|
|
Hello Lister, Can anyone please offer me suggestions for Javier
Javier F. Adur, Ph.D. Research Center in Optical and Fotonic (CEPOF) Biomedical Lasers Application Laboratory IFGW - UNICAMP - BRASIL
|
|
Steffen Dietzel |
|
|
Hello Javier,
I only came across your e-mail today. And since you apparently didn't get a lot of answers so far... The easiest test specimen is probably some interface of water and glas or water and plastic: big Ri-mismatch = strong THG signal. I use water in a petri dish or on a glass slide to check that our excitation beam is well centered, since it provides a uniform signal. If you need a signal in xy-direction, you should be able to get it by adding a coverslip to the water, so that you can image the edge of the cover slip. If you draw lines on the surface with a black marker pen, it will help to focus in ordinary brightfield illumination, so that it is easier to find the focal plane in THG. If you want to get more fancy, you can use a thin muscle tissue (striated muscle gives nice signals, see our paper J. Biomed. Opt., Vol. 15, 026017 (2010); doi:10.1117/1.3374337) or a fat-water emulsion between coverslip and slide, such as condensed milk or sunscreen, you get the idea. (yes, we've tried that too :-) ) Hope it helps, Steffen At 19:22 05.08.2010, you wrote: >Hello Lister, >Hope all are doing well. > > <?xml:namespace prefix = o ns = "urn:schemas-microsoft-com:office:office" /> > >Can anyone please offer me suggestions for >a test sample (a handy)for imaging THG in transmission? > >Cheers >Javier > > > > Javier F. Adur, Ph.D. > >Research Center in Optical and Fotonic (CEPOF) > > Biomedical Lasers Application Laboratory > > IFGW - UNICAMP - BRASIL > -- --------------------------------------------------------------------------------------------------- Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
|
|
Agoust 12,
Steffen, thank you very much for your response. The information was very nice. Javier.
Javier F. Adur, Ph.D. Research Center in Optical and Fotonic (CEPOF) Biomedical Lasers Application Laboratory IFGW - UNICAMP - BRASIL
> Date: Wed, 11 Aug 2010 16:41:33 +0200 > From: [hidden email] > Subject: Re: Convenient samples for THG > To: [hidden email] > > Hello Javier, > > I only came across your e-mail today. And since > you apparently didn't get a lot of answers so far... > > The easiest test specimen is probably some > interface of water and glas or water and plastic: > big Ri-mismatch = strong THG signal. > > I use water in a petri dish or on a glass slide > to check that our excitation beam is well > centered, since it provides a uniform signal. If > you need a signal in xy-direction, you should be > able to get it by adding a coverslip to the > water, so that you can image the edge of the cover slip. > > If you draw lines on the surface with a black > marker pen, it will help to focus in ordinary > brightfield illumination, so that it is easier to find the focal plane in THG. > > If you want to get more fancy, you can use a thin > muscle tissue (striated muscle gives nice > signals, see our paper J. Biomed. Opt., Vol. 15, > 026017 (2010); doi:10.1117/1.3374337) or a > fat-water emulsion between coverslip and slide, > such as condensed milk or sunscreen, you get the > idea. (yes, we've tried that too :-) ) > > Hope it helps, > > Steffen > > > At 19:22 05.08.2010, you wrote: > > >Hello Lister, > >Hope all are doing well. > > > > <?xml:namespace prefix = o ns = "urn:schemas-microsoft-com:office:office" /> > > > >Can anyone please offer me suggestions for > >a test sample (a handy)for imaging THG in transmission? > > > >Cheers > >Javier > > > > > > > > Javier F. Adur, Ph.D. > > > >Research Center in Optical and Fotonic (CEPOF) > > > > Biomedical Lasers Application Laboratory > > > > IFGW - UNICAMP - BRASIL > > > > -- > --------------------------------------------------------------------------------------------------- > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Walter-Brendel-Zentrum für experimentelle Medizin (WBex) > Head of light microscopy > > Mail room: > Marchioninistr. 15, D-81377 München > > Building location: > Marchioninistr. 27, München-Großhadern |
|
|
In reply to this post by Steffen Dietzel
-- Julio Vazquez Fred Hutchinson Cancer Research Center 1100 Fairview Ave. N., mailstop DE-512 Seattle, WA 98109-1024 Tel: Office: 206-667-1215/ Lab: 206-667-4205 |
| Free forum by Nabble | Edit this page |