DIC Image

classic Classic list List threaded Threaded
♦
♦
26 messages Options
12
Charu Tanwar Charu Tanwar
Reply | Threaded
Open this post in threaded view
♦
♦
|

DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies for
two different proteins and trying to see localization.However, Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my scope
checked and that is in good health. I tried this with other scope and the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced such a
weird problem before. Even confocal microscope refuses to give a good DIC
image (same scope is giving very good DIC images for RBC'S and Candida). I
do see two different expressions in my cells but i am not getting a good DIC
image, sometimes i do not see anything in DIC channel but simultaneously i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India

 

Peter Gabriel Pitrone Peter Gabriel Pitrone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

Hello Charu,

Have you set up Koehler illumination?

Pete

On May 6, 2010, at 08:27 AM, Charu Tanwar wrote:

> Dear List
> I am doing an experiment with HeLa cells staining them with antibodies for
> two different proteins and trying to see localization.However, Fluorescence
> is not my problem but i am not able to visualize my cells in bright
> field.......Well, i have to really put my eyes on so much of strain to
> visualize them in bright field and get their morphology (Cells appear to
> become very thin and barely gives boundary demarcation)  . I got my scope
> checked and that is in good health. I tried this with other scope and the
> problem remains same. I have to focus my cells through epifluorescence
> illumination (i.e. by looking at the expression). I have never faced such a
> weird problem before. Even confocal microscope refuses to give a good DIC
> image (same scope is giving very good DIC images for RBC'S and Candida). I
> do see two different expressions in my cells but i am not getting a good DIC
> image, sometimes i do not see anything in DIC channel but simultaneously i
> see good expression.
> I foresee that this is not a scope problem but either with the cells or
> staining protocol. However, i have changed my staining protocol 4 times
> (Fixation, time incubation for Ab's, no. of washes and blocking) but the
> story is same.When i visualize my cells before starting the experiment,
> cells look healthy and appear fine.
> Please help
> Charu Tanwar
>
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India
>
>
>
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar
Charu,

My guess is that your fixation is permeabilising your cells so
effectively - and maybe extracting some lipids - that there isn't too
much left to generate DIC contrast.  As Peter suggests, meticulously
checking the setup of your microscope - Koehler, polarizer, analyser etc
may help.  But phase contrast might be a better option than DIC -
generally it does better with very thin samples.  Phase lenses come in
two forms low (L) and high (H) absorbance - if you have a choice an H
lens will do better with a weakly diffracting object.  

                                                Guy

Please help fight breast cancer by sponsoring my
run in the Sydney Half Marathon on May 16th.
http://www.everydayhero.com.au/guy_cox
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Charu Tanwar
Sent: Thursday, 6 May 2010 4:27 PM
To: [hidden email]
Subject: DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies
for
two different proteins and trying to see localization.However,
Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my
scope
checked and that is in good health. I tried this with other scope and
the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced
such a
weird problem before. Even confocal microscope refuses to give a good
DIC
image (same scope is giving very good DIC images for RBC'S and Candida).
I
do see two different expressions in my cells but i am not getting a good
DIC
image, sometimes i do not see anything in DIC channel but simultaneously
i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India

 
 

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
04:26:00
Charu Tanwar Charu Tanwar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Peter Gabriel Pitrone
Hello Peter
Koehler illumination is set. This is a problem with this experiment only, i could get very good DIC with yeast and RBC's.
Thanks
Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Thu, 6/5/10, Peter Pitrone <[hidden email]> wrote:

From: Peter Pitrone <[hidden email]>
Subject: Re: DIC Image
To: [hidden email]
Date: Thursday, 6 May, 2010, 12:11 PM

Hello Charu,

Have you set up Koehler illumination?

Pete

On May 6, 2010, at 08:27 AM, Charu Tanwar wrote:

> Dear List
> I am doing an experiment with HeLa cells staining them with antibodies for
> two different proteins and trying to see localization.However, Fluorescence
> is not my problem but i am not able to visualize my cells in bright
> field.......Well, i have to really put my eyes on so much of strain to
> visualize them in bright field and get their morphology (Cells appear to
> become very thin and barely gives boundary demarcation)  . I got my scope
> checked and that is in good health. I tried this with other scope and the
> problem remains same. I have to focus my cells through epifluorescence
> illumination (i.e. by looking at the expression). I have never faced such a
> weird problem before. Even confocal microscope refuses to give a good DIC
> image (same scope is giving very good DIC images for RBC'S and Candida). I
> do see two different expressions in my cells but i am not getting a good DIC
> image, sometimes i do not see anything in DIC channel but simultaneously i
> see good expression.
> I foresee that this is not a scope problem but either with the cells or
> staining protocol. However, i have changed my staining protocol 4 times
> (Fixation, time incubation for Ab's, no. of washes and blocking) but the
> story is same.When i visualize my cells before starting the experiment,
> cells look healthy and appear fine.
> Please help
> Charu Tanwar
>
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India
>
>
>

Jean-Yves Tinevez-3 Jean-Yves Tinevez-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar



On May 6, 2010, at 9:37 AM, charu tanwar wrote:

Hello Peter
Koehler illumination is set. This is a problem with this experiment only, i could get very good DIC with yeast and RBC's.
Thanks
Charu


Hi Charu,

Could it be because you are using plastic dishes or plastic covers? 
They disrupt light polarization, which you need for proper DIC. If this is the case, just switch to glass.
Best
jy



--
Jean-Yves Tinevez
PFID - Imagopole
Institut Pasteur
25-28, rue du Docteur Roux
75724 Paris cedex 15 
France
tel: +33 1 40 61 31 77




Charu Tanwar Charu Tanwar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

No i grow my cells on coverslips and do my experiment and mount the coverslip on a glass slide.

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Thu, 6/5/10, Jean-Yves Tinevez <[hidden email]> wrote:

From: Jean-Yves Tinevez <[hidden email]>
Subject: Re: DIC Image
To: [hidden email]
Date: Thursday, 6 May, 2010, 1:42 PM




On May 6, 2010, at 9:37 AM, charu tanwar wrote:

Hello Peter
Koehler illumination is set. This is a problem with this experiment only, i could get very good DIC with yeast and RBC's.
Thanks
Charu


Hi Charu,

Could it be because you are using plastic dishes or plastic covers? 
They disrupt light polarization, which you need for proper DIC. If this is the case, just switch to glass.
Best
jy



--
Jean-Yves Tinevez
PFID - Imagopole
Institut Pasteur
25-28, rue du Docteur Roux
75724 Paris cedex 15 
France
tel: +33 1 40 61 31 77





Charu Tanwar Charu Tanwar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Guy Cox-2
Hi
I also suspected the same when i face this problem for the first time. I did little bit of modification in my fixation. What i am doing now is fixing my cells with 100% chilled methanol and keep the cells at 4 degrees for 10min. Then i do permeabilisation with 0.3% triton X-100. i need to do this step in order to stain my protein which is inside nucleus.
May be this is becoming too much for the cells.
Thanks
Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:

From: Guy Cox <[hidden email]>
Subject: Re: DIC Image
To: [hidden email]
Date: Thursday, 6 May, 2010, 12:15 PM

Charu,

My guess is that your fixation is permeabilising your cells so
effectively - and maybe extracting some lipids - that there isn't too
much left to generate DIC contrast.  As Peter suggests, meticulously
checking the setup of your microscope - Koehler, polarizer, analyser etc
may help.  But phase contrast might be a better option than DIC -
generally it does better with very thin samples.  Phase lenses come in
two forms low (L) and high (H) absorbance - if you have a choice an H
lens will do better with a weakly diffracting object. 

                                                Guy

Please help fight breast cancer by sponsoring my
run in the Sydney Half Marathon on May 16th.
http://www.everydayhero.com.au/guy_cox
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net


-----Original Message-----
From: Confocal Microscopy List [mailto:CONFOCALMICROSCOPY@...]
On Behalf Of Charu Tanwar
Sent: Thursday, 6 May 2010 4:27 PM
To: CONFOCALMICROSCOPY@...
Subject: DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies
for
two different proteins and trying to see localization.However,
Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my
scope
checked and that is in good health. I tried this with other scope and
the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced
such a
weird problem before. Even confocal microscope refuses to give a good
DIC
image (same scope is giving very good DIC images for RBC'S and Candida).
I
do see two different expressions in my cells but i am not getting a good
DIC
image, sometimes i do not see anything in DIC channel but simultaneously
i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India




No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
04:26:00

Eric Scarfone Eric Scarfone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

alcohol will extract a lot of things and flatten your cells, Adehyde is a better fix fo preserving morphology and 3D structure.

However if you have good images with other cells and same protocol it is puzzling.

Stick to fluorescence then and use counterstaining (membrane, cytoplasm, nucleus depending on what you look for) to visualize cell structure!.

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: charu tanwar <[hidden email]>
Date: Thursday, May 6, 2010 10:25 am
Subject: Re: DIC Image
To: [hidden email]

> Hi
> I also suspected the same when i face this problem for the first
> time. I did little bit of modification in my fixation. What i am
> doing now is fixing my cells with 100% chilled methanol and keep
> the cells at 4 degrees for 10min. Then i do permeabilisation with


> 0.3% triton X-100. i need to do this step in order to stain my
> protein which is inside nucleus.
> May be this is becoming too much for the cells.
> Thanks
> Charu
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi 110067
> India.
>
> --- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
>
>
> From: Guy Cox <[hidden email]>
> Subject: Re: DIC Image
> To: [hidden email]
> Date: Thursday, 6 May, 2010, 12:15 PM
>
>
> Charu,
>
> My guess is that your fixation is permeabilising your cells so
> effectively - and maybe extracting some lipids - that there isn't too
> much left to generate DIC contrast.  As Peter suggests, meticulously

> checking the setup of your microscope - Koehler, polarizer,
> analyser etc
> may help.  But phase contrast might be a better option than DIC -
> generally it does better with very thin samples.  Phase lenses
> come in
> two forms low (L) and high (H) absorbance - if you have a choice
> an H
> lens will do better with a weakly diffracting object. 
>
>                                                 Guy
>
> Please help fight breast cancer by sponsoring my
> run in the Sydney Half Marathon on May 16th.
> http://www.everydayhero.com.au/guy_cox
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Australian Centre for Microscopy & Microanalysis,

> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>              Mobile 0413 281 861
> ______________________________________________
>       http://www.guycox.net
>
>
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]]On Behalf Of Charu Tanwar
> Sent: Thursday, 6 May 2010 4:27 PM
> To: [hidden email]
> Subject: DIC Image
>
> Dear List
> I am doing an experiment with HeLa cells staining them with antibodies
> for
> two different proteins and trying to see localization.However,
> Fluorescence
> is not my problem but i am not able to visualize my cells in bright
> field.......Well, i have to really put my eyes on so much of

> strain to
> visualize them in bright field and get their morphology (Cells
> appear to
> become very thin and barely gives boundary demarcation)  . I got my
> scope
> checked and that is in good health. I tried this with other scope and
> the
> problem remains same. I have to focus my cells through epifluorescence
> illumination (i.e. by looking at the expression). I have never faced
> such a
> weird problem before. Even confocal microscope refuses to give a good
> DIC
> image (same scope is giving very good DIC images for RBC'S and
> Candida).I
> do see two different expressions in my cells but i am not getting
> a good
> DIC
> image, sometimes i do not see anything in DIC channel but
> simultaneouslyi
> see good expression.
> I foresee that this is not a scope problem but either with the

> cells or
> staining protocol. However, i have changed my staining protocol 4
> times(Fixation, time incubation for Ab's, no. of washes and
> blocking) but the
> story is same.When i visualize my cells before starting the
> experiment,cells look healthy and appear fine.
> Please help
> Charu Tanwar
>
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India
>
>
>
>
> No virus found in this incoming message.
> Checked by AVG - www.avg.com
> Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> 05/06/1004:26:00
>
>
>

Johannes Helm Johannes Helm
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar
Hi, Charu,

from what you write it, indeed, seems that you have a specimen related
problem.

