Arvydas Matiukas |
![]() |
Our Coherent Chameleon laser which was ocassionally
used for 2p or 2p-DAPI excitation on LSM510 system
recently began performing unstably, sometimes even
not getting "modelocked".
The laser is 4 years old but was used only 1000 hrs.
However, Coherent engineer states that lasing crystals
may have degraded and suggest to replace them for $15-30k
with one more year warranty.
I would appreciate any advice/comment on cheaper options
and laser lifetime.
Thanks,
Arvydas
*************
Arvydas Matiukas, Ph.D.
Director of Confocal&Two-Photon Imaging Core Facility Department of Pharmacology SUNY Upstate Medical University 766 Irving Ave., WH 3159 Syracuse, NY 13210 tel.: 315-464-7997 fax: 315-464-8014 email: [hidden email] |
![]() ![]() |
Craig Brideau |
![]() |
If you've only put 1000 hours on it, I doubt its your crystals unless your laser is in a very 'dirty' environment; dust in the cavity and the like. There's supposed to be a filtered air purge though to keep particles out of the laser to prevent this.
It's possible there is a bad coating on an optic somewhere that has failed, the green pump laser may be going, or one of any of a number of things could be wrong. My advise to you would be to have the laser thoroughly checked over before you start buying replacement parts. Craig On Tue, Nov 11, 2008 at 4:11 PM, Arvydas Matiukas <[hidden email]> wrote:
|
![]() ![]() |
Haberman, Ann |
![]() |
In reply to this post by Arvydas Matiukas
Hi Arvydas,
Have you recently adjusting any components within the microscope? Laser feedback into the laser cavity can cause this as well. Ann Haberman >Content-type: text/html; charset=US-ASCII >Content-description: HTML > >Our Coherent Chameleon laser which was ocassionally >used for 2p or 2p-DAPI excitation on LSM510 system >recently began performing unstably, sometimes even >not getting "modelocked". > >The laser is 4 years old but was used only 1000 hrs. >However, Coherent engineer states that lasing crystals >may have degraded and suggest to replace them for $15-30k >with one more year warranty. > >I would appreciate any advice/comment on cheaper options >and laser lifetime. > >Thanks, >Arvydas >************* > > > > >Arvydas Matiukas, Ph.D. >Director of Confocal&Two-Photon Imaging Core Facility >Department of Pharmacology >SUNY Upstate Medical University >766 Irving Ave., WH 3159 >Syracuse, NY 13210 >tel.: 315-464-7997 >fax: 315-464-8014 >email: [hidden email] -- Ann Haberman, PhD Department of Laboratory Medicine Yale University School of Medicine 1 Gilbert St. TAC S541 New Haven, CT 06510 203-785-7349 203-785-5415 (fax) [hidden email] |
![]() ![]() |
Csúcs Gábor |
![]() |
In reply to this post by Arvydas Matiukas
Dear Arvydas,
Unfortunately I have no good advice, just an own experience: The same thing happened to us, our laser had even less then 1000 hours, was only 2.5 years old and was not in a dirty environment. The repair cost 40.000 USD. Cheers Gabor |
![]() ![]() |
Scientist Two Photon Spectroscopy |
![]() |
In reply to this post by Arvydas Matiukas
Hi all,
I have the same experience with both of our chameleon lasers. After a period of time the laser don´t modelock anymore. You can try to smash with your hand on the case to cause the hard aperture to vibrate. It will modelock again. But it only works for three weeks. After than you can pay 40.000 $ for a repair. The laser will have than the half of lifetime before again. This is very disappointing. I also did not get a detailed repair description, only "Reapir : xxx$" I guess the repair shop did not know the reason for the modelocking problems either. In my view the cavity design especially the control of the piezo mirrors is the main cause for the modelocking problems. It is not stable at all, especially when it shipped in fragile wooden box all over the world. Now we are using fs-lasers from another manufacturer. I hope you are not so disappointed at all. The lasers are susceptible to failure. You have to take it into account. Regards, Bill |
![]() ![]() |
LANCE RODENKIRCH |
![]() |
In reply to this post by Arvydas Matiukas
Hi Arvydas,
