FRET using fixed cells

classic Classic list List threaded Threaded
♦
♦
12 messages Options
Syed J Khundmiri Syed J Khundmiri
Reply | Threaded
Open this post in threaded view
♦
♦
|

FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins for
FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville
Gary G. Li Gary G. Li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed Khundmiri,


The basic requirements for FRET are: 1) The biomolecules tagged with
fluorophores are located in the nanometer range; 2) The acceptor absorption
spectrum overlaps with the donor emission spectrum.   There are some
commonly used donor-acceptor pairs, including AlexaFluor dyes.


TIRF is a high sensitivity fluorescence technique with which you can image
cellular structures located in the range 0-150nm above the coverslip, so
your cells should be flat and adherent to the coverslip.


 Generally, you can use FRET and TIRF for both live cells and fixed cells.


Gary G. Li, Ph.D.

Johns Hopkins School of Medicine
Baltimore, MD 21224

On Fri, Dec 3, 2010 at 10:17 AM, Syed Khundmiri <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>
> I am a novice in the field. I use live cells transfected with tagged
> proteins for
> FRET and TIRF microscopy. My question is can one use fixed cells for FRET
> using AlexaFluor antibodies?? If yes, please guide me to some references
> and
> how to.
>
> Another question is has someone used TAT-fusion peptides labelled with
> AlexaFluor for TIRF and FRET??
>
> Best Regards,
>
> Syed Khundmiri
> University of Louisville
>
Vitaly Boyko Vitaly Boyko
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

In reply to this post by Syed J Khundmiri
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Syed,

If the fluorescent tag does not cause mis-localization or aggregation of a
protein of interest, I recommend to use CeFP (Dave Piston) and the "late
edition" of the standard EYFP (with the F64L substitution) as the fluorescence
tags, and 440nm diode (or 427/10 nm Semrock excitation filter), 514 nm Argon
line (504/12 nm exciter, Semrock) and 472/30nm and 542/27nm emission filters,
also Semrock. TIRF or wide-field is the best optical systems for the sensitized
emission FRET.

Good luck,

Vitaly
301-515-7833




________________________________
From: Syed Khundmiri <[hidden email]>
To: [hidden email]
Sent: Fri, December 3, 2010 10:17:19 AM
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins
for

FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville



     
Vitaly Boyko Vitaly Boyko
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

In reply to this post by Syed J Khundmiri
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

PS Live cells would give you better and reproducible data than the fixed
material.

Dear Syed,

If the fluorescent tag does not cause mis-localization or aggregation of a
protein of interest, I recommend to use CeFP (Dave Piston) and the "late
edition" of the standard EYFP (with the F64L substitution) as the fluorescence
tags, and 440nm diode (or 427/10 nm Semrock excitation filter), 514 nm Argon
line (504/12 nm exciter, Semrock) and 472/30nm and 542/27nm emission filters,
also Semrock. TIRF or wide-field is the best optical systems for the sensitized
emission FRET.

Good luck,

Vitaly
301-515-7833



________________________________
From: Syed Khundmiri <[hidden email]>
To: [hidden email]
Sent: Fri, December 3, 2010 10:17:19 AM
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins
for

FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville



     
lgelman lgelman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

In reply to this post by Syed J Khundmiri
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed,

We used cells labeled with antibodies coupled to Alexas for FRET with success. The Alexa pairs we used were 488&555 and 568&633.

Best regards,

Laurent.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Syed Khundmiri
Sent: vendredi 3 décembre 2010 16:17
To: [hidden email]
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins for
FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville
Periasamy, Ammasi (ap3t) Periasamy, Ammasi (ap3t)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear All
It has been widely used fixed cell for FRET. We recently found out that the E% decreases for the fixed cells versus live cell. We also found out that the fixative reduces the E%. you can clearly see the change in E% in the lifetime FRET imaging. We did not see much change in E% if you collect the data within few days after the cells are fixed. We used both formaldehyde and methanol as fixative. We are finalizing the data and plan to submit soon for a journal and I will post later as soon as it published.
We also train the participants in our annual hands on training FRET workshop during March in the FLIM-FRET system live versus fixed cell.
Not a good idea to use fixed cell for FRET or any fixed sample in the TIRF system. You are working with evanescent wave in the TIRF system with nanometer depth and it is require a alive specimen.
Hope this helps
Ammasi

Prof. Ammasi Periasamy
Director, Keck Center for Cellular Imaging (KCCI)
Biology, Gilmer Hall (064), 485 McCormick Rd
University of Virginia
Charlottesville, VA 22904
Voice: 434-243-7602 (Office); 982-4869 (lab)
Fax:434-982-5210; Email:[hidden email]
http://www.kcci.virginia.edu/Contact/peri.php
************************
10th Annual Workshop on FRET Microscopy, March 8-13, 2011
http://www.kcci.virginia.edu/workshop/workshop2011/
One Day Symposium to honor Prof. Theodor Förster's 100th Birth Day, March 10, 2011
http://www.kcci.virginia.edu/symposium/
*************************

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gelman, Laurent
Sent: Tuesday, December 07, 2010 6:51 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed,

We used cells labeled with antibodies coupled to Alexas for FRET with success. The Alexa pairs we used were 488&555 and 568&633.