What I would try (though I cannot guarantee that this will help):

a)
Make sure the field diaphragm ("F diaphragm") is opened not any further
then necessary to cover the entire field of view (assuming the condenser
is properly centered).

b)
At the price of a decrease in resolution, close the aperture diaphragm ("A
diaphragm") of the condenser. This will make everything a "little bit
darker" and since it will reduce the NA of the condenser, it will also, as
written, reduce resolution, but it will, on the other hand enhance
contrast.

c)
In biological preparations, like your cells, the main source of scattering
usually is not Rayleigh but Mie scattering. Nevertheless, there still is a
square relationsship between wavelength and scattering cross section.
Normally, you would prefer to use light of the near IR for thick
preparations in order to maximize penetration possibilities.
It might, however, be an idea to try the opposite in your case. If you can
get hold of a panchromatic green filter or even blue filter - they are /
were (?) usually  standard items delivered together with any new
microscope - to be placed in the ray path, it's a five minutes test to
check whether this will help to increase your contrast. If you do not have
a green filter or blue filter available, try anyway the DLF ("Day Light
Filter", sometimes known as "conversion filter"), which has a reduced red
transmission in order to change the spectral distribution of the light
source (halogen bulb 12V, 100W) to match the spectral distribution of
solar light.

d)
If you can make a digital image (8bits or 12bits), you can offline - on
powerful systems online - change the display properties. A non-linear,
possibly logarithmic, display might help to increase contrast.

Best wishes,

Johannes



> Dear List
> I am doing an experiment with HeLa cells staining them with antibodies for
> two different proteins and trying to see localization.However,
> Fluorescence
> is not my problem but i am not able to visualize my cells in bright
> field.......Well, i have to really put my eyes on so much of strain to
> visualize them in bright field and get their morphology (Cells appear to
> become very thin and barely gives boundary demarcation)  . I got my scope
> checked and that is in good health. I tried this with other scope and the
> problem remains same. I have to focus my cells through epifluorescence
> illumination (i.e. by looking at the expression). I have never faced such
> a
> weird problem before. Even confocal microscope refuses to give a good DIC
> image (same scope is giving very good DIC images for RBC'S and Candida). I
> do see two different expressions in my cells but i am not getting a good
> DIC
> image, sometimes i do not see anything in DIC channel but simultaneously i
> see good expression.
> I foresee that this is not a scope problem but either with the cells or
> staining protocol. However, i have changed my staining protocol 4 times
> (Fixation, time incubation for Ab's, no. of washes and blocking) but the
> story is same.When i visualize my cells before starting the experiment,
> cells look healthy and appear fine.
> Please help
> Charu Tanwar
>
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India
>
>
>
>


--
P. Johannes Helm

Voice: (+47) 228 51159 (office)
Fax: (+47) 228 51499 (office)
JOEL B. SHEFFIELD JOEL B. SHEFFIELD
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar
Hi Charu,

This is probably due to the mounting medium that you are using.  In general, the best mounting media for high resolution fluorescence have a refractive index that is similar to that of cells.  Since DIC and phase contrast both depend on refractive index differences between, say, membranes and cytoplasm, or membranes and extracellular media, you are unable to visuallize the cells.  You might try diluting your mounting media, or mounting with a PBS/Glycerol (50/50) mix containing your own anti-fade, such as n-propyl-gallate.

Best of luck.

Joel


On Thu, May 6, 2010 at 2:27 AM, Charu Tanwar <[hidden email]> wrote:
Dear List
I am doing an experiment with HeLa cells staining them with antibodies for
two different proteins and trying to see localization.However, Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my scope
checked and that is in good health. I tried this with other scope and the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced such a
weird problem before. Even confocal microscope refuses to give a good DIC
image (same scope is giving very good DIC images for RBC'S and Candida). I
do see two different expressions in my cells but i am not getting a good DIC
image, sometimes i do not see anything in DIC channel but simultaneously i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India






--


Joel B. Sheffield, Ph.D
Department of Biology
Temple University
Philadelphia, PA 19122
Voice: 215 204 8839
e-mail: [hidden email]
URL:  http://astro.temple.edu/~jbs

Keith Morris Keith Morris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar

Hi,

 

Following on from Joels comments:

 

I’m using live HeLa cells at the moment and via the confocal the DIC image is adequate at best  [naturally depends on confluency]. Once fixed and mounted the cells look a lot worse under DIC– I’ve always assumed it was the anti-fade mounting media losing the cell outline, as it probably isn’t optimized for transmission light e.g. wrong refractive index or something [we use Vectashield liquid and Prolong Gold mostly, probably out of habit]. This is also the same for phase contrast [fixed cells looking poorer], although contrast/image quality is superior both live cell and after fixation/mounting. To be honest I’ve just thought ‘who cares’ and used a plasma membrane marker if necessary to delineate cells [Invitrogen wheatgerm/CellMask/Organelle Lights plasma membrane stains] or DAPI/fixed and Hoescht/Live  for cell number.

 

So I’ve never bothered improving the DIC image for fluorescence labeled fixed cells [our fixed flattened BPAE cells in particular are largely transparent under DIC, mounted in pro-long gold – they look great under fluorescence though]. For decent confocal transmission images we mostly use 20x phase contrast, and I only use tend to use 63x oil DIC on living cells. The confocal also seems to make a far better job of imaging phase contrast compared to DIC as well [as the confocal trans detector isn’t optimally placed]. Most of our confocal users don’t bother with DIC trans using the 63x oil objective unless there’s a structure in the fixed cell/tissue that responds well to DIC. You can [and probably should] adjust the confocal gain and offset to increase DIC contrast. Our wide-field Nikon microscope [that’s ‘better’ for DIC, imaged via a CCD camera] can apparently be fitted with  a selection of DIC options [three different sliders] and one is for ‘low contrast thin specimens’ [i.e. cells], so possibly a change of DIC optics as well as mountant might help as well.

 

e.g. http://www.microscopy-news.com/download-center/Cat_Appl_of_DIC.pdf

So no help on improving the situation I’m afraid, I just don’t find it a problem I guess.

 

Regards

 

Keith

 

---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford  OX3 7BN,
United Kingdom.

Telephone:  +44 (0)1865 287568
Email:  [hidden email]

Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of charu tanwar
Sent: 06 May 2010 09:26
To: [hidden email]
Subject: Re: DIC Image

 

Hi

I also suspected the same when i face this problem for the first time. I did little bit of modification in my fixation. What i am doing now is fixing my cells with 100% chilled methanol and keep the cells at 4 degrees for 10min. Then i do permeabilisation with 0.3% triton X-100. i need to do this step in order to stain my protein which is inside nucleus.

May be this is becoming too much for the cells.

Thanks

Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:


From: Guy Cox <[hidden email]>
Subject: Re: DIC Image
To: [hidden email]
Date: Thursday, 6 May, 2010, 12:15 PM

Charu,

My guess is that your fixation is permeabilising your cells so
effectively - and maybe extracting some lipids - that there isn't too
much left to generate DIC contrast.  As Peter suggests, meticulously
checking the setup of your microscope - Koehler, polarizer, analyser etc
may help.  But phase contrast might be a better option than DIC -
generally it does better with very thin samples.  Phase lenses come in
two forms low (L) and high (H) absorbance - if you have a choice an H
lens will do better with a weakly diffracting object. 

                                                Guy

Please help fight breast cancer by sponsoring my
run in the Sydney Half Marathon on May 16th.
http://www.everydayhero.com.au/guy_cox
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net


-----Original Message-----
From: Confocal Microscopy List [mailto:CONFOCALMICROSCOPY@...]
On Behalf Of Charu Tanwar
Sent: Thursday, 6 May 2010 4:27 PM
To: CONFOCALMICROSCOPY@...
Subject: DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies
for
two different proteins and trying to see localization.However,
Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my
scope
checked and that is in good health. I tried this with other scope and
the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced
such a
weird problem before. Even confocal microscope refuses to give a good
DIC
image (same scope is giving very good DIC images for RBC'S and Candida).
I
do see two different expressions in my cells but i am not getting a good
DIC
image, sometimes i do not see anything in DIC channel but simultaneously
i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India




No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
04:26:00

 

Axel Kurt Preuss Axel Kurt Preuss
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

When physics fails let biology speak! Best to let your cells grow on  fixed (dead) cell layers of DIC "friendly" cells  (e.g simply dried in your sterile hood and then washed once with  80 percent of physiological  buffer ).
The dead cells will give your probe cells a nice DIC friendly environment. Once your cells have grown to the right condition proceed as usual


Hope that helps

Axel

Axel K Preuss PhD, A*Star IMCB-Central Imaging Facility

On May 6, 2010, at 8:12 PM, "Keith Morris" <[hidden email]<mailto:[hidden email]>> wrote:

Hi,

Following on from Joels comments:

I’m using live HeLa cells at the moment and via the confocal the DIC image is adequate at best  [naturally depends on confluency]. Once fixed and mounted the cells look a lot worse under DIC– I’ve always assumed it was the anti-fade mounting media losing the cell outline, as it probably isn’t optimized for transmission light e.g. wrong refractive index or something [we use Vectashield liquid and Prolong Gold mostly, probably out of habit]. This is also the same for phase contrast [fixed cells looking poorer], although contrast/image quality is superior both live cell and after fixation/mounting. To be honest I’ve just thought ‘who cares’ and used a plasma membrane marker if necessary to delineate cells [Invitrogen wheatgerm/CellMask/Organelle Lights plasma membrane stains] or DAPI/fixed and Hoescht/Live  for cell number.

