|
Stanislav Vitha |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hallo, I would appreciate a pointer to a paper or a book chapter that provides the theoretical background/explanation for the fact that a DIC prism will degrade resolution of the fluorescence image. My search in archives of this list brought up quite a few discussions on this topic, but I was looking for a more theoretical treatment of the issue. I have a little puzzle with our Fluoview FV1000 system (for those not familiar with Olympus setup: these is a single DIC prism behind the objective; the condenser turret then has several different DIC prisms, specific to each objective - but that is not relevant for epifluorescnece). A expected, having the DIC prism behind the objective degrades the confocal image with all objectives, but surprisingly the degradation is much worse with our 20x/0.85 oil than is with the 60x/1.2 WI or 100x/1.4 oil objectives. I expected the opposite, the amount of shear introduced by the prism (a fraction of a micrometer) should have been more noticeable with higher resolution objectives. It makes it very easy to demonstrate the effect of DIC prism on confocal image quality to new users during training, but I would sure like to know what is happening. The second puzzle is that inserting the DIC prism in the optical path causes a shift in focus (at least several micrometers, very noticeable). In my naive thinking, I assumed that inserting a flat piece of glas at normal orientation into the infininty space would not cause any focus shift, at least for imaging sample area close to the center of the optical axis. Any ideas? I am in discussion with Olympus and thay are very receptive, but I thought I would ask here as well. Thanks in advance! Stan Vitha Microscopy and Imaging Center Texas A&M University College Station, TX |
|
Johannes Helm |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Stanislav. You might have a look at http://micro.magnet.fsu.edu/primer/techniques/dic/dichome.html Depending on what your concept of "theoretical" is, this might be the source of information you are looking for - or not. The wording "behind" the objective in your text is a little bit un-clear. What you mean, I suppose, is the following: In a classical transmitted light configuration, you will have the light source, a collector lens, the field diaphragm, then the DIC polarizer, the condenser DIC prism - also known as the "main DIC prism" -, the condenser with its front lens, the condenser immersion medium (air, water, oil, or glycerine), the microscope slide, the specimen in its mounting medium, the cover slip, the immersion medium for the objective - e.g. air, water, oil, glycerine -, the objective, the objective DIC prism - also known as "auxiliary DIC prism" -, the DIC analyzer and finally the observation or imaging optics. What you describe as the "DIC prism behind the objective" most probably is the auxiliary DIC prism, i.e. the DIC prism on the image side of the objective. While a number of microscope producers have this auxiliary DIC prism mounted so that you can adjust it for optimizing your contrast image ("DIC slider"), in the Olympus systems an adjustment is done in a slightly different way; Olympus provides a lambda/4 plate as an additional element between DIC polarizer and condenser DIC prism. The user then adjusts the orientation of the DIC polarizer in order to change the contrast settings. Also, optimal DIC observation sometimes may request a rotatable specimen stage centered in the optical axis of the microscope. The name DIC, "D"IC, "D"ifferential Interference Contrast, delivers the explanation to your question. The DIC prism, also known as "Nomarski prism", named after its inventor Jerzy Nomarski, provides a "d"ifferential splitting of the ray paths of the light in the microscope. If you do confocal laser scanning microscopy and have a well polarized laser beam - normally, a laser beam generated by a "typical CLSM laser" will be polarized per se, but it might nevertheless be sent through the DIC analyzer (which is nothing else but a polarizer) -, then that beam will be split into two beams, which are laterally displaced "differentially". The lateral separation will hence be small, smaller than the lateral resolution limit of the objective when used in the wide field observation mode. Other types of interference contrast microscopes - e.g. Jamin Lebedeff - will generate two beams with a macroscopic separation; then