Eric Shelden |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** All: I recently had a chance to look at a 40X water immersion 1.1NA Corr lens on our Leica SP5 using both conventional and multiphoton illumination. Theoretically, it should have given better resolution and better brightness than our 20X 0.7 Corr multi-immersion lens. re, with two different specimens and preparations it provided only slightly better resolution and prequired higher gain settings to achieve saturation with the same laser power settings. I was wondering if there was a reasonable explanation, and if anyone has had experience with the 40X lens in question. Thanks, Eric |
|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Eric, When you did your comparison, did you change the digital sampling rate (number of pixels 512x512, 1024x1024, etc.) or the scan zoom settings between the two objectives, or were the settings all the same for both objective tests? How are you judging resolution? Are you looking at the FWHM of sub-diffraction sized fluorescent beads in XY and Z? What are the test specimens? What medium are they embedded in (dry/air, water, oil mounting, etc.). Is there a coverslip present or not? Answers to these questions might point out some key factors that might explain the result you got. John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2015-03-13, at 1:19 PM, Eric Shelden wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > All: > I recently had a chance to look at a 40X water immersion 1.1NA Corr lens > on our Leica SP5 using both conventional and multiphoton illumination. > Theoretically, it should have given better resolution and better brightness > than our 20X 0.7 Corr multi-immersion lens. re, with two different > specimens and preparations it provided only slightly better resolution and > prequired higher gain settings to achieve saturation with the same laser > power settings. I was wondering if there was a reasonable explanation, and > if anyone has had experience with the 40X lens in question. > Thanks, > Eric |
|
Paul Rigby-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Eric, Unless you have a significant problem with your test objective (someone has dropped it), I would suspect that the differences you observed were more likely due to incorrect specimen preparation. If you image a single fluorescent bead, is there evidence of any spherical aberration? (Is the shape of the PSF identical above and below the bead focal plane?) If you are using a coverslip, how thick is that coverslip (is it = 0.17mm thick?). Have you adjusted any correction collar on your objectives to achieve optimal imaging? For the WI objective, is your sample embedded in water (or does the mounting medium have a RI = 1.33)? These issues of correct specimen preparation are critical, especially for higher NA objectives. Hope this helps. Cheers Paul Assoc. Prof. Paul Rigby Optical Microscopy Specialist Centre for Microscopy, Characterisation & Analysis (M519) The University of Western Australia 35 Stirling Highway Crawley WA 6007 Australia -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos Sent: Saturday, 14 March 2015 2:01 AM To: [hidden email] Subject: Re: Comparing objective lens performance ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Eric, When you did your comparison, did you change the digital sampling rate (number of pixels 512x512, 1024x1024, etc.) or the scan zoom settings between the two objectives, or were the settings all the same for both objective tests? How are you judging resolution? Are you looking at the FWHM of sub-diffraction sized fluorescent beads in XY and Z? What are the test specimens? What medium are they embedded in (dry/air, water, oil mounting, etc.). Is there a coverslip present or not? Answers to these questions might point out some key factors that might explain the result you got. John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2015-03-13, at 1:19 PM, Eric Shelden wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > All: > I recently had a chance to look at a 40X water immersion 1.1NA Corr > lens on our Leica SP5 using both conventional and multiphoton illumination. > Theoretically, it should have given better resolution and better > brightness than our 20X 0.7 Corr multi-immersion lens. re, with two > different specimens and preparations it provided only slightly better > resolution and prequired higher gain settings to achieve saturation > with the same laser power settings. I was wondering if there was a > reasonable explanation, and if anyone has had experience with the 40X lens in question. > Thanks, > Eric |
|
Zdenek Svindrych-2 |
|
|
In reply to this post by Eric Shelden
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Eric, bear in mind, that the same power settings do not mean the same excitation powers. The two objectives have very different diameters of the apertures at the back focal plane... Best, zdenek ---------- Původní zpráva ---------- Od: Eric Shelden <[hidden email]> Komu: [hidden email] Datum: 13. 3. 2015 13:21:55 Předmět: Comparing objective lens performance "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** All: I recently had a chance to look at a 40X water immersion 1.1NA Corr lens on our Leica SP5 using both conventional and multiphoton illumination. Theoretically, it should have given better resolution and better brightness than our 20X 0.7 Corr multi-immersion lens. re, with two different specimens and preparations it provided only slightly better resolution and prequired higher gain settings to achieve saturation with the same laser power settings. I was wondering if there was a reasonable explanation, and if anyone has had experience with the 40X lens in question. Thanks, Eric" |
|
|
In reply to this post by Eric Shelden
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Am 13.03.2015 14:21 schrieb "Eric Shelden" <[hidden email]>: >40X water immersion 1.1NA Corr lens > should have given better resolution and better brightness > than our 20X 0.7 Corr multi-immersion lens. > it provided only slightly better resolution and > prequired higher gain settings to achieve saturation with the same laser > power settings. I was wondering if there was a reasonable explanation Eric, you can measure the angle of light emitted from small but bright objects in your specimen or of beads. xz scan through a couple of beads or objects. The sinus of the full angle multiplied with the refractive index of your media gives the NA of the detection system. To make sure to achieve full resolution you also need to make sure to overfill the back aperture of the objective. There is a beam expander that should be set to the largest number. Make sure to also adapt the magnification to integrate photons from the same area. Without that you would expect a lower intensity for the 40x, because you integrate from a smaller area. Hope it helps, Jens Visiting Scientist @ Center for Technological Development in Health (CDTS), Oswaldo Cruz Foundation (Fiocruz), Ministry of Health, Rio de Janeiro, Brazil |