So I’ve never bothered improving the DIC image for fluorescence labeled fixed cells [our fixed flattened BPAE cells in particular are largely transparent under DIC, mounted in pro-long gold – they look great under fluorescence though]. For decent confocal transmission images we mostly use 20x phase contrast, and I only use tend to use 63x oil DIC on living cells. The confocal also seems to make a far better job of imaging phase contrast compared to DIC as well [as the confocal trans detector isn’t optimally placed]. Most of our confocal users don’t bother with DIC trans using the 63x oil objective unless there’s a structure in the fixed cell/tissue that responds well to DIC. You can [and probably should] adjust the confocal gain and offset to increase DIC contrast. Our wide-field Nikon microscope [that’s ‘better’ for DIC, imaged via a CCD camera] can apparently be fitted with  a selection of DIC options [three different sliders] and one is for ‘low contrast thin specimens’ [i.e. cells], so possibly a change of DIC optics as well as mountant might help as well.

e.g. <http://www.microscopy-news.com/download-center/Cat_Appl_of_DIC.pdf> http://www.microscopy-news.com/download-center/Cat_Appl_of_DIC.pdf

So no help on improving the situation I’m afraid, I just don’t find it a problem I guess.

Regards

Keith

---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford  OX3 7BN,
United Kingdom.

Telephone:  +44 (0)1865 287568
Email:  <mailto:[hidden email]> [hidden email]<mailto:[hidden email]>
Web-pages: <http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy> http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of charu tanwar
Sent: 06 May 2010 09:26
To: <mailto:[hidden email]> [hidden email]<mailto:[hidden email]>
Subject: Re: DIC Image

Hi
I also suspected the same when i face this problem for the first time. I did little bit of modification in my fixation. What i am doing now is fixing my cells with 100% chilled methanol and keep the cells at 4 degrees for 10min. Then i do permeabilisation with 0.3% triton X-100. i need to do this step in order to stain my protein which is inside nucleus.
May be this is becoming too much for the cells.
Thanks
Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.

--- On Thu, 6/5/10, Guy Cox <<mailto:[hidden email]>[hidden email]<mailto:[hidden email]>> wrote:

From: Guy Cox <<mailto:[hidden email]>[hidden email]<mailto:[hidden email]>>
Subject: Re: DIC Image
To: <mailto:[hidden email]> [hidden email]<mailto:[hidden email]>
Date: Thursday, 6 May, 2010, 12:15 PM
Charu,

My guess is that your fixation is permeabilising your cells so
effectively - and maybe extracting some lipids - that there isn't too
much left to generate DIC contrast.  As Peter suggests, meticulously
checking the setup of your microscope - Koehler, polarizer, analyser etc
may help.  But phase contrast might be a better option than DIC -
generally it does better with very thin samples.  Phase lenses come in
two forms low (L) and high (H) absorbance - if you have a choice an H
lens will do better with a weakly diffracting object.

                                                Guy

Please help fight breast cancer by sponsoring my
run in the Sydney Half Marathon on May 16th.
<http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      <http://www.guycox.net/> http://www.guycox.net


-----Original Message-----
From: Confocal Microscopy List [mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...>[hidden email]<mailto:[hidden email]>]
On Behalf Of Charu Tanwar
Sent: Thursday, 6 May 2010 4:27 PM
To: <http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...> [hidden email]<mailto:[hidden email]>
Subject: DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies
for
two different proteins and trying to see localization.However,
Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my
scope
checked and that is in good health. I tried this with other scope and
the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced
such a
weird problem before. Even confocal microscope refuses to give a good
DIC
image (same scope is giving very good DIC images for RBC'S and Candida).
I
do see two different expressions in my cells but i am not getting a good
DIC
image, sometimes i do not see anything in DIC channel but simultaneously
i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India




No virus found in this incoming message.
Checked by AVG - <http://www.avg.com> www.avg.com<http://www.avg.com>
Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
04:26:00



Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

What you might want to say is:

When we cannot explain it by physics, let biology speak but not to explain.

On 5/6/10 1:56 PM, Axel Kurt Preuss wrote:
> When physics fails let biology speak!  Best to let your cells grow on  fixed (dead) cell layers of DIC
"friendly" cells  (e.g simply dried in your sterile hood and then washed
once with  80 percent of physiological  buffer ).

> The dead cells will give your probe cells a nice DIC friendly environment. Once your cells have grown to the right condition proceed as usual
>
>
> Hope that helps
>
> Axel
>
> Axel K Preuss PhD, A*Star IMCB-Central Imaging Facility
>
> On May 6, 2010, at 8:12 PM, "Keith Morris"<[hidden email]<mailto:[hidden email]>>  wrote:
>
> Hi,
>
> Following on from Joels comments:
>
> I’m using live HeLa cells at the moment and via the confocal the DIC image is adequate at best  [naturally depends on confluency]. Once fixed and mounted the cells look a lot worse under DIC– I’ve always assumed it was the anti-fade mounting media losing the cell outline, as it probably isn’t optimized for transmission light e.g. wrong refractive index or something [we use Vectashield liquid and Prolong Gold mostly, probably out of habit]. This is also the same for phase contrast [fixed cells looking poorer], although contrast/image quality is superior both live cell and after fixation/mounting. To be honest I’ve just thought ‘who cares’ and used a plasma membrane marker if necessary to delineate cells [Invitrogen wheatgerm/CellMask/Organelle Lights plasma membrane stains] or DAPI/fixed and Hoescht/Live  for cell number.
>
> So I’ve never bothered improving the DIC image for fluorescence labeled fixed cells [our fixed flattened BPAE cells in particular are largely transparent under DIC, mounted in pro-long gold – they look great under fluorescence though]. For decent confocal transmission images we mostly use 20x phase contrast, and I only use tend to use 63x oil DIC on living cells. The confocal also seems to make a far better job of imaging phase contrast compared to DIC as well [as the confocal trans detector isn’t optimally placed]. Most of our confocal users don’t bother with DIC trans using the 63x oil objective unless there’s a structure in the fixed cell/tissue that responds well to DIC. You can [and probably should] adjust the confocal gain and offset to increase DIC contrast. Our wide-field Nikon microscope [that’s ‘better’ for DIC, imaged via a CCD camera] can apparently be fitted with  a selection of DIC options [three different sliders] and one is for ‘low contras
t thin specimens’ [i.e. cells], so possibly a change of DIC optics as well as mountant might help as well.

>
> e.g.<http://www.microscopy-news.com/download-center/Cat_Appl_of_DIC.pdf>  http://www.microscopy-news.com/download-center/Cat_Appl_of_DIC.pdf
>
> So no help on improving the situation I’m afraid, I just don’t find it a problem I guess.
>
> Regards
>
> Keith
>
> ---------------------------------------------------------------------------
> Dr Keith J. Morris,
> Molecular Cytogenetics and Microscopy Core,
> Laboratory 00/069 and 00/070,
> The Wellcome Trust Centre for Human Genetics,
> Roosevelt Drive,
> Oxford  OX3 7BN,
> United Kingdom.
>
> Telephone:  +44 (0)1865 287568
> Email:<mailto:[hidden email]>  [hidden email]<mailto:[hidden email]>
> Web-pages:<http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy>  http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy
>
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of charu tanwar
> Sent: 06 May 2010 09:26
> To:<mailto:[hidden email]>  [hidden email]<mailto:[hidden email]>
> Subject: Re: DIC Image
>
> Hi
> I also suspected the same when i face this problem for the first time. I did little bit of modification in my fixation. What i am doing now is fixing my cells with 100% chilled methanol and keep the cells at 4 degrees for 10min. Then i do permeabilisation with 0.3% triton X-100. i need to do this step in order to stain my protein which is inside nucleus.
> May be this is becoming too much for the cells.
> Thanks
> Charu
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi 110067
> India.
>
> --- On Thu, 6/5/10, Guy Cox<<mailto:[hidden email]>[hidden email]<mailto:[hidden email]>>  wrote:
>
> From: Guy Cox<<mailto:[hidden email]>[hidden email]<mailto:[hidden email]>>
> Subject: Re: DIC Image
> To:<mailto:[hidden email]>  [hidden email]<mailto:[hidden email]>
> Date: Thursday, 6 May, 2010, 12:15 PM
> Charu,
>
> My guess is that your fixation is permeabilising your cells so
> effectively - and maybe extracting some lipids - that there isn't too
> much left to generate DIC contrast.  As Peter suggests, meticulously
> checking the setup of your microscope - Koehler, polarizer, analyser etc
> may help.  But phase contrast might be a better option than DIC -
> generally it does better with very thin samples.  Phase lenses come in
> two forms low (L) and high (H) absorbance - if you have a choice an H
> lens will do better with a weakly diffracting object.
>
>                                                  Guy
>
> Please help fight breast cancer by sponsoring my
> run in the Sydney Half Marathon on May 16th.
> <http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Australian Centre for Microscopy&  Microanalysis,
> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>               Mobile 0413 281 861
> ______________________________________________
>        <http://www.guycox.net/>  http://www.guycox.net
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...>[hidden email]<mailto:[hidden email]>]
> On Behalf Of Charu Tanwar
> Sent: Thursday, 6 May 2010 4:27 PM
> To:<http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...>  [hidden email]<mailto:[hidden email]>
> Subject: DIC Image
>
> Dear List
> I am doing an experiment with HeLa cells staining them with antibodies
> for
> two different proteins and trying to see localization.However,
> Fluorescence
> is not my problem but i am not able to visualize my cells in bright
> field.......Well, i have to really put my eyes on so much of strain to
> visualize them in bright field and get their morphology (Cells appear to
> become very thin and barely gives boundary demarcation)  . I got my
> scope
> checked and that is in good health. I tried this with other scope and
> the
> problem remains same. I have to focus my cells through epifluorescence
> illumination (i.e. by looking at the expression). I have never faced
> such a
> weird problem before. Even confocal microscope refuses to give a good
> DIC
> image (same scope is giving very good DIC images for RBC'S and Candida).
> I
> do see two different expressions in my cells but i am not getting a good
> DIC
> image, sometimes i do not see anything in DIC channel but simultaneously
> i
> see good expression.
> I foresee that this is not a scope problem but either with the cells or
> staining protocol. However, i have changed my staining protocol 4 times
> (Fixation, time incubation for Ab's, no. of washes and blocking) but the
> story is same.When i visualize my cells before starting the experiment,
> cells look healthy and appear fine.
> Please help
> Charu Tanwar
>
>
> CHARU TANWAR
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India
>
>
>
>
> No virus found in this incoming message.
> Checked by AVG -<http://www.avg.com>  www.avg.com<http://www.avg.com>
> Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
> 04:26:00
>
>
>
> Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Charu Tanwar
Re: DIC Image
Hi Charu,

Well, you seem to have thought of most of the suggested solutions already.

