Michael Moore |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm a new member to the confocal listserv Can someone explain GDD in regards to multiphoton imaging? And how is GDD adjusted. We currently have a Coherent chameleon vision II that has GDD. I have been told conflicting instructions on how the GDD is adjusted or used for imaging. Maybe I just don't understand how GDD works. for one, is GDD adjusted for each objective used? is it independent of the objective? Thanks to all who reply. Mike |
![]() ![]() |
Michael Giacomelli |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Mike, GDD is the group delay dispersion, the difference in time between different wavelengths of light when passing through a system. It represents the tendency of shorter wavelength components of a pulse to travel at a different speed through glass than longer wavelengths. Its is the temporal analog of chromatic aberration if you are familiar with that concept. The GDD depends on: 1) The amount and type of glass in the system (and therefore on the objectives used) 2) The wavelength of light (typical glasses have lower GDD as one approaches 1.3 um). 3) The length of the pulse of your laser (short pulses are very sensitive to GDD) In your case, your chameleon has a relatively long pulse length (140 fs), and will therefore not be very sensitive to even relatively large amounts of GDD. Personally, I do not bother compensating GDD on our Chameleon because I found that it had a very minor effect on image contrast (although I am using scan optics that are optimized for low GDD). However, since you have GDD correction built in, you might as well use it. Take a slide filled with the fluorophore of your choice and use your preferred objective + wavelength. Then adjust GDD until the image brightness is maximized. I recommend using low power and concentrated fluorophore to minimize photobleaching. You can repeat this for each objective and wavelength if you are determined, but the difference will likely be insignificant unless you are using very widely spaced wavelengths or extremely dispersive objectives. Hope that helps, Mike On Wed, Nov 12, 2014 at 3:44 PM, Michael Moore <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm a new member to the confocal listserv > > Can someone explain GDD in regards to multiphoton imaging? > And how is GDD adjusted. > > We currently have a Coherent chameleon vision II that has GDD. > > I have been told conflicting instructions on how the GDD is adjusted or > used for imaging. > > Maybe I just don't understand how GDD works. > > for one, is GDD adjusted for each objective used? is it independent of the > objective? > > Thanks to all who reply. > > Mike > |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Michael Moore
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** That is a question requiring a fairly detailed answer. I have a short paper on it I wrote for Photonics West 2009 that I've passed along. Hopefully that will get you started, and I recommend you skim my references. Craig Brideau On Wed, Nov 12, 2014 at 1:44 PM, Michael Moore <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm a new member to the confocal listserv > > Can someone explain GDD in regards to multiphoton imaging? > And how is GDD adjusted. > > We currently have a Coherent chameleon vision II that has GDD. > > I have been told conflicting instructions on how the GDD is adjusted or > used for imaging. > > Maybe I just don't understand how GDD works. > > for one, is GDD adjusted for each objective used? is it independent of the > objective? > > Thanks to all who reply. > > Mike > |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Michael Giacomelli
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Note that short pulses can also increase photodamage, so you may have to find a balance between pulse-width, power, signal, and damage. Craig On Wed, Nov 12, 2014 at 2:21 PM, Michael Giacomelli <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Mike, > > GDD is the group delay dispersion, the difference in time between different > wavelengths of light when passing through a system. It represents the > tendency of shorter wavelength components of a pulse to travel at a > different speed through glass than longer wavelengths. Its is the temporal > analog of chromatic aberration if you are familiar with that concept. > > The GDD depends on: > > 1) The amount and type of glass in the system (and therefore on the > objectives used) > 2) The wavelength of light (typical glasses have lower GDD as one > approaches 1.3 um). > 3) The length of the pulse of your laser (short pulses are very sensitive > to GDD) > > In your case, your chameleon has a relatively long pulse length (140 fs), > and