http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, It's just a theoretical question. Can multiphoton microscopes be classified as confocal microscopes? Is there any reference published about this type of microscope classification? Thank you very much in advance -- Juan Luis Ribas Servicio de Microscopía Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b 41012 Sevilla Tfno: 954559983 |
|
Michael Weber-4 |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Juan, a multiphoton system is not confocal. For which purpose do you need the classification? Michael Juan Luis Ribas wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Dear all, > It's just a theoretical question. Can multiphoton microscopes be > classified as confocal microscopes? > Is there any reference published about this type of microscope > classification? > > Thank you very much in advance |
|
Beat Ludin |
|
|
In reply to this post by jlribas
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal That's a question of definition. IMHO, a MP microscope can usually be used both in confocal (when imaging with descanned detectors behind a pinhole) or non-confocal modes (when imaging with descanned detectors). Beat At 10:27 23-07-2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear all, >It's just a theoretical question. Can >multiphoton microscopes be classified as confocal microscopes? >Is there any reference published about this type of microscope classification? > >Thank you very much in advance > >-- >Juan Luis Ribas >Servicio de Microscopía >Centro de Investigación, Tecnología e Innovación >Universidad de Sevilla >Av. Reina Mercedes 4b >41012 Sevilla > >Tfno: 954559983 |
|
George McNamara |
|
|
In reply to this post by jlribas
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Juan, confocal refers to conjugate pinhole. So on a MP-only microscope with non-descanned detector(s): no. However, most MP systems are mounted on a confocal microscope and the confocal detectors are often used instead of the NDD's. The confocal detector pinhole(s) are usually adjustable (typically to 1 Airy unit, based on the objective lens NA), but even a "wide open" pinhole may be much much smaller than if there was no pinhole aperture diaphragm in place (an NDD condition!). So, most MP microscopes are multiphoton+confocal. When writing the methods section, be sure to specify what light path is used and whether using NDD or confocal detector, and if the latter, the number of Airy units for the pinhole size (there are no laws forcing you to use 1 Airy unit, in fact 2 Airy units will typically double the amount of light collected). George p.s. on an MP system, the confocal detectors reject stray light, so if the room is not dark (an MP scope needs to have a black shroud!) the confocal detectors will usually have a lower background than the NDD's. A simple test is to shine a flashlight at the specimen while scanning with either. At 04:27 AM 7/23/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear all, >It's just a theoretical question. Can >multiphoton microscopes be classified as confocal microscopes? >Is there any reference published about this type of microscope classification? > >Thank you very much in advance > >-- >Juan Luis Ribas >Servicio de Microscopía >Centro de Investigación, Tecnología e Innovación >Universidad de Sevilla >Av. Reina Mercedes 4b >41012 Sevilla > >Tfno: 954559983 George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (Analytical Imaging Core Facility) |
|
|
In reply to this post by jlribas
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal However, taking into account all these factors, an MP microscope is 'perfectly' confocal, whereas no 'real' confocal microscope can ever be. (Unless you have some way of making an infinitely small pinhole.) See: Guy Cox & Colin Sheppard, 2004. Practical limits of resolution in confocal and non-linear microscopy. Microscopy Research & Technique, 63, 18-22 Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara Sent: Monday, 28 July 2008 9:30 AM To: [hidden email] Subject: Re: Classification of confocal microscopes Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Juan, confocal refers to conjugate pinhole. So on a MP-only microscope with non-descanned detector(s): no. However, most MP systems are mounted on a confocal microscope and the confocal detectors are often used instead of the NDD's. The confocal detector pinhole(s) are usually adjustable (typically to 1 Airy unit, based on the objective lens NA), but even a "wide open" pinhole may be much much smaller than if there was no pinhole aperture diaphragm in place (an NDD condition!). So, most MP microscopes are multiphoton+confocal. When writing the methods section, be sure to specify what light path is used and whether using NDD or confocal detector, and if the latter, the number of Airy units for the pinhole size (there are no laws forcing you to use 1 Airy unit, in fact 2 Airy units will typically double the amount of light collected). George p.s. on an MP system, the confocal detectors reject stray light, so if the room is not dark (an MP scope needs to have a black shroud!) the confocal detectors will usually have a lower background than the NDD's. A simple test is to shine a flashlight at the specimen while scanning with either. At 04:27 AM 7/23/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear all, >It's just a theoretical question. Can >multiphoton microscopes be classified as confocal microscopes? >Is there any reference published about this type of microscope classification? > >Thank you very much in advance > >-- >Juan Luis Ribas >Servicio de Microscopía >Centro de Investigación, Tecnología e Innovación Universidad de Sevilla >Av. Reina Mercedes 4b >41012 Sevilla > >Tfno: 954559983 George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (Analytical Imaging Core Facility) No virus found in this incoming message. Checked by AVG. Version: 7.5.526 / Virus Database: 270.5.6/1576 - Release Date: 27/07/2008 4:16 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.526 / Virus Database: 270.5.6/1576 - Release Date: 27/07/2008 4:16 PM |
|
Beat Ludin |
|
|
In reply to this post by jlribas
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Oops, that should have been "...or non-confocal modes (when imaging with NON-descanned detectors).", of course. Sorry for the confusion Beat At 10:27 23-07-2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear all, >It's just a theoretical question. Can >multiphoton microscopes be classified as confocal microscopes? >Is there any reference published about this type of microscope classification? > >Thank you very much in advance > >-- >Juan Luis Ribas >Servicio de Microscopía >Centro de Investigación, Tecnología e Innovación >Universidad de Sevilla >Av. Reina Mercedes 4b >41012 Sevilla > >Tfno: 954559983 |