If, as Guy suggests, the problem is that there is really not enough left of your sample to produce a detectable RI difference, (which seems strange as video-enhanced DIC used to give good images of single microtubules, or even single actin filaments.) another possibility is Darkfield imaging.

As any "survivor" of the UBC Course will tell you, darkfield is my favorite transmitted light imaging method, and it will really produce the most contrast for a given quantity of structural material present. The only problem is that you must use a darkfield condenser with a higher NA than that of your objective and you need to align the condenser with great care. Although this isn't always easy, the results can be really superb.

If you don't have a super high-NA darkfield condenser (>0.9 NA. i.e., one that you have to "oil" to the slide), you might find an objective with an iris diaphragm that will still allow you to make the objective NA lower than that of the illumination. While this will reduce the resolution a bit, it will definitely allow you to see the edges of the cells.

In confocal one can get a similar image using the backscattered light signal, however, the signal from the cell will be overwhelmed by the much larger reflection from the glass-water interface and this will make it hard to see cellular details within 3-4 micrometers of the interface.

Good luck,

Jim Pawley

*********************************************************************************
Prof. James B. Pawley,                                     Ph.  608-263-3147 
Room 223, Zoology Research Building,                                  FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706                                [hidden email]
3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/                Applications still being accepted
               "If it ain't diffraction, it must be statistics." Anon.

Hi
I also suspected the same when i face this problem for the first time. I did little bit of modification in my fixation. What i am doing now is fixing my cells with 100% chilled methanol and keep the cells at 4 degrees for 10min. Then i do permeabilisation with 0.3% triton X-100. i need to do this step in order to stain my protein which is inside nucleus.
May be this is becoming too much for the cells.
Thanks
Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:

From: Guy Cox <[hidden email]>
Subject: Re: DIC Image
To: [hidden email]
Date: Thursday, 6 May, 2010, 12:15 PM
Charu,

My guess is that your fixation is permeabilising your cells so
effectively - and maybe extracting some lipids - that there isn't too
much left to generate DIC contrast.  As Peter suggests, meticulously
checking the setup of your microscope - Koehler, polarizer, analyser etc
may help.  But phase contrast might be a better option than DIC -
generally it does better with very thin samples.  Phase lenses come in
two forms low (L) and high (H) absorbance - if you have a choice an H
lens will do better with a weakly diffracting object. 

                                                Guy

Please help fight breast cancer by sponsoring my
run in the Sydney Half Marathon on May 16th.
http://www.everydayhero.com.au/guy_cox
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net


-----Original Message-----
From: Confocal Microscopy List [mailto:CONFOCALMICROSCOPY@...]
On Behalf Of Charu Tanwar
Sent: Thursday, 6 May 2010 4:27 PM
To: CONFOCALMICROSCOPY@...
Subject: DIC Image

Dear List
I am doing an experiment with HeLa cells staining them with antibodies
for
two different proteins and trying to see localization.However,
Fluorescence
is not my problem but i am not able to visualize my cells in bright
field.......Well, i have to really put my eyes on so much of strain to
visualize them in bright field and get their morphology (Cells appear to
become very thin and barely gives boundary demarcation)  . I got my
scope
checked and that is in good health. I tried this with other scope and
the
problem remains same. I have to focus my cells through epifluorescence
illumination (i.e. by looking at the expression). I have never faced
such a
weird problem before. Even confocal microscope refuses to give a good
DIC
image (same scope is giving very good DIC images for RBC'S and Candida).
I
do see two different expressions in my cells but i am not getting a good
DIC
image, sometimes i do not see anything in DIC channel but simultaneously
i
see good expression.
I foresee that this is not a scope problem but either with the cells or
staining protocol. However, i have changed my staining protocol 4 times
(Fixation, time incubation for Ab's, no. of washes and blocking) but the
story is same.When i visualize my cells before starting the experiment,
cells look healthy and appear fine.
Please help
Charu Tanwar


CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India




No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date: 05/06/10
04:26:00


-- 
Eric Scarfone Eric Scarfone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

Jim,
High NA condenser/objectives come with pretty short working distance that imposes mounting samples between 2 coverslip. At least that's the requirement for highly efficient DIC that imposes perfect kölher alignment I believe. That + the oil on both sides have made this practice less and less popular among students and facility managers (although not for the same reasons). 

Is darkfield more tolerant than DIC?

Cheers

Eric (13 years survivor, had to count twice to make sure, GOSH!!!)

 

 

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: James Pawley <[hidden email]>
Date: Thursday, May 6, 2010 3:42 pm
Subject: Re: DIC Image
To: [hidden email]

> Hi Charu,
>
> Well, you seem to have thought of most of the suggested solutions
> already.
> If, as Guy suggests, the problem is that there is really not
> enough
> left of your sample to produce a detectable RI difference, (which
> seems strange as video-enhanced DIC used to give good images of
> single microtubules, or even single actin filaments.) another


> possibility is Darkfield imaging.
>
> As any "survivor" of the UBC Course will tell you, darkfield is my
> favorite transmitted light imaging method, and it will really
> produce
> the most contrast for a given quantity of structural material
> present. The only problem is that you must use a darkfield
> condenser
> with a higher NA than that of your objective and you need to align
> the condenser with great care. Although this isn't always easy,
> the
> results can be really superb.
>
> If you don't have a super high-NA darkfield condenser (>0.9 NA.
> i.e.,
> one that you have to "oil" to the slide), you might find an
> objective
> with an iris diaphragm that will still allow you to make the
> objective NA lower than that of the illumination. While this will
> reduce the resolution a bit, it will definitely allow you to see

> the
> edges of the cells.
>
> In confocal one can get a similar image using the backscattered
> light
> signal, however, the signal from the cell will be overwhelmed by
> the
> much larger reflection from the glass-water interface and this
> will
> make it hard to see cellular details within 3-4 micrometers of the
> interface.
>
> Good luck,
>
> Jim Pawley
>
> *********************************************************************************
> Prof. James B. Pawley,
> Ph. 608-263-3147
> Room 223, Zoology Research Building,
> FAX 608-265-5315
> 1117 Johnson Ave., Madison, WI, 53706
> [hidden email]
> 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
> Vancouver Canada
> Info: http://www.3dcourse.ubc.ca/ Applications

> still
> being accepted
> "If it ain't diffraction, it must be statistics." Anon.
>
> >Hi
> >I also suspected the same when i face this problem for the first
> >time. I did little bit of modification in my fixation. What i am
> >doing now is fixing my cells with 100% chilled methanol and keep
> the
> >cells at 4 degrees for 10min. Then i do permeabilisation with
> 0.3%
> >triton X-100. i need to do this step in order to stain my protein
> >which is inside nucleus.
> >May be this is becoming too much for the cells.
> >Thanks
> >Charu
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi 110067
> >India.
> >
> >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:

> >
> >
> >From: Guy Cox <[hidden email]>
> >Subject: Re: DIC Image
> >To: [hidden email]
> >Date: Thursday, 6 May, 2010, 12:15 PM
> >
> >Charu,
> >
> >My guess is that your fixation is permeabilising your cells so
> >effectively - and maybe extracting some lipids - that there isn't too
> >much left to generate DIC contrast. As Peter suggests, meticulously
> >checking the setup of your microscope - Koehler, polarizer,
> analyser etc
> >may help. But phase contrast might be a better option than DIC -
> >generally it does better with very thin samples. Phase lenses
> come in
> >two forms low (L) and high (H) absorbance - if you have a choice
> an H
> >lens will do better with a weakly diffracting object.

> >
> > Guy
> >
> >Please help fight breast cancer by sponsoring my
> >run in the Sydney Half Marathon on May 16th.
> ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox
> >______________________________________________
> >Associate Professor Guy Cox, MA, DPhil(Oxon)
> >Australian Centre for Microscopy & Microanalysis,
> >Madsen Building F09, University of Sydney, NSW 2006
> >
> >Phone +61 2 9351 3176 Fax +61 2 9351 7682
> > Mobile 0413 281 861
> >______________________________________________
> > <http://www.guycox.net/>http://www.guycox.net
> >
> >
> >-----Original Message-----
> >From: Confocal Microscopy List
> >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]]

> >On Behalf Of Charu Tanwar
> >Sent: Thursday, 6 May 2010 4:27 PM
> >To:
> ><http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]
> >Subject: DIC Image
> >
> >Dear List
> >I am doing an experiment with HeLa cells staining them with
> antibodies>for
> >two different proteins and trying to see localization.However,
> >Fluorescence
> >is not my problem but i am not able to visualize my cells in bright
> >field.......Well, i have to really put my eyes on so much of
> strain to
> >visualize them in bright field and get their morphology (Cells
> appear to
> >become very thin and barely gives boundary demarcation) . I got my

> >scope
> >checked and that is in good health. I tried this with other scope and
> >the
> >problem remains same. I have to focus my cells through
> epifluorescence>illumination (i.e. by looking at the expression).
> I have never faced
> >such a
> >weird problem before. Even confocal microscope refuses to give a good
> >DIC
> >image (same scope is giving very good DIC images for RBC'S and
> Candida).>I
> >do see two different expressions in my cells but i am not getting
> a good
> >DIC
> >image, sometimes i do not see anything in DIC channel but
> simultaneously>i
> >see good expression.
> >I foresee that this is not a scope problem but either with the
> cells or
> >staining protocol. However, i have changed my staining protocol 4

> times>(Fixation, time incubation for Ab's, no. of washes and
> blocking) but the
> >story is same.When i visualize my cells before starting the
> experiment,>cells look healthy and appear fine.
> >Please help
> >Charu Tanwar
> >
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi
> >India
> >
> >
> >
> >
> >No virus found in this incoming message.
> >Checked by AVG - www.avg.com
> >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> 05/06/10>04:26:00
>
>
> --
>

Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

I am embarrassed to say I don’t know whether high NA condensers requiring oil also require mounting your specimens on 0.17 mm thick coverslips. Is this true?  

 

 

Thomas E. Phillips, Ph.D

Professor of Biological Sciences

Director, Molecular Cytology Core

2 Tucker Hall

University of Missouri

Columbia, MO 65211-7400

573-882-4712 (office)

573-882-0123 (fax)

[hidden email]

 

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Eric Scarfone
Sent: Thursday, May 06, 2010 10:01 AM
To: [hidden email]
Subject: Re: DIC Image

 

Jim,
High NA condenser/objectives come with pretty short working distance that imposes mounting samples between 2 coverslip. At least that's the requirement for highly efficient DIC that imposes perfect kölher alignment I believe. That + the oil on both sides have made this practice less and less popular among students and facility managers (although not for the same reasons). 

Is darkfield more tolerant than DIC?

Cheers

Eric (13 years survivor, had to count twice to make sure, GOSH!!!)

