I have a cell dying problem in my microscope when I take pictures of them overnight or longer. I suspect the humidity is failing and I want to monitor that, CO2 and temperature.(It is not photodamage. Cells die in areas away from the objective.)
Can anyone suggest a supplier of recording sensors that I can get next to the well slides on my stage? Thank you! Paul Herzmark Specialist [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 (510) 643-9500 fax |
![]() ![]() |
Goodhouse, Joseph G. |
![]() |
Paul,
Could you describe in detail your
heating system. Do you overlay your cells with mineral oil? This prevents
evaporation which will alter your salt concentration.
I use a Recordable Temp sensor from Fisher
Scientific. Cat No 15-007-26. Inexpensive and works well.
Joe Goodhouse
Visit us at http://www.molbio1.princeton.edu/facility/confocal/ From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Herzmark Sent: Thursday, February 12, 2009 2:24 PM To: [hidden email] Subject: CO2, humidity, temp sensors Can anyone suggest a supplier of recording sensors that I can get next to the well slides on my stage? Thank you! Paul Herzmark Specialist [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 (510) 643-9500 fax |
![]() ![]() |
Paul Herzmark |
![]() |
I knew I would not get away with out details! I have an Invivo (company) plexiglass incubator box around my inverted microscope. 5% CO2 is piped inside to a small space (about 50 ml) around the multiwell-Labtech (company)-coverslip. The small space is created by a glass plate with a frame around it that goes over the coverslip.The CO2 passes into the small space via special tubing.The tubing goes under incubator-warmed water where it is supposed to pick up water vapor without bubbling. 1. The whole incubator, with most of the microscope is warmed to 37 degrees 2. Only the small space surrounding the cells is humidified and has CO2. On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. <[hidden email]> wrote:
|
![]() ![]() |
Goodhouse, Joseph G. |
![]() |
My Guess is still evaporation of media is killing your
cells.. Overlay the cell media in each compartment with mineral oil.
It stops all evaporation but allows CO2 gas exchange to occur to maintain
pH.. I do this with all my chamber systems and it works great. No
need to humidify flowing air or gas.
Joe Goodhouse
Visit us at http://www.molbio1.princeton.edu/facility/confocal/ From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Herzmark Sent: Thursday, February 12, 2009 4:36 PM To: [hidden email] Subject: Re: CO2, humidity, temp sensors I knew I would not get away with out details! I have an Invivo (company) plexiglass incubator box around my inverted microscope. 5% CO2 is piped inside to a small space (about 50 ml) around the multiwell-Labtech (company)-coverslip. The small space is created by a glass plate with a frame around it that goes over the coverslip.The CO2 passes into the small space via special tubing.The tubing goes under incubator-warmed water where it is supposed to pick up water vapor without bubbling. 1. The whole incubator, with most of the microscope is warmed to 37 degrees 2. Only the small space surrounding the cells is humidified and has CO2. On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G.
<[hidden email]> wrote:
|
![]() ![]() |
Paul Herzmark |
![]() |
Joe,
Thank you for the suggestion. I also suspect the evaporation. I don't understand how mineral oil will stop water vapor from passing but allow CO2 exchange. Any idea? Paul
On Thu, Feb 12, 2009 at 2:23 PM, Goodhouse, Joseph G. <[hidden email]> wrote:
|
![]() |
In reply to this post by Paul Herzmark
Paul,
We have a similar system and have been using it successfully after spending some time to tune the 5% CO2 flow rate until the pH stayed constant (judged simply by the colour of the medium), and after placing several beakers of water within the 37C area to humidify the warm air and thus keep the evaporation rate low. This way experiments can be done overnight w/o cells dying (appr. 17 hours).
