Adam Glaser |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everyone, I was wondering if anyone had experience or knew if water dipping objectives (specifically the Zeiss W Plan-Apochromat) were designed for immersion in DI water or PBS. Or if they have been designed to perform well in both refractive indices? Thanks! Adam |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm working with this lens on in-vivo. We use both and I don't see much difference. But as I also work with a temp controlled chamber so most of the time I use ultrasound gel. |
![]() ![]() |
Sylvie Le Guyader |
![]() |
In reply to this post by Adam Glaser
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Adam Do you find a difference in refraction index between PBS and water? I think they are both 1.33. If they have the same refraction index then the objective should perform identically with both. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Blickagången 16, Room 7362 (lab)/7840 (office) 14157 Huddinge, Sweden mobile: +46 (0) 73 733 5008 LCI website LCI microscopy blog -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Adam Glaser Sent: Friday, 7 February, 2020 06:42 To: [hidden email] Subject: Dipping objectives - water or PBS? ***** To join, leave or search the confocal microscopy listserv, go to: https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=Us8B9MW12KnO0ZkHigDnV7dRMhaBFioHskNRbV2z6yg%3D&reserved=0 Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=ow3oJfNxg%2FZ19nC57cb0RBwfQXdTgDRmEN7bxbNDgpA%3D&reserved=0 and include the link in your posting. ***** Hi everyone, I was wondering if anyone had experience or knew if water dipping objectives (specifically the Zeiss W Plan-Apochromat) were designed for immersion in DI water or PBS. Or if they have been designed to perform well in both refractive indices? Thanks! Adam När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>. Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>. |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The salt doesn't make that much of a difference. One detail is to remind users to rinse the lens off with pure water after using PBS. I made a 'house call' to a lab that was having problems using one of their patch rigs. After checking all their filters and prisms I happened to look down at the objective and found a perfect window of dried salt sitting over the face of the objective. One lengthy freshwater soak later and everything was fine... Craig On Fri, Feb 7, 2020 at 2:01 AM Sylvie Le Guyader <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Adam > > Do you find a difference in refraction index between PBS and water? I > think they are both 1.33. > If they have the same refraction index then the objective should perform > identically with both. > > Med vänlig hälsning / Best regards > Sylvie > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Blickagången 16, > Room 7362 (lab)/7840 (office) > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI website > LCI microscopy blog > > -----Original Message----- > From: Confocal Microscopy List <[hidden email]> On > Behalf Of Adam Glaser > Sent: Friday, 7 February, 2020 06:42 > To: [hidden email] > Subject: Dipping objectives - water or PBS? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=Us8B9MW12KnO0ZkHigDnV7dRMhaBFioHskNRbV2z6yg%3D&reserved=0 > Post images on > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=ow3oJfNxg%2FZ19nC57cb0RBwfQXdTgDRmEN7bxbNDgpA%3D&reserved=0 > and include the link in your posting. > ***** > > Hi everyone, > I was wondering if anyone had experience or knew if water dipping > objectives (specifically the Zeiss W Plan-Apochromat) were designed for > immersion in DI water or PBS. Or if they have been designed to perform well > in both refractive indices? > Thanks! > Adam > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI > kommer att behandla dina personuppgifter. Här finns information om hur KI > behandlar personuppgifter< > https://ki.se/medarbetare/integritetsskyddspolicy>. > > > Sending email to Karolinska Institutet (KI) will result in KI processing > your personal data. You can read more about KI’s processing of personal > data here<https://ki.se/en/staff/data-protection-policy>. > |
![]() ![]() |
