Deconvolution of Transmission Images

classic Classic list List threaded Threaded
♦
♦
4 messages Options
Richard Cole Richard Cole
Reply | Threaded
Open this post in threaded view
♦
♦
|

Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You are correct, transmitted images are formed completely differently than
fluorescent images, therefore the algorithms are separate.  Having said
that, we use Autoquant for both types of images.  For DIC and Phase image we
have also used a simple summed projection for qualitative purposes.

 

Rich

 

 

Richard Cole
Research Scientist V
Director: Advanced Light Microscopy Core Unit
Wadsworth Center

 

Research Assistant Professor
Dept. of Biomedical Sciences
School of Public Health State University of New York


P.O. Box 509 Albany N.Y. 12201-0509
518-474-7048 Phone
518-474-4430 Fax

 

Email  <mailto:[hidden email]> [hidden email]

Website  <http://www.wadsworth.org/cores/alm/index.htm>
www.wadsworth.org/cores/alm/index.htm

 

 



IMPORTANT NOTICE: This e-mail and any attachments may contain
confidential or sensitive information which is, or may be, legally
privileged or otherwise protected by law from further disclosure.  It
is intended only for the addressee.  If you received this in error or
from someone who was not authorized to send it to you, please do not
distribute, copy or use it or any attachments.  Please notify the
sender immediately by reply e-mail and delete this from your
system. Thank you for your cooperation.
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You cannot really deconvolve transmission images, because the image of a point is not independent of its neighbours, so each point is not in fact convolved with a psf.  You can however isolate in-focus information from out of focus information and there are several algorithms for this.  The basic idea is than in-focus detail will change a lot if we blur it whereas out of focus information won't change much.  So you can extract 3D information, or construct an extended-focus image, but it isn't deconvolution.  

                                                              Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Richard Cole
Sent: Friday, 23 March 2012 1:32 AM
To: [hidden email]
Subject: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You are correct, transmitted images are formed completely differently than fluorescent images, therefore the algorithms are separate.  Having said that, we use Autoquant for both types of images.  For DIC and Phase image we have also used a simple summed projection for qualitative purposes.

 

Rich

 

 

Richard Cole
Research Scientist V
Director: Advanced Light Microscopy Core Unit Wadsworth Center

 

Research Assistant Professor
Dept. of Biomedical Sciences
School of Public Health State University of New York


P.O. Box 509 Albany N.Y. 12201-0509
518-474-7048 Phone
518-474-4430 Fax

 

Email  <mailto:[hidden email]> [hidden email]

Website  <http://www.wadsworth.org/cores/alm/index.htm>
www.wadsworth.org/cores/alm/index.htm

 

 



IMPORTANT NOTICE: This e-mail and any attachments may contain confidential or sensitive information which is, or may be, legally privileged or otherwise protected by law from further disclosure.  It is intended only for the addressee.  If you received this in error or from someone who was not authorized to send it to you, please do not distribute, copy or use it or any attachments.  Please notify the sender immediately by reply e-mail and delete this from your system. Thank you for your cooperation.
Jim Mansfield Jim Mansfield
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Guy,

Maybe I'm just being overly fussy, or maybe I am not understanding some part
of using deconvolution to resolve focal planes, but of course one can do
real and actual deconvolution on brightfield images. The DeltaVision
products all do widefield (i.e., non-confocal) deconvolution (albeit in FL)
and that is very similar to brightfield deconvolution. Besides, confocal
images are not completely independent; real point-spread functions do go to
infinity on both sides of the focal plane (follow the direct light path
through the pinhole from source to detector).

Generating the PSF for brightfield deconvolution is much more difficult than
for confocals, and so it is considerably more problematic than for a
confocal, and there can be serious errors generated that way, but of course
one can, in fact, really deconvolve brightfield images!

Jim


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Guy Cox
Sent: Thursday, March 22, 2012 2:58 PM
To: [hidden email]
Subject: Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You cannot really deconvolve transmission images, because the image of a
point is not independent of its neighbours, so each point is not in fact
convolved with a psf.  You can however isolate in-focus information from out
of focus information and there are several algorithms for this.  The basic
idea is than in-focus detail will change a lot if we blur it whereas out of
focus information won't change much.  So you can extract 3D information, or
construct an extended-focus image, but it isn't deconvolution.  

                                                              Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Richard Cole
Sent: Friday, 23 March 2012 1:32 AM
To: [hidden email]
Subject: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You are correct, transmitted images are formed completely differently than
fluorescent images, therefore the algorithms are separate.  Having said
that, we use Autoquant for both types of images.  For DIC and Phase image we
have also used a simple summed projection for qualitative purposes.