one beam, the reference beam, will bypass the entire microscope on a ray path outside the microscope frame, but this is not of any interest here. What hence happens when you try to do CLSM with a fluorescent specimen - fluorescent light assumed to be un-polarized - and have a DIC prism in the ray path is that you will experience what are a kind of super imposed "double images" if you have a sufficiently large zooming factor so that you approach the resolution limit with your pixelization. The story is a little bit more complicated in case of reflected light CLS microscopy because of possible polarization conservation effects, but let's keep this out of the discussion here for simplicity. So, in order to avoid these effects, take out the secondary DIC prism from the ray path when CLSMing. NOW, there are some microscope objective holders, e.g. on a BX51WI used by cell physiologists, e.g. Olympus item code WI-SRE2, in which a secondary DIC prism, e.g. WI-DICTHRA2, is inserted so that you cannot easilry remove it from the ray path. There does, however, exist a version of WI-SRE2, named "FV10-SRE", if I recall correctly, where you have a slider, which will permit you to easily insert or extract the DIC prism from the ray path. Unfortunately, this FV10-SRE seemingly is not sold a million times per month or so, hence it is not really cheap. But it might help. Best wishes, Johannes -- P. Johannes Helm, M.Sc. PhD Seniorengineer CMBN University of Oslo Institute of Basic Medical Science Department of Anatomy Postboks 1105 - Blindern NO-0317 Oslo Voice: +47 228 51159 Fax: +47 228 51499 WWW: folk.uio.no/jhelm |
|
Stanislav Vitha |
|
|
In reply to this post by Stanislav Vitha
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Johannes, Thank you for your detailed explanation. I should clarify that our confocal system has an inverted IX-81 microscope. By "behind the objective" I meant on the opposite side of the objective than the specimen is (the convention, I believe, is that when taking pictures, the object is in front of the lens, and photographer is behind the lens. The image is formed behind the lens, which is also the side where we find the back focal plane). Since I am collecting confocal epifluorescnce image simultaneously with recording a DIC image on the transmitted detector, the degradation of fluorescence images relates to the objective DIC prism. The condenser DIC prism is not in the optical path of the epifluorescence imaging setup. I prefer to take the DIC prism out of the optical path when doing fluorescence imaging, but there are situations where I need DIC + fluorescence simultaneously. My main question was not why I get degradation of resolution in the fluorescence channel (that I expected), but rather why is the degradation worse for a lower-resolution objective (20x/0.85 oil imm versus 100x/1.4 oil imm). The "image doubling" due to the shear of the objective DIC prism should be more noticeable with the high-resolution lens. Stan Vitha On Thu, 7 Apr 2011 22:42:00 +0200, P. Johannes Helm <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear Stanislav. > > >You might have a look at > >http://micro.magnet.fsu.edu/primer/techniques/dic/dichome.html > >Depending on what your concept of "theoretical" is, this might be the >source of information you are looking for - or not. > >The wording "behind" the objective in your text is a little bit un-clear. >What you mean, I suppose, is the following: > >In a classical transmitted light configuration, you will have the light >source, a collector lens, the field diaphragm, then the DIC polarizer, the >condenser DIC prism - also known as the "main DIC prism" -, the condenser >with its front lens, the condenser immersion medium (air, water, oil, or >glycerine), the microscope slide, the specimen in its mounting medium, >the cover slip, the immersion medium for the objective - e.g. air, water, >oil, glycerine -, the objective, the objective DIC prism - also known as >"auxiliary DIC prism" -, the DIC analyzer and finally the observation or >imaging optics. > >What you describe as the "DIC prism behind the objective" most probably is >the auxiliary DIC prism, i.e. the DIC prism on the image side of the >objective. While a number of microscope producers have this auxiliary DIC >prism mounted so that you can adjust it for optimizing your contrast image >("DIC slider"), in the Olympus systems an adjustment is done in a slightly >different way; Olympus provides a lambda/4 