Doube, Michael |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Phil, no idea about percentages but in my hands a single blue high power LED like this one, is 'enough' for practical application, when set directly in place of a 50W Hg lamp, on a heatsink in the lamphouse: http://uk.rs-online.com/web/p/visible-leds/7393548/ (no commercial interest) Provided of course that you only want blueish excitation... You can also run them off very simple and cheap switching boxes, if you are not afraid of a soldering iron. Michael On 05/11/13 18:00, Philip Oshel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > All, > > I had this question put to me by a new faculty member, and don't have a > ready answer: > "Is there a ballpark percentage for how much less bright an LED vs a > standard mercury lamp light?" > This is for regular epifluorescence, not confocal. > > This is in the realm of arm-waving over a picture of beer (a good, dark > stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part > of the spectrum is used, and all that. Personally, I'd think the answer > is more like, "Doesn't matter, the dimmer system is still too bright to > use all the available light and not damage the specimen." But ... ? > > Phil -- Dr Michael Doube BPhil BVSc PhD MRCVS Lecturer, Comparative Biomedical Sciences The Royal Veterinary College London NW1 0TU United Kingdom +44 (0)20 7121 1903 [RVC Logo - link to RVC Website]<http://www.rvc.ac.uk> [Twitter icon - link to RVC (Official) Twitter] <http://twitter.com/RoyalVetCollege> [Facebook icon - link to RVC (Official) Facebook] <http://www.facebook.com/theRVC> [YouTube icon - link to RVC YouTube] <http://www.youtube.com/user/RoyalVetsLondon?feature=mhee> [Pinterest icon - link to RVC Pinterest] <http://pinterest.com/royalvetcollege/> [Instagram icon - link to RVC Instagram] <http://instagram.com/royalvetcollege> This message, together with any attachments, is intended for the stated addressee(s) only and may contain privileged or confidential information. Any views or opinions presented are solely those of the author and do not necessarily represent those of the Royal Veterinary College (RVC). If you are not the intended recipient, please notify the sender and be advised that you have received this message in error and that any use, dissemination, forwarding, printing, or copying is strictly prohibited. Unless stated expressly in this email, this email does not create, form part of, or vary any contractual or unilateral obligation. Email communication cannot be guaranteed to be secure or error free as information could be intercepted, corrupted, amended, lost, destroyed, incomplete or contain viruses. Therefore, we do not accept liability for any such matters or their consequences. Communication with us by email will be taken as acceptance of the risks inherent in doing so. |
|
Kurt Thorn |
|
|
In reply to this post by Armstrong, Brian
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** My sense is also that the new LED sources are brighter than Hg lamps, except possibly at the very brightest mercury peaks (365, 546 nm). For GFP, they are for sure brighter. True LEDs have been somewhat dim in the 560 nm range, which is why Lumencor uses an LED-pumped phosphor for those wavelengths. This gives very bright 560 emission. Kurt On 11/5/2013 10:21 AM, Armstrong, Brian wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Phil I would say that the new LED sources and Light Engines will be brighter than mercury sources. For example I compared an Exfo to a SOLA light engine and the light engine is brighter. The new Lumen Dynamics XLED should provide much better illumination than a mercury source. > Hopefully, mercury sources will soon be a thing of the past. > Cheers, > > Brian D Armstrong PhD > Associate Research Professor > Director, Light Microscopy Core > Beckman Research Institute > City of Hope > Dept of Neuroscience > 1450 E Duarte Rd > Duarte, CA 91010 > 626-256-4673 x62872 > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philip Oshel > Sent: Tuesday, November 05, 2013 9:57 AM > To: [hidden email] > Subject: Brightness difference Hg vs LED > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > All, > > I had this question put to me by a new faculty member, and don't have a > ready answer: > "Is there a ballpark percentage for how much less bright an LED vs a > standard mercury lamp light?" > This is for regular epifluorescence, not confocal. > > This is in the realm of arm-waving over a picture of beer (a good, dark > stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part > of the spectrum is used, and all that. Personally, I'd think the answer > is more like, "Doesn't matter, the dimmer system is still too bright to > use all the available light and not damage the specimen." But ... ? > > Phil |
|
Julio Vazquez |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** A quick comparison in our lab gave the following: 488 nm (FITC) band of an LED-based light engine (Lumencor): 40 mW 300 W Xenon Sutter illuminator (probably our brightest light source) with FITC EX filter: 70 mW both measured at the exit of the fiber or light guide. I agree that LED based light engines are probably at least as good (power-wise) as most lamps, plus they have other benefits. Charts such as the one found here should be a good guide regarding relative power: http://lumencor.com/wp-content/uploads/2011/09/SpectraX-LE-User-Manual1.pdf Julio Vazquez, PhD Fred Hutchinson Cancer Research Center Seattle, WA 98109 http://www.fhcrc.org/en.html On Nov 5, 2013, at 1:50 PM, Kurt Thorn wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > My sense is also that the new LED sources are brighter than Hg lamps, except possibly at the very brightest mercury peaks (365, 546 nm). For GFP, they are for sure brighter. True LEDs have been somewhat dim in the 560 nm range, which is why Lumencor uses an LED-pumped phosphor for those wavelengths. This gives very bright 560 emission. > > Kurt > > On 11/5/2013 10:21 AM, Armstrong, Brian wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi Phil I would say that the new LED sources and Light Engines will be brighter than mercury sources. For example I compared an Exfo to a SOLA light engine and the light engine is brighter. The new Lumen Dynamics XLED should provide much better illumination than a mercury source. >> Hopefully, mercury sources will soon be a thing of the past. >> Cheers, >> >> Brian D Armstrong PhD >> Associate Research Professor >> Director, Light Microscopy Core >> Beckman Research Institute >> City of Hope >> Dept of Neuroscience >> 1450 E Duarte Rd >> Duarte, CA 91010 >> 626-256-4673 x62872 >> >> >> >> -----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philip Oshel >> Sent: Tuesday, November 05, 2013 9:57 AM >> To: [hidden email] >> Subject: Brightness difference Hg vs LED >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> All, >> >> I had this question put to me by a new faculty member, and don't have a >> ready answer: >> "Is there a ballpark percentage for how much less bright an LED vs a >> standard mercury lamp light?" >> This is for regular epifluorescence, not confocal. >> >> This is in the realm of arm-waving over a picture of beer (a good, dark >> stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part >> of the spectrum is used, and all that. Personally, I'd think the answer >> is more like, "Doesn't matter, the dimmer system is still too bright to >> use all the available light and not damage the specimen." But ... ? >> >> Phil |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** How much of the 70 mW is from UV or IR leaking through? _________________________________________ Michael Cammer, Assistant Research Scientist Skirball Institute of Biomolecular Medicine Lab: (212) 263-3208 Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Julio Vazquez [[hidden email]] Sent: Tuesday, November 05, 2013 5:09 PM To: [hidden email] Subject: Re: Brightness difference Hg vs LED A quick comparison in our lab gave the following: 488 nm (FITC) band of an LED-based light engine (Lumencor): 40 mW 300 W Xenon Sutter illuminator (probably our brightest light source) with FITC EX filter: 70 mW both measured at the exit of the fiber or light guide. I agree that LED based light engines are probably at least as good (power-wise) as most lamps, plus they have other benefits. Charts such as the one found here should be a good guide regarding relative power: http://lumencor.com/wp-content/uploads/2011/09/SpectraX-LE-User-Manual1.pdf Julio Vazquez, PhD Fred Hutchinson Cancer Research Center Seattle, WA 98109 http://www.fhcrc.org/en.html ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