 

 

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: James Pawley <[hidden email]>
Date: Thursday, May 6, 2010 3:42 pm
Subject: Re: DIC Image
To: [hidden email]

> Hi Charu,
>
> Well, you seem to have thought of most of the suggested solutions
> already.
> If, as Guy suggests, the problem is that there is really not
> enough
> left of your sample to produce a detectable RI difference, (which
> seems strange as video-enhanced DIC used to give good images of
> single microtubules, or even single actin filaments.) another
> possibility is Darkfield imaging.
>
> As any "survivor" of the UBC Course will tell you, darkfield is my
> favorite transmitted light imaging method, and it will really
> produce
> the most contrast for a given quantity of structural material
> present. The only problem is that you must use a darkfield
> condenser
> with a higher NA than that of your objective and you need to align
> the condenser with great care. Although this isn't always easy,
> the
> results can be really superb.
>
> If you don't have a super high-NA darkfield condenser (>0.9 NA.
> i.e.,
> one that you have to "oil" to the slide), you might find an
> objective
> with an iris diaphragm that will still allow you to make the
> objective NA lower than that of the illumination. While this will
> reduce the resolution a bit, it will definitely allow you to see
> the
> edges of the cells.
>
> In confocal one can get a similar image using the backscattered
> light
> signal, however, the signal from the cell will be overwhelmed by
> the
> much larger reflection from the glass-water interface and this
> will
> make it hard to see cellular details within 3-4 micrometers of the
> interface.
>
> Good luck,
>
> Jim Pawley
>
> *********************************************************************************
> Prof. James B. Pawley,
> Ph. 608-263-3147
> Room 223, Zoology Research Building,
> FAX 608-265-5315
> 1117 Johnson Ave., Madison, WI, 53706
> [hidden email]
> 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
> Vancouver Canada
> Info: http://www.3dcourse.ubc.ca/ Applications
> still
> being accepted
> "If it ain't diffraction, it must be statistics." Anon.
>
> >Hi
> >I also suspected the same when i face this problem for the first
> >time. I did little bit of modification in my fixation. What i am
> >doing now is fixing my cells with 100% chilled methanol and keep
> the
> >cells at 4 degrees for 10min. Then i do permeabilisation with
> 0.3%
> >triton X-100. i need to do this step in order to stain my protein
> >which is inside nucleus.
> >May be this is becoming too much for the cells.
> >Thanks
> >Charu
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi 110067
> >India.
> >
> >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
> >
> >
> >From: Guy Cox <[hidden email]>
> >Subject: Re: DIC Image
> >To: [hidden email]
> >Date: Thursday, 6 May, 2010, 12:15 PM
> >
> >Charu,
> >
> >My guess is that your fixation is permeabilising your cells so
> >effectively - and maybe extracting some lipids - that there isn't too
> >much left to generate DIC contrast. As Peter suggests, meticulously
> >checking the setup of your microscope - Koehler, polarizer,
> analyser etc
> >may help. But phase contrast might be a better option than DIC -
> >generally it does better with very thin samples. Phase lenses
> come in
> >two forms low (L) and high (H) absorbance - if you have a choice
> an H
> >lens will do better with a weakly diffracting object.
> >
> > Guy
> >
> >Please help fight breast cancer by sponsoring my
> >run in the Sydney Half Marathon on May 16th.
> ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox
> >______________________________________________
> >Associate Professor Guy Cox, MA, DPhil(Oxon)
> >Australian Centre for Microscopy & Microanalysis,
> >Madsen Building F09, University of Sydney, NSW 2006
> >
> >Phone +61 2 9351 3176 Fax +61 2 9351 7682
> > Mobile 0413 281 861
> >______________________________________________
> > <http://www.guycox.net/>http://www.guycox.net
> >
> >
> >-----Original Message-----
> >From: Confocal Microscopy List
> >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]]
> >On Behalf Of Charu Tanwar
> >Sent: Thursday, 6 May 2010 4:27 PM
> >To:
> ><http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]
> >Subject: DIC Image
> >
> >Dear List
> >I am doing an experiment with HeLa cells staining them with
> antibodies>for
> >two different proteins and trying to see localization.However,
> >Fluorescence
> >is not my problem but i am not able to visualize my cells in bright
> >field.......Well, i have to really put my eyes on so much of
> strain to
> >visualize them in bright field and get their morphology (Cells
> appear to
> >become very thin and barely gives boundary demarcation) . I got my
> >scope
> >checked and that is in good health. I tried this with other scope and
> >the
> >problem remains same. I have to focus my cells through
> epifluorescence>illumination (i.e. by looking at the expression).
> I have never faced
> >such a
> >weird problem before. Even confocal microscope refuses to give a good
> >DIC
> >image (same scope is giving very good DIC images for RBC'S and
> Candida).>I
> >do see two different expressions in my cells but i am not getting
> a good
> >DIC
> >image, sometimes i do not see anything in DIC channel but
> simultaneously>i
> >see good expression.
> >I foresee that this is not a scope problem but either with the
> cells or
> >staining protocol. However, i have changed my staining protocol 4
> times>(Fixation, time incubation for Ab's, no. of washes and
> blocking) but the
> >story is same.When i visualize my cells before starting the
> experiment,>cells look healthy and appear fine.
> >Please help
> >Charu Tanwar
> >
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi
> >India
> >
> >
> >
> >
> >No virus found in this incoming message.
> >Checked by AVG - www.avg.com
> >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> 05/06/10>04:26:00
>
>
> --
>

mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Eric Scarfone

Charu - Have you tried closing the condenser aperture?

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Eric Scarfone
Sent: Thursday, May 06, 2010 11:01 AM
To: [hidden email]
Subject: Re: DIC Image

 

Jim,
High NA condenser/objectives come with pretty short working distance that imposes mounting samples between 2 coverslip. At least that's the requirement for highly efficient DIC that imposes perfect kölher alignment I believe. That + the oil on both sides have made this practice less and less popular among students and facility managers (although not for the same reasons). 

Is darkfield more tolerant than DIC?

Cheers

Eric (13 years survivor, had to count twice to make sure, GOSH!!!)

 

 

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: James Pawley <[hidden email]>
Date: Thursday, May 6, 2010 3:42 pm
Subject: Re: DIC Image
To: [hidden email]

> Hi Charu,
>
> Well, you seem to have thought of most of the suggested solutions
> already.
> If, as Guy suggests, the problem is that there is really not
> enough
> left of your sample to produce a detectable RI difference, (which
> seems strange as video-enhanced DIC used to give good images of
> single microtubules, or even single actin filaments.) another
> possibility is Darkfield imaging.
>
> As any "survivor" of the UBC Course will tell you, darkfield is my
> favorite transmitted light imaging method, and it will really
> produce
> the most contrast for a given quantity of structural material
> present. The only problem is that you must use a darkfield
> condenser
> with a higher NA than that of your objective and you need to align
> the condenser with great care. Although this isn't always easy,
> the
> results can be really superb.
>
> If you don't have a super high-NA darkfield condenser (>0.9 NA.
> i.e.,
> one that you have to "oil" to the slide), you might find an
> objective
> with an iris diaphragm that will still allow you to make the
> objective NA lower than that of the illumination. While this will
> reduce the resolution a bit, it will definitely allow you to see
> the
> edges of the cells.
>
> In confocal one can get a similar image using the backscattered
> light
> signal, however, the signal from the cell will be overwhelmed by
> the
> much larger reflection from the glass-water interface and this
> will
> make it hard to see cellular details within 3-4 micrometers of the
> interface.
>
> Good luck,
>
> Jim Pawley
>
> *********************************************************************************
> Prof. James B. Pawley,
> Ph. 608-263-3147
> Room 223, Zoology Research Building,
> FAX 608-265-5315
> 1117 Johnson Ave., Madison, WI, 53706
> [hidden email]
> 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
> Vancouver Canada
> Info: http://www.3dcourse.ubc.ca/ Applications
> still
> being accepted
> "If it ain't diffraction, it must be statistics." Anon.
>
> >Hi
> >I also suspected the same when i face this problem for the first
> >time. I did little bit of modification in my fixation. What i am
> >doing now is fixing my cells with 100% chilled methanol and keep
> the
> >cells at 4 degrees for 10min. Then i do permeabilisation with
> 0.3%
> >triton X-100. i need to do this step in order to stain my protein
> >which is inside nucleus.
> >May be this is becoming too much for the cells.
> >Thanks
> >Charu
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi 110067
> >India.
> >
> >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
> >
> >
> >From: Guy Cox <[hidden email]>
> >Subject: Re: DIC Image
> >To: [hidden email]
> >Date: Thursday, 6 May, 2010, 12:15 PM
> >
> >Charu,
> >
> >My guess is that your fixation is permeabilising your cells so
> >effectively - and maybe extracting some lipids - that there isn't too
> >much left to generate DIC contrast. As Peter suggests, meticulously
> >checking the setup of your microscope - Koehler, polarizer,
> analyser etc
> >may help. But phase contrast might be a better option than DIC -
> >generally it does better with very thin samples. Phase lenses
> come in
> >two forms low (L) and high (H) absorbance - if you have a choice
> an H
> >lens will do better with a weakly diffracting object.
> >
> > Guy
> >
> >Please help fight breast cancer by sponsoring my
> >run in the Sydney Half Marathon on May 16th.
> ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox
> >______________________________________________
> >Associate Professor Guy Cox, MA, DPhil(Oxon)
> >Australian Centre for Microscopy & Microanalysis,
> >Madsen Building F09, University of Sydney, NSW 2006
> >
> >Phone +61 2 9351 3176 Fax +61 2 9351 7682
> > Mobile 0413 281 861
> >______________________________________________
> > <http://www.guycox.net/>http://www.guycox.net
> >
> >
> >-----Original Message-----
> >From: Confocal Microscopy List
> >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]]
> >On Behalf Of Charu Tanwar
> >Sent: Thursday, 6 May 2010 4:27 PM
> >To:
> ><http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]
> >Subject: DIC Image
> >
> >Dear List
> >I am doing an experiment with HeLa cells staining them with
> antibodies>for
> >two different proteins and trying to see localization.However,
> >Fluorescence
> >is not my problem but i am not able to visualize my cells in bright
> >field.......Well, i have to really put my eyes on so much of
> strain to
> >visualize them in bright field and get their morphology (Cells
> appear to
> >become very thin and barely gives boundary demarcation) . I got my
> >scope
> >checked and that is in good health. I tried this with other scope and
> >the
> >problem remains same. I have to focus my cells through
> epifluorescence>illumination (i.e. by looking at the expression).
> I have never faced
> >such a
> >weird problem before. Even confocal microscope refuses to give a good
> >DIC
> >image (same scope is giving very good DIC images for RBC'S and
> Candida).>I
> >do see two different expressions in my cells but i am not getting
> a good
> >DIC
> >image, sometimes i do not see anything in DIC channel but
> simultaneously>i
> >see good expression.
> >I foresee that this is not a scope problem but either with the
> cells or
> >staining protocol. However, i have changed my staining protocol 4
> times>(Fixation, time incubation for Ab's, no. of washes and
> blocking) but the
> >story is same.When i visualize my cells before starting the
> experiment,>cells look healthy and appear fine.
> >Please help
> >Charu Tanwar
> >
> >
> >CHARU TANWAR
> >Imaging Specialist
> >Advanced Instrumentation Research Facility
> >Jawaharlal Nehru University
> >New Delhi
> >India
> >
> >
> >
> >
> >No virus found in this incoming message.
> >Checked by AVG - www.avg.com
> >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> 05/06/10>04:26:00
>
>
> --
>

Eric Scarfone Eric Scarfone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

In reply to this post by Phillips, Thomas E.
Philip, the problem arises when you combine two short working distance lenses above and below the sample! If you want both of them to be focused on the same plane (ie Köhler illumination) you have no other way than to use coveslips on both sides!