 

Rich

 

 

Richard Cole
Research Scientist V
Director: Advanced Light Microscopy Core Unit Wadsworth Center

 

Research Assistant Professor
Dept. of Biomedical Sciences
School of Public Health State University of New York


P.O. Box 509 Albany N.Y. 12201-0509
518-474-7048 Phone
518-474-4430 Fax

 

Email  <mailto:[hidden email]> [hidden email]

Website  <http://www.wadsworth.org/cores/alm/index.htm>
www.wadsworth.org/cores/alm/index.htm

 

 



IMPORTANT NOTICE: This e-mail and any attachments may contain confidential
or sensitive information which is, or may be, legally privileged or
otherwise protected by law from further disclosure.  It is intended only for
the addressee.  If you received this in error or from someone who was not
authorized to send it to you, please do not distribute, copy or use it or
any attachments.  Please notify the sender immediately by reply e-mail and
delete this from your system. Thank you for your cooperation.
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

One can only regard an image as the convolution of an object with a PSF if each point is imaged independently.  Abbe showed that this is not so in regular widefield imaging.  It IS so in fluorescence (Rayleigh imaging).  In DIC it is even more obviously not so, since we are combining two beams, one through the point and one through its close neighbourhood, to form the image.  As I said in my previous post, that doesn't mean we can't use algorithmic techniques to isolate in-focus information, but this is not deconvolution, it's image processing.  In particular, it cannot improve resolution which, in principle, deconvolution can - Superresolution Three-Dimensional Images of Fluorescence in Cells with Minimal Light Exposure: Walter A. Carrington, Ronald M. Lynch, Edwin D. W. Moore, Gerrit Isenberg, Kevin E. Fogarty, Fredric S. Fay.   Science, 268, (Jun. 9, 1995), pp. 1483-1487.  I accept that most commercial software does not seek to give super-resolution, but we still need to distinguish what is deconvolution and what is not.  

                                                                                                                                 Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jim Mansfield
Sent: Friday, 23 March 2012 7:08 AM
To: [hidden email]
Subject: Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Guy,

Maybe I'm just being overly fussy, or maybe I am not understanding some part of using deconvolution to resolve focal planes, but of course one can do real and actual deconvolution on brightfield images. The DeltaVision products all do widefield (i.e., non-confocal) deconvolution (albeit in FL) and that is very similar to brightfield deconvolution. Besides, confocal images are not completely independent; real point-spread functions do go to infinity on both sides of the focal plane (follow the direct light path through the pinhole from source to detector).

Generating the PSF for brightfield deconvolution is much more difficult than for confocals, and so it is considerably more problematic than for a confocal, and there can be serious errors generated that way, but of course one can, in fact, really deconvolve brightfield images!

Jim


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: Thursday, March 22, 2012 2:58 PM
To: [hidden email]
Subject: Re: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You cannot really deconvolve transmission images, because the image of a point is not independent of its neighbours, so each point is not in fact convolved with a psf.  You can however isolate in-focus information from out of focus information and there are several algorithms for this.  The basic idea is than in-focus detail will change a lot if we blur it whereas out of focus information won't change much.  So you can extract 3D information, or construct an extended-focus image, but it isn't deconvolution.  

                                                              Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Richard Cole
Sent: Friday, 23 March 2012 1:32 AM
To: [hidden email]
Subject: Deconvolution of Transmission Images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You are correct, transmitted images are formed completely differently than fluorescent images, therefore the algorithms are separate.  Having said that, we use Autoquant for both types of images.  For DIC and Phase image we have also used a simple summed projection for qualitative purposes.

 

Rich

 

 

Richard Cole
Research Scientist V
Director: Advanced Light Microscopy Core Unit Wadsworth Center

 

Research Assistant Professor
Dept. of Biomedical Sciences
School of Public Health State University of New York


P.O. Box 509 Albany N.Y. 12201-0509
518-474-7048 Phone
518-474-4430 Fax

 

Email  <mailto:[hidden email]> [hidden email]

Website  <http://www.wadsworth.org/cores/alm/index.htm>
www.wadsworth.org/cores/alm/index.htm

 

 



IMPORTANT NOTICE: This e-mail and any attachments may contain confidential or sensitive information which is, or may be, legally privileged or otherwise protected by law from further disclosure.  It is intended only for the addressee.  If you received this in error or from someone who was not authorized to send it to you, please do not distribute, copy or use it or any attachments.  Please notify the sender immediately by reply e-mail and delete this from your system. Thank you for your cooperation.