|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello everybody, A simple question related to ZEN in an Apotome system. I noticed that when you save a configuration in the "Experiment manager", let's say "transmited light, no apotome" and you reload it, ZEN, automatically put the system in apotome mode. However, if you load an image with "transmited light, no apotome" and push the "reuse" function, ZEN put exactly this configuration and do not active the apotome mode. Has anybody found a solution for this bug? Thank you very much in advance Best regards Juan Luis -- Juan Luis Ribas Servicio de Microscopía Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b 41012 Sevilla Spain |
|
Quintanar, Loren |
|
|
Hi Juan,
1) Which ZEN version do you have? We are running ZEN 2 Blue with hotfix 2.0.0.6. on a Zeiss Observer Z1. 2) Did you initially create and SAVE the particular experiment profile while ApoTome was activated? (By activated I mean the power supply switched on, the ApoTome pushed into the microscope, and then ZEN loaded properly afterward). My gut feeling is that ZEN probably "saved" the engaged ApoTome state while you created a profiled called "X, no ApoTome." ZEN has the uncanny ability to remember and save very small minutiae within the software when you create any experimental profile. It also has the uncanny ability to reuse exactly any picture's settings, regardless of the hardware setup (Like if the ApoTome was engaged at the time you click "reuse settings"). Hope this helps. Loren Quintanar Research Associate II Light Microscopy Core Facility Beckman Research Institute City of Hope x62874 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Juan Luis Ribas Sent: Tuesday, March 15, 2016 11:43 AM To: [hidden email] Subject: Failure in Experiment manager.- Apotome ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello everybody, A simple question related to ZEN in an Apotome system. I noticed that when you save a configuration in the "Experiment manager", let's say "transmited light, no apotome" and you reload it, ZEN, automatically put the system in apotome mode. However, if you load an image with "transmited light, no apotome" and push the "reuse" function, ZEN put exactly this configuration and do not active the apotome mode. Has anybody found a solution for this bug? Thank you very much in advance Best regards Juan Luis -- Juan Luis Ribas Servicio de Microscopía Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b 41012 Sevilla Spain --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Loren, The problem is solved very easily. There is an option under "Channels" that says "Disable for apotome". If you click in, the system doesn't use that part of the hardware for acquisition. Thank you very much Have a good day Best regards Juan Luis Ribas El 15/03/2016 a las 22:17, Quintanar, Loren escribió: > Hi Juan, > > 1) Which ZEN version do you have? We are running ZEN 2 Blue with hotfix 2.0.0.6. on a Zeiss Observer Z1. > 2) Did you initially create and SAVE the particular experiment profile while ApoTome was activated? (By activated I mean the power supply switched on, the ApoTome pushed into the microscope, and then ZEN loaded properly afterward). > > My gut feeling is that ZEN probably "saved" the engaged ApoTome state while you created a profiled called "X, no ApoTome." ZEN has the uncanny ability to remember and save very small minutiae within the software when you create any experimental profile. It also has the uncanny ability to reuse exactly any picture's settings, regardless of the hardware setup (Like if the ApoTome was engaged at the time you click "reuse settings"). > > Hope this helps. > > Loren Quintanar > Research Associate II > Light Microscopy Core Facility > Beckman Research Institute > City of Hope > x62874 > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Juan Luis Ribas > Sent: Tuesday, March 15, 2016 11:43 AM > To: [hidden email] > Subject: Failure in Experiment manager.- Apotome > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello everybody, > A simple question related to ZEN in an Apotome system. > I noticed that when you save a configuration in the "Experiment manager", let's say "transmited light, no apotome" and you reload it, ZEN, automatically put the system in apotome mode. However, if you load an image with "transmited light, no apotome" and push the "reuse" > function, ZEN put exactly this configuration and do not active the apotome mode. > Has anybody found a solution for this bug? > > Thank you very much in advance > Best regards > Juan Luis > > > -- > Juan Luis Ribas > Servicio de Microscopía > Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b > 41012 Sevilla > Spain > > > --------------------------------------------------------------------- > *SECURITY/CONFIDENTIALITY WARNING: > This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) > --------------------------------------------------------------------- > -- Juan Luis Ribas Servicio de Microscopía Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b 41012 Sevilla Tfnos: 954559983 // 628220681 // 12068 http://investigacion.us.es/scisi/sgi/servicios/microscopia/equipamiento |
| Free forum by Nabble | Edit this page |