Best regards,

Laurent.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Syed Khundmiri
Sent: vendredi 3 décembre 2010 16:17
To: [hidden email]
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins for
FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville
Ignatius, Mike-2 Ignatius, Mike-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

To add to the issues in comparing fixed and live cell results, unrelated to reagent choice, this November's Microscopy Today issue has an article on AFM imaging of live and fixed cells by A. Berquand et al, page 8-14.

U2-OS cells shrunk from 14 microns in height to 5 microns after gluteraldehyde fixation.  Not much change in lateral spacing, but a ~65% change (collapse) in Z needs to be considered in relating fixed specimen FRET results to live cell results.  

Mike Ignatius

Molecular Probes/LifeTechnologies

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Periasamy, Ammasi (ap3t)
Sent: Tuesday, December 07, 2010 6:46 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear All
It has been widely used fixed cell for FRET. We recently found out that the E% decreases for the fixed cells versus live cell. We also found out that the fixative reduces the E%. you can clearly see the change in E% in the lifetime FRET imaging. We did not see much change in E% if you collect the data within few days after the cells are fixed. We used both formaldehyde and methanol as fixative. We are finalizing the data and plan to submit soon for a journal and I will post later as soon as it published.
We also train the participants in our annual hands on training FRET workshop during March in the FLIM-FRET system live versus fixed cell.
Not a good idea to use fixed cell for FRET or any fixed sample in the TIRF system. You are working with evanescent wave in the TIRF system with nanometer depth and it is require a alive specimen.
Hope this helps
Ammasi

Prof. Ammasi Periasamy
Director, Keck Center for Cellular Imaging (KCCI)
Biology, Gilmer Hall (064), 485 McCormick Rd
University of Virginia
Charlottesville, VA 22904
Voice: 434-243-7602 (Office); 982-4869 (lab)
Fax:434-982-5210; Email:[hidden email]
http://www.kcci.virginia.edu/Contact/peri.php
************************
10th Annual Workshop on FRET Microscopy, March 8-13, 2011
http://www.kcci.virginia.edu/workshop/workshop2011/
One Day Symposium to honor Prof. Theodor Förster's 100th Birth Day, March 10, 2011
http://www.kcci.virginia.edu/symposium/
*************************

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gelman, Laurent
Sent: Tuesday, December 07, 2010 6:51 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed,

We used cells labeled with antibodies coupled to Alexas for FRET with success. The Alexa pairs we used were 488&555 and 568&633.

Best regards,

Laurent.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Syed Khundmiri
Sent: vendredi 3 décembre 2010 16:17
To: [hidden email]
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins for
FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville
Syed J Khundmiri Syed J Khundmiri
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thank you all for your valuable suggestions.
 
We are using neutrophils for our studies. They are suspension cells and do not attach to the glass slides unless fixed with paraformaldehyde. However, if we incubate the cells for an hour they will settle down and attach loosely to collagen coated plates. They do not survive more than few hours and therefore we cannot do any transfections. However, we know that they take up antibodies. If I let them settle down and do not fix them with paraformaldehyde, then is it possible to do FRET/TIRF in these cells using fluorescent antibodies???
 
Syed
 
 
Syed Khundmiri, Ph.D.
University of Louisville
 

>>> "Ignatius, Mike" <[hidden email]> 12/7/2010 12:50 PM >>>
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

To add to the issues in comparing fixed and live cell results, unrelated to reagent choice, this November's Microscopy Today issue has an article on AFM imaging of live and fixed cells by A. Berquand et al, page 8-14.

U2-OS cells shrunk from 14 microns in height to 5 microns after gluteraldehyde fixation.  Not much change in lateral spacing, but a ~65% change (collapse) in Z needs to be considered in relating fixed specimen FRET results to live cell results.  

Mike Ignatius

Molecular Probes/LifeTechnologies

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Periasamy, Ammasi (ap3t)
Sent: Tuesday, December 07, 2010 6:46 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear All
It has been widely used fixed cell for FRET. We recently found out that the E% decreases for the fixed cells versus live cell. We also found out that the fixative reduces the E%. you can clearly see the change in E% in the lifetime FRET imaging. We did not see much change in E% if you collect the data within few days after the cells are fixed. We used both formaldehyde and methanol as fixative. We are finalizing the data and plan to submit soon for a journal and I will post later as soon as it published.
We also train the participants in our annual hands on training FRET workshop during March in the FLIM-FRET system live versus fixed cell.
Not a good idea to use fixed cell for FRET or any fixed sample in the TIRF system. You are working with evanescent wave in the TIRF system with nanometer depth and it is require a alive specimen.
Hope this helps
Ammasi

Prof. Ammasi Periasamy
Director, Keck Center for Cellular Imaging (KCCI)
Biology, Gilmer Hall (064), 485 McCormick Rd
University of Virginia
Charlottesville, VA 22904
Voice: 434-243-7602 (Office); 982-4869 (lab)
Fax:434-982-5210; Email:[hidden email]
http://www.kcci.virginia.edu/Contact/peri.php
************************
10th Annual Workshop on FRET Microscopy, March 8-13, 2011
http://www.kcci.virginia.edu/workshop/workshop2011/
One Day Symposium to honor Prof. Theodor Förster's 100th Birth Day, March 10, 2011
http://www.kcci.virginia.edu/symposium/
*************************

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gelman, Laurent
Sent: Tuesday, December 07, 2010 6:51 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed,

We used cells labeled with antibodies coupled to Alexas for FRET with success. The Alexa pairs we used were 488&555 and 568&633.

