|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Imagers, I am trying to image DY650 (1P) and Hoechst with MP (710nm), surprisingly the MP seems to be bleaching the DY650 even in sequential scans. Anyone else experiencing this? I don't see any 2P spectra for the DY series? Thanks Richard Cole Research Scientist V Director: Advanced Light Microscopy & Image Analysis Core Wadsworth Center Research Assistant Professor Dept. of Biomedical Sciences School of Public Health State University of New York 120 New Scotland Avenue, Albany N.Y. 12208 518-474-7048 Phone 518-408-1730 Fax |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Richard, I've seen this with some far red dyes, such as mitotrackers. It's most likely due to single-photon absorption of your MP laser wavelength. When you look at single-photon absorption curve of DY-650, it's nearly zero at 700-710 nm, but not exactly zero. Together with the fact that the MP laser light is typically three orders of magnitude stronger than a corresponding 1P laser (e.g. 633 nm), it can cause rapid photobleaching. Obvious solution is to take the DY-650 image (or stack) first and then the MP image. You should also try longer MP wavelength (e.g. 760 nm), that will still excite Hoechst, but may not bleach the DY-650 so quickly... Good luck! Best, zdenek -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth email: [hidden email] ---------- Původní e-mail ---------- Od: Cole, Richard W (HEALTH) <[hidden email]> Komu: [hidden email] Datum: 12. 2. 2018 12:08:31 Předmět: DY650 & multiphoton "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Imagers, I am trying to image DY650 (1P) and Hoechst with MP (710nm), surprisingly the MP seems to be bleaching the DY650 even in sequential scans. Anyone else experiencing this? I don't see any 2P spectra for the DY series? Thanks Richard Cole Research Scientist V Director: Advanced Light Microscopy & Image Analysis Core Wadsworth Center Research Assistant Professor Dept. of Biomedical Sciences School of Public Health State University of New York 120 New Scotland Avenue, Albany N.Y. 12208 518-474-7048 Phone 518-408-1730 Fax " |
| Free forum by Nabble | Edit this page |