To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello everybody, We would need to stain living S. Pombe cells imaged on a fluorescence microscope to segment thereafter during image analysis automatically the cells. Staining cell wall and/or membrane would be perfect. Has someone any specific experience with yeast and/or Pombe? Note that we need an organic dye and cannot express any recombinant protein (we are screening a yeast library). Thanks for your advices! Best regards, Laurent. _____________________________________________ Laurent Gelman, PhD Head of Facility for Advanced Imaging and Microscopy (Light Microscopy) Friedrich Miescher Institut WRO 1066.2.16 Maulbeerstrasse 66 CH-4058 Basel +41 (0)61 696 43 38 +41 (0)79 618 73 69 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Laurent - Calcofluor or other cellulose-binders are typical, caveat being I don't work with yeasts.. A membrane stain called FUN has also been used. See papers with Ghannoum in the author line. Rob On Aug 15, 2013, at 5:46 AM, Gelman, Laurent wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everybody, > > > We would need to stain living S. Pombe cells imaged on a > fluorescence microscope to segment thereafter during image analysis > automatically the cells. > Staining cell wall and/or membrane would be perfect. Has someone any > specific experience with yeast and/or Pombe? > > Note that we need an organic dye and cannot express any recombinant > protein (we are screening a yeast library). > > Thanks for your advices! > > Best regards, > > > Laurent. > > > > _____________________________________________ > Laurent Gelman, PhD > Head of Facility for Advanced Imaging and Microscopy > (Light Microscopy) > > Friedrich Miescher Institut > WRO 1066.2.16 > Maulbeerstrasse 66 > CH-4058 Basel > +41 (0)61 696 43 38 > +41 (0)79 618 73 69 Robert J. Palmer Jr., Ph.D. Microbial Receptors Section Laboratory of Cell and Developmental Biology Natl Inst Dental Craniofacial Res - Natl Insts Health Bldg 30, Room 207 30 Convent Drive Bethesda MD 20892 ph 301-594-0025 fax 301-402-0396 |
![]() ![]() |
Graham Wright-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Laurent, How about FM4-46 (it will stain the membranes, initially the plasma membrane)? One potential problem though, depending in the rate of endocytosis in your cells, will be that it will be internalised over time if the cells will be in it for a while before imaging. It worked well with Neurospora crassa for us. Regards Graham -- Graham Wright, PhD Institute of Medical Biology, A*STAR Singapore On 15 Aug, 2013, at 20:50, "Rob Palmer" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Laurent - > Calcofluor or other cellulose-binders are typical, caveat being I > don't work with yeasts.. A membrane stain called FUN has also been > used. See papers with Ghannoum in the author line. > Rob > > On Aug 15, 2013, at 5:46 AM, Gelman, Laurent wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello everybody, >> >> >> We would need to stain living S. Pombe cells imaged on a >> fluorescence microscope to segment thereafter during image analysis >> automatically the cells. >> Staining cell wall and/or membrane would be perfect. Has someone any >> specific experience with yeast and/or Pombe? >> >> Note that we need an organic dye and cannot express any recombinant >> protein (we are screening a yeast library). >> >> Thanks for your advices! >> >> Best regards, >> >> >> Laurent. >> >> >> >> _____________________________________________ >> Laurent Gelman, PhD >> Head of Facility for Advanced Imaging and Microscopy >> (Light Microscopy) >> >> Friedrich Miescher Institut >> WRO 1066.2.16 >> Maulbeerstrasse 66 >> CH-4058 Basel >> +41 (0)61 696 43 38 >> +41 (0)79 618 73 69 > > Robert J. Palmer Jr., Ph.D. > Microbial Receptors Section > Laboratory of Cell and Developmental Biology > Natl Inst Dental Craniofacial Res - Natl Insts Health > Bldg 30, Room 207 > 30 Convent Drive > Bethesda MD 20892 > ph 301-594-0025 > fax 301-402-0396 |
![]() ![]() |
Theresa Swayne |
![]() |
In reply to this post by lgelman
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Laurent, Fluorescent Concanavalin A and Calcofluor both work well for S. cerevisiae. Calcofluor is the cell wall dye used in the FUN-1 (Live/Dead) yeast kit. The Ohya and Morishita labs used FITC-ConA to generate the Saccharomyces cerevisiae Morphological Database, http://scmd.gi.k.u-tokyo.ac.jp/datamine/. If you are doing long-term imaging, ConA might be better as you could choose wavelengths that would avoid UV exposure. Hope this helps! Theresa On Aug 15, 2013, at 5:46 AM, "Gelman, Laurent" <[hidden email]> wrote: > > We would need to stain living S. Pombe cells imaged on a fluorescence microscope to segment thereafter during image analysis automatically the cells. > Staining cell wall and/or membrane would be perfect. Has someone any specific experience with yeast and/or Pombe? > > Note that we need an organic dye and cannot express any recombinant protein (we are screening a yeast library). > > ------------------------------------ Theresa C. Swayne, Ph.D. Manager, Confocal and Specialized Microscopy Shared Resource Herbert Irving Comprehensive Cancer Center, Columbia University 1130 Saint Nicholas Ave, 222A New York, NY 10032 212-851-4613 [hidden email] http://hiccc.columbia.edu/research/sharedresources/confocal |
Free forum by Nabble | Edit this page |