|
Serge Kaddoura |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We recently tried to use an sCMOS Andor Zyla 4.2 camera with Micromanager 1.4.23 + Andor SDK 3.13.30034 (inside Micromanager folder), on the same computer with Nikon NIS 5.02.00 build 1266 + Andor SDK 3.13.30001, but Micromanager crashes when trying to connect the camera through the Hardware configuration wizard! We replaced the Andor SDK under the Micromanager folder to 3.13.30001 so that it matches NIS, but still had the same crash with Micromanager when trying to connect the camera through the Hardware configuration wizard! Interestingly, or luckily, the sCMOS Andor Zyla 4.2 is not available under Micromanager but still available from NIS! Has anyone else run into a similar issue? Any thoughts, suggestions, ideas would be much appreciated... Thanks in advance. Kind Regards, Serge Kaddoura. Tel: 703-763-0698 nanoscaleLABS.com<http://www.nanoscalelabs.com/welcome.html> - providing a harmonious ensemble of technologies to better orchestrate your research! primo, the first multi-protein photopatterning solution, based on Light Induced Molecular Adsorption of Protein! eOne, a new class of low-cost, high-accuracy ion channel and nanopore current amplifier! SPRm 200, the world's first commercial SPR Microscopy: a powerful technique for the simultaneous measurement of phenotypical changes of the sample via bright field and binding strength and kinetics via SPR. |
|
Kyle Michael Douglass |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Serge, On 04/05/2018 02:58 AM, Serge Kaddoura wrote: > We recently tried to use an sCMOS Andor Zyla 4.2 camera with Micromanager 1.4.23 + Andor SDK 3.13.30034 (inside Micromanager folder), on the same computer with Nikon NIS 5.02.00 build 1266 + Andor SDK 3.13.30001, but Micromanager crashes when trying to connect the camera through the Hardware configuration wizard! > Have you asked this question on the Micro-Manager mailing list? If not, I highly recommend doing so because Andor's 3rd Party Support team regularly monitors it for questions concerning Andor hardware and Micro-Manager. I am not certain whether they check this list for tech support issues. https://micro-manager.org/wiki/Micro-Manager%20Community In the mean time, I would suggest verifying that you haven't unknowingly installed two BitFlow driver versions, which is entirely possible because it appears that you have two different Andor SDK's installed on the same machine. Cheers, and see you on the Micro-Manager list ;) Kyle -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch |
| Free forum by Nabble | Edit this page |