will therefore not be very sensitive to even relatively large amounts > of GDD. Personally, I do not bother compensating GDD on our Chameleon > because I found that it had a very minor effect on image contrast (although > I am using scan optics that are optimized for low GDD). > > However, since you have GDD correction built in, you might as well use it. > Take a slide filled with the fluorophore of your choice and use your > preferred objective + wavelength. Then adjust GDD until the image > brightness is maximized. I recommend using low power and concentrated > fluorophore to minimize photobleaching. You can repeat this for each > objective and wavelength if you are determined, but the difference will > likely be insignificant unless you are using very widely spaced wavelengths > or extremely dispersive objectives. > > Hope that helps, > Mike > > > On Wed, Nov 12, 2014 at 3:44 PM, Michael Moore <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > I'm a new member to the confocal listserv > > > > Can someone explain GDD in regards to multiphoton imaging? > > And how is GDD adjusted. > > > > We currently have a Coherent chameleon vision II that has GDD. > > > > I have been told conflicting instructions on how the GDD is adjusted or > > used for imaging. > > > > Maybe I just don't understand how GDD works. > > > > for one, is GDD adjusted for each objective used? is it independent of > the > > objective? > > > > Thanks to all who reply. > > > > Mike > > > |
![]() ![]() |
Mark Cannell-2 |
![]() |
In reply to this post by Michael Moore
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Have a read of this, I think it gives a relatively easy to understand description of the problem and its solution: Soeller, C. & Cannell, M.B., 1996. Construction of a two-photon microscope and optimisation of illumination pulse duration. Pflügers Archiv : European journal of physiology, 432(3), pp.555–561. HTH Cheers On 12/11/2014, at 8:44 pm, Michael Moore <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm a new member to the confocal listserv > > Can someone explain GDD in regards to multiphoton imaging? > And how is GDD adjusted. > > We currently have a Coherent chameleon vision II that has GDD. > > I have been told conflicting instructions on how the GDD is adjusted or > used for imaging. > > Maybe I just don't understand how GDD works. > > for one, is GDD adjusted for each objective used? is it independent of the > objective? > > Thanks to all who reply. > > Mike Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Really good early work. I basically redesigned the prisms but used the same compressor design as Soeller. The key change I made for my own work was to use a higher-index glass so the prisms didn't have to be so far apart to achieve a given -GVD. The trick was to balance out the Abbe value for the bandwidth of the laser so I induced mainly second order but not much third order dispersion. Craig On Wed, Nov 12, 2014 at 4:06 PM, Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi > > Have a read of this, I think it gives a relatively easy to understand > description of the problem and its solution: > Soeller, C. & Cannell, M.B., 1996. Construction of a two-photon microscope > and optimisation of illumination pulse duration. Pflügers Archiv : European > journal of physiology, 432(3), pp.555–561. > > HTH > > Cheers > > > > On 12/11/2014, at 8:44 pm, Michael Moore <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > I'm a new member to the confocal listserv > > > > Can someone explain GDD in regards to multiphoton imaging? > > And how is GDD adjusted. > > > > We currently have a Coherent chameleon vision II that has GDD. > > > > I have been told conflicting instructions on how the GDD is adjusted or > > used for imaging. > > > > Maybe I just don't understand how GDD works. > > > > for one, is GDD adjusted for each objective used? is it independent of > the > > objective? > > > > Thanks to all who reply. > > > > Mike > > Mark B. Cannell Ph.D. FRSNZ > Professor of Cardiac Cell Biology > School of Physiology & Pharmacology > Medical Sciences Building > University of Bristol > Bristol > BS8 1TD UK > > [hidden email] > |
![]() ![]() |
Debora Keller |
![]() |
In reply to this post by Michael Moore
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, Following up on the effect of pulse-width and photo-toxicity, I would also have a look at the following paper from the Beaurepaire lab: Débarre D, Olivier N, Supatto W, Beaurepaire E (2014) Mitigating Phototoxicity during Multiphoton Microscopy of Live Drosophila Embryos in the 1.01.2 µm Wavelength Range. PLoS ONE 9(8): e104250. doi:10.1371/journal.pone.0104250 Principles should be applicable to other organisms (me thinks) Debbi |