I would check my baseplate temperature. Main menu/Pump laser menu/verdi laser status/baseplate temp. The baseplate should be 33 C. Are you using the original Thermotek pump that was shipped with the unit? Coherent advises to change the chiller water monthly. Can you achieve modelock by doing a pzt recovery? Main menu/chameleon status/PZT recovery. If this is a temperature problem, you should get error codes number 44 and 50 on the front interface. I have 6K hours on a 5 year old Chameleon but I have been having similar issues since July. To other Chameleon operators out there, how often are they doing PZT recoveries? It would be best to start the laser at a known calibration point such as 800nm and then tune to the desired wavelength. A standard Chameleon has 23 calibrated points across the tuning range. The system will first attempt to modelock at the desired wavelength. Should it not modelock on the first try, it will tune to the next closest calibrated point, modelock, then tune back to the desired wavelength so that a modelock pulse is obtained. Lance Rodenkirch W.M Keck Laboratory for Biological Imaging University of Wisconsin School of Medicine 1300 University Ave Rm 172 MSC Madison, WI 53706 http://www.keck.bioimaging.wisc.edu/ Phone: 608.265.5651 Fax: 608.265.5536 ----- Original Message ----- From: Arvydas Matiukas <[hidden email]> Date: Tuesday, November 11, 2008 5:12 pm Subject: Coherent Chameleon laser lifetime To: [hidden email] > Our Coherent Chameleon laser which was ocassionally > used for 2p or 2p-DAPI excitation on LSM510 system > recently began performing unstably, sometimes even > not getting "modelocked". > > The laser is 4 years old but was used only 1000 hrs. > However, Coherent engineer states that lasing crystals > may have degraded and suggest to replace them for $15-30k > with one more year warranty. > > I would appreciate any advice/comment on cheaper options > and laser lifetime. > > Thanks, > Arvydas > ************* > > > > > Arvydas Matiukas, Ph.D. > Director of Confocal&Two-Photon Imaging Core Facility > Department of Pharmacology > SUNY Upstate Medical University > 766 Irving Ave., WH 3159 > Syracuse, NY 13210 > tel.: 315-464-7997 > fax: 315-464-8014 > email: [hidden email] > |
![]() ![]() |
G. Esteban Fernandez |
![]() |
We had modelock problems for some time but the laser has been very reliable (or at least predictable) since we've been setting the chiller temp. to 27°C at start-up. This is higher than the what the laser should be operating at and may give a diode overcurrent fault, so we only use that temp. for start-up. Once it modelocks we turn the chiller back down to 23°C. Of course it doesn't like the temp. change but it seems less harsh than smashing it with your hand :) Contact me off-list for the dirty details of how we keep modelock without having to spend $40k.
-Esteban
On Wed, Nov 12, 2008 at 12:37 PM, LANCE RODENKIRCH <[hidden email]> wrote:
Hi Arvydas, -- G. Esteban Fernandez, Ph.D. Associate Director Molecular Cytology Core Facility University of Missouri 120 Bond Life Sciences Center Columbia, MO 65211 http://www.biotech.missouri.edu/mcc/ 573-882-4895 573-884-9395 fax |
![]() ![]() |
Arvydas Matiukas |
![]() |
Hi all,
thanks for the suggestions: after checking/setting chiller/baseplate temperatures our Chameleon laser
again is stable as usual.
Actually Chameleon manual states that proper temperature settings are important, and now based on my own experience
I can confirm that they are REALLY important. Fortunately the laser performance has nothing
to do with crystal degradation as follows from the replies and my enquiries to Ti:Sa laser owners.
Arvydas
We had modelock problems for some time but the laser has been very reliable (or at least predictable) since we've been setting the chiller temp. to 27°C at start-up. This is higher than the what the laser should be operating at and may give a diode overcurrent fault, so we only use that temp. for start-up. Once it modelocks we turn the chiller back down to 23°C. Of course it doesn't like the temp. change but it seems less harsh than smashing it with your hand :) Contact me off-list for the dirty details of how we keep modelock without having to spend $40k.
-Esteban
On Wed, Nov 12, 2008 at 12:37 PM, LANCE RODENKIRCH <[hidden email]> wrote: Hi Arvydas, -- G. Esteban Fernandez, Ph.D. Associate Director Molecular Cytology Core Facility University of Missouri 120 Bond Life Sciences Center Columbia, MO 65211 http://www.biotech.missouri.edu/mcc/ 573-882-4895 573-884-9395 fax |
![]() ![]() |
Craig Brideau |
![]() |
Hooray! Glad it was a simple solution for you. Every time there's a hiccup like that with an expensive laser system it can be stressful for everyone!
Craig On Tue, Nov 18, 2008 at 10:44 AM, Arvydas Matiukas <[hidden email]> wrote:
|