George McNamara |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Phil, Current LED light sources can be brighter and (should have) more stable light output (and "instant" on/off, and less heat output and less ozone and no chance of the bulb exploding ... "do not look at top of arc lamp with remaining eye". Also many LEDs have precise - and reproducible - voltage control. Purchase price will eventually be made up in total cost of ownership. Brighter light sources enable selection of narrower wavelength range, for example, at the excitation peak of the desired fluorophore (and hopefully minima of unwanted fluorophores), leaving more room for emission wavelength range. George On 11/5/2013 11:56 AM, Philip Oshel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > All, > > I had this question put to me by a new faculty member, and don't have > a ready answer: > "Is there a ballpark percentage for how much less bright an LED vs a > standard mercury lamp light?" > This is for regular epifluorescence, not confocal. > > This is in the realm of arm-waving over a picture of beer (a good, > dark stout), ignoring brands, how old the Hg bulb is, ex/em cubes, > which part of the spectrum is used, and all that. Personally, I'd > think the answer is more like, "Doesn't matter, the dimmer system is > still too bright to use all the available light and not damage the > specimen." But ... ? > > Phil -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/26/ |
|
Mark Cannell-2 |
|
|
In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Almost none, the Sutter uses common narrow band interference filters that are well blocked outside the pass band. As you raised this question, I wonder how much heat comes from a LED source -the chip gets very hot in high power applications and about half the power consumption of a LED is waste heat… Cheers M On 6/11/2013, at 1:48 am, Cammer, Michael <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > How much of the 70 mW is from UV or IR leaking through? > _________________________________________ > Michael Cammer, Assistant Research Scientist > Skirball Institute of Biomolecular Medicine > Lab: (212) 263-3208 Cell: (914) 309-3270 > > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on behalf of Julio Vazquez [[hidden email]] > Sent: Tuesday, November 05, 2013 5:09 PM > To: [hidden email] > Subject: Re: Brightness difference Hg vs LED > > > A quick comparison in our lab gave the following: > > 488 nm (FITC) band of an LED-based light engine (Lumencor): 40 mW > 300 W Xenon Sutter illuminator (probably our brightest light source) with FITC EX filter: 70 mW > > both measured at the exit of the fiber or light guide. > > I agree that LED based light engines are probably at least as good (power-wise) as most lamps, plus they have other benefits. Charts such as the one found here should be a good guide regarding relative power: > > http://lumencor.com/wp-content/uploads/2011/09/SpectraX-LE-User-Manual1.pdf > > > Julio Vazquez, PhD > Fred Hutchinson Cancer Research Center > Seattle, WA 98109 > > http://www.fhcrc.org/en.html > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. > ================================= Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Watkins, Simon C |
|
|
In reply to this post by Julio Vazquez
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We actually did a lot of work on this a year or so back. Specifically with the Lumencor Spectra X. There is a lot more power with these newer diode sources. These numbers are using Chroma filter sets, the effort was spent because we wanted to maximize speed (see our PLOS one paper last year on imaging danio blood flow). Essentially the folks at Chroma came up with a design which only involves two cubes, a dapi/fitc(GFP)/tritc/cy5 cube and a cyan/yellow/red fp cube. So as the exciter filters are part of the light source we could maximize switching speed for most expts (no cube change), using triggered acquisition and a sub arrayed CMOS detector this can be hundreds of multicolor images/second All powers are measured at the end of the light guide with a Coherent power meter. The open power is total light coming down the guide with no filter in place, filtered is self explanatory as is the filter wavelength, powers are total mW not/unit area. Hope this helps Nominal wavelength Open Power filtered Filter Wavelength 390nm 220 185 395/25 440nm 322 190 440/20 475nm 191 147 470/24 514nm 91 70 508/24 550nm 366 220 550/15 575nm 800 230 575/22 632nm 265 208 640/30 Simon Watkins Ph.D Professor and Vice Chair Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Bsts 225 3550 terrace st Pittsburgh PA 15261 Www.cbi.pitt.edu <http://Www.cbi.pitt.edu/> 412-352-2277 On 11/5/13, 5:09 PM, "Julio Vazquez" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >A quick comparison in our lab gave the following: > >488 nm (FITC) band of an LED-based light engine (Lumencor): 40 mW >300 W Xenon Sutter illuminator (probably our brightest light source) with >FITC EX filter: 70 mW > >both measured at the exit of the fiber or light guide. > >I agree that LED based light engines are probably at least as good >(power-wise) as most lamps, plus they have other benefits. Charts such as >the one found here should be a good guide regarding relative power: > >http://lumencor.com/wp-content/uploads/2011/09/SpectraX-LE-User-Manual1.pd >f > > >Julio Vazquez, PhD >Fred Hutchinson Cancer Research Center >Seattle, WA 98109 > >http://www.fhcrc.org/en.html > > > > >On Nov 5, 2013, at 1:50 PM, Kurt Thorn wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> My sense is also that the new LED sources are brighter than Hg lamps, >>except possibly at the very brightest mercury peaks (365, 546 nm). For >>GFP, they are for sure brighter. True LEDs have been somewhat dim in >>the 560 nm range, which is why Lumencor uses an LED-pumped phosphor for >>those wavelengths. This gives very bright 560 emission. >> >> Kurt >> >> On 11/5/2013 10:21 AM, Armstrong, Brian wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi Phil I would say that the new LED sources and Light Engines will be >>>brighter than mercury sources. For example I compared an Exfo to a SOLA >>>light engine and the light engine is brighter. The new Lumen Dynamics >>>XLED should provide much better illumination than a mercury source. >>> Hopefully, mercury sources will soon be a thing of the past. >>> Cheers, >>> >>> Brian D Armstrong PhD >>> Associate Research Professor >>> Director, Light Microscopy Core >>> Beckman Research Institute >>> City of Hope >>> Dept of Neuroscience >>> 1450 E Duarte Rd >>> Duarte, CA 91010 >>> 626-256-4673 x62872 >>> >>> >>> >>> -----Original Message----- >>> From: Confocal Microscopy List >>>[mailto:[hidden email]] On Behalf Of Philip Oshel >>> Sent: Tuesday, November 05, 2013 9:57 AM >>> To: [hidden email] >>> Subject: Brightness difference Hg vs LED >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> All, >>> >>> I had this question put to me by a new faculty member, and don't have a >>> ready answer: >>> "Is there a ballpark percentage for how much less bright an LED vs a >>> standard mercury lamp light?" >>> This is for regular epifluorescence, not confocal. >>> >>> This is in the realm of arm-waving over a picture of beer (a good, dark >>> stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which >>>part >>> of the spectrum is used, and all that. Personally, I'd think the answer >>> is more like, "Doesn't matter, the dimmer system is still too bright to >>> use all the available light and not damage the specimen." But ... ? >>> >>> Phil |