Adam Glaser |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone. I also now see online that PBS should have a RI of 1.335. If this is true then I agree it is a very small difference. Measuring some PBS in our lab on a refractometer I got a reading of 1.341 which seemed like it could start to have an impact. I will try to contact Zeiss and see what they designed the objective for, otherwise I'll just go with DI water for now! On Fri, Feb 7, 2020 at 10:08 AM Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > The salt doesn't make that much of a difference. One detail is to remind > users to rinse the lens off with pure water after using PBS. I made a > 'house call' to a lab that was having problems using one of their patch > rigs. After checking all their filters and prisms I happened to look down > at the objective and found a perfect window of dried salt sitting over the > face of the objective. One lengthy freshwater soak later and everything was > fine... > > Craig > > On Fri, Feb 7, 2020 at 2:01 AM Sylvie Le Guyader <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Adam > > > > Do you find a difference in refraction index between PBS and water? I > > think they are both 1.33. > > If they have the same refraction index then the objective should perform > > identically with both. > > > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader, PhD > > Live Cell Imaging Facility Manager > > Karolinska Institutet- Bionut Dpt > > Blickagången 16, > > Room 7362 (lab)/7840 (office) > > 14157 Huddinge, Sweden > > mobile: +46 (0) 73 733 5008 > > LCI website > > LCI microscopy blog > > > > -----Original Message----- > > From: Confocal Microscopy List <[hidden email]> On > > Behalf Of Adam Glaser > > Sent: Friday, 7 February, 2020 06:42 > > To: [hidden email] > > Subject: Dipping objectives - water or PBS? > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > > > > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=Us8B9MW12KnO0ZkHigDnV7dRMhaBFioHskNRbV2z6yg%3D&reserved=0 > > Post images on > > > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=ow3oJfNxg%2FZ19nC57cb0RBwfQXdTgDRmEN7bxbNDgpA%3D&reserved=0 > > and include the link in your posting. > > ***** > > > > Hi everyone, > > I was wondering if anyone had experience or knew if water dipping > > objectives (specifically the Zeiss W Plan-Apochromat) were designed for > > immersion in DI water or PBS. Or if they have been designed to perform > well > > in both refractive indices? > > Thanks! > > Adam > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att > KI > > kommer att behandla dina personuppgifter. Här finns information om hur KI > > behandlar personuppgifter< > > https://ki.se/medarbetare/integritetsskyddspolicy>. > > > > > > Sending email to Karolinska Institutet (KI) will result in KI processing > > your personal data. You can read more about KI’s processing of personal > > data here<https://ki.se/en/staff/data-protection-policy>. > > > |
![]() ![]() |
Mark Sonders |
![]() |
In reply to this post by Adam Glaser
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A hint following on Craig's spur from the original PBS refractive index question.... Caked-on salt films on water immersion objectives that have formed from repeated exposures to PBS buffers appear be removed by dipping them into 0.1 N acetic acid solution for ~10 min and rinsing with no obvious deleterious effect. -=-Mark Department of Psychiatry Columbia University Division of Molecular Therapeutics New York State Psychiatric Institute New York, NY 10032 |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks Mark! I'll give that a try when I next encounter salt films. It would probably be quicker than just a DI water soak. Craig On Sat, Feb 8, 2020 at 7:43 AM Mark Sonders <[hidden email]> wrote: > A hint following on Craig's spur from the original PBS refractive index > question.... > > Caked-on salt films on water immersion objectives that have formed from > repeated exposures to PBS buffers appear be removed by dipping them into > 0.1 N acetic acid solution for ~10 min and rinsing with no obvious > deleterious effect. > > -=-Mark > > Department of Psychiatry > Columbia University > Division of Molecular Therapeutics > New York State Psychiatric Institute > New York, NY 10032 > > > |
![]() ![]() |
Zdenek Svindrych-2 |
![]() |
In reply to this post by Adam Glaser
