lechristophe |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi,
I have some small FluoSpheres (20 nm far-red, carboxylate-modified) that I would like to deposit on polylysine-coated glass coverslips (or non-coated ones if better) in order to calibrate my STORM microscope (similar to PSF measurement, but denser to calibrate Z astigmatism). I usually do it with 100 nm tetraspeck beads, which readily adhere to PLL-coated coverslips (I use 1/50-1/500 depending on the density I want to obtain). With the FluoSphere I don't get any specific signal. Should I use them less diluted (they are "2% solid"), or is there an adhesion problem (my understanding is that carboxylate-modified are negatively charged and won't adhere to PLL)? Do you have a specific protocol for these FluoSpheres? Thanks, Christophe |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Christophe,
carboxylated beads typically bind very well to just glass if in PBS. However if used in water they do not stick. Best, Jens
|
In reply to this post by lechristophe
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Would it be possible to simply dry them on a coverslip? You don’t need liquid above the coverslip to image what is on its surface Mike Model From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Christophe Leterrier ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Hi, I have some small FluoSpheres (20 nm far-red, carboxylate-modified) that I would like to deposit on polylysine-coated glass coverslips (or non-coated ones if better) in order to calibrate my STORM microscope (similar to PSF measurement,
but denser to calibrate Z astigmatism). I usually do it with 100 nm tetraspeck beads, which readily adhere to PLL-coated coverslips (I use 1/50-1/500 depending on the density I want to obtain). With the FluoSphere I don't get any specific signal. Should I use them less diluted
(they are "2% solid"), or is there an adhesion problem (my understanding is that carboxylate-modified are negatively charged and won't adhere to PLL)? Do you have a specific protocol for these FluoSpheres? Thanks, Christophe |
Ewers, Helge |
In reply to this post by lechristophe
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Christophe,
the problem with the very small spheres is that they are extremely strongly charged. For example they are useless for SPT as well. We never got anything to work with them. I would suggest using bigger ones, or maybe the new ones from Gattaquant.
I have ordered some, but not received them yet, I’ll let you know if they work well.
For the far red region, also gold rods work well. And they are in the 20 nm range.
Cheers,
Helge
--
Dr. Helge Ewers Senior Lecturer King's College London Randall Division of Cell and Molecular Biophysics 2nd Floor, Hodgkin Building Guy's Campus London SE1 1UL United Kingdom [hidden email] +44 20 7848 6441 tel
|
In reply to this post by lechristophe
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Christophe I’m playing catch up here, however I’ve spent quite a lot of time on this very question. I discovered that drying the spheres onto
the coverslip (which as Michael suggests works well) has an issue for 3D. The spheres are literally on the coverslip but for good 3D calibration you would want the spheres slightly above the coverslip. I’ve tried a variety of ways of doing this with success. 1.
Dilute the spheres in low mounting point agarose which is transparent and mount these on the coverslip. 2.
Mix the spheres in gelatin. I used fluorescent gelatin (Oregon green 488) which was useful for finding focus. 3.
Dilute the spheres in PDMS and cure the PDMS. This test sample will then likely last a really long time. I tried this sometime
ago and it worked and I want to go back to this procedure now as we’ll have a sample we can return to time and again for alignment 4.
If all else fails Gattaquant Nanoruler are also good for this job. Another trick which you may find useful – we certainly did for setting up 2 colour STORM – is mixing the tetraspeck beads with the
much smaller beads (maybe a 1:5 ratio?). Its very difficult to find 20nm beads by eye and its useful to find a good place in the bead lawn for setting up the system with. Hope this helps
Dr Ann Wheeler E: [hidden email] From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Christophe Leterrier ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Hi, I have some small FluoSpheres (20 nm far-red, carboxylate-modified) that I would like to deposit on polylysine-coated glass coverslips (or non-coated ones if better) in order to calibrate my STORM microscope (similar to PSF measurement,
but denser to calibrate Z astigmatism). I usually do it with 100 nm tetraspeck beads, which readily adhere to PLL-coated coverslips (I use 1/50-1/500 depending on the density I want to obtain). With the FluoSphere I don't get any specific signal. Should I use them less diluted
(they are "2% solid"), or is there an adhesion problem (my understanding is that carboxylate-modified are negatively charged and won't adhere to PLL)? Do you have a specific protocol for these FluoSpheres? Thanks, Christophe ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336.
Ann Wheeler
Head of Advanced Imaging Facility Institute of Genetics and Molecular Medicine University of Edinburgh United Kingdom |
Free forum by Nabble | Edit this page |