http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello listers, I am wondering the difference between HV and gain in the operational software of FV1000. Please educate me. Thanks very much in advance. Ohkyung |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello Ohkyung, The HV is the current supplied to the PMT. The Gain is the amplification of the signal from the PMT. hope this helps. best James Beals [hidden email] 734.936-2051 205 Zina Pitcher Place 2038 MBNI Molecular and Behavioral Neuroscience Institute University of Michigan Ann Arbor, Mi 48109 On Sep 3, 2008, at 11:28 AM, Ohkyung Kwon wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hello listers, > > I am wondering the difference between HV and gain in the > operational software of FV1000. Please educate me. Thanks very much > in advance. > > Ohkyung |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Ohkyung, James, I admittedly have never used an FV1000, but I am surprised that the HV refers to current rather than PMT operating voltage (which drives the current). Current generated in a PMT is exponentially sensitive to the voltage applied and each tube will have an optimal voltage (usually several hundred) that gives a linear current output over a range of light intensities. Gain, on the other other hand, would be normally pertinent to amplification of the PMT current output to something suitable for the A/D converters that digitize the signal. A clarification would helpful. Thanks, Mario >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hello Ohkyung, >The HV is the current supplied to the PMT. >The Gain is the amplification of the signal from the PMT. >hope this helps. >best >James Beals >[hidden email] >734.936-2051 > >205 Zina Pitcher Place >2038 MBNI >Molecular and Behavioral Neuroscience Institute >University of Michigan >Ann Arbor, Mi >48109 > > > > > >On Sep 3, 2008, at 11:28 AM, Ohkyung Kwon wrote: > >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>Hello listers, >> >>I am wondering the difference between HV and gain in the >>operational software of FV1000. Please educate me. Thanks very much >>in advance. >> >>Ohkyung -- ________________________________________________________________________________ Mario M. Moronne, Ph.D. [hidden email] [hidden email] [hidden email] |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal On the FV1000 the gain is applied after the A/D conversion. This becoms evident if the histogram is studied closely on images collected at gain settings higher than 1.0. We confirmed this through dialogue with the local consultant and the manufacturer. I understand that other makes have a gain stage prior to this step, as Mario suggests. Regards, Barry At 05:16 a.m. 4/09/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Ohkyung, James, > >I admittedly have never used an FV1000, but I am surprised that the HV >refers to current rather than PMT operating voltage (which drives the >current). Current generated in a PMT is exponentially sensitive to the >voltage applied and each tube will have an optimal voltage (usually >several hundred) that gives a linear current output over a range of light >intensities. Gain, on the other other hand, would be normally pertinent to >amplification of the PMT current output to something suitable for the A/D >converters that digitize the signal. > >A clarification would helpful. Thanks, > >Mario > >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>Hello Ohkyung, >>The HV is the current supplied to the PMT. >>The Gain is the amplification of the signal from the PMT. >>hope this helps. >>best >>James Beals >>[hidden email] >>734.936-2051 >> >>205 Zina Pitcher Place >>2038 MBNI >>Molecular and Behavioral Neuroscience Institute >>University of Michigan >>Ann Arbor, Mi >>48109 >> >> >> >> >> >>On Sep 3, 2008, at 11:28 AM, Ohkyung Kwon wrote: >> >>>Search the CONFOCAL archive at >>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>>Hello listers, >>> >>>I am wondering the difference between HV and gain in the operational >>>software of FV1000. Please educate me. Thanks very much in advance. >>> >>>Ohkyung > > >-- >________________________________________________________________________________ >Mario M. Moronne, Ph.D. > >[hidden email] >[hidden email] >[hidden email] Dr Barry O'Brien Dept of Biological Sciences, University of Waikato Private Bag 3105 HAMILTON New Zealand Fax 0064 7 838 4324 Phone 0064 7 838 4179 |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal G'day Ohkyung, I'm only new to the FV1000, so if others disagree please correct me. HV refers to High Voltage and is the number of volts applied to the PMT. To add to your confusion this is sometimes also called "Gain" or "PMT Gain" by other manufacturers. As you and others have pointed out the FV1000 also has a "Gain" function, and this is an amplification of the PMT signal as previously pointed out. Generally speaking I think it is preferable to have this amplification gain switched off (ie. set to 1), and use only the HV to increase signal. I presume as the Amp. Gain feature is offered by Olympus, there must be situations where if the signal strength is low, and if increasing the HV further results in more noise (Lower signal/noise ratio (S/N) ), then using some Amplifier gain and a little less PMT HV may produce a better S/N ratio (ie. a less noisy image). I have not yet found that set of circumstances. Can anyone comment in what set of circumstances the Amp. Gain becomes useful? Cheers Steve Stephen H. Cody Microscopy Manager Central Resource for Advanced Microscopy Ludwig Institute for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 Australia Tel: 61 3 9341 3155 Fax: 61 3 9341 3104 email: [hidden email] www.ludwig.edu.au/labs/confocal.html www.ludwig.edu.au/confocal Tip: Learn how to receive reminders about you microscope booking: www.ludwig.edu.au/confocal/Local/Booking_Hint.htm -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Barry O'Brien Sent: Thursday, 4 September 2008 8:07 AM To: [hidden email] Subject: Re: FV1000 : HV vs gain Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal On the FV1000 the gain is applied after the A/D conversion. This becoms evident if the histogram is studied closely on images collected at gain settings higher than 1.0. We confirmed this through dialogue with the local consultant and the manufacturer. I understand that other makes have a gain stage prior to this step, as Mario suggests. Regards, Barry At 05:16 a.m. 4/09/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Ohkyung, James, > >I admittedly have never used an FV1000, but I am surprised that the HV >refers to current rather than PMT operating voltage (which drives the >current). Current generated in a PMT is exponentially sensitive to the >voltage applied and each tube will have an optimal voltage (usually >several hundred) that gives a linear current output over a range of >intensities. Gain, on the other other hand, would be normally pertinent to >amplification of the PMT current output to something suitable for the A/D >converters that digitize the signal. > >A clarification would helpful. Thanks, > >Mario > >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>Hello Ohkyung, >>The HV is the current supplied to the PMT. >>The Gain is the amplification of the signal from the PMT. >>hope this helps. >>best >>James Beals >>[hidden email] >>734.936-2051 >> >>205 Zina Pitcher Place >>2038 MBNI >>Molecular and Behavioral Neuroscience Institute >>University of Michigan >>Ann Arbor, Mi >>48109 >> >> >> >> >> >>On Sep 3, 2008, at 11:28 AM, Ohkyung Kwon wrote: >> >>>Search the CONFOCAL archive at >>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>>Hello listers, >>> >>>I am wondering the difference between HV and gain in the operational >>>software of FV1000. Please educate me. Thanks very much in advance. >>> >>>Ohkyung > > >-- >_______________________________________________________________________ >Mario M. Moronne, Ph.D. > >[hidden email] >[hidden email] >[hidden email] Dr Barry O'Brien Dept of Biological Sciences, University of Waikato Private Bag 3105 HAMILTON New Zealand Fax 0064 7 838 4324 Phone 0064 7 838 4179 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research. |
|
|
In reply to this post by Oky
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, I agree with Steve that the "Gain" on FV1000 should be left untouched (set to 1). And that is also what the Olympus specialist told me during the initial training on our system. If I understand correctly, the Gain control just multiplies the digitized pixel intensity by the set number (e.g., 1.25). This can be done during post-processing in image analysis software of your choice. Perhaps someone from Olympus could comment when using "gain" would be beneficial? When the signal is weak and would require high HV (above ~ 750V on our system) leading to high noise, I use lower HV and adjust the lookup table to make the image look