To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, I am looking for coverglass with anti-reflection coating. Are there any suppliers/ manufactures in US/ Canada. Thanks in advance for your help. Thomas A Thomas Abraham, PhD National Research Council-CNBC Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. 459, Stn 18 Chalk River, ON,K0J 1J0 (613) 584 8811 ext.6473 Fax: (613) 584 4040 |
|
Deron Walters |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > Hi all, I am looking for coverglass with anti-reflection coating. Are > there any suppliers/ manufactures in US/ Canada. Thanks in advance for > your help. > > Thomas A > > > Thomas Abraham, PhD > National Research Council-CNBC > Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. > 459, Stn 18 Chalk River, ON,K0J 1J0 > > (613) 584 8811 ext.6473 > Fax: (613) 584 4040 You might consider these ultra-thin windows: http://www.edmundoptics.com/onlinecatalog/displayproduct.cfm?productID=3241 You would need to use these with an objective that has coverslip thickness adjustment, since they are 0.20 mm thick instead of 0.17 mm. The index and dispersion may also be slightly off from normal coverslips. The price is quite high, but the optical quality and antireflection coating are excellent. No commercial interest. Good luck--Deron Deron Walters Staff Scientist, Physics Asylum Research |
|
j jureller |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You could try asking esco: http://www.escoproducts.com/ They sell coverslips and would probably be willing to add an AR coating for you. No commercial interest. We've used them in the past for odd custom optics (including super-thin windows) and they've always been good. regards justin On Wed, Mar 30, 2011 at 5:37 PM, Deron Walters <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > >> Hi all, I am looking for coverglass with anti-reflection coating. Are >> there any suppliers/ manufactures in US/ Canada. Thanks in advance for >> your help. >> >> Thomas A >> >> >> Thomas Abraham, PhD >> National Research Council-CNBC >> Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. >> 459, Stn 18 Chalk River, ON,K0J 1J0 >> >> (613) 584 8811 ext.6473 >> Fax: (613) 584 4040 > > > You might consider these ultra-thin windows: > > http://www.edmundoptics.com/onlinecatalog/displayproduct.cfm?productID=3241 > > You would need to use these with an objective that has coverslip thickness adjustment, since they are 0.20 mm thick instead of 0.17 mm. The index and dispersion may also be slightly off from normal coverslips. The price is quite high, but the optical quality and antireflection coating are excellent. > > No commercial interest. Good luck--Deron > > Deron Walters > Staff Scientist, Physics > Asylum Research > |
|
Sudipta Maiti |
|
|
In reply to this post by Deron Walters
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Just a word of caution while using AR coated coverslips with high NA lenses: In general, AR coating depends on angle of incidence, polarization of the light, and the refractive index of the medium. Sudipta On Wed, 30 Mar 2011 22:37:51 +0000, Deron Walters wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > > Hi all, I am looking for coverglass with anti-reflection coating. Are > > there any suppliers/ manufactures in US/ Canada. Thanks in advance for > > your help. > > > > Thomas A > > > > > > Thomas Abraham, PhD > > National Research Council-CNBC > > Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. > > 459, Stn 18 Chalk River, ON,K0J 1J0 > > > > (613) 584 8811 ext.6473 > > Fax: (613) 584 4040 > > You might consider these ultra-thin windows: > > http://www.edmundoptics.com/onlinecatalog/displayproduct.cfm?productID=3241 > > You would need to use these with an objective that has coverslip > thickness adjustment, since they are 0.20 mm thick instead of 0.17 > mm. The index and dispersion may also be slightly off from normal > coverslips. The price is quite high, but the optical quality and > antireflection coating are excellent. > > No commercial interest. Good luck--Deron > > Deron Walters > Staff Scientist, Physics > Asylum Research Dr. Sudipta Maiti Associate Professor Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.wetpaint.com |
|