plate as an additional element >between DIC polarizer and condenser DIC prism. The user then adjusts the >orientation of the DIC polarizer in order to change the contrast settings. >Also, optimal DIC observation sometimes may request a rotatable specimen >stage centered in the optical axis of the microscope. > >The name DIC, "D"IC, "D"ifferential Interference Contrast, delivers the >explanation to your question. The DIC prism, also known as "Nomarski >prism", named after its inventor Jerzy Nomarski, provides a "d"ifferential >splitting of the ray paths of the light in the microscope. If you do >confocal laser scanning microscopy and have a well polarized laser beam - >normally, a laser beam generated by a "typical CLSM laser" will be >polarized per se, but it might nevertheless be sent through the DIC >analyzer (which is nothing else but a polarizer) -, then that beam will be >split into two beams, which are laterally displaced "differentially". The >lateral separation will hence be small, smaller than the lateral >resolution limit of the objective when used in the wide field observation >mode. Other types of interference contrast microscopes - e.g. Jamin >Lebedeff - will generate two beams with a macroscopic separation; then one >beam, the reference beam, will bypass the entire microscope on a ray path >outside the microscope frame, but this is not of any interest here. >What hence happens when you try to do CLSM with a fluorescent specimen - >fluorescent light assumed to be un-polarized - and have a DIC prism in the >ray path is that you will experience what are a kind of super imposed >"double images" if you have a sufficiently large zooming factor so that >you approach the resolution limit with your pixelization. The story is a >little bit more complicated in case of reflected light CLS microscopy >because of possible polarization conservation effects, but let's keep this >out of the discussion here for simplicity. > >So, in order to avoid these effects, take out the secondary DIC prism from >the ray path when CLSMing. > >NOW, there are some microscope objective holders, e.g. on a BX51WI used >cell physiologists, e.g. Olympus item code WI-SRE2, in which a secondary >DIC prism, e.g. WI-DICTHRA2, is inserted so that you cannot easilry remove >it from the ray path. There does, however, exist a version of WI-SRE2, >named "FV10-SRE", if I recall correctly, where you have a slider, which >will permit you to easily insert or extract the DIC prism from the ray >path. > >Unfortunately, this FV10-SRE seemingly is not sold a million times per >month or so, hence it is not really cheap. But it might help. > >Best wishes, > >Johannes > > > >-- >P. Johannes Helm, M.Sc. PhD >Seniorengineer >CMBN >University of Oslo >Institute of Basic Medical Science >Department of Anatomy >Postboks 1105 - Blindern >NO-0317 Oslo > >Voice: +47 228 51159 >Fax: +47 228 51499 > >WWW: folk.uio.no/jhelm |
|
Johannes Helm |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, again, Stanislav, the "double" image, I think, might be prominent in any type of lens as soon as you approach the resolution limit with your pixelization, in other words, as soon as you "nyquist the image". Of course, the construction of the prism will also have a considerable influence on this matter. Best, Johannes > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Johannes, > Thank you for your detailed explanation. I should clarify that our > confocal > system has an inverted IX-81 microscope. > By "behind the objective" I meant on the opposite side of the objective > than > the specimen is (the convention, I believe, is that when taking pictures, > the > object is in front of the lens, and photographer is behind the lens. The > image > is formed behind the lens, which is also the side where we find the back > focal > plane). > > Since I am collecting confocal epifluorescnce image simultaneously with > recording a DIC image on the transmitted detector, the degradation of > fluorescence images relates to the objective DIC prism. The condenser DIC > prism is not in the optical path of the epifluorescence imaging setup. > > I prefer to take the DIC prism out of the optical path when doing > fluorescence > imaging, but there are situations where I need DIC + fluorescence > simultaneously. > > My main question was not why I get degradation of resolution in the > fluorescence channel (that I expected), but rather why is the degradation > worse for a lower-resolution