Hope this helps a bit. Zoltan On Thu, Feb 12, 2009 at 7:23 PM, Paul Herzmark <[hidden email]> wrote: I have a cell dying problem in my microscope when I take pictures of them overnight or longer. I suspect the humidity is failing and I want to monitor that, CO2 and temperature.(It is not photodamage. Cells die in areas away from the objective.) -- Zoltan Cseresnyes Facility manager, Imaging Suite |
![]() |
Hi,
Temperature and %rh probes. It is useful to measure temperature and relative humidity in your incubator and there is a superb company that produces excellent probes for doing this on the cheap: http://www.picotech.com/temperature-humidity.html Price depends on accuracy: +/-0.3oC or +/-0.01oC, the former being fine for most. I use the £135 three probe Th-03 system thats accurate to +/-0.3oC, although theres the more expensive [£400 single] platinum resistance probe thats accurate to +/- 0.01oC or more. Temp measurement is a complex thing with probes, e.g. conduction heat loss down the probe etc but they do give a useful indication of the temp where the cells are rather what an external control box says. Pico also make a very cheap software oscilloscope and %rh probes. My PhD is in fuel and combustion, so thermocouples and stuff are second nature to me, but these Pico data loggers are great value, and simple to use [although the TH-03 probes are a bit large for some situations]. Evaporation Yes you can use mineral oil for long term time-lapse, but some worry about gas exchange and toxic effects [well I do]. A preferred option would be special gas not water permeable membrane lids/seals for your small Petri dish to virtually eliminate evaporation: e.g. Zeiss Foil Cover and CultFoil Foil Covers that extend the incubation time considerably by reducing evaporation. A special foil (CultFoil), gripped between two thin steel components, replaces the usual cover of the culture vessel. The foil effectively reduces the amount of water vapor that is able to escape. Gas exchange is not affected. But they arent as cheap as mineral oil. Generally we just humidified the air in the incubators to reduce evaporation rates. For short term [2-4 hours] I used to just put a beaker filled with water and overhanging tissues into the incubator [if its a large one], although you need to change it daily to prevent it getting smelly. However for long term time-lapse we always used the Zeiss [PeCon] humidifiers that fitted in-line to air circulation system [cost about £500 each], and these worked very well. All our systems were Zeiss/Leica and so PeCon friendly. For CO2 we trusted our fantastically expensive variable% Zeiss CO2 controllers [or the 5% CO2 label on the BOC CO2 cylinder]. Humidifier: http://www.pe-con.de/pecon/index.htm search for relative humidity Regards Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ ________________________________________ From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Zoltan Cseresnyes Sent: 12 February 2009 23:22 To: [hidden email] Subject: Re: CO2, humidity, temp sensors Paul, We have a similar system and have been using it successfully after spending some time to tune the 5% CO2 flow rate until the pH stayed constant (judged simply by the colour of the medium), and after placing several beakers of water within the 37C area to humidify the warm air and thus keep the evaporation rate low. This way experiments can be done overnight w/o cells dying (appr. 17 hours). Hope this helps a bit. Zoltan On Thu, Feb 12, 2009 at 7:23 PM, Paul Herzmark <[hidden email]> wrote: I have a cell dying problem in my microscope when I take pictures of them overnight or longer. I suspect the humidity is failing and I want to monitor that, CO2 and temperature.(It is not photodamage. Cells die in areas away from the objective.) Can anyone suggest a supplier of recording sensors that I can get next to the well slides on my stage? Thank you! Paul Herzmark Specialist [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 (510) 643-9500 fax -- Zoltan Cseresnyes Facility manager, Imaging Suite |
![]() ![]() |
Brotchie, Daniel |
![]() |
In reply to this post by Paul Herzmark
Hi, I guess not applicable to Paul’s micro environment, but SenseAir
sell a meter that measures CO2, temperature and humidity simultaneously - cost
about 200 USD: http://www.senseair.se/includes/products/psense_rh.php Where you have an time-lapse incubator, but have CO2 issues/no
CO2 cylinder you can try CO2 independent media: or use CO2 generating tablets (baking powder): http://www.oxoid.com/UK/blue/prod_detail/prod_detail.asp?pr=BR0039&c=UK&lang=EN NCI Daniel ------------------------------------------------ Daniel Brotchie Ophthalmology School of Clinical Sciences University Clinical Departments The Duncan Building Daulby Street LIVERPOOL L69 3GA UK Tel: +44 (0) 151 706 4017 Fax: +44 (0) 151 706 5934 E-mail [hidden email] From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of Paul Herzmark I have a cell dying problem in my microscope when I take
pictures of them overnight or longer. I suspect the humidity is failing and I
want to monitor that, CO2 and temperature.(It is not photodamage. Cells die in
areas away from the objective.) |
![]() |
In reply to this post by Goodhouse, Joseph G.