|
James Pawley |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, Lots of good answers already but I think that it is important to remember that Hg has a lot of 10x peaks in the visible part of its spectrum. So an answer with one bandpass filter may not indicate a general trend. One also needs to know a bit about the optics used. Critical illumination (source focused onto the image plane) will usually be brighter than Kohler, but whereas critical illumination that focuses the brightest "plasma ball" of the arc into the centre of the imaged area can be relatively uniform over a small field-of-view, it is not so clear what will happen when one images an LED into the image plane (It will depend on the construction of the LED). (Note: The actual ball of a 100 w Hg source is usually about 150µm in diam. just near the electrode. Assuming both the collector lens and the objective are of about the same, high NA, the most efficient optics to convey this to the focus plane will do so at a magnification of 1:1. Of course the field of view of the objective (in µm) will vary inversely with its magnification, but the area that can be properly sampled with a given CCD/sCMOS will not vary: a 1000x1000 array of 0.1µm pixels will be about 140µm across its diagonal). However each manufacturer has made different compromises in terms of the magnification of their epi-illumination system (at very least to also accommodate low-magnification use) and some may utilize more of the light leaving the original light source (arc or fibre) than others. The idea of measuring the output at the microscope end of a fibre-optic seems sensible as long as this is how you will illuminate your sample on your exact setup. However, such fibers have an NA (angle at which the light leaves them) and so not all the light leaving them necessarily makes it to the image plane. For instance, a given system may under- or overfill the entrance pupil of a given objective. As one can never make the light brighter (in photons/second/µm*2) simply by focusing it, if the end of the fibre is much larger than 150 µm (in the example above) then some of the light leaving it must inevitably be lost somewhere before it reaches that part of the focus plane covered by the 1000x1000 image sensor. My preference would be to measure the light leaving the objective once the field diaphragm has been set to repeatable diameter (say 100µm diam. at the image plane). Of course, you can only set the diaphragm properly if the scope is set to Kohler. Once it is set (to standardize the light path to that point) you can still tweak and condenser focus to approximate critical illumination. (i.e., make the image as bright as possible). The adjustment is inevitably a bit of a "fudge" because, as the arc is a 3D source rather than the planar object imagined by the Kohler Illumination diagram, it "cannot be focused into a plane". Of course, there is still a problem: You probably want to use a hi-NA objective but above NA 0.5 more and more of the high-NA light will reflect back into the objective from its front surface. Rays at >NA 1.0 will not escape into the air at all. Efforts to couple the sensor of your photometer to the objective will a drop of immersion oil will only work if there is no air-gap between the sensor window and the sensitive element. The options are: 1) To couple a small, strong plano-convex lens onto the back of the microscope slide with immersion oil to make the light beam less divergent or 2) To set up using an NA 0.75 air objective and hope that the difference in NA isn't too important or 3) Measure the fluorescent light signal at the CCD from a thin, uniform layer of fluorescent dye (It should be thin so that you don't end up focusing too far inside the layer, where SA and absorption may be variables you don't want). So now you see why just measuring the output of the fibre seems easier. Hope that this isn't too confusing. I am really theoretically very pro-LED (faster, cooler, just the light you want etc). Indeed, I think that Chapter 3 in the Handbook was one of the first places where LED microscope sources were discussed in any depth. I would just like to see a few more variables nailed down. Cheers, Jim Pawley > > >Hi Phil, > >Current LED light sources can be brighter and >(should have) more stable light output (and >"instant" on/off, and less heat output and less >ozone and no chance of the bulb exploding ... >"do not look at top of arc lamp with remaining >eye". Also many LEDs have precise - and >reproducible - voltage control. Purchase price >will eventually be made up in total cost of >ownership. > >Brighter light sources enable selection of >narrower wavelength range, for example, at the >excitation peak of the desired fluorophore (and >hopefully minima of unwanted fluorophores), >leaving more room for emission wavelength range. > >George > >On 11/5/2013 11:56 AM, Philip Oshel wrote: >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>***** >> >>All, >> >>I had this question put to me by a new faculty >>member, and don't have a ready answer: >>"Is there a ballpark percentage for how much >>less bright an LED vs a standard mercury lamp >>light?" >>This is for regular epifluorescence, not confocal. >> >>This is in the realm of arm-waving over a >>picture of beer (a good, dark stout), ignoring >>brands, how old the Hg bulb is, ex/em cubes, >>which part of the spectrum is used, and all >>that. Personally, I'd think the answer is more >>like, "Doesn't matter, the dimmer system is >>still too bright to use all the available light >>and not damage the specimen." But ... ? >> >>Phil > > >-- > > > >George McNamara, Ph.D. >Single Cells Analyst >L.J.N. Cooper Lab >University of Texas M.D. Anderson Cancer Center >Houston, TX 77054 >Tattletales http://works.bepress.com/gmcnamara/26/ -- James and Christine Pawley, 5446 Burley Place (PO Box 2348), Sechelt, BC, Canada, V0N3A0, Phone 604-885-0840, email <[hidden email]> NEW! NEW! AND DIFFERENT Cell (when I remember to turn it on!) 1-604-989-6146 |
|
Claire Brown |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We have been testing a number of the solid state light sources. We do a lot of live cell imaging and they are all way too bright for live cells. We either add more ND filters so the cells don't die or we turn them way down in % output if that is an option on the light source. The companies seem to be making them brighter and brighter. I would love to know what applications people have that need all this light? We usually advise people to use less light and longer exposure times even with fixed samples. That being said we were able to get institutional support to replace our mercury sources as part of a sustainability program at McGill in a Mercury Free Microscopy initiative. This will save us lots of time and money in replacing the mercury bulbs all the time in addition to a reduction in mercury. Depending on how the sources are used they can also save a lot in power consumption too. We have been more interested in stability of the light sources on different time scales and they really perform well. Variability is in the single percentage range or less on most time scales. I would be happy to provide more information to anyone offline if they want to contact me. Sincerely, Claire |
