We have recently had preliminary feedback from a number of independent
sources that show much reduced bleaching when a sample is excited using an
LED source rather than a Hg bulb. These tests, carried out with identical
powers in the excitation bandpass region, showed that imaging could be
carried out for up to three times longer.
On live tests cells were seen to be still living happily after being
exposed for twice the time it took to kill the cells under Hg excitation.
We are still gathering information on this and intend to publish in due
course.  
In the meantime I will be happy to discuss offline and would be interested
in hearing from anyone who has seen similar results.

Best Regards,

Gerry

Gerard Whoriskey
Development Engineer
CoolLED Ltd
CIL House
Charlton Road
Andover
Hampshire
SP10 3JL
 
Mob: 07789535762
Tel: +44 (0) 1264 321321
Dir: +44 (0)1264 320984
web site: www.coolled.com
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I don't understand this. The only explanation I can think of is that
that the excitation wavelengths are _not_ the same in the two cases.(if
they were and the power is the same the photon flux is the same).  Any
other comments/views?

Regards.

Gerard Whoriskey wrote:

> Commercial interest.
>
> We have recently had preliminary feedback from a number of independent
> sources that show much reduced bleaching when a sample is excited using an
> LED source rather than a Hg bulb. These tests, carried out with identical
> powers in the excitation bandpass region, showed that imaging could be
> carried out for up to three times longer.
> On live tests cells were seen to be still living happily after being
> exposed for twice the time it took to kill the cells under Hg excitation.
> We are still gathering information on this and intend to publish in due
> course.
>  
Ignatius, Mike Ignatius, Mike
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

My understanding is that it is critical that all UV light be
removed/blocked as Hg bulbs produce lots of UV.  Diodes don't have this
concern.

Not all scopes can claim 100% UV block produced by Hg or xenon bulbs.
But when it is done (for example AP/Delta Vision Scopes), cell viability
along with dye stability is greatly enhanced.

Mike Ignatius


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Mark Cannell
Sent: Tuesday, March 11, 2008 2:02 PM
To: [hidden email]
Subject: Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I don't understand this. The only explanation I can think of is that
that the excitation wavelengths are _not_ the same in the two cases.(if
they were and the power is the same the photon flux is the same).  Any
other comments/views?

Regards.

Gerard Whoriskey wrote:
> Commercial interest.
>
> We have recently had preliminary feedback from a number of independent

> sources that show much reduced bleaching when a sample is excited
using an
> LED source rather than a Hg bulb. These tests, carried out with
identical
> powers in the excitation bandpass region, showed that imaging could be

> carried out for up to three times longer.
> On live tests cells were seen to be still living happily after being
> exposed for twice the time it took to kill the cells under Hg
excitation.
> We are still gathering information on this and intend to publish in
due
> course.
>  
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

In reply to this post by Mark Cannell
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I think we'd need to be told a whole lot more -
what the dye was, what the bandpass was, etc.
A mercury lamp has very powerful spectral lines,
so (especially in the green) the measured power
is not spread across the spectral range but
concentrated at one wavelength.  If this happens
to hit a particular transition the effect on the
fluorochrome could be different from excitation
by a broad band.

                                       Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell
Sent: Wednesday, 12 March 2008 8:02 AM
To: [hidden email]
Subject: Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I don't understand this. The only explanation I can think of is that that the excitation wavelengths are _not_ the same in the two cases.(if they were and the power is the same the photon flux is the same).  Any other comments/views?

Regards.

Gerard Whoriskey wrote:

> Commercial interest.
>
> We have recently had preliminary feedback from a number of independent
> sources that show much reduced bleaching when a sample is excited
> using an LED source rather than a Hg bulb. These tests, carried out
> with identical powers in the excitation bandpass region, showed that
> imaging could be carried out for up to three times longer.
> On live tests cells were seen to be still living happily after being
> exposed for twice the time it took to kill the cells under Hg excitation.
> We are still gathering information on this and intend to publish in
> due course.
>  

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.518 / Virus Database: 269.21.7/1324 - Release Date: 10/03/2008 7:27 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.518 / Virus Database: 269.21.7/1325 - Release Date: 11/03/2008 1:41 PM
 
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

In reply to this post by Mark Cannell
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>I don't understand this. The only explanation I can think of is that
>that the excitation wavelengths are _not_ the same in the two
>cases.(if they were and the power is the same the photon flux is the
>same).  Any other comments/views?
>
>Regards.
>
>Gerard Whoriskey wrote:
>>Commercial interest.
>>
>>We have recently had preliminary feedback from a number of
>>independent sources that show much reduced bleaching when a sample
>>is excited using an LED source rather than a Hg bulb. These tests,
>>carried out with identical powers in the excitation bandpass
>>region, showed that imaging could be carried out for up to three
>>times longer.
>>On live tests cells were seen to be still living happily after
>>being exposed for twice the time it took to kill the cells under Hg
>>excitation. We are still gathering information on this and intend
>>to publish in due course.

