Hello Lister,
Hope all are doing well.

 

Can anyone please offer me suggestions for
a test sample (a handy)for imaging THG in transmission?

Cheers
Javier


                    

                    Javier F. Adur, Ph.D.

Research Center in Optical and Fotonic (CEPOF)

     Biomedical Lasers Application Laboratory

           IFGW - UNICAMP - BRASIL   

  

             



Steffen Dietzel Steffen Dietzel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient samples for THG

Hello Javier,

I only came across your e-mail today. And since
you apparently didn't get a lot of answers so far...

The easiest test specimen is probably some
interface of water and glas or water and plastic:
big Ri-mismatch = strong THG signal.

I use water in a petri dish or on a glass slide
to check that our excitation beam is well
centered, since it provides a uniform signal. If
you need a signal in xy-direction, you should be
able to get it by adding a coverslip to the
water, so that you can image the edge of the cover slip.

If you draw lines on the surface with a black
marker pen, it will help to focus in ordinary
brightfield illumination, so that it is easier to find the focal plane in THG.

If you want to get more fancy, you can use a thin
muscle tissue (striated muscle gives nice
signals, see our paper J. Biomed. Opt., Vol. 15,
026017 (2010); doi:10.1117/1.3374337) or a
fat-water emulsion between coverslip and slide,
such as condensed milk or sunscreen, you get the
idea. (yes, we've tried that too :-) )

Hope it helps,

Steffen


At 19:22 05.08.2010, you wrote:

>Hello Lister,
>Hope all are doing well.
>
>  <?xml:namespace prefix = o ns = "urn:schemas-microsoft-com:office:office" />
>
>Can anyone please offer me suggestions for
>a test sample (a handy)for imaging THG in transmission?
>
>Cheers
>Javier
>
>
>
>                     Javier F. Adur, Ph.D.
>
>Research Center in Optical and Fotonic (CEPOF)
>
>      Biomedical Lasers Application Laboratory
>
>            IFGW - UNICAMP - BRASIL
>

--
---------------------------------------------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Walter-Brendel-Zentrum für experimentelle Medizin (WBex)
Head of light microscopy

Mail room:
Marchioninistr. 15, D-81377 München

Building location:
Marchioninistr. 27,  München-Großhadern
Javier Adur Javier Adur
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient samples for THG

Agoust 12,
Steffen, thank you very much for your response.
The information was very nice.
Javier.  


                    

                    Javier F. Adur, Ph.D.

Research Center in Optical and Fotonic (CEPOF)

     Biomedical Lasers Application Laboratory

           IFGW - UNICAMP - BRASIL   

  

             




 

> Date: Wed, 11 Aug 2010 16:41:33 +0200
> From: [hidden email]
> Subject: Re: Convenient samples for THG
> To: [hidden email]
>
> Hello Javier,
>
> I only came across your e-mail today. And since
> you apparently didn't get a lot of answers so far...
>
> The easiest test specimen is probably some
> interface of water and glas or water and plastic:
> big Ri-mismatch = strong THG signal.
>
> I use water in a petri dish or on a glass slide
> to check that our excitation beam is well
> centered, since it provides a uniform signal. If
> you need a signal in xy-direction, you should be
> able to get it by adding a coverslip to the
> water, so that you can image the edge of the cover slip.
>
> If you draw lines on the surface with a black
> marker pen, it will help to focus in ordinary
> brightfield illumination, so that it is easier to find the focal plane in THG.
>
> If you want to get more fancy, you can use a thin
> muscle tissue (striated muscle gives nice
> signals, see our paper J. Biomed. Opt., Vol. 15,
> 026017 (2010); doi:10.1117/1.3374337) or a
> fat-water emulsion between coverslip and slide,
> such as condensed milk or sunscreen, you get the
> idea. (yes, we've tried that too :-) )
>
> Hope it helps,
>
> Steffen
>
>
> At 19:22 05.08.2010, you wrote:
>
> >Hello Lister,
> >Hope all are doing well.
> >
> > <?xml:namespace prefix = o ns = "urn:schemas-microsoft-com:office:office" />
> >
> >Can anyone please offer me suggestions for
> >a test sample (a handy)for imaging THG in transmission?
> >
> >Cheers
> >Javier
> >
> >
> >
> > Javier F. Adur, Ph.D.
> >
> >Research Center in Optical and Fotonic (CEPOF)
> >
> > Biomedical Lasers Application Laboratory
> >
> > IFGW - UNICAMP - BRASIL
> >
>
> --
> ---------------------------------------------------------------------------------------------------
> Steffen Dietzel, PD Dr. rer. nat
> Ludwig-Maximilians-Universität München
> Walter-Brendel-Zentrum für experimentelle Medizin (WBex)
> Head of light microscopy
>
> Mail room:
> Marchioninistr. 15, D-81377 München
>
> Building location:
> Marchioninistr. 27, München-Großhadern
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

unsubscribe

In reply to this post by Steffen Dietzel

--
Julio Vazquez
Fred Hutchinson Cancer Research Center
1100 Fairview Ave. N.,  mailstop DE-512
Seattle, WA 98109-1024

Tel: Office: 206-667-1215/ Lab: 206-667-4205