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: "Phillips, Thomas E." <[hidden email]>
Date: Thursday, May 6, 2010 5:11 pm
Subject: Re: DIC Image
To: [hidden email]

> I am embarrassed to say I don't know whether high NA condensers
> requiring oil also require mounting your specimens on 0.17 mm

> thick coverslips. Is this true?
>
>
> Thomas E. Phillips, Ph.D
> Professor of Biological Sciences
> Director, Molecular Cytology Core
> 2 Tucker Hall
> University of Missouri
> Columbia, MO 65211-7400
> 573-882-4712 (office)
> 573-882-0123 (fax)
> [hidden email]<mailto:[hidden email]>
>
> http://www.biology.missouri.edu/faculty/phillips.html
> http://www.biotech.missouri.edu/mcc/
>
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of Eric Scarfone
> Sent: Thursday, May 06, 2010 10:01 AM
> To: [hidden email]
> Subject: Re: DIC Image
>
>
> Jim,
> High NA condenser/objectives come with pretty short working
> distance that imposes mounting samples between 2 coverslip. At

> least that's the requirement for highly efficient DIC that imposes
> perfect kölher alignment I believe. That + the oil on both sides
> have made this practice less and less popular among students and
> facility managers (although not for the same reasons).
>
> Is darkfield more tolerant than DIC?
>
> Cheers
>
> Eric (13 years survivor, had to count twice to make sure, GOSH!!!)
>
>
>
>
>
> Eric Scarfone, PhD, CNRS,
> Center for Hearing and communication Research
> Department of Clinical Neuroscience
> Karolinska Institutet
>
> Postal Address:
> CFH, M1:02
> Karolinska Hospital,
> SE-171 76 Stockholm, Sweden
>
> Work: +46 (0)8-517 79343,
> Cell: +46 (0)70 888 2352
> Fax: +46 (0)8-301876
>
> email: [hidden email]

> http://www.ki.se/cfh/
>
>
> ----- Original Message -----
> From: James Pawley <[hidden email]>
> Date: Thursday, May 6, 2010 3:42 pm
> Subject: Re: DIC Image
> To: [hidden email]
>
> > Hi Charu,
> >
> > Well, you seem to have thought of most of the suggested solutions
> > already.
> > If, as Guy suggests, the problem is that there is really not
> > enough
> > left of your sample to produce a detectable RI difference, (which
> > seems strange as video-enhanced DIC used to give good images of
> > single microtubules, or even single actin filaments.) another
> > possibility is Darkfield imaging.
> >
> > As any "survivor" of the UBC Course will tell you, darkfield is my
> > favorite transmitted light imaging method, and it will really

> > produce
> > the most contrast for a given quantity of structural material
> > present. The only problem is that you must use a darkfield
> > condenser
> > with a higher NA than that of your objective and you need to align
> > the condenser with great care. Although this isn't always easy,
> > the
> > results can be really superb.
> >
> > If you don't have a super high-NA darkfield condenser (>0.9 NA.
> > i.e.,
> > one that you have to "oil" to the slide), you might find an
> > objective
> > with an iris diaphragm that will still allow you to make the
> > objective NA lower than that of the illumination. While this will
> > reduce the resolution a bit, it will definitely allow you to see
> > the
> > edges of the cells.

> >
> > In confocal one can get a similar image using the backscattered
> > light
> > signal, however, the signal from the cell will be overwhelmed by
> > the
> > much larger reflection from the glass-water interface and this
> > will
> > make it hard to see cellular details within 3-4 micrometers of the
> > interface.
> >
> > Good luck,
> >
> > Jim Pawley
> >
> >
> *********************************************************************************> Prof. James B. Pawley,
> > Ph. 608-263-3147
> > Room 223, Zoology Research Building,
> > FAX 608-265-5315
> > 1117 Johnson Ave., Madison, WI, 53706
> > [hidden email]
> > 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
> > Vancouver Canada
> > Info: http://www.3dcourse.ubc.ca/ Applications

> > still
> > being accepted
> > "If it ain't diffraction, it must be statistics." Anon.
> >
> > >Hi
> > >I also suspected the same when i face this problem for the first
> > >time. I did little bit of modification in my fixation. What i am
> > >doing now is fixing my cells with 100% chilled methanol and keep
> > the
> > >cells at 4 degrees for 10min. Then i do permeabilisation with
> > 0.3%
> > >triton X-100. i need to do this step in order to stain my protein
> > >which is inside nucleus.
> > >May be this is becoming too much for the cells.
> > >Thanks
> > >Charu
> > >
> > >CHARU TANWAR
> > >Imaging Specialist
> > >Advanced Instrumentation Research Facility
> > >Jawaharlal Nehru University

> > >New Delhi 110067
> > >India.
> > >
> > >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
> > >
> > >
> > >From: Guy Cox <[hidden email]>
> > >Subject: Re: DIC Image
> > >To: [hidden email]
> > >Date: Thursday, 6 May, 2010, 12:15 PM
> > >
> > >Charu,
> > >
> > >My guess is that your fixation is permeabilising your cells so
> > >effectively - and maybe extracting some lipids - that there
> isn't too
> > >much left to generate DIC contrast. As Peter suggests, meticulously
> > >checking the setup of your microscope - Koehler, polarizer,
> > analyser etc
> > >may help. But phase contrast might be a better option than DIC -
> > >generally it does better with very thin samples. Phase lenses

> > come in
> > >two forms low (L) and high (H) absorbance - if you have a choice
> > an H
> > >lens will do better with a weakly diffracting object.
> > >
> > > Guy
> > >
> > >Please help fight breast cancer by sponsoring my
> > >run in the Sydney Half Marathon on May 16th.
> >
> ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox> >______________________________________________
> > >Associate Professor Guy Cox, MA, DPhil(Oxon)
> > >Australian Centre for Microscopy & Microanalysis,
> > >Madsen Building F09, University of Sydney, NSW 2006
> > >
> > >Phone +61 2 9351 3176 Fax +61 2 9351 7682
> > > Mobile 0413 281 861
> > >______________________________________________

> > > <http://www.guycox.net/>http://www.guycox.net
> > >
> > >
> > >-----Original Message-----
> > >From: Confocal Microscopy List
> >
> >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]]> >On Behalf Of Charu Tanwar
> > >Sent: Thursday, 6 May 2010 4:27 PM
> > >To:
> >
> ><http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]> >Subject: DIC Image
> > >
> > >Dear List
> > >I am doing an experiment with HeLa cells staining them with
> > antibodies>for
> > >two different proteins and trying to see localization.However,
> > >Fluorescence
> > >is not my problem but i am not able to visualize my cells in bright

> > >field.......Well, i have to really put my eyes on so much of
> > strain to
> > >visualize them in bright field and get their morphology (Cells
> > appear to
> > >become very thin and barely gives boundary demarcation) . I got my
> > >scope
> > >checked and that is in good health. I tried this with other
> scope and
> > >the
> > >problem remains same. I have to focus my cells through
> > epifluorescence>illumination (i.e. by looking at the expression).
> > I have never faced
> > >such a
> > >weird problem before. Even confocal microscope refuses to give
> a good
> > >DIC
> > >image (same scope is giving very good DIC images for RBC'S and
> > Candida).>I
> > >do see two different expressions in my cells but i am not getting

> > a good
> > >DIC
> > >image, sometimes i do not see anything in DIC channel but
> > simultaneously>i
> > >see good expression.
> > >I foresee that this is not a scope problem but either with the
> > cells or
> > >staining protocol. However, i have changed my staining protocol 4
> > times>(Fixation, time incubation for Ab's, no. of washes and
> > blocking) but the
> > >story is same.When i visualize my cells before starting the
> > experiment,>cells look healthy and appear fine.
> > >Please help
> > >Charu Tanwar
> > >
> > >
> > >CHARU TANWAR
> > >Imaging Specialist
> > >Advanced Instrumentation Research Facility
> > >Jawaharlal Nehru University

> > >New Delhi
> > >India
> > >
> > >
> > >
> > >
> > >No virus found in this incoming message.
> > >Checked by AVG - www.avg.com
> > >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> > 05/06/10>04:26:00
> >
> >
> > --
> >
>
Tobias Baskin Tobias Baskin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

THis is true if you make a condenser from an
objective lens. But I believe that high NA
condensers (like the dark field type that Jim
Pawley mentioned) have a working distance long
enough to accept a slide. I have been able to
fill the rear focal plane of a 1.25 NA Nomarksi
objective by oiling the high NA condenser to the
slide (I suppose might have been even better  by
using a coverslip sandwich, but I don't think
that is essential).