brighter, or even better, switch to photon counting (in that case, you definitely have to tweak the LUT to see anything in the image). Stan Dr. Stanislav Vitha Microscopy and Imaging Center Texas A&M University BSBW 119 College Station, TX 77843-2257 http://microscopy.tamu.edu On Thu, 4 Sep 2008 12:15:34 +1000, Stephen Cody <[hidden email]> wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >G'day Ohkyung, > >I'm only new to the FV1000, so if others disagree please correct me. > >HV refers to High Voltage and is the number of volts applied to the PMT. >To add to your confusion this is sometimes also called "Gain" or "PMT >Gain" by other manufacturers. > >As you and others have pointed out the FV1000 also has a "Gain" >function, and this is an amplification of the PMT signal as previously >pointed out. > >Generally speaking I think it is preferable to have this amplification >gain switched off (ie. set to 1), and use only the HV to increase >signal. I presume as the Amp. Gain feature is offered by Olympus, there >must be situations where if the signal strength is low, and if >increasing the HV further results in more noise (Lower signal/noise >ratio (S/N) ), then using some Amplifier gain and a little less PMT HV >may produce a better S/N ratio (ie. a less noisy image). I have not yet >found that set of circumstances. Can anyone comment in what set of >circumstances the Amp. Gain becomes useful? > >Cheers >Steve > >Stephen H. Cody >Microscopy Manager >Central Resource for Advanced Microscopy >Ludwig Institute for Cancer Research >PO Box 2008 Royal Melbourne Hospital >Victoria, 3050 >Australia >Tel: 61 3 9341 3155 Fax: 61 3 9341 3104 >email: [hidden email] >www.ludwig.edu.au/labs/confocal.html >www.ludwig.edu.au/confocal > >Tip: Learn how to receive reminders about you microscope booking: >www.ludwig.edu.au/confocal/Local/Booking_Hint.htm > >-----Original Message----- >From: Confocal Microscopy List [mailto:[hidden email]] >Behalf Of Barry O'Brien >Sent: Thursday, 4 September 2008 8:07 AM >To: [hidden email] >Subject: Re: FV1000 : HV vs gain > >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >On the FV1000 the gain is applied after the A/D conversion. This becoms >evident if the histogram is studied closely on images collected at gain >settings higher than 1.0. We confirmed this through dialogue with the >local consultant and the manufacturer. >I understand that other makes have a gain stage prior to this step, as >Mario suggests. > >Regards, > >Barry > > >At 05:16 a.m. 4/09/2008, you wrote: >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>Ohkyung, James, >> >>I admittedly have never used an FV1000, but I am surprised that the HV >>refers to current rather than PMT operating voltage (which drives the >>current). Current generated in a PMT is exponentially sensitive to the >>voltage applied and each tube will have an optimal voltage (usually >>several hundred) that gives a linear current output over a range of >light >>intensities. Gain, on the other other hand, would be normally pertinent >to >>amplification of the PMT current output to something suitable for the >A/D >>converters that digitize the signal. >> >>A clarification would helpful. Thanks, >> >>Mario >> >>>Search the CONFOCAL archive at >>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>>Hello Ohkyung, >>>The HV is the current supplied to the PMT. >>>The Gain is the amplification of the signal from the PMT. >>>hope this helps. >>>best >>>James Beals >>>[hidden email] >>>734.936-2051 >>> >>>205 Zina Pitcher Place >>>2038 MBNI >>>Molecular and Behavioral Neuroscience Institute >>>University of Michigan >>>Ann Arbor, Mi >>>48109 >>> >>> >>> >>> >>> >>>On Sep 3, 2008, at 11:28 AM, Ohkyung Kwon wrote: >>> >>>>Search the CONFOCAL archive at >>>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>>> >>>>Hello listers, >>>> >>>>I am wondering the difference between HV and gain in the operational >>>>software of FV1000. Please educate me. Thanks very much in advance. >>>> >>>>Ohkyung >> >> >>-- >>________________________________________________________________ >_________ >>Mario M. Moronne, Ph.D. >> >>[hidden email] >>[hidden email] >>[hidden email] > >Dr Barry O'Brien >Dept of Biological Sciences, >University of Waikato >Private Bag 3105 >HAMILTON >New Zealand > >Fax 0064 7 838 4324 >Phone 0064 7 838 4179 > > >This communication is intended only for the named recipient and may contain Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research. |