Need to check that both systems have UV absorption filters or
optics... If you have quartz optics, and no UV filter, then you are
relying on the blocking in the barrier filter, which is not always
enough.

LEDs have no UV but Hg does.

Jim Pawley
--
               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/             Applications due by March 15, 2008
               "If it ain't diffraction, it must be statistics." Anon.
Stanislav Vitha Stanislav Vitha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

In reply to this post by Gerard Whoriskey-3
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:  
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the CoolLED system, and it is not operating in strobe mode.  It is continuous illumination.  The UV question is interesting.  Is there an easy way to test if there is UV leaking through the epi filter set with the mercury burner?  Dave

On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:   
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Meet about specs

Hi Dave,
 
Got the RFP.  Thanks.
 
Any chance of getting together at the end of next week (Thu. or Fri.) to discuss CSU options?
 
Gary
Ignatius, Mike Ignatius, Mike
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

In reply to this post by Knecht, David
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Pulsed Illumination:   Prior to Stefan Hell's truly elegant paper this paper discussed making a cheaper version of pulsed diode to reduce photo-fading/toxicity.
 
Stroboscopic illumination using light-emitting diodes reduces phototoxicity in fluorescence cell imaging.
by Nishigaki T, Wood CD, Shiba K, Baba SA, Darszon A. in <A href="javascript:AL_get(this, 'jour', 'Biotechniques.');">Biotechniques. 2006 Aug;41(2):191-7.
 
The effects weren't as dramatic, but the benefits were very clear and at a fraction of the cost. 
 
It is worth noting as well that the spinning discs, fast scanners, fast galvos, etc, effectively deliver a decent form of the T-Relaxation light form that Dr. Hell recommends in his paper. 
 
As for UV leak, J. Nordberg at U Mass had an interesting ASCB abstract/poster in 2006.  UV was effecting his experiments, so he looked at the amount of leak.  My apologies to Dr. Nordberg if I misrepresent his findings, but my notes say that i) photon flux from UV light vs tungsten was 30,000 times greater.  (Any of us failing to have filter block in place when we look through the eye pieces with an Hg lamp on know this personally.)  Still from my notes, so these numbers need verification from the pros out there, but standard filters block 99% or 2 orders of magnitude of light while optimized filters can cut another 4 orders more.  I liken this to watching soccer/football under the lights at night - like a tungsten lamp on a slide.  Imagine then if the light was turned up 300 times (30,000 less 99%), this is what the extent of UV leak might look like to our cells.  We don't see it as it is UV, but they (and the dyes) feel it.
 
Mike Ignatius,
 
Molecular Probes/Invitrogen


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht
Sent: Wednesday, March 12, 2008 2:56 PM
To: [hidden email]
Subject: Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the CoolLED system, and it is not operating in strobe mode.  It is continuous illumination.  The UV question is interesting.  Is there an easy way to test if there is UV leaking through the epi filter set with the mercury burner?  Dave

On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:   
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Benoist, Christophe Benoist, Christophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Meet about specs

In reply to this post by Gary Laevsky-2
Fascinating...

        -----Original Message-----
        From: Confocal Microscopy List on behalf of Gary Laevsky
        Sent: Wed 3/12/2008 6:23 PM
        To: [hidden email]
        Cc:
        Subject: Meet about specs
       
       

        Hi Dave,
       
        Got the RFP.  Thanks.
       
        Any chance of getting together at the end of next week (Thu. or Fri.) to discuss CSU options?
       
        Gary
       

Mario-2 Mario-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

In reply to this post by Knecht, David
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: Fluorophore bleaching by excitation light sources
Dave,

Depends on what you mean by easy. To narrow the problem a bit, the important issue is how much UV reaches the sample. Any other filter is more or less irrelevant with respects to cell viability, unless there is so much back scattered UV that extra background noise must be averaged out with extra exposures.