|
Gerard Whoriskey |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Phil and others, Answering this question can be a bit complex as there is so much variation in power across the Hg spectrum. There is also some variation across the LED/solid state hybrid spectrum through it is not so extreme. In general Hg wins on the sharp peaks but for some time LEDs have won in the Hg troughs, the main one being between 435nm and 546nm. In most cases users should now be getting similar results to a new Hg across the spectrum and more than adequate power for most general wide-field fluorescence work. If we consider the spectrum split up into 4 bands, violet, blue, green-yellow and red-NIR we can explain comparisons in a bit more detail. LEDs in the blue from 440-500nm have been stronger than the Hg for quite some time as in the red-NIR region. Both these regions are in Hg troughs. The metal halide bulbs have done a good job filling up the blue trough and its maybe only over the last couple of years that LEDs are now beating the metal halide in this region. The violet region which I would put between 365-440nm has been adequate for some time, most people commenting that DAPI is always too bright. There has been a lot of progress on power over this region with 365nm now being more than good enough in most cases. The green-yellow region has been the most problematic area. Lumencor were the first to deliver good power in this region and stood out for some time with the phosphor rod technology they use to cover this region. Other LED technologies have caught up on this now, so bright green should now be expected from other vendors. At CoolLED we have seen our green-yellow region improve by as much as 6 times over the past year. This is all great news however there still can be disappointment from LED systems and I believe this is down to the limited spectrum. This means that if filters and LED peaks do not match up then results will be poor. With bulbs users did not need to be so conscious of filter sets but with LEDs it is crucial. For example a user may purchase a broad spectrum LED/solid state light source to find that no DAPI is visible as no 365nm is available. In such cases a new filter set is needed. You may also specify an LED system with 460nm or 470nm that is great for GFP with most cubes. This can fall down however when a multiband filter set is used which pushes the GFP excitation passband up to around 490nm. Ideally one would like both the 470nm and 490nm included in the LED system. Addressing this problem has been a major issue for LED system manufacturers. Some get around this by offering swappable LED modules but this brings other issues like swapping time and ensuring combining dichroics are correct. Just adding more wavelengths is another option but this has a big overhead with LED systems as each new source requires its own drive electronics, collimating optics and combining dichroics. Adding more wavelengths in the conventional way has an increasingly negative impact on overall power mainly due to the dichroic losses. This has limited the number of peaks in a system typically to around 4 but up to 6 or 7 at most. The good news is we have now addressed this limitation with a novel approach at CoolLED and will be showing our 16 selectable wavelength system, 365- 770nm, at Neuroscience next week. Im happy to explain more on this to anyone that is interested. Kind regards, Gerry Technical Manager CoolLED |
|
Rosemary.White |
|
|
In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes! I agree! These sources are very bright, even the lowest setting is usually too much for our fluorochromes and can even damage plant cells. We also had to get ND filters. And ditto, stability and lifetime are the two key issues, stability more than anything. Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 M 61 2 420 972 028 E [hidden email] On 7/11/13 2:40 AM, "Claire Brown" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >We have been testing a number of the solid state light sources. We do a >lot >of live cell imaging and they are all way too bright for live cells. We >either add more ND filters so the cells don't die or we turn them way down >in % output if that is an option on the light source. The companies seem >to >be making them brighter and brighter. I would love to know what >applications >people have that need all this light? We usually advise people to use less >light and longer exposure times even with fixed samples. > >That being said we were able to get institutional support to replace our >mercury sources as part of a sustainability program at McGill in a Mercury >Free Microscopy initiative. This will save us lots of time and money in >replacing the mercury bulbs all the time in addition to a reduction in >mercury. Depending on how the sources are used they can also save a lot in >power consumption too. > >We have been more interested in stability of the light sources on >different >time scales and they really perform well. Variability is in the single >percentage range or less on most time scales. I would be happy to provide >more information to anyone offline if they want to contact me. > >Sincerely, > >Claire |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello to all, I work for Prior Scientific but this is not meant as a commercial response. It is an offer to loan equipment if someone wanted to collect data with it and share it with this list community. If someone (should be a non-commercial entity) wanted to spend some time employing a method of testing as the full illumination system supplied from the manufacturer along with the recommended filter sets etc. and not just the particular light source standing on its own, I would be willing to help by offering to loan our spectral power meter and software to them. We purpose developed this tool to characterize the light exiting the objective lens for these very reasons. (The data sheet for our LumaSpec 800 is online at our website if you want to look at it.) This thread has several great comments, information, and suggestions. I certainly agree that measuring off the scope has meaning, but ultimately the characterization of brightness (and other factors such as stability) to eventually choose one illuminator over the other for the person's application ought to be determined on the microscope tested as a system. You could have the most powerful, most versatile illuminator on the planet on the bench right behind your microscope but it might not perform as you expect. In other words a car with 600 horsepower and square wheels doesn't get you very far. Something as simple as the optical coupler / adapter to the microscope can attribute to light loss and have a huge effect on the final output exiting your chosen objective lens. There are a host of other things that can also play a role that have been mentioned by others such as simple alignment issues, incorrect/misaligned/ unmatched filter sets etc. so what begins outside the scope eventually changes quite a bit on its way to your sample for a variety of reasons. So if you want to give this a whirl please let me know and drop me an email. I will also be at Neuroscience next week (and Cell Bio next month as well) in our booth if you would like to speak to me directly about this. Thanks and Regards, Dennis Dennis Doherty National Sales Manager Prior Scientific, Inc. 80 Reservoir Park Drive Rockland, MA 02370 [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Wednesday, November 06, 2013 3:00 PM To: [hidden email] Subject: Re: Brightness difference Hg vs LED ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes! I agree! These sources are very bright, even the lowest setting is usually too much for our fluorochromes and can even damage plant cells. We also had to get ND filters. And ditto, stability and lifetime are the two key issues, stability more than anything. Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 M 61 2 420 972 028 E [hidden email] On 7/11/13 2:40 AM, "Claire Brown" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >We have been testing a number of the solid state light sources. We do a >lot of live cell imaging and they are all way too bright for live >cells. We either add more ND filters so the cells don't die or we turn >them way down in % output if that is an option on the light source. The >companies seem to be making them brighter and brighter. I would love to >know what applications people have that need all this light? We usually >advise people to use less light and longer exposure times even with >fixed samples. > >That being said we were able to get institutional support to replace >our mercury sources as part of a sustainability program at McGill in a >Mercury Free Microscopy initiative. This will save us lots of time and >money in replacing the mercury bulbs all the time in addition to a >reduction in mercury. Depending on how the sources are used they can >also save a lot in power consumption too. > >We have been more interested in stability of the light sources on >different time scales and they really perform well. Variability is in >the single percentage range or less on most time scales. I would be >happy to provide more information to anyone offline if they want to >contact me. > >Sincerely, > >Claire This email and any files transmitted with it are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you have received this email in error please notify the sender immediately and delete this e-mail. Please note that any views or opinions presented in this email are solely those of the author and do not necessarily represent those of the company. WARNING: Although the company has taken reasonable precautions to ensure no viruses are present in this email, the company cannot accept responsibility for any loss or damage arising from the use of this email or attachments. |
|
Benjamin Freiberg |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** All, Having been involved in the lighting industry from a commercial standpoint I wanted to respond to this thread in an unbiased and independent experienced voice. As has been previously mentioned, there are many areas of the spectrum in which solid-state solutions have surpassed HBO in intensity at the sample plane. I completely agree with the assessments that there are many components that go into building an illuminator and therefore many sources of light loss in the systems on the market. Key to any measurement of power is how much light is delivered to the sample, so if anyone does take up the offer to compare light sources, measurement of mW//cm^2 at the sample plane will provide the best indication of the total performance of a light source on a microscope. That being said, one must make sure that the excitation filters used with solid-state sources are matched to the peak of the source and not simply use off the shelf filters and expect them to perform optimally. There has been mention in this thread of the gaps produced when using multiple sources and combining them into a shared collimated beam. As has been stated, in general, increasing the number of input sources decreases the overall throughput of light in the system as source combining optics can be lossy and will cut the overall intensity of any single source that could be coupled to the microscope. One solution that has not been mentioned in this thread is the 120-LED from Lumen Dynamics. The 120 LED overcomes the limits of combining multiple sources by using a high intensity white LED source. This product, according to the LDGI website, covers wavelengths from 370nm all the way out to 700+ nm with reasonable intensity. While HBO may be brighter at some wavelengths versus the 120-LED, rarely are these sources, HBO or solid-state, used at max power especially in live cell applications. I would suggest that the 120-LED be included in whatever tests come from this thread as it is a unique product from the standpoint of how the white light is generated for fluorescence applications versus the other solutions on the market that combine many sources. As for the overall performance of all solid state products mentioned thus far in this thread, one should consider the following: every mid to high end solid state source produces multiple times more power than 175W xenon at the sample plane at all wavelengths greater than 360nm and all the way out to the NIR. If anyone has experience using xenon as their primary source, I hope this puts the current state of solid-state illuminator intensity into perspective. Bottom line is for most applications, the benefits in cost of ownership, stability and overall function have the solid-state solutions greatly out-perform their HBO counterparts with an additional benefit being that solid-state sources are a green solution and they eliminate the threat of mercury contamination due to improper disposal or explosion of mercury burners. Ben On Tue, 5 Nov 2013 12:56:42 -0500, Philip Oshel <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >All, > >I had this question put to me by a new faculty member, and don't have a >ready answer: >"Is there a ballpark percentage for how much less bright an LED vs a >standard mercury lamp light?" >This is for regular epifluorescence, not confocal. > >This is in the realm of arm-waving over a picture of beer (a good, dark >stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part >of the spectrum is used, and all that. Personally, I'd think the answer >is more like, "Doesn't matter, the dimmer system is still too bright to >use all the available light and not damage the specimen." But ... ? > >Phil >-- >Philip Oshel >Microscopy Facility Supervisor >Biology Department >024C Brooks Hall >Central Michigan University >Mt. Pleasant, MI 48859 >(989) 774-3576 |
|
|
In reply to this post by James Pawley