objective (20x/0.85 oil imm versus 100x/1.4 > oil > imm). The "image doubling" due to the shear of the objective DIC prism > should > be more noticeable with the high-resolution lens. > > Stan Vitha > > On Thu, 7 Apr 2011 22:42:00 +0200, P. Johannes Helm > <[hidden email]> wrote: > >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>***** >> >>Dear Stanislav. >> >> >>You might have a look at >> >>http://micro.magnet.fsu.edu/primer/techniques/dic/dichome.html >> >>Depending on what your concept of "theoretical" is, this might be the >>source of information you are looking for - or not. >> >>The wording "behind" the objective in your text is a little bit un-clear. >>What you mean, I suppose, is the following: >> >>In a classical transmitted light configuration, you will have the light >>source, a collector lens, the field diaphragm, then the DIC polarizer, >> the >>condenser DIC prism - also known as the "main DIC prism" -, the condenser >>with its front lens, the condenser immersion medium (air, water, oil, or >>glycerine), the microscope slide, the specimen in its mounting medium, >>the cover slip, the immersion medium for the objective - e.g. air, water, >>oil, glycerine -, the objective, the objective DIC prism - also known as >>"auxiliary DIC prism" -, the DIC analyzer and finally the observation or >>imaging optics. >> >>What you describe as the "DIC prism behind the objective" most probably >> is >>the auxiliary DIC prism, i.e. the DIC prism on the image side of the >>objective. While a number of microscope producers have this auxiliary DIC >>prism mounted so that you can adjust it for optimizing your contrast >> image >>("DIC slider"), in the Olympus systems an adjustment is done in a >> slightly >>different way; Olympus provides a lambda/4 plate as an additional element >>between DIC polarizer and condenser DIC prism. The user then adjusts the >>orientation of the DIC polarizer in order to change the contrast >> settings. >>Also, optimal DIC observation sometimes may request a rotatable specimen >>stage centered in the optical axis of the microscope. >> >>The name DIC, "D"IC, "D"ifferential Interference Contrast, delivers the >>explanation to your question. The DIC prism, also known as "Nomarski >>prism", named after its inventor Jerzy Nomarski, provides a >> "d"ifferential >>splitting of the ray paths of the light in the microscope. If you do >>confocal laser scanning microscopy and have a well polarized laser beam - >>normally, a laser beam generated by a "typical CLSM laser" will be >>polarized per se, but it might nevertheless be sent through the DIC >>analyzer (which is nothing else but a polarizer) -, then that beam will >> be >>split into two beams, which are laterally displaced "differentially". The >>lateral separation will hence be small, smaller than the lateral >>resolution limit of the objective when used in the wide field observation >>mode. Other types of interference contrast microscopes - e.g. Jamin >>Lebedeff - will generate two beams with a macroscopic separation; then >> one >>beam, the reference beam, will bypass the entire microscope on a ray path >>outside the microscope frame, but this is not of any interest here. >>What hence happens when you try to do CLSM with a fluorescent specimen - >>fluorescent light assumed to be un-polarized - and have a DIC prism in >> the >>ray path is that you will experience what are a kind of super imposed >>"double images" if you have a sufficiently large zooming factor so that >>you approach the resolution limit with your pixelization. The story is a >>little bit more complicated in case of reflected light CLS microscopy >>because of possible polarization conservation effects, but let's keep >> this >>out of the discussion here for simplicity. >> >>So, in order to avoid these effects, take out the secondary DIC prism >> from >>the ray path when CLSMing. >> >>NOW, there are some microscope objective holders, e.g. on a BX51WI used > by >>cell physiologists, e.g. Olympus item code WI-SRE2, in which a secondary >>DIC prism, e.g. WI-DICTHRA2, is inserted so that you cannot easilry >> remove >>it from the ray path. There does, however, exist a version of WI-SRE2, >>named "FV10-SRE", if I recall correctly, where you have a slider, which >>will permit you to easily insert or extract the DIC prism from the ray >>path. >> >>Unfortunately, this FV10-SRE seemingly is not sold a million times per >>month or so, hence it is not really cheap. But it might help. >> >>Best wishes, >> >>Johannes >> >> >> >>-- >>P. Johannes Helm, M.Sc. PhD >>Seniorengineer >>CMBN >>University of Oslo >>Institute of Basic Medical Science >>Department of Anatomy >>Postboks 1105 - Blindern >>NO-0317 Oslo >> >>Voice: +47 228 51159 >>Fax: +47 228 51499 >> >>WWW: folk.uio.no/jhelm > -- P. Johannes Helm Voice: (+47) 228 51159 (office) Fax: (+47) 228 51499 (office) |