Over time, the mineral oil overlay will basically void the medium of
all lipophilic components (vitamin E, for instance). This may or may not be a problem, but you need to be aware of it and do the proper controls. If you want to check for evaporation, just weigh your sample before and after the imaging. Beat At 23:23 12-02-2009, you wrote: >My Guess is still evaporation of media is killing your >cells.. Overlay the cell media in each compartment with mineral >oil. It stops all evaporation but allows CO2 gas exchange to occur >to maintain pH.. I do this with all my chamber systems and it works >great. No need to humidify flowing air or gas. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.princeton.edu/facility/confocal/ > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 4:36 PM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors > > >I knew I would not get away with out details! > >I have an Invivo (company) plexiglass incubator box around my >inverted microscope. 5% CO2 is piped inside to a small space (about >50 ml) around the multiwell-Labtech (company)-coverslip. The small >space is created by a glass plate with a frame around it that goes >over the coverslip.The CO2 passes into the small space via special >tubing.The tubing goes under incubator-warmed water where it is >supposed to pick up water vapor without bubbling. > >1. The whole incubator, with most of the microscope is warmed to 37 degrees >2. Only the small space surrounding the cells is humidified and has CO2. > > > > >On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. ><<mailto:[hidden email]>[hidden email]> wrote: >Paul, > Could you describe in detail your heating system. Do you > overlay your cells with mineral oil? This prevents evaporation > which will alter your salt concentration. > > I use a Recordable Temp sensor from Fisher Scientific. Cat No > 15-007-26. Inexpensive and works well. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.<http://princeton.edu/facility/confocal/>princeton.edu/facility/confocal/ > > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 2:24 PM >To: <mailto:[hidden email]>[hidden email] >Subject: CO2, humidity, temp sensors > >I have a cell dying problem in my microscope when I take pictures of >them overnight or longer. I suspect the humidity is failing and I >want to monitor that, CO2 and temperature.(It is not photodamage. >Cells die in areas away from the objective.) > >Can anyone suggest a supplier of recording sensors that I can get >next to the well slides on my stage? > >Thank you! > >Paul Herzmark >Specialist ><mailto:[hidden email]>[hidden email] > >Department of Molecular and Cell Biology >479 Life Science Addition >University of California, Berkeley >Berkeley, CA 94720-3200 >(510) 643-9603 >(510) 643-9500 fax > |
![]() ![]() |
James Denegre |
![]() |
In our lab we equilibrate the mineral oil with the medium to prevent uptake of components, and have successfully imaged under mineral oil for five continuous days.
James Denegre, Ph.D. Associate Research Scientist Manager, Light Microscopy The Jackson Laboratory Bar Harbor, ME 04609 207.288.6321 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Beat Ludin Sent: Wednesday, February 25, 2009 5:05 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors Over time, the mineral oil overlay will basically void the medium of all lipophilic components (vitamin E, for instance). This may or may not be a problem, but you need to be aware of it and do the proper controls. If you want to check for evaporation, just weigh your sample before and after the imaging. Beat At 23:23 12-02-2009, you wrote: >My Guess is still evaporation of media is killing your >cells.. Overlay the cell media in each compartment with mineral >oil. It stops all evaporation but allows CO2 gas exchange to occur >to maintain pH.. I do this with all my chamber systems and it works >great. No need to humidify flowing air or gas. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.princeton.edu/facility/confocal/ > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 4:36 PM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors > > >I knew I would not get away with out details! > >I have an Invivo (company) plexiglass incubator box around my >inverted microscope. 5% CO2 is piped inside to a small space (about >50 ml) around the multiwell-Labtech (company)-coverslip. The small >space is created by a glass plate with a frame around it that goes >over the coverslip.The CO2 passes into the small space via special >tubing.The tubing goes under incubator-warmed water where it is >supposed to pick up water vapor without bubbling. > >1. The whole incubator, with most of the microscope is warmed to 37 degrees >2. Only the small space surrounding the cells is humidified and has CO2. > > > > >On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. ><<mailto:[hidden email]>[hidden email]> wrote: >Paul, > Could you describe in detail your heating system. Do you > overlay your cells with mineral oil? This prevents evaporation > which will alter your salt concentration. > > I use a Recordable Temp sensor from Fisher Scientific. Cat No > 15-007-26. Inexpensive and works well. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.<http://princeton.edu/facility/confocal/>princeton.edu/facility/confocal/ > > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 2:24 PM >To: <mailto:[hidden email]>[hidden email] >Subject: CO2, humidity, temp sensors > >I have a cell dying problem in my microscope when I take pictures of >them overnight or longer. I suspect the humidity is failing and I >want to monitor that, CO2 and temperature.(It is not photodamage. >Cells die in areas away from the objective.) > >Can anyone suggest a supplier of recording sensors that I can get >next to the well slides on my stage? > >Thank you! > >Paul Herzmark >Specialist ><mailto:[hidden email]>[hidden email] > >Department of Molecular and Cell Biology >479 Life Science Addition >University of California, Berkeley >Berkeley, CA 94720-3200 >(510) 643-9603 >(510) 643-9500 fax > |
![]() ![]() |
Armstrong, Brian |
![]() |
Hi James, can you elaborate on your technique " equilibrate the mineral
oil with the medium"? Thanks, Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag ing/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of James Denegre Sent: Thursday, February 26, 2009 8:10 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors and mu\ineral oil In our lab we equilibrate the mineral oil with the medium to prevent uptake of components, and have successfully imaged under mineral oil for five continuous days. James Denegre, Ph.D. Associate Research Scientist Manager, Light Microscopy The Jackson Laboratory Bar Harbor, ME 04609 207.288.6321 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Beat Ludin Sent: Wednesday, February 25, 2009 5:05 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors Over time, the mineral oil overlay will basically void the medium of all lipophilic components (vitamin E, for instance). This may or may not be a problem, but you need to be aware of it and do the proper controls. If you want to check for evaporation, just weigh your sample before and after the imaging. Beat At 23:23 12-02-2009, you wrote: >My Guess is still evaporation of media is killing your >cells.. Overlay the cell media in each compartment with mineral >oil. It stops all evaporation but allows CO2 gas exchange to occur >to maintain pH.. I do this with all my chamber systems and it works >great. No need to humidify flowing air or gas. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.princeton.edu/facility/confocal/ > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 4:36 PM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors > > >I knew I would not get away with out details! > >I have an Invivo (company) plexiglass incubator box around my >inverted microscope. 5% CO2 is piped inside to a small space (about >50 ml) around the multiwell-Labtech (company)-coverslip. The small >space is created by a glass plate with a frame around it that goes >over the coverslip.The CO2 passes into the small space via special >tubing.The tubing goes under incubator-warmed water where it is >supposed to pick up water vapor without bubbling. > >1. The whole incubator, with most of the microscope is warmed to 37 >2. Only the small space surrounding the cells is humidified and has CO2. > > > > >On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. ><<mailto:[hidden email]>[hidden email]> wrote: >Paul, > Could you describe in detail your heating system. Do you > overlay your cells with mineral oil? This prevents evaporation > which will alter your salt concentration. > > I use a Recordable Temp sensor from Fisher Scientific. Cat No > 15-007-26. Inexpensive and works well. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.<http://princeton.edu/facility/c > > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 2:24 PM >To: <mailto:[hidden email]>[hidden email] U >Subject: CO2, humidity, temp sensors > >I have a cell dying problem in my microscope when I take pictures of >them overnight or longer. I suspect the humidity is failing and I >want to monitor that, CO2 and temperature.(It is not photodamage. >Cells die in areas away from the objective.) > >Can anyone suggest a supplier of recording sensors that I can get >next to the well slides on my stage? > >Thank you! > >Paul Herzmark >Specialist ><mailto:[hidden email]>[hidden email] > >Department of Molecular and Cell Biology >479 Life Science Addition >University of California, Berkeley >Berkeley, CA 94720-3200 >(510) 643-9603 >(510) 643-9500 fax > --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