        Tobias

>Philip, the problem arises when you combine two
>short working distance lenses above and below
>the sample! If you want both of them to be
>focused on the same plane (ie Köhler
>illumination) you have no other way than to use
>coveslips on both sides!
>
>Eric Scarfone, PhD, CNRS,
>Center for Hearing and communication Research
>Department of Clinical Neuroscience
>Karolinska Institutet
>
>Postal Address:
>CFH, M1:02
>Karolinska Hospital,
>SE-171 76 Stockholm, Sweden
>
>Work: +46 (0)8-517 79343,
>Cell: +46 (0)70 888 2352
>Fax: +46 (0)8-301876
>
>email: [hidden email]
>http://www.ki.se/cfh/
>
>
>----- Original Message -----
>From: "Phillips, Thomas E." <[hidden email]>
>Date: Thursday, May 6, 2010 5:11 pm
>Subject: Re: DIC Image
>To: [hidden email]
>
>>  I am embarrassed to say I don't know whether high NA condensers
>>  requiring oil also require mounting your specimens on 0.17 mm
>>  thick coverslips. Is this true?
>>
>>
>>  Thomas E. Phillips, Ph.D
>>  Professor of Biological Sciences
>>  Director, Molecular Cytology Core
>>  2 Tucker Hall
>>  University of Missouri
>>  Columbia, MO 65211-7400
>>  573-882-4712 (office)
>>  573-882-0123 (fax)
>>  [hidden email]<mailto:[hidden email]>
>>
>>  http://www.biology.missouri.edu/faculty/phillips.html
>>  http://www.biotech.missouri.edu/mcc/
>>
>>  From: Confocal Microscopy List
>>  [mailto:[hidden email]] On Behalf Of Eric Scarfone
>>  Sent: Thursday, May 06, 2010 10:01 AM
>>  To: [hidden email]
>>  Subject: Re: DIC Image
>>
>>
>>  Jim,
>>  High NA condenser/objectives come with pretty short working
>>  distance that imposes mounting samples between 2 coverslip. At
>>  least that's the requirement for highly efficient DIC that imposes
>>  perfect kölher alignment I believe. That + the oil on both sides
>>  have made this practice less and less popular among students and
>>  facility managers (although not for the same reasons).
>>
>>  Is darkfield more tolerant than DIC?
>>
>>  Cheers
>>
>>  Eric (13 years survivor, had to count twice to make sure, GOSH!!!)
>>
>>
>>
>>
>>
>>  Eric Scarfone, PhD, CNRS,
>>  Center for Hearing and communication Research
>>  Department of Clinical Neuroscience
>>  Karolinska Institutet
>>
>>  Postal Address:
>>  CFH, M1:02
>>  Karolinska Hospital,
>>  SE-171 76 Stockholm, Sweden
>>
>>  Work: +46 (0)8-517 79343,
>>  Cell: +46 (0)70 888 2352
>>  Fax: +46 (0)8-301876
>>
>>  email: [hidden email]
>>  http://www.ki.se/cfh/
>>
>>
>>  ----- Original Message -----
>>  From: James Pawley <[hidden email]>
>>  Date: Thursday, May 6, 2010 3:42 pm
>>  Subject: Re: DIC Image
>>  To: [hidden email]
>>
>>  > Hi Charu,
>>  >
>>  > Well, you seem to have thought of most of the suggested solutions
>>  > already.
>>  > If, as Guy suggests, the problem is that there is really not
>>  > enough
>>  > left of your sample to produce a detectable RI difference, (which
>>  > seems strange as video-enhanced DIC used to give good images of
>>  > single microtubules, or even single actin filaments.) another
>>  > possibility is Darkfield imaging.
>>  >
>>  > As any "survivor" of the UBC Course will tell you, darkfield is my
>>  > favorite transmitted light imaging method, and it will really
>>  > produce
>>  > the most contrast for a given quantity of structural material
>>  > present. The only problem is that you must use a darkfield
>>  > condenser
>>  > with a higher NA than that of your objective and you need to align
>>  > the condenser with great care. Although this isn't always easy,
>>  > the
>>  > results can be really superb.
>>  >
>>  > If you don't have a super high-NA darkfield condenser (>0.9 NA.
>  > > i.e.,
>>  > one that you have to "oil" to the slide), you might find an
>>  > objective
>>  > with an iris diaphragm that will still allow you to make the
>>  > objective NA lower than that of the illumination. While this will
>>  > reduce the resolution a bit, it will definitely allow you to see
>>  > the
>>  > edges of the cells.
>>  >
>>  > In confocal one can get a similar image using the backscattered
>>  > light
>>  > signal, however, the signal from the cell will be overwhelmed by
>>  > the
>>  > much larger reflection from the glass-water interface and this
>>  > will
>>  > make it hard to see cellular details within 3-4 micrometers of the
>>  > interface.
>>  >
>>  > Good luck,
>>  >
>>  > Jim Pawley
>>  >
>>  >
>>
>>*********************************************************************************>
>>Prof. James B. Pawley,
>>  > Ph. 608-263-3147
>>  > Room 223, Zoology Research Building,
>>  > FAX 608-265-5315
>>  > 1117 Johnson Ave., Madison, WI, 53706
>>  > [hidden email]
>>  > 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
>>  > Vancouver Canada
>>  > Info: http://www.3dcourse.ubc.ca/ Applications
>>  > still
>>  > being accepted
>>  > "If it ain't diffraction, it must be statistics." Anon.
>>  >
>>  > >Hi
>>  > >I also suspected the same when i face this problem for the first
>>  > >time. I did little bit of modification in my fixation. What i am
>>  > >doing now is fixing my cells with 100% chilled methanol and keep
>>  > the
>>  > >cells at 4 degrees for 10min. Then i do permeabilisation with
>>  > 0.3%
>>  > >triton X-100. i need to do this step in order to stain my protein
>>  > >which is inside nucleus.
>>  > >May be this is becoming too much for the cells.
>>  > >Thanks
>>  > >Charu
>>  > >
>>  > >CHARU TANWAR
>>  > >Imaging Specialist
>>  > >Advanced Instrumentation Research Facility
>>  > >Jawaharlal Nehru University
>>  > >New Delhi 110067
>>  > >India.
>>  > >
>>  > >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
>>  > >
>>  > >
>>  > >From: Guy Cox <[hidden email]>
>>  > >Subject: Re: DIC Image
>>  > >To: [hidden email]
>>  > >Date: Thursday, 6 May, 2010, 12:15 PM
>>  > >
>>  > >Charu,
>>  > >
>>  > >My guess is that your fixation is permeabilising your cells so
>>  > >effectively - and maybe extracting some lipids - that there
>>  isn't too
>>  > >much left to generate DIC contrast. As Peter suggests, meticulously
>>  > >checking the setup of your microscope - Koehler, polarizer,
>>  > analyser etc
>>  > >may help. But phase contrast might be a better option than DIC -
>>  > >generally it does better with very thin samples. Phase lenses
>>  > come in
>>  > >two forms low (L) and high (H) absorbance - if you have a choice
>>  > an H
>>  > >lens will do better with a weakly diffracting object.
>>  > >
>>  > > Guy
>>  > >
>>  > >Please help fight breast cancer by sponsoring my
>>  > >run in the Sydney Half Marathon on May 16th.
>>  >
>>  ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox> >______________________________________________
>>  > >Associate Professor Guy Cox, MA, DPhil(Oxon)
>>  > >Australian Centre for Microscopy & Microanalysis,
>>  > >Madsen Building F09, University of Sydney, NSW 2006
>>  > >
>>  > >Phone +61 2 9351 3176 Fax +61 2 9351 7682
>>  > > Mobile 0413 281 861
>>  > >______________________________________________
>>  > > <http://www.guycox.net/>http://www.guycox.net
>>  > >
>>  > >
>>  > >-----Original Message-----
>>  > >From: Confocal Microscopy List
>>  >
>>  >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...>[hidden email]]> >On
>>Behalf Of Charu Tanwar
>>  > >Sent: Thursday, 6 May 2010 4:27 PM
>>  > >To:
>>  >
>>  ><http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...>[hidden email]> >Subject:
>>DIC Image
>>  > >
>>  > >Dear List
>>  > >I am doing an experiment with HeLa cells staining them with
>>  > antibodies>for
>>  > >two different proteins and trying to see localization.However,
>>  > >Fluorescence
>>  > >is not my problem but i am not able to visualize my cells in bright
>>  > >field.......Well, i have to really put my eyes on so much of
>>  > strain to
>>  > >visualize them in bright field and get their morphology (Cells
>  > > appear to
>>  > >become very thin and barely gives boundary demarcation) . I got my
>>  > >scope
>>  > >checked and that is in good health. I tried this with other
>>  scope and
>>  > >the
>>  > >problem remains same. I have to focus my cells through
>>  > epifluorescence>illumination (i.e. by looking at the expression).
>>  > I have never faced
>>  > >such a
>>  > >weird problem before. Even confocal microscope refuses to give
>>  a good
>>  > >DIC
>>  > >image (same scope is giving very good DIC images for RBC'S and
>>  > Candida).>I
>>  > >do see two different expressions in my cells but i am not getting
>>  > a good
>>  > >DIC
>>  > >image, sometimes i do not see anything in DIC channel but
>>  > simultaneously>i
>>  > >see good expression.
>>  > >I foresee that this is not a scope problem but either with the
>>  > cells or
>>  > >staining protocol. However, i have changed my staining protocol 4
>>  > times>(Fixation, time incubation for Ab's, no. of washes and
>>  > blocking) but the
>>  > >story is same.When i visualize my cells before starting the
>>  > experiment,>cells look healthy and appear fine.
>>  > >Please help
>>  > >Charu Tanwar
>>  > >
>>  > >
>>  > >CHARU TANWAR
>>  > >Imaging Specialist
>>  > >Advanced Instrumentation Research Facility
>>  > >Jawaharlal Nehru University
>>  > >New Delhi
>>  > >India
>>  > >
>>  > >
>>  > >
>>  > >
>>  > >No virus found in this incoming message.
>>  > >Checked by AVG - www.avg.com
>>  > >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
>>  > 05/06/10>04:26:00
>>  >
>>  >
>>  > --
>>  >
>>


--
       _      ____          __   ____
      /  \   /          / \    /   \ \        Tobias I. Baskin
     /   /  /          /   \   \      \         Biology Department
    /_ /   __      /__ \   \       \__    611 N. Pleasant St.
   /      /          /       \   \       \        University of Massachusetts
  /      /          /         \   \       \    Amherst, MA, 01003
/      / ___   /           \   \__/  \ ____
www.bio.umass.edu/biology/baskin
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243
Eric Scarfone Eric Scarfone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DIC Image

I agree for most cases this extreme setting is not required, luckily enough.
But since Charu seems to have a particularly difficult sample, he might need to make use of those high angled light rays. 
But again, Dark field might be more tolerant than DIC

Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet

Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden

Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876

email: [hidden email]
http://www.ki.se/cfh/


----- Original Message -----
From: Tobias Baskin <[hidden email]>
Date: Thursday, May 6, 2010 5:58 pm
Subject: Re: DIC Image
To: [hidden email]

> THis is true if you make a condenser from an
> objective lens. But I believe that high NA

> condensers (like the dark field type that Jim
> Pawley mentioned) have a working distance long
> enough to accept a slide. I have been able to
> fill the rear focal plane of a 1.25 NA Nomarksi
> objective by oiling the high NA condenser to the
> slide (I suppose might have been even better by
> using a coverslip sandwich, but I don't think
> that is essential).
>
> Tobias
>
> >Philip, the problem arises when you combine two
> >short working distance lenses above and below
> >the sample! If you want both of them to be
> >focused on the same plane (ie Köhler
> >illumination) you have no other way than to use
> >coveslips on both sides!
> >
> >Eric Scarfone, PhD, CNRS,
> >Center for Hearing and communication Research
> >Department of Clinical Neuroscience

