Peter Werner |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** At our facility, we have a spinning-disk confocal system, with an incubator that surrounds the entire stage and substage area. The question I have concerns the type of CO2 I should be using for the incubator. I was told by someone from the UC Berkeley Biological Imaging Facility that I should be careful about the grade of carbon dioxide that I use - apparently, it should be high purity and as low as possible in H2O. The reason is, even if the cell cultures themselves will be fine with a lower purity of CO2, CO2 forms carbonic acid in reaction with water, and the acidified water content in the chamber atmosphere can be damaging to the instrument over the long term, hence, the less H2O mixed in with the CO2, the better. If this is the case, that would imply that "Instrument Grade" or "Bone Dry" grades would be what I need to use. I was wondering if anybody else knew of this recommendation and followed this practice. It makes sense to me why one would want to minimize carbonic acid formation in environments the confocal system is exposed to. On the other hand, it does entail some expense, not so much in terms of the gas itself, but in terms of special cylinders for high-purity CO2, which must be rented, or purchased at $500+ per gas cylinder. Let me know, Peter G. Werner Instructional Assistant/Lab Technician, Microscopy, Merritt College SEM/AFM Lab Technician, Ohlone College [hidden email] |
|
Kurt Thorn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Something seems weird here, since a live cell incubator needs to be humidified to keep the cells from drying out. Typically, only the area around the sample is kept in a 5% CO2 atmosphere, and this is also humidified to close to 100%, so any water in the CO2 tank is negligible compared to the amount of water added by the humidification system. Even without this, unless you're in the desert, the air is often 50% or greater relative humidity, so as soon as you mix the CO2 with air, there's a fair amount of water present.... Kurt On 7/22/2014 4:42 PM, Peter Werner wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > At our facility, we have a spinning-disk confocal system, with an > incubator that surrounds the entire stage and substage area. The > question I have concerns the type of CO2 I should be using for the > incubator. I was told by someone from the UC Berkeley Biological > Imaging Facility that I should be careful about the grade of carbon > dioxide that I use - apparently, it should be high purity and as low > as possible in H2O. The reason is, even if the cell cultures > themselves will be fine with a lower purity of CO2, CO2 forms carbonic > acid in reaction with water, and the acidified water content in the > chamber atmosphere can be damaging to the instrument over the long > term, hence, the less H2O mixed in with the CO2, the better. If this > is the case, that would imply that "Instrument Grade" or "Bone Dry" > grades would be what I need to use. > > I was wondering if anybody else knew of this recommendation and > followed this practice. It makes sense to me why one would want to > minimize carbonic acid formation in environments the confocal system > is exposed to. On the other hand, it does entail some expense, not so > much in terms of the gas itself, but in terms of special cylinders for > high-purity CO2, which must be rented, or purchased at $500+ per gas > cylinder. > > Let me know, > Peter G. Werner > Instructional Assistant/Lab Technician, Microscopy, Merritt College > SEM/AFM Lab Technician, Ohlone College > [hidden email] > > -- Kurt Thorn Director, Nikon Imaging Center http://nic.ucsf.edu/blog/ |
|
Loralei Dewe-3 |
|
|
In reply to this post by Peter Werner
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** It seems to me that if this is a problem use the culture medium that has phenol red so you can see the pH change occuring. In all my years, I've never had to order specific CO for microscopy. Loralei On Jul 22, 2014 4:45 PM, "Peter Werner" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > At our facility, we have a spinning-disk confocal system, with an > incubator that surrounds the entire stage and substage area. The question I > have concerns the type of CO2 I should be using for the incubator. I was > told by someone from the UC Berkeley Biological Imaging Facility that I > should be careful about the grade of carbon dioxide that I use - > apparently, it should be high purity and as low as possible in H2O. The > reason is, even if the cell cultures themselves will be fine with a lower > purity of CO2, CO2 forms carbonic acid in reaction with water, and the > acidified water content in the chamber atmosphere can be damaging to the > instrument over the long term, hence, the less H2O mixed in with the CO2, > the better. If this is the case, that would imply that "Instrument Grade" > or "Bone Dry" grades would be what I need to use. > > I was wondering if anybody else knew of this recommendation and followed > this practice. It makes sense to me why one would want to minimize carbonic > acid formation in environments the confocal system is exposed to. On the > other hand, it does entail some expense, not so much in terms of the gas > itself, but in terms of special cylinders for high-purity CO2, which must > be rented, or purchased at $500+ per gas cylinder. > > Let me know, > Peter G. Werner > Instructional Assistant/Lab Technician, Microscopy, Merritt College > SEM/AFM Lab Technician, Ohlone College > [hidden email] > |
|
Steffen Dietzel |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Am 23.07.2014 02:30, schrieb Loralei Dewe: > It seems to me that if this is a problem use the culture medium that has > phenol red so you can see the pH change occuring. But that is not something to be universally recommended, since phenol red is fluorescent and can cause quite a background. I'd restrict its use to testing purposes. Or a small volume right next to your actual sample. Steffen -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
|
Craig Brideau |
|
|
In reply to this post by Loralei Dewe-3
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I agree that special CO is unnecessary. As Kurt says, the atmospheric water is probably higher anyway. The one thing we have done in the past is swap out carbogen (5% CO2 balance O2) for a 5% CO2 balance air mixture, because we found 95% O2 was harsh on a particular sample. Craig Brideau On Tue, Jul 22, 2014 at 6:30 PM, Loralei Dewe <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > It seems to me that if this is a problem use the culture medium that has > phenol red so you can see the pH change occuring. In all my years, I've > never had to order specific CO for microscopy. > > Loralei > On Jul 22, 2014 4:45 PM, "Peter Werner" <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > At our facility, we have a spinning-disk confocal system, with an > > incubator that surrounds the entire stage and substage area. The > question I > > have concerns the type of CO2 I should be using for the incubator. I was > > told by someone from the UC Berkeley Biological Imaging Facility that I > > should be careful about the grade of carbon dioxide that I use - > > apparently, it should be high purity and as low as possible in H2O. The > > reason is, even if the cell cultures themselves will be fine with a lower > > purity of CO2, CO2 forms carbonic acid in reaction with water, and the > > acidified water content in the chamber atmosphere can be damaging to the > > instrument over the long term, hence, the less H2O mixed in with the CO2, > > the better. If this is the case, that would imply that "Instrument Grade" > > or "Bone Dry" grades would be what I need to use. > > > > I was wondering if anybody else knew of this recommendation and followed > > this practice. It makes sense to me why one would want to minimize > carbonic > > acid formation in environments the confocal system is exposed to. On the > > other hand, it does entail some expense, not so much in terms of the gas > > itself, but in terms of special cylinders for high-purity CO2, which must > > be rented, or purchased at $500+ per gas cylinder. > > > > Let me know, > > Peter G. Werner > > Instructional Assistant/Lab Technician, Microscopy, Merritt College > > SEM/AFM Lab Technician, Ohlone College > > [hidden email] > > > |
|
Armstrong, Brian |
|
|
Hi Peter, I agree with the consensus here. I think that our CO2 costs around $3 (US$) per small bottle. You should be careful however with the pressure you deliver to the mixing unit as high pressures can damage the sensor unit. This usually happens the upon the initial installation. To avoid this you can open your flow with the regulator attached and adjust the flow rate prior to attaching it to your system. Also important is having a good filter on the line.
Best of luck, Brian D Armstrong PhD Associate Research Professor Director, Light Microscopy Core Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, July 23, 2014 10:53 AM To: [hidden email] Subject: Re: CO2 grade for on-scope incubator? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I agree that special CO is unnecessary. As Kurt says, the atmospheric water is probably higher anyway. The one thing we have done in the past is swap out carbogen (5% CO2 balance O2) for a 5% CO2 balance air mixture, because we found 95% O2 was harsh on a particular sample. Craig Brideau On Tue, Jul 22, 2014 at 6:30 PM, Loralei Dewe <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > It seems to me that if this is a problem use the culture medium that has > phenol red so you can see the pH change occuring. In all my years, I've > never had to order specific CO for microscopy. > > Loralei > On Jul 22, 2014 4:45 PM, "Peter Werner" <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > At our facility, we have a spinning-disk confocal system, with an > > incubator that surrounds the entire stage and substage area. The > question I > > have concerns the type of CO2 I should be using for the incubator. I was > > told by someone from the UC Berkeley Biological Imaging Facility that I > > should be careful about the grade of carbon dioxide that I use - > > apparently, it should be high purity and as low as possible in H2O. The > > reason is, even if the cell cultures themselves will be fine with a lower > > purity of CO2, CO2 forms carbonic acid in reaction with water, and the > > acidified water content in the chamber atmosphere can be damaging to the > > instrument over the long term, hence, the less H2O mixed in with the CO2, > > the better. If this is the case, that would imply that "Instrument Grade" > > or "Bone Dry" grades would be what I need to use. > > > > I was wondering if anybody else knew of this recommendation and followed > > this practice. It makes sense to me why one would want to minimize > carbonic > > acid formation in environments the confocal system is exposed to. On the > > other hand, it does entail some expense, not so much in terms of the gas > > itself, but in terms of special cylinders for high-purity CO2, which must > > be rented, or purchased at $500+ per gas cylinder. > > > > Let me know, > > Peter G. Werner > > Instructional Assistant/Lab Technician, Microscopy, Merritt College > > SEM/AFM Lab Technician, Ohlone College > > [hidden email] > > > --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |
|
Teng-Leong Chew-2 |
|
|
In reply to this post by Peter Werner
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Peter, Kurt is right. You need humidity to minimize evaporation from the culture medium or that will drastically alter the osmolarity of the medium very quickly. That¹s why there is a tray of water at the bottom of your regular cell culture incubator. Apparently the effect from the carbonic acid level is miniscule. You don¹t need the bone dry CO2. Regards, Leong -- Teng-Leong Chew, PhD Director, Advanced Imaging Center Howard Hughes Medical Institute at Janelia Ashburn, VA 20147 On 7/22/14, 7:42 PM, "Peter Werner" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >At our facility, we have a spinning-disk confocal system, with an >incubator that surrounds the entire stage and substage area. The >question I have concerns the type of CO2 I should be using for the >incubator. I was told by someone from the UC Berkeley Biological Imaging >Facility that I should be careful about the grade of carbon dioxide that >I use - apparently, it should be high purity and as low as possible in >H2O. The reason is, even if the cell cultures themselves will be fine >with a lower purity of CO2, CO2 forms carbonic acid in reaction with >water, and the acidified water content in the chamber atmosphere can be >damaging to the instrument over the long term, hence, the less H2O mixed >in with the CO2, the better. If this is the case, that would imply that >"Instrument Grade" or "Bone Dry" grades would be what I need to use. > >I was wondering if anybody else knew of this recommendation and followed >this practice. It makes sense to me why one would want to minimize >carbonic acid formation in environments the confocal system is exposed >to. On the other hand, it does entail some expense, not so much in terms >of the gas itself, but in terms of special cylinders for high-purity >CO2, which must be rented, or purchased at $500+ per gas cylinder. > >Let me know, >Peter G. Werner >Instructional Assistant/Lab Technician, Microscopy, Merritt College >SEM/AFM Lab Technician, Ohlone College >[hidden email] |
|
Cole, Richard W (HEALTH) |
|
|
In reply to this post by Peter Werner
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Not to belabor the point but. I have been using stage incubators for ~30 years without any "high purity" gases. I use both the type that requires premixed, i.e. 5% and the type that mixes at the scope and have never seen any effect on the scopes Cheers Rich Richard Cole Research Scientist V Director: Advanced Light Microscopy & Image Analysis Core Wadsworth Center Research Assistant Professor Dept. of Biomedical Sciences School of Public Health State University of New York P.O. Box 509 Albany N.Y. 12201-0509 518-474-7048 Phone 518-473-2895 Fax Website www.wadsworth.org/cores/alm/index.htm |
|
|
In reply to this post by Peter Werner
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Peter, EMBL operates such incubators for 10+ years using technical grade C02 without any problems. We have even saturated the air in the incubator with water in order to counteract evaporation of culture media for long time lapse acquisitions. You can use pre-mixed gas, as mentioned, and it is better to mix CO2 with air or with N2. A low partial pressure of O2 helps avoiding photobleaching and facilitates photo-switching of some GFPs. Kind regards, Jens Visiting Scientist @ Center for Technological Development in Health (CDTS), Oswaldo Cruz Foundation (Fiocruz), Ministry of Health, Rio de Janeiro, Brazil. http://br.linkedin.com/pub/jens-rietdorf/6/4a3/189/ Skype jens.rietdorf On Tue, Jul 22, 2014 at 8:42 PM, Peter Werner <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > At our facility, we have a spinning-disk confocal system, with an > incubator that surrounds the entire stage and substage area. The question I > have concerns the type of CO2 I should be using for the incubator. I was > told by someone from the UC Berkeley Biological Imaging Facility that I > should be careful about the grade of carbon dioxide that I use - > apparently, it should be high purity and as low as possible in H2O. The > reason is, even if the cell cultures themselves will be fine with a lower > purity of CO2, CO2 forms carbonic acid in reaction with water, and the > acidified water content in the chamber atmosphere can be damaging to the > instrument over the long term, hence, the less H2O mixed in with the CO2, > the better. If this is the case, that would imply that "Instrument Grade" > or "Bone Dry" grades would be what I need to use. > > I was wondering if anybody else knew of this recommendation and followed > this practice. It makes sense to me why one would want to minimize carbonic > acid formation in environments the confocal system is exposed to. On the > other hand, it does entail some expense, not so much in terms of the gas > itself, but in terms of special cylinders for high-purity CO2, which must > be rented, or purchased at $500+ per gas cylinder. > > Let me know, > Peter G. Werner > Instructional Assistant/Lab Technician, Microscopy, Merritt College > SEM/AFM Lab Technician, Ohlone College > [hidden email] > |