![]() |
Here is the protocol in use at The Jackson Laboratory for producing mineral oil for use in in vitro fertilizations and embryo culture:
1. In a sterile bottle add 1000mL of oil and 100mL tissue culture medium. 2. Stir oil/medium for two days at room temperature. 3. Let oil/medium settle for one day. 4. Decant medium the next day. 5. Filter the oil using a 0.8цL filter (Fisher #09-740-30), aliquot into 250 ml units. The oil is stable for two months at room temp. Use it to layer over cells in medium, to prevent evaporation. It will exchange gases, allowing CO2 to maintain the proper pH (we have verified this using a pH indicator, and moving dishes in and out of CO2 incubators. James Denegre, Ph.D. Associate Research Scientist Manager, Light Microscopy The Jackson Laboratory Bar Harbor, ME 04609 207.288.6321 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Armstrong, Brian Sent: Thursday, February 26, 2009 11:46 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors and mu\ineral oil Hi James, can you elaborate on your technique " equilibrate the mineral oil with the medium"? Thanks, Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag ing/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of James Denegre Sent: Thursday, February 26, 2009 8:10 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors and mu\ineral oil In our lab we equilibrate the mineral oil with the medium to prevent uptake of components, and have successfully imaged under mineral oil for five continuous days. James Denegre, Ph.D. Associate Research Scientist Manager, Light Microscopy The Jackson Laboratory Bar Harbor, ME 04609 207.288.6321 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Beat Ludin Sent: Wednesday, February 25, 2009 5:05 AM To: [hidden email] Subject: Re: CO2, humidity, temp sensors Over time, the mineral oil overlay will basically void the medium of all lipophilic components (vitamin E, for instance). This may or may not be a problem, but you need to be aware of it and do the proper controls. If you want to check for evaporation, just weigh your sample before and after the imaging. Beat At 23:23 12-02-2009, you wrote: >My Guess is still evaporation of media is killing your >cells.. Overlay the cell media in each compartment with mineral >oil. It stops all evaporation but allows CO2 gas exchange to occur >to maintain pH.. I do this with all my chamber systems and it works >great. No need to humidify flowing air or gas. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.princeton.edu/facility/confocal/ > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 4:36 PM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors > > >I knew I would not get away with out details! > >I have an Invivo (company) plexiglass incubator box around my >inverted microscope. 5% CO2 is piped inside to a small space (about >50 ml) around the multiwell-Labtech (company)-coverslip. The small >space is created by a glass plate with a frame around it that goes >over the coverslip.The CO2 passes into the small space via special >tubing.The tubing goes under incubator-warmed water where it is >supposed to pick up water vapor without bubbling. > >1. The whole incubator, with most of the microscope is warmed to 37 >2. Only the small space surrounding the cells is humidified and has CO2. > > > > >On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. ><<mailto:[hidden email]>[hidden email]> wrote: >Paul, > Could you describe in detail your heating system. Do you > overlay your cells with mineral oil? This prevents evaporation > which will alter your salt concentration. > > I use a Recordable Temp sensor from Fisher Scientific. Cat No > 15-007-26. Inexpensive and works well. > > >Joe Goodhouse >Confocal Core Lab Manager >Dept. of Molecular Biology >Princeton University >609-258-5432 > >Visit us at ><http://www.molbio/>http://www.molbio1.<http://princeton.edu/facility/c > > > > >---------- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Paul Herzmark >Sent: Thursday, February 12, 2009 2:24 PM >To: <mailto:[hidden email]>[hidden email] U >Subject: CO2, humidity, temp sensors > >I have a cell dying problem in my microscope when I take pictures of >them overnight or longer. I suspect the humidity is failing and I >want to monitor that, CO2 and temperature.(It is not photodamage. >Cells die in areas away from the objective.) > >Can anyone suggest a supplier of recording sensors that I can get >next to the well slides on my stage? > >Thank you! > >Paul Herzmark >Specialist ><mailto:[hidden email]>[hidden email] > >Department of Molecular and Cell Biology >479 Life Science Addition >University of California, Berkeley >Berkeley, CA 94720-3200 >(510) 643-9603 >(510) 643-9500 fax > --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