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** For once I pretty much totally agree with Jim! It is just impossible to compare something as peaky as the Hg spectrum with a continuum source. If one of the Hg peaks matches the excitation you need - Bingo! Hg is fantastic in the near UV at 360, and even better in the violet at ~405 and ~470. It is pathetic in the blue (480-500) but brilliant at 550. Obviously it isn't as specific as a laser source, but you still have to think hard about where your peaks are. LED sources are much broader, and therefore more versatile. As for critical illumination - can you actually do it with a commercial epi illuminator? Guy Guy Cox, Honorary Associate Professor School of Medical Sciences Australian Centre for Microscopy and Microanalysis, Madsen, F09, University of Sydney, NSW 2006 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of James Pawley Sent: Thursday, 7 November 2013 1:06 AM To: [hidden email] Subject: Re: Brightness difference Hg vs LED ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, Lots of good answers already but I think that it is important to remember that Hg has a lot of 10x peaks in the visible part of its spectrum. So an answer with one bandpass filter may not indicate a general trend. One also needs to know a bit about the optics used. Critical illumination (source focused onto the image plane) will usually be brighter than Kohler, but whereas critical illumination that focuses the brightest "plasma ball" of the arc into the centre of the imaged area can be relatively uniform over a small field-of-view, it is not so clear what will happen when one images an LED into the image plane (It will depend on the construction of the LED). (Note: The actual ball of a 100 w Hg source is usually about 150µm in diam. just near the electrode. Assuming both the collector lens and the objective are of about the same, high NA, the most efficient optics to convey this to the focus plane will do so at a magnification of 1:1. Of course the field of view of the objective (in µm) will vary inversely with its magnification, but the area that can be properly sampled with a given CCD/sCMOS will not vary: a 1000x1000 array of 0.1µm pixels will be about 140µm across its diagonal). However each manufacturer has made different compromises in terms of the magnification of their epi-illumination system (at very least to also accommodate low-magnification use) and some may utilize more of the light leaving the original light source (arc or fibre) than others. The idea of measuring the output at the microscope end of a fibre-optic seems sensible as long as this is how you will illuminate your sample on your exact setup. However, such fibers have an NA (angle at which the light leaves them) and so not all the light leaving them necessarily makes it to the image plane. For instance, a given system may under- or overfill the entrance pupil of a given objective. As one can never make the light brighter (in photons/second/µm*2) simply by focusing it, if the end of the fibre is much larger than 150 µm (in the example above) then some of the light leaving it must inevitably be lost somewhere before it reaches that part of the focus plane covered by the 1000x1000 image sensor. My preference would be to measure the light leaving the objective once the field diaphragm has been set to repeatable diameter (say 100µm diam. at the image plane). Of course, you can only set the diaphragm properly if the scope is set to Kohler. Once it is set (to standardize the light path to that point) you can still tweak and condenser focus to approximate critical illumination. (i.e., make the image as bright as possible). The adjustment is inevitably a bit of a "fudge" because, as the arc is a 3D source rather than the planar object imagined by the Kohler Illumination diagram, it "cannot be focused into a plane". Of course, there is still a problem: You probably want to use a hi-NA objective but above NA 0.5 more and more of the high-NA light will reflect back into the objective from its front surface. Rays at >NA 1.0 will not escape into the air at all. Efforts to couple the sensor of your photometer to the objective will a drop of immersion oil will only work if there is no air-gap between the sensor window and the sensitive element. The options are: 1) To couple a small, strong plano-convex lens onto the back of the microscope slide with immersion oil to make the light beam less divergent or 2) To set up using an NA 0.75 air objective and hope that the difference in NA isn't too important or 3) Measure the fluorescent light signal at the CCD from a thin, uniform layer of fluorescent dye (It should be thin so that you don't end up focusing too far inside the layer, where SA and absorption may be variables you don't want). So now you see why just measuring the output of the fibre seems easier. Hope that this isn't too confusing. I am really theoretically very pro-LED (faster, cooler, just the light you want etc). Indeed, I think that Chapter 3 in the Handbook was one of the first places where LED microscope sources were discussed in any depth. I would just like to see a few more variables nailed down. Cheers, Jim Pawley > > >Hi Phil, > >Current LED light sources can be brighter and (should have) more stable >light output (and "instant" on/off, and less heat output and less ozone >and no chance of the bulb exploding ... >"do not look at top of arc lamp with remaining eye". Also many LEDs >have precise - and reproducible - voltage control. Purchase price will >eventually be made up in total cost of ownership. > >Brighter light sources enable selection of narrower wavelength range, >for example, at the excitation peak of the desired fluorophore (and >hopefully minima of unwanted fluorophores), leaving more room for >emission wavelength range. > >George > >On 11/5/2013 11:56 AM, Philip Oshel wrote: >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>***** >> >>All, >> >>I had this question put to me by a new faculty member, and don't have >>a ready answer: >>"Is there a ballpark percentage for how much less bright an LED vs a >>standard mercury lamp light?" >>This is for regular epifluorescence, not confocal. >> >>This is in the realm of arm-waving over a picture of beer (a good, >>dark stout), ignoring brands, how old the Hg bulb is, ex/em cubes, >>which part of the spectrum is used, and all that. Personally, I'd >>think the answer is more like, "Doesn't matter, the dimmer system is >>still too bright to use all the available light and not damage the >>specimen." But ... ? >> >>Phil > > >-- > > > >George McNamara, Ph.D. >Single Cells Analyst >L.J.N. Cooper Lab >University of Texas M.D. Anderson Cancer Center Houston, TX 77054 >Tattletales http://works.bepress.com/gmcnamara/26/ -- James and Christine Pawley, 5446 Burley Place (PO Box 2348), Sechelt, BC, Canada, V0N3A0, Phone 604-885-0840, email <[hidden email]> NEW! NEW! AND DIFFERENT Cell (when I remember to turn it on!) 1-604-989-6146 |
|
Judy Trogadis-2 |
|
|
In reply to this post by Benjamin Freiberg
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All This posting of a couple of weeks ago has suddenly become relevant for us. Does anyone know of a powerful continuous white light source in the 600-800nm range? Emphasis on the word 'powerful' because in most light sources there seems to be a steep drop in power above 650nm. Thank you, Judy Judy Trogadis Bio-Imaging Coordinator Keenan Research Centre, St. Michael's 209 Victoria Street Toronto M5B 1T8, Canada office: 416-864-6060 ext. 77612 imaging facility: ext. 77434 cell: 416-254-9330 [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ben Freiberg Sent: Thursday, November 07, 2013 10:59 PM To: [hidden email] Subject: Re: Brightness difference Hg vs LED ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** All, Having been involved in the lighting industry from a commercial standpoint I wanted to respond to this thread in an unbiased and independent experienced voice. As has been previously mentioned, there are many areas of the spectrum in which solid-state solutions have surpassed HBO in intensity at the sample plane. I completely agree with the assessments that there are many components that go into building an illuminator and therefore many sources of light loss in the systems on the market. Key to any measurement of power is how much light is delivered to the sample, so if anyone does take up the offer to compare light sources, measurement of mW//cm^2 at the sample plane will provide the best indication of the total performance of a light source on a microscope. That being said, one must make sure that the excitation filters used with solid-state sources are matched to the