Commercial Reply - Re: Problem with old MRC 1024


Dear Joe,

Firstly, Carl Zeiss still offer support and service (on a "best effort"
basis) for the MRC-1024 range of LSMs.

The error you report does not necessarily mean you have a problem with your
laser board, any break in the H8 communication chain will generate this
error message; as an example, if you disconnect the serial cable from the
computer and try to start LaserSharp (real hardware mode) you will get the
same error.  Therefore there may be a problem somewhere else in the comms
chain.

Additionally, this error is not connected to the VIS board, therefore
replacing this will not resolve the problem.

We will contact you directly to offer assistance in fault finding the
likely cause and then suggest a plan of action.

Yours sincerely,

Dino

Carl Zeiss MicroImaging GmbH
Carl Zeiss Gruppe/Carl Zeiss Group
Cell Science Products International Customer Support Group
Service Support BioSciences Jena (UK based)

D i n o   S h a r m a

International/UK Customer Support Enquiries - Tel: +44 1707 871 354

International/UK Customer Support Enquiries -
mailto:[hidden email]

http://www.zeiss.de/lsm

********************************************************
For UK service scheduling:

UK Customers Only: 01707 871 300

UK Customers Only - mailto:[hidden email]
********************************************************

Joe Shope <[hidden email]>
Sent by: Confocal Microscopy List <[hidden email]>


10/01/2012 20:47

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We have an old MRC 1024 that recently has stopped functioning. It gives the

error "Firmware cannot be downloaded to laser board". Has anyone else
experienced a problem like this or have any old vis & laser boards sitting
around?

Thanks,
joe


----------------------------------------
This message is intended for a particular addressee only and may contain business or company secrets. If you have received this email in error, please contact the sender and delete the message immediately. Any use of this email, including saving, publishing, copying, replication or forwarding of the message or the contents is not permitted.