CFP detection under 458 excitation

classic Classic list List threaded Threaded
♦
♦
7 messages Options
islutsky islutsky
Reply | Threaded
Open this post in threaded view
♦
♦
|

CFP detection under 458 excitation

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  

 

Aryeh Weiss Aryeh Weiss
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation

0.3% may not be a small signal compared to your fluorescence. I would
suggest that you use a filter with a cutoff
no higher than 470, and also that you use the HQ ot ET series filters
(which have very sharp skirts).

--aryeh

Inna Slutsky wrote:

>
> Recently we upgraded our Olympus FV-300 confocal system and purchased
> filters for CFP/YFP detection for FRET applications.
>
>  
>
> The problem is that the 465-495 nm band pass filter (Chroma) showed
> direct excitation of the 458 line (multi-line argon laser) and a heavy
> interference with direct reflection from the glass. Chroma claims that
> it only passes 0.3% of 458nm, and theoretically it is supposed to be
> OK for 458 nm excitation.
>
>  
>
> I wonder if someone has encountered a similar problem and may offer a
> better solution for detection of CFP signal (together with YFP under
> 514 nm excitation).
>
>  
>
> Thanks,
>
> Inna.
>
>  
>
>  
>
> --------------------------------------------------------
>
> Inna Slutsky, PhD
> Dept. of Physiology and Pharmacology
>
> Sackler Faculty of Medicine, room 511
> Tel Aviv University
> Ramat Aviv, Tel Aviv 69978, Israel
> phone: +972 3 640 6021
>
> e-mail: [hidden email]
>
>  
>
>  
>
>  
>


--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

Israel Society for Microscopy 2009 meeting website:
http://materials.technion.ac.il/ism/ISM2009.html
Monique Vasseur Monique Vasseur
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation

In reply to this post by islutsky

Which dichroic are you using?

 

Monique Vasseur

tél. (514) 343-6111 poste 5148

De : Confocal Microscopy List [mailto:[hidden email]] De la part de Inna Slutsky
Envoyé : 26 mai 2009 01:57
À : [hidden email]
Objet : CFP detection under 458 excitation

 

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  

 

islutsky islutsky
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation

458 / 514.

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vasseur Monique
Sent: Tuesday, May 26, 2009 4:03 PM
To: [hidden email]
Subject: Re: CFP detection under 458 excitation

 

Which dichroic are you using?

 

Monique Vasseur

tél. (514) 343-6111 poste 5148

De : Confocal Microscopy List [mailto:[hidden email]] De la part de Inna Slutsky
Envoyé : 26 mai 2009 01:57
À : [hidden email]
Objet : CFP detection under 458 excitation

 

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  

 

Monique Vasseur Monique Vasseur
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation

I meant the emission dichroic, the one to split emission between the PMTs.

 

Monique Vasseur

tél. (514) 343-6111 poste 5148

De : Confocal Microscopy List [mailto:[hidden email]] De la part de Inna Slutsky
Envoyé : 26 mai 2009 09:31
À : [hidden email]
Objet : Re: CFP detection under 458 excitation

 

458 / 514.

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vasseur Monique
Sent: Tuesday, May 26, 2009 4:03 PM
To: [hidden email]
Subject: Re: CFP detection under 458 excitation

 

Which dichroic are you using?

 

Monique Vasseur

tél. (514) 343-6111 poste 5148

De : Confocal Microscopy List [mailto:[hidden email]] De la part de Inna Slutsky
Envoyé : 26 mai 2009 01:57
À : [hidden email]
Objet : CFP detection under 458 excitation

 

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  

 

Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation

In reply to this post by islutsky
I'm not sure I understand the question.  What do you mean by "direct excitation of the 58nm line?"  This is presumably a barrier filter, for CFP, which you'd expect to excite with the 458nm line.  Do you mean that you are getting a reflection image back?  
 
Points to consider:
1.  Is your dichroic correct?
2.  Is your filter installed at its designed angle? 
3. Is your blocking system (AOTF?) for the 488nm laser line functioning properly?  It would only take a tiny leakage there to really screw things up. 
 