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Adam, btw, what's your RI reading for DI water? Just FYI, our routine PBS (137 mM NaCl, 2.7 mM KCl, 12 mM phosphate buffer pH 7.4) has a RI of 1.3355, and pure water gives me 1.3335 on my refractometer. zdenek On Sat, Feb 8, 2020 at 12:50 AM Adam Glaser <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thanks everyone. I also now see online that PBS should have a RI of 1.335. > If this is true then I agree it is a very small difference. Measuring some > PBS in our lab on a refractometer I got a reading of 1.341 which seemed > like it could start to have an impact. I will try to contact Zeiss and see > what they designed the objective for, otherwise I'll just go with DI water > for now! > > On Fri, Feb 7, 2020 at 10:08 AM Craig Brideau <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > The salt doesn't make that much of a difference. One detail is to remind > > users to rinse the lens off with pure water after using PBS. I made a > > 'house call' to a lab that was having problems using one of their patch > > rigs. After checking all their filters and prisms I happened to look down > > at the objective and found a perfect window of dried salt sitting over > the > > face of the objective. One lengthy freshwater soak later and everything > was > > fine... > > > > Craig > > > > On Fri, Feb 7, 2020 at 2:01 AM Sylvie Le Guyader < > [hidden email]> > > wrote: > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Hi Adam > > > > > > Do you find a difference in refraction index between PBS and water? I > > > think they are both 1.33. > > > If they have the same refraction index then the objective should > perform > > > identically with both. > > > > > > Med vänlig hälsning / Best regards > > > Sylvie > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > > Sylvie Le Guyader, PhD > > > Live Cell Imaging Facility Manager > > > Karolinska Institutet- Bionut Dpt > > > Blickagången 16, > > > Room 7362 (lab)/7840 (office) > > > 14157 Huddinge, Sweden > > > mobile: +46 (0) 73 733 5008 > > > LCI website > > > LCI microscopy blog > > > > > > -----Original Message----- > > > From: Confocal Microscopy List <[hidden email]> On > > > Behalf Of Adam Glaser > > > Sent: Friday, 7 February, 2020 06:42 > > > To: [hidden email] > > > Subject: Dipping objectives - water or PBS? > > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > > > > > > > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=Us8B9MW12KnO0ZkHigDnV7dRMhaBFioHskNRbV2z6yg%3D&reserved=0 > > > Post images on > > > > > > https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Csylvie.le.guyader%40KI.SE%7C7633ffe7aec949b2022208d7ab908db5%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637166510111401463&sdata=ow3oJfNxg%2FZ19nC57cb0RBwfQXdTgDRmEN7bxbNDgpA%3D&reserved=0 > > > and include the link in your posting. > > > ***** > > > > > > Hi everyone, > > > I was wondering if anyone had experience or knew if water dipping > > > objectives (specifically the Zeiss W Plan-Apochromat) were designed for > > > immersion in DI water or PBS. Or if they have been designed to perform > > well > > > in both refractive indices? > > > Thanks! > > > Adam > > > > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att > > KI > > > kommer att behandla dina personuppgifter. Här finns information om hur > KI > > > behandlar personuppgifter< > > > https://ki.se/medarbetare/integritetsskyddspolicy>. > > > > > > > > > Sending email to Karolinska Institutet (KI) will result in KI > processing > > > your personal data. You can read more about KI’s processing of personal > > > data here<https://ki.se/en/staff/data-protection-policy>. > > > > > > -- -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The error of measurement is usually greater than 0.001, and it depends on the wavelength http://www.philiplaven.com/p20.html Refractive index<http://www.philiplaven.com/p20.html> Refractive index of water. Most people would assume that the refractive index of water is known to a high degree of accuracy. However, as shown in Fig. 1, the published literature reveals significant differences in the values of refractive index of water for a given wavelength. www.philiplaven.com ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Zdenek Svindrych <[hidden email]> Sent: Monday, February 10, 2020 9:34 AM To: [hidden email] <[hidden email]> Subject: Re: Dipping objectives - water or PBS? ***** To join, leave or search the confocal microscopy listserv, go to: https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=5KA7aQsukPtd0ZYhtBR4ggTkNgkaDA6fKva740XgqCE%3D&reserved=0 Post images on https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=cTom0QQw3sOAuwvINDDv6pAwuPhPenNnux7NVqpAPFY%3D&reserved=0 and include the link in your posting. ***** Hi Adam, btw, what's your RI reading for DI water? Just FYI, our routine PBS (137 mM NaCl, 2.7 mM KCl, 12 mM phosphate buffer pH 7.4) has a RI of 1.3355, and pure water gives me 1.3335 on my refractometer. zdenek On Sat, Feb 