Ignoring the latter, one might deal with the former by using a test object that is excited by UV. There are quite a few options including UV sensitive inorganic phosphors that can be purchased as fine powders or disks. The latter compounds can have luminescent lifetimes that range from as short as 40 nsec. to milliseconds. The latter often use a lanthanide such as europium (red) or terbium (blue, green, red) with sharp emission lines. Some have broad emissions such as P31, a ZnS based phosphor. Another possibility would be some organic fluorophores including AMCA or well saturated nuclei stained with DAPI or Hoechst (say 3 ug dye /ml of cells).

A slide made with a phosphorescent disk will last for years. As I recall, some EM suppliers sell phosphorescent disks and related materials for use in cathodoluminescent detectors. I have used or made all of the above options in one form or another. A further option is to use a fiber optic couple micro spectrofluorometer. I like the USB ported version from Ocean Optics, Fl. (no financial interest). They can provide a cable that is capped with a collection lens that carries the excitation light from the microscope objective to the spectrometer which uses a grating and linear CCD. As I recall, there are a few grating options one of which provides sensitivity down near 220 nm on up to 700 nm.

Anyway, some of these approaches are crude but they will tell you if there is significant UV bleed through.

Mario


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the CoolLED system, and it is not operating in strobe mode.  It is continuous illumination.  The UV question is interesting.  Is there an easy way to test if there is UV leaking through the epi filter set with the mercury burner?  Dave

On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:   
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha

Dr. David Knecht  
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


-- 
________________________________________________________________________________
Mario M. Moronne, Ph.D.
Cairn research ltd Cairn research ltd
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fluorophore bleaching by excitation light sources

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: Fluorophore bleaching by excitation light sources

There is of course nothing magic about photons produced by LEDs rather than Hg or other arc lamps, so any apparent reduction in photobleaching when using LEDs for illumination must presumably relate to the spectral characteristics of the excitation light.  The spectral characteristics of Hg (as opposed to Xe) arc lamps are very uneven, so very effective filtering may be necessary to block out-of-band spectral peaks that might cause increased phtobleaching.

The Donnert et al paper is very interesting.  Basically their suggestion is that a small proportion of excited flurophores goes to a much longer-lived triplet state, with a lifetime of around a microsecond, rather than undergoing "normal" fluorescence emission with a lifetime of a few nanoseconds.  Therefore, as the light intensity is increased, more and more of the fluorophores are pushed into the triplet state on account of its much greater lifetime.  If (as is likely) the triplet state fluorophores are destroyed (i.e. bleached) if they absorb a second photon, you then have a nice model to explain why high light intensities cause disproportionately greater photobleaching (it's a square-law effect, as two photons are involved).

The Donnert et al paper used pulsed illumination as a rather neat way of showing this effect, but in our opinion pulsed illumination is unlikely to offer any SIGNIFICANT benefit over continuous illumination of the same average intensity.  The important criterion is that the illumination intensity should be such that on average any one fluorophore molecule is excited at a rate that is low compared with the lifetime of its triplet state, so that it never gets the chance to absorb a second photon.  Whether all the fluorophores are excited at the same time by a very brief pulse, or asynchronously by the same average light intensity, is unlikely to make a big difference in our opinion, but of course it would be worth checking this out experimentally.  Our OptoLED product is capable in principle of being switched at these sort of rates, but with pulsed illumination of any LED source (i.e. the LED off most of the time) you probably wouldn't get enough light, so in practice we think that you'd need laser illumination to check this one out.

Cairn OptoLED - http://www.cairn-research.co.uk/Products


Dr. Martin Thomas

Managing Director
Cairn Research Ltd
Graveney Road
Faversham
Kent, ME13 8UP
UK


www.cairn-research.co.uk
[hidden email]

Tel: + 44 (0)1795 590140
Fax: + 44 (0)1795 594510


>Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: Fluorophore
> bleaching by excitation light sources
>Dave,

>
>Depends on what you mean by easy. To narrow the problem a bit, the important
> issue is how much UV reaches the sample. Any other filter is more or less
> irrelevant with respects to cell viability, unless there is so muchback
> scattered UV that extra background noise must be averaged out with extra
> exposures.

>
>Ignoring the latter, one might deal with the former by using a test object that
> is excited by UV. There are quite a few options including UV sensitive
> inorganic phosphors that can be purchased as fine powders or disks.The latter
> compounds can have luminescent lifetimes that range from as short as 40 nsec.
> to milliseconds. The latter often use a lanthanide such as europium (red) or
> terbium (blue, green, red) with sharp emission lines. Some have broad
> emissions such as P31, a ZnS based phosphor. Another possibility would be some
> organic fluorophores including AMCA or well saturated nuclei stained with DAPI
> or Hoechst (say 3 ug dye /ml of cells).

>
>A slide made with a phosphorescent disk will last for years. As I recall, some
> EM suppliers sell phosphorescent disks and related materials for use in
> cathodoluminescent detectors. I have used or made all of the above options in
> one form or another. A further option is to use a fiber optic couple micro
> spectrofluorometer. I like the USB ported version from Ocean Optics, Fl. (no
> financial interest). They can provide a cable that is capped with a collection
> lens that carries the excitation light from the microscope objective to the
> spectrometer which uses a grating and linear CCD. As I recall, there are a few
> grating options one of which provides sensitivity down near 220 nm on up to
> 700 nm.

>
>Anyway, some of these approaches are crude but they will tell you if there is
> significant UV bleed through.

>
>Mario

>

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the
> CoolLED system, and it is not operating in strobe mode.  It is continuous
> illumination.  The UV question is interesting.  Is there aneasy way to test if
> there is UV leaking through the epi filter set with the mercury burner?  Dave
> On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Regarding the difference between LED illumination and Hg lamp -
> Perhaps if the illumination is not continuous but is set up for strobe
> operation of the LED, this could allow dark state relaxation an prolong
> the life of the fluorophore (I am not sure the same would apply to the
> life of the cell). - The time between light pulses should be more than one
> microscecond.
> reference:   
> Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
> in fluorescence microscopy through dark-state relaxation. Nature Methods -
> 4, 81 - 86 (2007)
>
>
> Stan Vitha
>
> Dr. David Knecht   Department of Molecular and Cell Biology Co-head Flow
> Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd.
> University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax)

>

> --  
>_______________________________________________________________________________
>_
>Mario M. Moronne, Ph.D.
>
>cell (510) 367-8497
>
> [hidden email]
> [hidden email]
> [hidden email]

----------------------- Original Message -----------------------
  
From: Mario Moronne [hidden email]
Date: Wed, 12 Mar 2008 17:17:55 -0700
Subject: Re: Fluorophore bleaching by excitation light sources
  
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dave,

Depends on what you mean by easy. To narrow the problem a bit, the important issue is how much UV reaches the sample. Any other filter is more or less irrelevant with respects to cell viability, unless there is so much back scattered UV that extra background noise must be averaged out with extra exposures.

Ignoring the latter, one might deal with the former by using a test object that is excited by UV. There are quite a few options including UV sensitive inorganic phosphors that can be purchased as fine powders or disks. The latter compounds can have luminescent lifetimes that range from as short as 40 nsec. to milliseconds. The latter often use a lanthanide such as europium (red) or terbium (blue, green, red) with sharp emission lines. Some have broad emissions such as P31, a ZnS based phosphor. Another possibility would be some organic fluorophores including AMCA or well saturated nuclei stained with DAPI or Hoechst (say 3 ug dye /ml of cells).

A slide made with a phosphorescent disk will last for years. As I recall, some EM suppliers sell phosphorescent disks and related materials for use in cathodoluminescent detectors. I have used or made all of the above options in one form or another. A further option is to use a fiber optic couple micro spectrofluorometer. I like the USB ported version from Ocean Optics, Fl. (no financial interest). They can provide a cable that is capped with a collection lens that carries the excitation light from the microscope objective to the spectrometer which uses a grating and linear CCD. As I recall, there are a few grating options one of which provides sensitivity down near 220 nm on up to 700 nm.

Anyway, some of these approaches are crude but they will tell you if there is significant UV bleed through.

Mario


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the CoolLED system, and it is not operating in strobe mode.  It is continuous illumination.  The UV question is interesting.  Is there an easy way to test if there is UV leaking through the epi filter set with the mercury burner?  Dave

On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:   
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha

Dr. David Knecht  
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


-- 
________________________________________________________________________________
Mario M. Moronne, Ph.D.