|
|
In reply to this post by Stanislav Vitha
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** This is just guessing, but the 20x NA 0.85 is very low mag for its NA. This may mean that the light rays are going through a greater width of the Wollaston prism. Guy Sponsor my next half-marathon on May 15th There's a special reason - find it out at http://www.everydayhero.com.au/Guy_Cox_4846 ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Stanislav Vitha Sent: Friday, 8 April 2011 5:28 PM To: [hidden email] Subject: Re: DIC prism, fluorescence and focus shift ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Johannes, Thank you for your detailed explanation. I should clarify that our confocal system has an inverted IX-81 microscope. By "behind the objective" I meant on the opposite side of the objective than the specimen is (the convention, I believe, is that when taking pictures, the object is in front of the lens, and photographer is behind the lens. The image is formed behind the lens, which is also the side where we find the back focal plane). Since I am collecting confocal epifluorescnce image simultaneously with recording a DIC image on the transmitted detector, the degradation of fluorescence images relates to the objective DIC prism. The condenser DIC prism is not in the optical path of the epifluorescence imaging setup. I prefer to take the DIC prism out of the optical path when doing fluorescence imaging, but there are situations where I need DIC + fluorescence simultaneously. My main question was not why I get degradation of resolution in the fluorescence channel (that I expected), but rather why is the degradation worse for a lower-resolution objective (20x/0.85 oil imm versus 100x/1.4 oil imm). The "image doubling" due to the shear of the objective DIC prism should be more noticeable with the high-resolution lens. Stan Vitha On Thu, 7 Apr 2011 22:42:00 +0200, P. Johannes Helm <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear Stanislav. > > >You might have a look at > >http://micro.magnet.fsu.edu/primer/techniques/dic/dichome.html > >Depending on what your concept of "theoretical" is, this might be the >source of information you are looking for - or not. > >The wording "behind" the objective in your text is a little bit >What you mean, I suppose, is the following: > >In a classical transmitted light configuration, you will have the light >source, a collector lens, the field diaphragm, then the DIC polarizer, the >condenser DIC prism - also known as the "main DIC prism" -, the condenser >with its front lens, the condenser immersion medium (air, water, oil, or >glycerine), the microscope slide, the specimen in its mounting medium, >the cover slip, the immersion medium for the objective - e.g. air, water, >oil, glycerine -, the objective, the objective DIC prism - also known as >"auxiliary DIC prism" -, the DIC analyzer and finally the observation or >imaging optics. > >What you describe as the "DIC prism behind the objective" most probably is >the auxiliary DIC prism, i.e. the DIC prism on the image side of the >objective. While a number of microscope producers have this auxiliary DIC >prism mounted so that you can adjust it for optimizing your contrast image >("DIC slider"), in the Olympus systems an adjustment is done in a slightly >different way; Olympus provides a lambda/4 plate as an additional element >between DIC polarizer and condenser DIC prism. The user then adjusts the >orientation of the DIC polarizer in order to change the contrast settings. >Also, optimal DIC observation sometimes may request a rotatable specimen >stage centered in the optical axis of the microscope. > >The name DIC, "D"IC, "D"ifferential Interference Contrast, delivers the >explanation to your question. The DIC prism, also known as "Nomarski >prism", named after its inventor Jerzy Nomarski, provides a "d"ifferential >splitting of the ray paths of the light in the microscope. If you do >confocal laser scanning microscopy and have a well polarized laser beam - >normally, a laser beam generated by a "typical CLSM laser" will