                                                              Guy
 
 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Inna Slutsky
Sent: Tuesday, 26 May 2009 3:57 PM
To: [hidden email]
Subject: CFP detection under 458 excitation

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]

 

  

 


Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM


Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM

Prashant Prabhat Prashant Prabhat
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: CFP detection under 458 excitation -"Commercial Response"

In reply to this post by islutsky
***************"Commercial Response" ******************
Dear Inna,
 
Ideally it is desired to provide very high degree of blocking (several orders of magnitude of blocking) of the unwanted laser lines to avoid interference patterns in the emission channel. Both the excitation filter and the emission filter contribute to this blocking of unwanted laser lines. For extremely demanding applications such as imaging of single molecules, it may be even desirable to provide additional blocking by using a suitable notch or a long pass filter in the emission channel (http://www.semrock.com/Data/Documents/PerfectingTIRF_0109_Reprint.pdf). This is generally not true for a dichroic mirror, as they are not specifically designed for blocking.
 
For your application, it may be worthwhile considering from following Semrock filter for imaging CFP & YFP.
 
1) CFP-2423A set (http://www.semrock.com/Catalog/SetDetails.aspx?SetBasePartID=21&CategoryID=16). Note that the emission filter of this filter set, FF01-483/32-25 is designed to provide >OD 6 blocking at 458 nm.
2) YFP-2427A set (http://www.semrock.com/Catalog/SetDetails.aspx?SetBasePartID=25&CategoryID=16). Note that the emission filter of this filter set, FF01-542/27-25 is designed to provide >OD 6 blocking at 514 nm.
3) Alternatively, a steep edge long pass filter in the emission channel, such as Semrock's LP02-458RS-25 (http://www.semrock.com/Catalog/Detail.aspx?FilterPartID=150&CategoryID=69), which is designed to provide > OD 6 blocking at 458 nm when used in conjunction with your existing filters should be able to eliminate interference patterns in this channel.
 
Please contact me offline at [hidden email], if you have questions.
 
Regards,
Prashant
 

Prashant Prabhat, Ph.D.

Applications Specialist

Semrock

A Unit of IDEX Corporation

3625 Buffalo Road, Suite 6

Rochester NY 14624

Email: [hidden email]

Phone: 585-594-7064

Toll Free: 866-SEMROCK

Fax: 585-594-7095

http://www.semrock.com

 

The Standard in Optical Filters for Biotech & Analytical Instrumentation

Hundreds of Thousands of Ion Beam Sputtered filters delivered - extensive inventory now!

The information contained in this message and any attachments may be privileged, confidential, and protected from disclosure. If the reader of this message is not the intended recipient, or any agent responsible for delivering this message to the intended recipient, you are hereby notified that any dissemination, distribution, or copying of this communication may be unlawful and therefore strictly prohibited. If you received this message in error, please reply to the message and delete it.

 

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Inna Slutsky
Sent: Tuesday, May 26, 2009 1:57 AM
To: [hidden email]
Subject: CFP detection under 458 excitation

Recently we upgraded our Olympus FV-300 confocal system and purchased filters for CFP/YFP detection for FRET applications.

 

The problem is that the 465-495 nm band pass filter (Chroma) showed direct excitation of the 458 line (multi-line argon laser) and a heavy interference with direct reflection from the glass. Chroma claims that it only passes 0.3% of 458nm, and theoretically it is supposed to be OK for 458 nm excitation.

 

I wonder if someone has encountered a similar problem and may offer a better solution for detection of CFP signal (together with YFP under 514 nm excitation).

 

Thanks,

Inna.

 

 

--------------------------------------------------------

Inna Slutsky, PhD
Dept. of Physiology and Pharmacology

Sackler Faculty of Medicine, room 511
Tel Aviv University
Ramat Aviv, Tel Aviv 69978, Israel
phone: +972 3 640 6021

e-mail: [hidden email]