peak of the source and not simply use off the shelf filters and expect them to perform optimally. There has been mention in this thread of the gaps produced when using multiple sources and combining them into a shared collimated beam. As has been stated, in general, increasing the number of input sources decreases the overall throughput of light in the system as source combining optics can be lossy and will cut the overall intensity of any single source that could be coupled to the microscope. One solution that has not been mentioned in this thread is the 120-LED from Lumen Dynamics. The 120 LED overcomes the limits of combining multiple sources by using a high intensity white LED source. This product, according to the LDGI website, covers wavelengths from 370nm all the way out to 700+ nm with reasonable intensity. While HBO may be brighter at some wavelengths versus the 120-LED, rarely are these sources, HBO or solid-state, used at max power especially in live cell applications. I would suggest that the 120-LED be included in whatever tests come from this thread as it is a unique product from the standpoint of how the white light is generated for fluorescence applications versus the other solutions on the market that combine many sources. As for the overall performance of all solid state products mentioned thus far in this thread, one should consider the following: every mid to high end solid state source produces multiple times more power than 175W xenon at the sample plane at all wavelengths greater than 360nm and all the way out to the NIR. If anyone has experience using xenon as their primary source, I hope this puts the current state of solid-state illuminator intensity into perspective. Bottom line is for most applications, the benefits in cost of ownership, stability and overall function have the solid-state solutions greatly out-perform their HBO counterparts with an additional benefit being that solid-state sources are a green solution and they eliminate the threat of mercury contamination due to improper disposal or explosion of mercury burners. Ben On Tue, 5 Nov 2013 12:56:42 -0500, Philip Oshel <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >All, > >I had this question put to me by a new faculty member, and don't have a >ready answer: >"Is there a ballpark percentage for how much less bright an LED vs a >standard mercury lamp light?" >This is for regular epifluorescence, not confocal. > >This is in the realm of arm-waving over a picture of beer (a good, dark >stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part >of the spectrum is used, and all that. Personally, I'd think the answer >is more like, "Doesn't matter, the dimmer system is still too bright to >use all the available light and not damage the specimen." But ... ? > >Phil >-- >Philip Oshel >Microscopy Facility Supervisor >Biology Department >024C Brooks Hall >Central Michigan University >Mt. Pleasant, MI 48859 >(989) 774-3576 |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** What sort of power are you talking about? Is a mW per nm 'lots' by the standards of what you want? Craig On Fri, Nov 22, 2013 at 8:33 AM, Judy Trogadis <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All > > This posting of a couple of weeks ago has suddenly become relevant for us. > > Does anyone know of a powerful continuous white light source in the > 600-800nm range? Emphasis on the word 'powerful' because in most light > sources there seems to be a steep drop in power above 650nm. > > Thank you, > Judy > > Judy Trogadis > Bio-Imaging Coordinator > Keenan Research Centre, St. Michael's > 209 Victoria Street > Toronto M5B 1T8, Canada > office: 416-864-6060 ext. 77612 > imaging facility: ext. 77434 > cell: 416-254-9330 > [hidden email] > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Ben Freiberg > Sent: Thursday, November 07, 2013 10:59 PM > To: [hidden email] > Subject: Re: Brightness difference Hg vs LED > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > All, > > Having been involved in the lighting industry from a commercial standpoint > I > wanted to respond to this thread in an unbiased and independent experienced > voice. > > As has been previously mentioned, there are many areas of the spectrum in > which solid-state solutions have surpassed HBO in intensity at the sample > plane. I completely agree with the assessments that there are many > components that go into building an illuminator and therefore many sources > of light loss in the systems on the market. Key to any measurement of > power > is how much light is delivered to the sample, so if anyone does take up the > offer to compare light sources, measurement of mW//cm^2 at the sample plane > will provide the best indication of the total performance of a light source > on a microscope. That being said, one must make sure that the excitation > filters used with solid-state sources are matched to the peak of the source > and not simply use off the shelf filters and expect them to perform > optimally. > > There has been mention in this thread of the gaps produced when using > multiple sources and combining them into a shared collimated beam. As has > been stated, in general, increasing the number of input sources decreases > the overall throughput of light in the system as source combining optics > can > be lossy and will cut the overall intensity of any single source that could > be coupled to the microscope. One solution that has not been mentioned in > this thread is the 120-LED from Lumen Dynamics. The 120 LED overcomes the > limits of combining multiple sources by using a high intensity white LED > source. This product, according to the LDGI website, covers wavelengths > from 370nm all the way out to 700+ nm with reasonable intensity. While HBO > may be brighter at some wavelengths versus the 120-LED, rarely are these > sources, HBO or solid-state, used at max power especially in live cell > applications. > > I would suggest that the 120-LED be included in whatever tests come from > this thread as it is a unique product from the standpoint of how the white > light is generated for fluorescence applications versus the other solutions > on the market that combine many sources. > > As for the overall performance of all solid state products mentioned thus > far in this thread, one should consider the following: every mid to high > end > solid state source produces multiple times more power than 175W xenon at > the > sample plane at all wavelengths greater than 360nm and all the way out to > the NIR. If anyone has experience using xenon as their primary source, I > hope this puts the current state of solid-state illuminator intensity into > perspective. Bottom line is for most applications, the benefits in cost > of > ownership, stability and overall function have the solid-state solutions > greatly out-perform their HBO counterparts with an additional benefit being > that solid-state sources are a green solution and they eliminate the threat > of mercury contamination due to improper disposal or explosion of mercury > burners. > > Ben > > > > > > > > > > > > > > > > On Tue, 5 Nov 2013 12:56:42 -0500, Philip Oshel <[hidden email]> > wrote: > > >***** > >To join, leave or search the confocal microscopy listserv, go to: > >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >***** > > > >All, > > > >I had this question put to me by a new faculty member, and don't have a > >ready answer: > >"Is there a ballpark percentage for how much less bright an LED vs a > >standard mercury lamp light?" > >This is for regular epifluorescence, not confocal. > > > >This is in the realm of arm-waving over a picture of beer (a good, dark > >stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part > >of the spectrum is used, and all that. Personally, I'd think the answer > >is more like, "Doesn't matter, the dimmer system is still too bright to > >use all the available light and not damage the specimen." But ... ? > > > >Phil > >-- > >Philip Oshel > >Microscopy Facility Supervisor > >Biology Department > >024C Brooks Hall > >Central Michigan University > >Mt. Pleasant, MI 48859 > >(989) 774-3576 > |