![]() |
James -
Have you verified that this protocol is sufficient to really equilibrate the mineral oil, i.e. that the composition of medium is no longer affect by contact with the equilibrated mineral oil? Or are you just going by cell survival/growth, which is certainly a good indicator but doesn't necessarily mean that the cells are affected in a non-lethal way? Also, still need to consider the effects of the mineral oil if you want to add compounds during the experiment. BTW, I don't mean to question that a mineral oil overlay is suitable for many applications, but as always, it's good to be aware of possible side-effects. Beat At 23:17 26-02-2009, you wrote: >Here is the protocol in use at The Jackson >Laboratory for producing mineral oil for use in >in vitro fertilizations and embryo culture: > > >1. In a sterile bottle add 1000mL of oil and 100mL tissue culture medium. >2. Stir oil/medium for two days at room temperature. >3. Let oil/medium settle for one day. >4. Decant medium the next day. >5. Filter the oil using a 0.8ÃL filter >(Fisher #09-740-30), aliquot into 250 ml units. > >The oil is stable for two months at room temp. > >Use it to layer over cells in medium, to prevent >evaporation. It will exchange gases, allowing >CO2 to maintain the proper pH (we have verified >this using a pH indicator, and moving dishes in and out of CO2 incubators. > > > >James Denegre, Ph.D. >Associate Research Scientist >Manager, Light Microscopy >The Jackson Laboratory >Bar Harbor, ME 04609 >207.288.6321 > >-----Original Message----- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Armstrong, Brian >Sent: Thursday, February 26, 2009 11:46 AM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors and mu\ineral oil > >Hi James, can you elaborate on your technique " equilibrate the mineral >oil with the medium"? >Thanks, > >Brian D Armstrong PhD >Light Microscopy Core Manager >Beckman Research Institute >City of Hope >Dept of Neuroscience >1450 E Duarte Rd >Duarte, CA 91010 >626-256-4673 x62872 >http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag >ing/Pages/default.aspx > >-----Original Message----- >From: Confocal Microscopy List [mailto:[hidden email]] >On Behalf Of James Denegre >Sent: Thursday, February 26, 2009 8:10 AM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors and mu\ineral oil > >In our lab we equilibrate the mineral oil with the medium to prevent >uptake of components, and have successfully imaged under mineral oil for >five continuous days. > >James Denegre, Ph.D. >Associate Research Scientist >Manager, Light Microscopy >The Jackson Laboratory >Bar Harbor, ME 04609 >207.288.6321 > >-----Original Message----- >From: Confocal Microscopy List [mailto:[hidden email]] >On Behalf Of Beat Ludin >Sent: Wednesday, February 25, 2009 5:05 AM >To: [hidden email] >Subject: Re: CO2, humidity, temp sensors > >Over time, the mineral oil overlay will basically void the medium of >all lipophilic components (vitamin E, for instance). This may or may >not be a problem, but you need to be aware of it and do the proper >controls. > >If you want to check for evaporation, just weigh your sample before >and after the imaging. > >Beat > >At 23:23 12-02-2009, you wrote: > >My Guess is still evaporation of media is killing your > >cells.. Overlay the cell media in each compartment with mineral > >oil. It stops all evaporation but allows CO2 gas exchange to occur > >to maintain pH.. I do this with all my chamber systems and it works > >great. No need to humidify flowing air or gas. > > > > > >Joe Goodhouse > >Confocal Core Lab Manager > >Dept. of Molecular Biology > >Princeton University > >609-258-5432 > > > >Visit us at > ><http://www.molbio/>http://www.molbio1.princeton.edu/facility/confocal/ > > > > > > > >---------- > >From: Confocal Microscopy List > >[mailto:[hidden email]] On Behalf Of Paul Herzmark > >Sent: Thursday, February 12, 2009 4:36 PM > >To: [hidden email] > >Subject: Re: CO2, humidity, temp sensors > > > > > >I knew I would not get away with out details! > > > >I have an Invivo (company) plexiglass incubator box around my > >inverted microscope. 