be >polarized per se, but it might nevertheless be sent through the DIC >analyzer (which is nothing else but a polarizer) -, then that beam will be >split into two beams, which are laterally displaced "differentially". The >lateral separation will hence be small, smaller than the lateral >resolution limit of the objective when used in the wide field observation >mode. Other types of interference contrast microscopes - e.g. Jamin >Lebedeff - will generate two beams with a macroscopic separation; then one >beam, the reference beam, will bypass the entire microscope on a ray path >outside the microscope frame, but this is not of any interest here. >What hence happens when you try to do CLSM with a fluorescent specimen - >fluorescent light assumed to be un-polarized - and have a DIC prism in the >ray path is that you will experience what are a kind of super imposed >"double images" if you have a sufficiently large zooming factor so that >you approach the resolution limit with your pixelization. The story is a >little bit more complicated in case of reflected light CLS microscopy >because of possible polarization conservation effects, but let's keep this >out of the discussion here for simplicity. > >So, in order to avoid these effects, take out the secondary DIC prism from >the ray path when CLSMing. > >NOW, there are some microscope objective holders, e.g. on a BX51WI used by >cell physiologists, e.g. Olympus item code WI-SRE2, in which a secondary >DIC prism, e.g. WI-DICTHRA2, is inserted so that you cannot easilry remove >it from the ray path. There does, however, exist a version of WI-SRE2, >named "FV10-SRE", if I recall correctly, where you have a slider, which >will permit you to easily insert or extract the DIC prism from the ray >path. > >Unfortunately, this FV10-SRE seemingly is not sold a million times per >month or so, hence it is not really cheap. But it might help. > >Best wishes, > >Johannes > > > >-- >P. Johannes Helm, M.Sc. PhD >Seniorengineer >CMBN >University of Oslo >Institute of Basic Medical Science >Department of Anatomy >Postboks 1105 - Blindern >NO-0317 Oslo > >Voice: +47 228 51159 >Fax: +47 228 51499 > >WWW: folk.uio.no/jhelm |
|
Shalin Mehta |
|
|
In reply to this post by Stanislav Vitha
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Stan, On Fri, Apr 8, 2011 at 3:28 PM, Stanislav Vitha <[hidden email]> wrote: > > My main question was not why I get degradation of resolution in the > fluorescence channel (that I expected), but rather why is the degradation > worse for a lower-resolution objective (20x/0.85 oil imm versus 100x/1.4 oil > imm). The "image doubling" due to the shear of the objective DIC prism should > be more noticeable with the high-resolution lens I happened to have gone through this exercise when measuring the shear in our DIC setup. We tried to recap what we understood in the appendix of our paper (http://www.mshalin.com/blog/wp-content/uploads/2010/09/mehtaao2010-samplefreecalibrationdic.pdf). Here is an attempt at understanding the specific case that you described: Nomarski prism is an angular polarizing beam-splitting devices. i.e., it splits the incoming light into two polarizations that travel at different angles. The angular split caused by the Nomarski prism (placed in the back focal plane) amounts to spatial shear in the specimen plane. The relationship between angular shear (a) and spatial shear (s) is rather simple: s=tan(a)*f. Where f is the focal length of the objective. This fact can be seen easily from the geometry of fig. 11(d) in above paper. Now the image doubling is caused when the shear is large with respect to the resolution. To judge the doubling, we should normalize both sides of the above equation by lambda/NA. Then, we have s/(lambda/NA) = tan(a) * f/(lambda/NA). Therefore, when s/(lambda/NA) or s*NA is large (for given wavelength) we will have image doubling. In the Olympus setup there is only one DIC prism on objective side, i.e., the angular shear is the same for all objectives. Therefore, the objectives for which f*NA is large will give rise to larger s*NA and hence image doubling. The product of focal length and NA is large for 20X 0.85 than for 100X1.4. I compute the product of focal length and NA below. Objective's magnification (M) = tube length (ft)/ objective's focal length (f). Therefore, f=ft/M. Therefore, f*NA = ft*(NA/M). For, 20X 0.85: f*NA= 0.0425*ft. For, 100x 1.4: f*NA= 0.014*ft. So the 'image doubling' for 20X 0.85 is almost 3 times more than for 100x 1.4. Now, it is interesting to see how this relates with Guy's intuitive observation. NA of the objective = radius of the objective BFP (d)/ objective focal length (f). So, r=f*NA. As we have seen above, the size of the BFP (ie., f*NA) for 20X 0.85 is 3 times larger than for 100x1.4. So as Guy says, rays collected by 20X 0.85 pass through larger (3 times larger) surface of the Wollaston prism. Best Shalin http://mshalin.com Research Associate Bioimaging Lab, Block-E3A, #7-10 Div of Bioengineering, NUS Singapore 117574 (M) +65-84330724 |