|
Kurt Thorn |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We set up such a system recently using a xenon arc lamp modified for long-wavelength output from Sutter. At 740 nm we get about 50 mW out of a 10x objective in a 13 nm bandwidth. I don't know if that qualifies as powerful or not, but for epi-fluorescence imaging, it's pretty bright. More details on my blog: http://nic.ucsf.edu/blog/?p=813 On a related note, does anyone have some numbers for the power densities required to saturate common fluorophores? I've been curious about using LED strobe illumination for very high speed imaging and I'm wondering how high you can push the power density before you saturate the dyes. Kurt On 11/22/2013 7:33 AM, Judy Trogadis wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All > > This posting of a couple of weeks ago has suddenly become relevant for us. > > Does anyone know of a powerful continuous white light source in the 600-800nm range? Emphasis on the word 'powerful' because in most light sources there seems to be a steep drop in power above 650nm. > > Thank you, > Judy > > Judy Trogadis > Bio-Imaging Coordinator > Keenan Research Centre, St. Michael's > 209 Victoria Street > Toronto M5B 1T8, Canada > office: 416-864-6060 ext. 77612 > imaging facility: ext. 77434 > cell: 416-254-9330 > [hidden email] > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ben Freiberg > Sent: Thursday, November 07, 2013 10:59 PM > To: [hidden email] > Subject: Re: Brightness difference Hg vs LED > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > All, > > Having been involved in the lighting industry from a commercial standpoint I > wanted to respond to this thread in an unbiased and independent experienced > voice. > > As has been previously mentioned, there are many areas of the spectrum in > which solid-state solutions have surpassed HBO in intensity at the sample > plane. I completely agree with the assessments that there are many > components that go into building an illuminator and therefore many sources > of light loss in the systems on the market. Key to any measurement of power > is how much light is delivered to the sample, so if anyone does take up the > offer to compare light sources, measurement of mW//cm^2 at the sample plane > will provide the best indication of the total performance of a light source > on a microscope. That being said, one must make sure that the excitation > filters used with solid-state sources are matched to the peak of the source > and not simply use off the shelf filters and expect them to perform optimally. > > There has been mention in this thread of the gaps produced when using > multiple sources and combining them into a shared collimated beam. As has > been stated, in general, increasing the number of input sources decreases > the overall throughput of light in the system as source combining optics can > be lossy and will cut the overall intensity of any single source that could > be coupled to the microscope. One solution that has not been mentioned in > this thread is the 120-LED from Lumen Dynamics. The 120 LED overcomes the > limits of combining multiple sources by using a high intensity white LED > source. This product, according to the LDGI website, covers wavelengths > from 370nm all the way out to 700+ nm with reasonable intensity. While HBO > may be brighter at some wavelengths versus the 120-LED, rarely are these > sources, HBO or solid-state, used at max power especially in live cell > applications. > > I would suggest that the 120-LED be included in whatever tests come from > this thread as it is a unique product from the standpoint of how the white > light is generated for fluorescence applications versus the other solutions > on the market that combine many sources. > > As for the overall performance of all solid state products mentioned thus > far in this thread, one should consider the following: every mid to high end > solid state source produces multiple times more power than 175W xenon at the > sample plane at all wavelengths greater than 360nm and all the way out to > the NIR. If anyone has experience using xenon as their primary source, I > hope this puts the current state of solid-state illuminator intensity into > perspective. Bottom line is for most applications, the benefits in cost of > ownership, stability and overall function have the solid-state solutions > greatly out-perform their HBO counterparts with an additional benefit being > that solid-state sources are a green solution and they eliminate the threat > of mercury contamination due to improper disposal or explosion of mercury > burners. > > Ben > > > > > > > > > > > > > > > > On Tue, 5 Nov 2013 12:56:42 -0500, Philip Oshel <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> All, >> >> I had this question put to me by a new faculty member, and don't have a >> ready answer: >> "Is there a ballpark percentage for how much less bright an LED vs a >> standard mercury lamp light?" >> This is for regular epifluorescence, not confocal. >> >> This is in the realm of arm-waving over a picture of beer (a good, dark >> stout), ignoring brands, how old the Hg bulb is, ex/em cubes, which part >> of the spectrum is used, and all that. Personally, I'd think the answer >> is more like, "Doesn't matter, the dimmer system is still too bright to >> use all the available light and not damage the specimen." But ... ? >> >> Phil >> -- >> Philip Oshel >> Microscopy Facility Supervisor >> Biology Department >> 024C Brooks Hall >> Central Michigan University >> Mt. Pleasant, MI 48859 >> (989) 774-3576 > |
|
Steffen Dietzel |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Not quite sure about powerful, but here are my two cents: For live cell imaging we often use a regular 12V 100W halogen lamp to avoid the large UV-peaks of the Hg-lamp, which are pretty much impossible to filter out completely. In the green range, this usually results in exposure times about 3x as long as with the Hg lamp. Doesn't bother us since the process we are looking at is slow. Anyway, the halogen lamp should be more powerful the more you get towards the infrared, up to well over 800 nm, see e.g. http://zeiss-campus.magnet.fsu.edu/articles/lightsources/tungstenhalogen.html So you might want to give it a try. The other day I used a red DNA stain (Cy5-Channel), 40x air objective, with the lamp only on 6V and the signal was still pretty bright. So maybe that would do for your purpose? Since you probably have all the parts already, you might want to give it a try. Steffen On 22.11.2013 16:33, Judy Trogadis wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All > > This posting of a couple of weeks ago has suddenly become relevant for us. > > Does anyone know of a powerful continuous white light source in the 600-800nm range? Emphasis on the word 'powerful' because in most light sources there seems to be a steep drop in power above 650nm. > > Thank you, > Judy > > Judy Trogadis > Bio-Imaging Coordinator > Keenan Research Centre, St. Michael's > 209 Victoria Street > Toronto M5B 1T8, Canada > office: 416-864-6060 ext. 77612 > imaging facility: ext. 77434 > cell: 416-254-9330 > [hidden email] > > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |