Nicholas Geisse |
|
|
I was wondering if there is a simple answer to this question. I am looking for data regarding the quality and/or efficacy of optical abberation corrections in various types of miroscope objectives. Particularly, I am interested in the linearity of the magnification from the central optical axis to the extremes of the field of view of objectives as well as flat field corrections. Obviously this will vary between different objective classes and wavelengths, but I can’t seem to find any hard data on magnification distortions across a field of view for any objectives. Does anyone know of a good source for these types of data, or of an approved method of measuring these distortions, if any? Thanks nick |
|
|
This will vary from objective to
objective. I use a calibration slide (stage micrometer) and measure the spacing at the
edges and at the centre. Do remember that in a confocal
non-linearity of the scan is also a possibility, so it’s best to check
in widefield. Also, the coupling lenses for your camera may not be perfect, so if
you get strange results check them against an eyepiece micrometer. For flatness (or otherwise) of field a
very good sample is a coverslip very lightly smoked over a match-flame. Make a few scratches to
give some coarse detail for easy location of
the focal plane. Mount it on a slide with a drop of oil. Then check
the focal shift between the centre and the edge. This sample is also
very good for showing up chromatic aberration. You will probably be surprised by what you
see when you do these tests!
Guy Optical Imaging Techniques in Cell Biology From:
Hey everybody, |
|
Masoumi, Zahra (GE Healthcare) |
|
|
To do flat field correction of image please see the poster
presented at the 11th Annual Conference of the Society for Biomolecular
Screening, Geneva, Switzerland, Sep 11-15, 2005
Regards,
Zahra From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Tuesday, December 02, 2008 1:49 AM To: [hidden email] Subject: Re: Flat field correction of objectives This will vary from
objective to objective. I use a calibration slide (stage
micrometer) and measure
the spacing at the edges and at the centre. Do remember that in a
confocal non-linearity of the scan is also a possibility, so it’s
best to check in widefield. Also, the coupling lenses
for your camera may not
be perfect, so if you get strange results check them against an
eyepiece micrometer. For flatness (or
otherwise) of field a very good sample is a coverslip
very lightly smoked over a
match-flame. Make a few scratches to give some coarse detail for
easy location of the focal plane. Mount it on a
slide with a drop of
oil. Then check the focal shift between the centre
and the edge.
This sample is also very good for showing up
chromatic aberration. You will probably be
surprised by what you see when you do these tests!
Guy Optical Imaging Techniques in Cell