|
Stanislav Vitha |
|
|
In reply to this post by Stanislav Vitha
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Shalin (and Guy), thank you so much for the explanation and the article, it is exactly what I was looking for. It now makes sense. I have to disclose that there may be one additional factor that contributes to image degradation with the 20x objective (and has something to do with the large diameter of the BFP) - I looked at the DIC prism and discovered that there is an oil droplet close to he edge of the prism. Generous application of immersion oil by our users must have caused the immersion oil to run down, seep through the thread and drip on the prism. So I think the oil droplet would degrade image more severely for objectives with large BFP. Now off to do some cleaning. Stan Dr. Stanislav Vitha Microscopy and Imaging Center Texas A&M University BSBW 119 College Station, TX 77843-2257 http://microscopy.tamu.edu On Sun, 10 Apr 2011 22:31:06 +0800, Shalin Mehta <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello Stan, > >On Fri, Apr 8, 2011 at 3:28 PM, Stanislav Vitha <[hidden email]> wrote: >> >> My main question was not why I get degradation of resolution in the >> fluorescence channel (that I expected), but rather why is the degradation >> worse for a lower-resolution objective (20x/0.85 oil imm versus 100x/1.4 oil >> imm). The "image doubling" due to the shear of the objective DIC prism should >> be more noticeable with the high-resolution lens > >I happened to have gone through this exercise when measuring the shear >in our DIC setup. >We tried to recap what we understood in the appendix of our paper >(http://www.mshalin.com/blog/wp- content/uploads/2010/09/mehtaao2010-samplefreecalibrationdic.pdf). > >Here is an attempt at understanding the specific case that you described: > >Nomarski prism is an angular polarizing beam-splitting devices. i.e., >it splits the incoming light into two polarizations that travel at >different angles. The angular split caused by the Nomarski prism >(placed in the back focal plane) amounts to spatial shear in the >specimen plane. The relationship between angular shear (a) and spatial >shear (s) is rather simple: s=tan(a)*f. Where f is the focal length of >the objective. This fact can be seen easily from the geometry of fig. >11(d) in above paper. Now the image doubling is caused when the shear >is large with respect to the resolution. To judge the doubling, we >should normalize both sides of the above equation by lambda/NA. Then, >we have s/(lambda/NA) = tan(a) * f/(lambda/NA). Therefore, when >s/(lambda/NA) or s*NA is large (for given wavelength) we will have >image doubling. > >In the Olympus setup there is only one DIC prism on objective side, >i.e., the angular shear is the same for all objectives. Therefore, the >objectives for which f*NA is large will give rise to larger s*NA and >hence image doubling. > >The product of focal length and NA is large for 20X 0.85 than for >100X1.4. I compute the product of focal length and NA below. > >Objective's magnification (M) = tube length (ft)/ objective's focal >Therefore, f=ft/M. >Therefore, f*NA = ft*(NA/M). >For, 20X 0.85: f*NA= 0.0425*ft. >For, 100x 1.4: f*NA= 0.014*ft. > >So the 'image doubling' for 20X 0.85 is almost 3 times more than for 100x 1.4. > >Now, it is interesting to see how this relates with Guy's intuitive observation. > >NA of the objective = radius of the objective BFP (d)/ objective focal >length (f). >So, r=f*NA. As we have seen above, the size of the BFP (ie., f*NA) >for 20X 0.85 is 3 times larger than for 100x1.4. So as Guy says, rays >collected by 20X 0.85 pass through larger (3 times larger) surface of >the Wollaston prism. > >Best >Shalin > >http://mshalin.com > >Research Associate >Bioimaging Lab, Block-E3A, #7-10 >Div of Bioengineering, NUS Singapore 117574 >(M) +65-84330724 |
| Free forum by Nabble | Edit this page |