Michael Weber-4 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear list, I am also looking for coverslips with AR coating and would be happy if you share your experiences. Ideal for my purpose would be round 18mm coverslips with 0.17mm thickness. best, Michael On Mar 31, 2011, at 9:52 AM, Sudipta Maiti wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Just a word of caution while using AR coated coverslips with high NA lenses: > In general, AR coating depends on angle of incidence, polarization of the > light, and the refractive index of the medium. > Sudipta > On Wed, 30 Mar 2011 22:37:51 +0000, Deron Walters wrote >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >>> Hi all, I am looking for coverglass with anti-reflection coating. Are >>> there any suppliers/ manufactures in US/ Canada. Thanks in advance for >>> your help. >>> >>> Thomas A >>> >>> >>> Thomas Abraham, PhD >>> National Research Council-CNBC >>> Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. >>> 459, Stn 18 Chalk River, ON,K0J 1J0 >>> >>> (613) 584 8811 ext.6473 >>> Fax: (613) 584 4040 >> >> You might consider these ultra-thin windows: >> >> http://www.edmundoptics.com/onlinecatalog/displayproduct.cfm?productID=3241 >> >> You would need to use these with an objective that has coverslip >> thickness adjustment, since they are 0.20 mm thick instead of 0.17 >> mm. The index and dispersion may also be slightly off from normal >> coverslips. The price is quite high, but the optical quality and >> antireflection coating are excellent. >> >> No commercial interest. Good luck--Deron >> >> Deron Walters >> Staff Scientist, Physics >> Asylum Research > > > Dr. Sudipta Maiti > Associate Professor > Dept. of Chemical Sciences > Tata Institute of Fundamental Research > Homi Bhabha Raod, Colaba, Mumbai 400005 > Ph. 91-22-2278-2716 / 2539 > Fax: 91-22-2280-4610 > alternate e-mail: [hidden email] > url: biophotonics.wetpaint.com _____________ Michael Weber PhD Student, Huisken lab Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstrasse 108, 01307 Dresden Tel. 0351/2102837 http://www.mpi-cbg.de/research/research-groups/jan-huisken |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** AR coatings are fairly expensive, so I think you'd be hard-pressed to find somebody who would make them for coverslips since coverslips are usually considered disposable. That said, if you have the cash you could get some custom-coated. A run would probably cost you in the $2000-4000 dollar range though... If you plan on re-using them you'd also want them made from a fairly sturdy material. Craig On Wed, Jun 6, 2012 at 3:43 AM, Michael Weber <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear list, > > I am also looking for coverslips with AR coating and would be happy if you > share your experiences. Ideal for my purpose would be round 18mm coverslips > with 0.17mm thickness. > > best, > Michael > > > On Mar 31, 2011, at 9:52 AM, Sudipta Maiti wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Just a word of caution while using AR coated coverslips with high NA > lenses: > > In general, AR coating depends on angle of incidence, polarization of the > > light, and the refractive index of the medium. > > Sudipta > > On Wed, 30 Mar 2011 22:37:51 +0000, Deron Walters wrote > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >>> Hi all, I am looking for coverglass with anti-reflection coating. Are > >>> there any suppliers/ manufactures in US/ Canada. Thanks in advance for > >>> your help. > >>> > >>> Thomas A > >>> > >>> > >>> Thomas Abraham, PhD > >>> National Research Council-CNBC > >>> Steacie Institute for Molecular Sciences Chalk River Laboratories Bldg. > >>> 459, Stn 18 Chalk River, ON,K0J 1J0 > >>> > >>> (613) 584 8811 ext.6473 > >>> Fax: (613) 584 4040 > >> > >> You might consider these ultra-thin windows: > >> > >> > http://www.edmundoptics.com/onlinecatalog/displayproduct.cfm?productID=3241 > >> > >> You would need to use these with an objective that has coverslip > >> thickness adjustment, since they are 0.20 mm thick instead of 0.17 > >> mm. The index and dispersion may also be slightly off from normal > >> coverslips. The price is quite high, but the optical quality and > >> antireflection coating are excellent. > >> > >> No commercial interest. Good luck--Deron > >> > >> Deron Walters > >> Staff Scientist, Physics > >> Asylum Research > > > > > > Dr. Sudipta Maiti > > Associate Professor > > Dept. of Chemical Sciences > > Tata Institute of Fundamental Research > > Homi Bhabha Raod, Colaba, Mumbai 400005 > > Ph. 91-22-2278-2716 / 2539 > > Fax: 91-22-2280-4610 > > alternate e-mail: [hidden email] > > url: biophotonics.wetpaint.com > > _____________ > > Michael Weber > PhD Student, Huisken lab > Max Planck Institute of Molecular Cell Biology and Genetics > Pfotenhauerstrasse 108, 01307 Dresden > Tel. 0351/2102837 > > http://www.mpi-cbg.de/research/research-groups/jan-huisken > |