> >Karolinska Institutet
> >
> >Postal Address:
> >CFH, M1:02
> >Karolinska Hospital,
> >SE-171 76 Stockholm, Sweden
> >
> >Work: +46 (0)8-517 79343,
> >Cell: +46 (0)70 888 2352
> >Fax: +46 (0)8-301876
> >
> >email: [hidden email]
> >http://www.ki.se/cfh/
> >
> >
> >----- Original Message -----
> >From: "Phillips, Thomas E." <[hidden email]>
> >Date: Thursday, May 6, 2010 5:11 pm
> >Subject: Re: DIC Image
> >To: [hidden email]
> >
> >> I am embarrassed to say I don't know whether high NA condensers
> >> requiring oil also require mounting your specimens on 0.17 mm
> >> thick coverslips. Is this true?
> >>
> >>
> >> Thomas E. Phillips, Ph.D

> >> Professor of Biological Sciences
> >> Director, Molecular Cytology Core
> >> 2 Tucker Hall
> >> University of Missouri
> >> Columbia, MO 65211-7400
> >> 573-882-4712 (office)
> >> 573-882-0123 (fax)
> >> [hidden email]<mailto:[hidden email]>
> >>
> >> http://www.biology.missouri.edu/faculty/phillips.html
> >> http://www.biotech.missouri.edu/mcc/
> >>
> >> From: Confocal Microscopy List
> >> [mailto:[hidden email]] On Behalf Of Eric
> Scarfone>> Sent: Thursday, May 06, 2010 10:01 AM
> >> To: [hidden email]
> >> Subject: Re: DIC Image
> >>
> >>
> >> Jim,
> >> High NA condenser/objectives come with pretty short working

> >> distance that imposes mounting samples between 2 coverslip. At
> >> least that's the requirement for highly efficient DIC that imposes
> >> perfect kölher alignment I believe. That + the oil on both sides
> >> have made this practice less and less popular among students and
> >> facility managers (although not for the same reasons).
> >>
> >> Is darkfield more tolerant than DIC?
> >>
> >> Cheers
> >>
> >> Eric (13 years survivor, had to count twice to make sure, GOSH!!!)
> >>
> >>
> >>
> >>
> >>
> >> Eric Scarfone, PhD, CNRS,
> >> Center for Hearing and communication Research
> >> Department of Clinical Neuroscience
> >> Karolinska Institutet

> >>
> >> Postal Address:
> >> CFH, M1:02
> >> Karolinska Hospital,
> >> SE-171 76 Stockholm, Sweden
> >>
> >> Work: +46 (0)8-517 79343,
> >> Cell: +46 (0)70 888 2352
> >> Fax: +46 (0)8-301876
> >>
> >> email: [hidden email]
> >> http://www.ki.se/cfh/
> >>
> >>
> >> ----- Original Message -----
> >> From: James Pawley <[hidden email]>
> >> Date: Thursday, May 6, 2010 3:42 pm
> >> Subject: Re: DIC Image
> >> To: [hidden email]
> >>
> >> > Hi Charu,
> >> >
> >> > Well, you seem to have thought of most of the suggested
> solutions>> > already.
> >> > If, as Guy suggests, the problem is that there is really not

> >> > enough
> >> > left of your sample to produce a detectable RI difference,
> (which>> > seems strange as video-enhanced DIC used to give good
> images of
> >> > single microtubules, or even single actin filaments.) another
> >> > possibility is Darkfield imaging.
> >> >
> >> > As any "survivor" of the UBC Course will tell you, darkfield
> is my
> >> > favorite transmitted light imaging method, and it will really
> >> > produce
> >> > the most contrast for a given quantity of structural material
> >> > present. The only problem is that you must use a darkfield
> >> > condenser
> >> > with a higher NA than that of your objective and you need to
> align>> > the condenser with great care. Although this isn't

> always easy,
> >> > the
> >> > results can be really superb.
> >> >
> >> > If you don't have a super high-NA darkfield condenser (>0.9 NA.
> > > > i.e.,
> >> > one that you have to "oil" to the slide), you might find an
> >> > objective
> >> > with an iris diaphragm that will still allow you to make the
> >> > objective NA lower than that of the illumination. While this
> will>> > reduce the resolution a bit, it will definitely allow
> you to see
> >> > the
> >> > edges of the cells.
> >> >
> >> > In confocal one can get a similar image using the backscattered
> >> > light
> >> > signal, however, the signal from the cell will be

> overwhelmed by
> >> > the
> >> > much larger reflection from the glass-water interface and this
> >> > will
> >> > make it hard to see cellular details within 3-4 micrometers
> of the
> >> > interface.
> >> >
> >> > Good luck,
> >> >
> >> > Jim Pawley
> >> >
> >> >
> >>
> >>*********************************************************************************>
> >>Prof. James B. Pawley,
> >> > Ph. 608-263-3147
> >> > Room 223, Zoology Research Building,
> >> > FAX 608-265-5315
> >> > 1117 Johnson Ave., Madison, WI, 53706
> >> > [hidden email]
> >> > 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,

> >> > Vancouver Canada
> >> > Info: http://www.3dcourse.ubc.ca/ Applications
> >> > still
> >> > being accepted
> >> > "If it ain't diffraction, it must be statistics." Anon.
> >> >
> >> > >Hi
> >> > >I also suspected the same when i face this problem for the
> first>> > >time. I did little bit of modification in my fixation.
> What i am
> >> > >doing now is fixing my cells with 100% chilled methanol and
> keep>> > the
> >> > >cells at 4 degrees for 10min. Then i do permeabilisation with
> >> > 0.3%
> >> > >triton X-100. i need to do this step in order to stain my
> protein>> > >which is inside nucleus.
> >> > >May be this is becoming too much for the cells.

> >> > >Thanks
> >> > >Charu
> >> > >
> >> > >CHARU TANWAR
> >> > >Imaging Specialist
> >> > >Advanced Instrumentation Research Facility
> >> > >Jawaharlal Nehru University
> >> > >New Delhi 110067
> >> > >India.
> >> > >
> >> > >--- On Thu, 6/5/10, Guy Cox <[hidden email]> wrote:
> >> > >
> >> > >
> >> > >From: Guy Cox <[hidden email]>
> >> > >Subject: Re: DIC Image
> >> > >To: [hidden email]
> >> > >Date: Thursday, 6 May, 2010, 12:15 PM
> >> > >
> >> > >Charu,
> >> > >
> >> > >My guess is that your fixation is permeabilising your cells so

> >> > >effectively - and maybe extracting some lipids - that there
> >> isn't too
> >> > >much left to generate DIC contrast. As Peter suggests,
> meticulously>> > >checking the setup of your microscope -
> Koehler, polarizer,
> >> > analyser etc
> >> > >may help. But phase contrast might be a better option than
> DIC -
> >> > >generally it does better with very thin samples. Phase lenses
> >> > come in
> >> > >two forms low (L) and high (H) absorbance - if you have a
> choice>> > an H
> >> > >lens will do better with a weakly diffracting object.
> >> > >
> >> > > Guy
> >> > >
> >> > >Please help fight breast cancer by sponsoring my

> >> > >run in the Sydney Half Marathon on May 16th.
> >> >
> >>
> ><http://www.everydayhero.com.au/guy_cox>http://www.everydayhero.com.au/guy_cox> >______________________________________________
> >> > >Associate Professor Guy Cox, MA, DPhil(Oxon)
> >> > >Australian Centre for Microscopy & Microanalysis,
> >> > >Madsen Building F09, University of Sydney, NSW 2006
> >> > >
> >> > >Phone +61 2 9351 3176 Fax +61 2 9351 7682
> >> > > Mobile 0413 281 861
> >> > >______________________________________________
> >> > > <http://www.guycox.net/>http://www.guycox.net
> >> > >
> >> > >
> >> > >-----Original Message-----
> >> > >From: Confocal Microscopy List

> >> >
> >>
> >[mailto:<http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]]> >On
> >>Behalf Of Charu Tanwar
> >> > >Sent: Thursday, 6 May 2010 4:27 PM
> >> > >To:
> >> >
> >>
> ><http://in.mc83.mail.yahoo.com/mc/compose?to=[hidden email]>[hidden email]> >Subject:
> >>DIC Image
> >> > >
> >> > >Dear List
> >> > >I am doing an experiment with HeLa cells staining them with
> >> > antibodies>for
> >> > >two different proteins and trying to see localization.However,
> >> > >Fluorescence
> >> > >is not my problem but i am not able to visualize my cells

> in bright
> >> > >field.......Well, i have to really put my eyes on so much of
> >> > strain to
> >> > >visualize them in bright field and get their morphology (Cells
> > > > appear to
> >> > >become very thin and barely gives boundary demarcation) . I
> got my
> >> > >scope
> >> > >checked and that is in good health. I tried this with other
> >> scope and
> >> > >the
> >> > >problem remains same. I have to focus my cells through
> >> > epifluorescence>illumination (i.e. by looking at the
> expression).>> > I have never faced
> >> > >such a
> >> > >weird problem before. Even confocal microscope refuses to give
> >> a good

> >> > >DIC
> >> > >image (same scope is giving very good DIC images for RBC'S and
> >> > Candida).>I
> >> > >do see two different expressions in my cells but i am not
> getting>> > a good
> >> > >DIC
> >> > >image, sometimes i do not see anything in DIC channel but
> >> > simultaneously>i
> >> > >see good expression.
> >> > >I foresee that this is not a scope problem but either with the
> >> > cells or
> >> > >staining protocol. However, i have changed my staining
> protocol 4
> >> > times>(Fixation, time incubation for Ab's, no. of washes and
> >> > blocking) but the
> >> > >story is same.When i visualize my cells before starting the
> >> > experiment,>cells look healthy and appear fine.

> >> > >Please help
> >> > >Charu Tanwar
> >> > >
> >> > >
> >> > >CHARU TANWAR
> >> > >Imaging Specialist
> >> > >Advanced Instrumentation Research Facility
> >> > >Jawaharlal Nehru University
> >> > >New Delhi
> >> > >India
> >> > >
> >> > >
> >> > >
> >> > >
> >> > >No virus found in this incoming message.
> >> > >Checked by AVG - www.avg.com
> >> > >Version: 9.0.819 / Virus Database: 271.1.1/2854 - Release Date:
> >> > 05/06/10>04:26:00
> >> >
> >> >
> >> > --
> >> >
> >>
>
>

> --
> _ ____ __ ____
> / \ / / \ / \ \ Tobias I. Baskin
> / / / / \ \ \ Biology Department
> /_ / __ /__ \ \ \__ 611 N. Pleasant St.
> / / / \ \ \ University of
> Massachusetts / / / \ \ \
> Amherst, MA, 01003
> / / ___ / \ \__/ \ ____
> www.bio.umass.edu/biology/baskin
> Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243
>
12