5% CO2 is piped inside to a small space (about > >50 ml) around the multiwell-Labtech (company)-coverslip. The small > >space is created by a glass plate with a frame around it that goes > >over the coverslip.The CO2 passes into the small space via special > >tubing.The tubing goes under incubator-warmed water where it is > >supposed to pick up water vapor without bubbling. > > > >1. The whole incubator, with most of the microscope is warmed to 37 >degrees > >2. Only the small space surrounding the cells is humidified and has >CO2. > > > > > > > > > >On Thu, Feb 12, 2009 at 11:30 AM, Goodhouse, Joseph G. > ><<mailto:[hidden email]>[hidden email]> wrote: > >Paul, > > Could you describe in detail your heating system. Do you > > overlay your cells with mineral oil? This prevents evaporation > > which will alter your salt concentration. > > > > I use a Recordable Temp sensor from Fisher Scientific. Cat No > > 15-007-26. Inexpensive and works well. > > > > > >Joe Goodhouse > >Confocal Core Lab Manager > >Dept. of Molecular Biology > >Princeton University > >609-258-5432 > > > >Visit us at > ><http://www.molbio/>http://www.molbio1.<http://princeton.edu/facility/c >onfocal/>princeton.edu/facility/confocal/ > > > > > > > > > >---------- > >From: Confocal Microscopy List > >[mailto:[hidden email]] On Behalf Of Paul Herzmark > >Sent: Thursday, February 12, 2009 2:24 PM > >To: ><mailto:[hidden email]>[hidden email] >U > >Subject: CO2, humidity, temp sensors > > > >I have a cell dying problem in my microscope when I take pictures of > >them overnight or longer. I suspect the humidity is failing and I > >want to monitor that, CO2 and temperature.(It is not photodamage. > >Cells die in areas away from the objective.) > > > >Can anyone suggest a supplier of recording sensors that I can get > >next to the well slides on my stage? > > > >Thank you! > > > >Paul Herzmark > >Specialist > ><mailto:[hidden email]>[hidden email] > > > >Department of Molecular and Cell Biology > >479 Life Science Addition > >University of California, Berkeley > >Berkeley, CA 94720-3200 > >(510) 643-9603 > >(510) 643-9500 fax > > > > >--------------------------------------------------------------------- > >SECURITY/CONFIDENTIALITY WARNING: >This message and any attachments are intended >solely for the individual or entity to which >they are addressed. This communication may >contain information that is privileged, >confidential, or exempt from disclosure under >applicable law (e.g., personal health >information, research data, financial >information). Because this e-mail has been sent >without encryption, individuals other than the >intended recipient may be able to view the >information, forward it to others or tamper with >the information without the knowledge or consent >of the sender. If you are not the intended >recipient, or the employee or person responsible >for delivering the message to the intended >recipient, any dissemination, distribution or >copying of the communication is strictly >prohibited. If you received the communication in >error, please notify the sender immediately by >replying to this message and deleting the >message and any accompanying files from your >system. If, due to the security risks, you do >not wish to receive further communications via >e-mail, please reply to this message and inform >the sender that you do not wish to receive further e-mail from the sender. >--------------------------------------------------------------------- |
![]() |
Beat,
Your points are well taken. For us, the test of the mineral oil has been the survival of two-cell mouse embryos to blastocyst stage in culture. The need to equilibrate mineral oil was discovered empirically by John Eppig (O'Brien MJ, Wigglesworth K and Eppig JJ (1993) Mouse oocyte and embryo culture. In Chapin RE and Heudel J (eds) Methods in Reproductive Toxicology, Vol 3B. Academic Press, New York. pp. 128-141). I am sure some cell lines may not benefit from the mineral oil overlay, and different media may not equilibrate the oil as well; but it is certainly inexpensive enough to give it a try. James Denegre, Ph.D. Associate Research Scientist Manager, Light Microscopy The Jackson Laboratory Bar Harbor, ME 04609 207.288.6321 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Beat Ludin Sent: Friday, February 27, 2009 2:23 AM To: [hidden email] Subject: Re: Mineral Oil James - Have you verified that this protocol is sufficient to really equilibrate the mineral oil, i.e. that the composition of medium is no longer affect by contact with the equilibrated mineral oil? Or are you just going by cell survival/growth, which is certainly a good indicator but doesn't necessarily mean that the cells are affected in a non-lethal way? Also, still need to consider the effects of the mineral oil if you want to add compounds during the experiment. BTW, I don't mean to question that a mineral oil overlay is suitable for many applications, but as always, it's good to be aware of possible side-effects. Beat --------------------------------- |