![]() ![]() |
Stephen Firth-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, I have both an LSM780 and LSM980 that I would like to keep at 37C continuously without having the main Zeiss controller ON. We are a multi-user Facility and I fear that laser/s and fluorescent light source will be left on when the microscope is not needed if we tell people to keep the main switch on. Is there a way to separate the Pecon heating control from the Zeiss system? I have contacted Pecon through Pecon.biz website twice over the last weeks and received no reply - is there a better contact available? Thanks in advance for your assistance, Stephen -- *Stephen Firth* | Manager | Monash Micro Imaging - Advanced Optical Microscopy 15 Innovation Walk, STRIP 1 Monash Biotechnology, Monash University, Clayton, VIC 3800, Australia Email: [hidden email] | TEL: +61 03 9902 0877 | MOB: 0400 695 425 *http://platforms.monash.edu/mmi <http://platforms.monash.edu/mmi>* |
![]() ![]() |
j.springfield |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Stephen, You can directly connect the Pecon controllers to the wall, all of ours are setup this way Best James Dr James Springfield Microscopy Facility Manager Institute for Molecular Bioscience Level 6 North | 306 Carmody Road (Building 80) | The University of Queensland | Brisbane Qld 4072 Australia T +61 7 3346 2390 | M +61 423 021 445 | F +61 7 3346 2101 | E [hidden email]<mailto:[hidden email]> Twitter @IMBmicroscopy Connect: imb.uq.edu.au<http://imb.uq.edu.au> | Subscribe | Facebook | Twitter This email (including any attached files) is intended solely for the addressee. It may contain private or confidential information. If you are not the intended addressee, you must take no action based on it, nor show a copy to anyone. Kindly notify the sender by reply email. Opinions and information in this email which do not relate to the official business of The University of Queensland shall be understood as neither given nor endorsed by the University. CRICOS Provider Number 00025B. On 27 Nov 2019, at 4:09 pm, Stephen Firth <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, I have both an LSM780 and LSM980 that I would like to keep at 37C continuously without having the main Zeiss controller ON. We are a multi-user Facility and I fear that laser/s and fluorescent light source will be left on when the microscope is not needed if we tell people to keep the main switch on. Is there a way to separate the Pecon heating control from the Zeiss system? I have contacted Pecon through Pecon.biz website twice over the last weeks and received no reply - is there a better contact available? Thanks in advance for your assistance, Stephen -- *Stephen Firth* | Manager | Monash Micro Imaging - Advanced Optical Microscopy 15 Innovation Walk, STRIP 1 Monash Biotechnology, Monash University, Clayton, VIC 3800, Australia Email: [hidden email] | TEL: +61 03 9902 0877 | MOB: 0400 695 425 *http://platforms.monash.edu/mmi <http://platforms.monash.edu/mmi>* |
![]() ![]() |
Dr. K N Ganesh |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Stephen Firth-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi We have been associated with Tokai Hit a Japanese company that makes precision stage top incubator and cage Incubators, They have stand alone controller for the incubator. Both performance and support is excellent. You may contact them as well. I heard that they recently started their office in US. NO COMMERCIAL INTEREST. Reg Ganesh On Wed, Nov 27, 2019 at 11:43 AM Stephen Firth < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi, > I have both an LSM780 and LSM980 that I would like to keep at 37C > continuously without having the main Zeiss controller ON. We are a > multi-user Facility and I fear that laser/s and fluorescent light source > will be left on when the microscope is not needed if we tell people to keep > the main switch on. > Is there a way to separate the Pecon heating control from the Zeiss system? > > I have contacted Pecon through Pecon.biz website twice over the last weeks > and received no reply - is there a better contact available? > Thanks in advance for your assistance, > Stephen > > -- > *Stephen Firth* | Manager | Monash Micro Imaging - Advanced Optical > Microscopy > 15 Innovation Walk, STRIP 1 Monash Biotechnology, Monash University, > Clayton, VIC 3800, Australia > Email: [hidden email] | TEL: +61 03 9902 0877 | MOB: 0400 695 > 425 > *http://platforms.monash.edu/mmi <http://platforms.monash.edu/mmi>* > ... [show rest of quote] -- With Best Regards, Ganesh Kadasoor |
![]() ![]() |
Rick Mannello |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Contains commercial content: Related to the issue of keeping stable microscope temperature control. There are new ways to do temperature control without having to worry about keeping the microscope, stages and objectives stabilized. The new VAHEAT system from Linnowave GmbH was developed by former researchers at Max Planck. VAHEAT maintains precise temperature control at the sample, in the field of view. No modification to the microscope is needed, the active device fits in the space of a conventional slide. North America: https://www.boselec.com/product-category/microscope-temperature-control-vaheat/ Europe/Asia: www.linnowave.com/vaheat/ Rick Mannello Boston Electronics www.boselec.com [hidden email] -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Ganesh Kadasoor Sent: Wednesday, November 27, 2019 2:20 AM To: [hidden email] Subject: Re: Continuous heating on Zeiss LSM780/LSM980 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi We have been associated with Tokai Hit a Japanese company that makes precision stage top incubator and cage Incubators, They have stand alone controller for the incubator. Both performance and support is excellent. You may contact them as well. I heard that they recently started their office in US. NO COMMERCIAL INTEREST. Reg Ganesh On Wed, Nov 27, 2019 at 11:43 AM Stephen Firth < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi, > I have both an LSM780 and LSM980 that I would like to keep at 37C > continuously without having the main Zeiss controller ON. We are a > multi-user Facility and I fear that laser/s and fluorescent light > source will be left on when the microscope is not needed if we tell > people to keep the main switch on. > Is there a way to separate the Pecon heating control from the Zeiss system? > > I have contacted Pecon through Pecon.biz website twice over the last > weeks and received no reply - is there a better contact available? > Thanks in advance for your assistance, Stephen > > -- > *Stephen Firth* | Manager | Monash Micro Imaging - Advanced Optical > Microscopy > 15 Innovation Walk, STRIP 1 Monash Biotechnology, Monash University, > Clayton, VIC 3800, Australia > Email: [hidden email] | TEL: +61 03 9902 0877 | MOB: 0400 > 695 > 425 > *http://platforms.monash.edu/mmi <http://platforms.monash.edu/mmi>* > ... [show rest of quote] -- With Best Regards, Ganesh Kadasoor |
![]() ![]() |
Stephen Firth-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thank you for the replies, on and off-list. I will go with James' advice and power the heating directly, rather than through the LSM main switch. Would still be good to talk to Pecon direct, if someone had a phone number I would be grateful. Cheers, Stephen -- *Stephen Firth* | Manager | Monash Micro Imaging - Advanced Optical Microscopy 15 Innovation Walk, STRIP 1 Monash Biotechnology, Monash University, Clayton, VIC 3800, Australia Email: [hidden email] | TEL: +61 03 9902 0877 | MOB: 0400 695 425 *http://platforms.monash.edu/mmi <http://platforms.monash.edu/mmi>* |
Free forum by Nabble | Edit this page |