Best regards,

Laurent.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Syed Khundmiri
Sent: vendredi 3 décembre 2010 16:17
To: [hidden email]
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins for
FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville
B. Prabhakar Pandian B. Prabhakar Pandian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Fluorescent bacteriophage

In reply to this post by Ignatius, Mike-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello Everybody,
                                 Does anyone have an idea of where I can
purchase fluorescent bacteriophages.

Thanks,

-Prabhakar
Kate Luby-Phelps Kate Luby-Phelps
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

In reply to this post by Syed J Khundmiri
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

It is possible to do TIRF with fixed cells as long as you keep them in PBS so that
the refractive index difference at the interface with the glass is appropriate for
TIRF. Don't know about TIRF-FRET with fixed cells. I defer to Peri on that.

Kate
ed@hpcf.upr.edu ed@hpcf.upr.edu
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Excellent.

Ed

********************************************************************

PLEASE NOTE: WHEN TRAVELING I USE

OTHER ELECTRONIC MAIL ACCOUNTS.

PLEASE SEND ALL YOUR REPLYS TO:

[hidden email].

 

Mailing Address:

Eduardo Rosa-Molinar, Ph.D.

Group Leader, Biological Imaging Group (BIG)

University of Puerto Rico-Rio Piedras

P.O. Box 21809 UPR Station

San Juan, Puerto Rico 00931-1809

Telephone: 787-764-0000; extension 1-3551

(voice mail only)

Facsimile:  888-762-7577

Electronic Mail:  mailto:[hidden email]@hpcf.upr.edu

Website:  http://pisces.cnnet.clu.edu/erm-lab

 

Shipping Address:

University of Puerto Rico-Rio Piedras

Biological Imaging Group (BIG)

Attention: Eduardo Rosa-Molinar, Ph.D.

Group Leader, Biological Imaging Group (BIG)

Esquina Barbosa y Gandara Avenue

Julio Garcia Bldg; Rm 118

Rio Piedras, Puerto Rico 00931

Telephone: 787-764-0000; extension 1-3551

(voice mail only)

Facsimile: 888-762-7577

Electronic Mail:  mailto:[hidden email]@hpcf.upr.edu

Website:  http://pisces.cnnet.clu.edu/erm-lab

 

Note:

This email and any files transmitted with it may contain information that is

PRIVILEDGE, CONFIDENTIAL, and exempt from disclosure under

applicable law.  It is intended only for the individual(s) or entity named
above.  

If you are not an intended recipient of this e-mail, you are hereby notified
that any

unauthorized use, dissemination or copying of this e-mail or the information

contained in it or attached to it is strictly prohibited.  If you have
received this email

in error, please delete it and immediately notify the person named above by
reply

email.

********************************************************************


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Kate Luby-Phelps
Sent: Wednesday, December 08, 2010 9:35 AM
To: [hidden email]
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

It is possible to do TIRF with fixed cells as long as you keep them in PBS
so that
the refractive index difference at the interface with the glass is
appropriate for
TIRF. Don't know about TIRF-FRET with fixed cells. I defer to Peri on that.

Kate
Vitaly Boyko Vitaly Boyko
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET using fixed cells

In reply to this post by lgelman
Suspension cells could be also transfected while in suspensio
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hy Syed,

Suspension cells could be also transfected while in suspension state and
immobilized on the coated coverglass immediately before imaging. Yes, antibody
based FRET works but it is much harder to get impressive and accurate
quantitative data with the antibody based FRET, as the FRET signal is usually
weak. However, if you would be happy with the 'YES FRET' and 'NO FRET' readout,
then the antibody based approach could be OK.

Cheers,

Vitaly
+1-301-515-7833




________________________________
From: "Gelman, Laurent" <[hidden email]>
To: [hidden email]
Sent: Tue, December 7, 2010 6:51:26 AM
Subject: Re: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Syed,

We used cells labeled with antibodies coupled to Alexas for FRET with success.
The Alexa pairs we used were 488&555 and 568&633.

Best regards,

Laurent.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Syed Khundmiri
Sent: vendredi 3 décembre 2010 16:17
To: [hidden email]
Subject: FRET using fixed cells

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

I am a novice in the field. I use live cells transfected with tagged proteins
for

FRET and TIRF microscopy. My question is can one use fixed cells for FRET
using AlexaFluor antibodies?? If yes, please guide me to some references and
how to.

Another question is has someone used TAT-fusion peptides labelled with
AlexaFluor for TIRF and FRET??

Best Regards,

Syed Khundmiri
University of Louisville