8, 2020 at 12:50 AM Adam Glaser <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=5KA7aQsukPtd0ZYhtBR4ggTkNgkaDA6fKva740XgqCE%3D&reserved=0 > Post images on https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=cTom0QQw3sOAuwvINDDv6pAwuPhPenNnux7NVqpAPFY%3D&reserved=0 and include the link in your posting. > ***** > > Thanks everyone. I also now see online that PBS should have a RI of 1.335. > If this is true then I agree it is a very small difference. Measuring some > PBS in our lab on a refractometer I got a reading of 1.341 which seemed > like it could start to have an impact. I will try to contact Zeiss and see > what they designed the objective for, otherwise I'll just go with DI water > for now! > > On Fri, Feb 7, 2020 at 10:08 AM Craig Brideau <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=5KA7aQsukPtd0ZYhtBR4ggTkNgkaDA6fKva740XgqCE%3D&reserved=0 > > Post images on https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=cTom0QQw3sOAuwvINDDv6pAwuPhPenNnux7NVqpAPFY%3D&reserved=0 and include the link in your > posting. > > ***** > > > > The salt doesn't make that much of a difference. One detail is to remind > > users to rinse the lens off with pure water after using PBS. I made a > > 'house call' to a lab that was having problems using one of their patch > > rigs. After checking all their filters and prisms I happened to look down > > at the objective and found a perfect window of dried salt sitting over > the > > face of the objective. One lengthy freshwater soak later and everything > was > > fine... > > > > Craig > > > > On Fri, Feb 7, 2020 at 2:01 AM Sylvie Le Guyader < > [hidden email]> > > wrote: > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=5KA7aQsukPtd0ZYhtBR4ggTkNgkaDA6fKva740XgqCE%3D&reserved=0 > > > Post images on https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=cTom0QQw3sOAuwvINDDv6pAwuPhPenNnux7NVqpAPFY%3D&reserved=0 and include the link in your > > posting. > > > ***** > > > > > > Hi Adam > > > > > > Do you find a difference in refraction index between PBS and water? I > > > think they are both 1.33. > > > If they have the same refraction index then the objective should > perform > > > identically with both. > > > > > > Med vänlig hälsning / Best regards > > > Sylvie > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > > Sylvie Le Guyader, PhD > > > Live Cell Imaging Facility Manager > > > Karolinska Institutet- Bionut Dpt > > > Blickagången 16, > > > Room 7362 (lab)/7840 (office) > > > 14157 Huddinge, Sweden > > > mobile: +46 (0) 73 733 5008 > > > LCI website > > > LCI microscopy blog > > > > > > -----Original Message----- > > > From: Confocal Microscopy List <[hidden email]> On > > > Behalf Of Adam Glaser > > > Sent: Friday, 7 February, 2020 06:42 > > > To: [hidden email] > > > Subject: Dipping objectives - water or PBS? > > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > > > > > > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=5KA7aQsukPtd0ZYhtBR4ggTkNgkaDA6fKva740XgqCE%3D&reserved=0 > > > Post images on > > > > > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064861033&sdata=cTom0QQw3sOAuwvINDDv6pAwuPhPenNnux7NVqpAPFY%3D&reserved=0 > > > and include the link in your posting. > > > ***** > > > > > > Hi everyone, > > > I was wondering if anyone had experience or knew if water dipping > > > objectives (specifically the Zeiss W Plan-Apochromat) were designed for > > > immersion in DI water or PBS. Or if they have been designed to perform > > well > > > in both refractive indices? > > > Thanks! > > > Adam > > > > > > > > > När du skickar e-post till Karolinska Institutet (KI) innebär detta att > > KI > > > kommer att behandla dina personuppgifter. Här finns information om hur > KI > > > behandlar personuppgifter< > > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fki.se%2Fmedarbetare%2Fintegritetsskyddspolicy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064871028&sdata=ety48jZn4daqsZBopNS2Q2tgKSNtgx0dWwlaYOBwfZ0%3D&reserved=0>. > > > > > > > > > Sending email to Karolinska Institutet (KI) will result in KI > processing > > > your personal data. You can read more about KI’s processing of personal > > > data here<https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fki.se%2Fen%2Fstaff%2Fdata-protection-policy&data=02%7C01%7Cmmodel%40KENT.EDU%7C86daf42d16d04892344508d7ae36e26f%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637169423064871028&sdata=3wIV6Y38qo7ikiPxve0zqs6rXGMSVAPK3HrdK6p5heQ%3D&reserved=0>. > > > > > > -- -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |