To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I was wondering if people could recommend far-red fluorophores for immunofluorescence that are excitable with a 633nm laser, I have tried Alexa633 with little success. Have other people had more luck with secondary antibodies conjugated to other fluorophores such as Cy5, DyLight649, Alexa647.....? Thanks in advance, Andrew |
Tamara Howard |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We've used DyLight 649 - works beautifully on our SP5. Tamara ................................................... Tamara Howard Dept. of Cell Biology & Physiology University of New Mexico Albuquerque, NM ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Andrew Lindsay [[hidden email]] Sent: Friday, July 12, 2013 8:29 AM To: [hidden email] Subject: Fluorophores for 633nm laser ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I was wondering if people could recommend far-red fluorophores for immunofluorescence that are excitable with a 633nm laser, I have tried Alexa633 with little success. Have other people had more luck with secondary antibodies conjugated to other fluorophores such as Cy5, DyLight649, Alexa647.....? Thanks in advance, Andrew |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We have had luck with Alexa 647, Dylight649, as well as APC conjugates. ________________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Andrew Lindsay <[hidden email]> Sent: Friday, July 12, 2013 10:29 AM To: [hidden email] Subject: Fluorophores for 633nm laser ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I was wondering if people could recommend far-red fluorophores for immunofluorescence that are excitable with a 633nm laser, I have tried Alexa633 with little success. Have other people had more luck with secondary antibodies conjugated to other fluorophores such as Cy5, DyLight649, Alexa647.....? Thanks in advance, Andrew |
Gabriel Lapointe-4 |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Andrew, Personally, Alexa647 is one of my favorite fluorophore. It is very stable and I haven't notice any significant cross talk with Alexa568. However, you have to remember that any far red dye are very hard to see with the naked eye (some say impossible) and that the quantum efficiency of a lot of cameras and PMT is usually much lower at 700nm than at 500nm. However, there is no auto-fluorescence in the far-red region and the 633nm is much less toxic so it is usually not an issue to increase exposure time or excitation power. Sincerely *Gabriel Lapointe, M.Sc.* Lab Manager / Microscopy Specialist Concordia University, Biology Department 7141 Sherbrooke St. West SP 534 Montréal QC H4B 1R6 Canada Lab : (514) 848-2424 x5988 Office : (514) 848-2424 x3008 Fax : (514) 848-2881 Cell : (514) 278-0247 [hidden email] cmac.concordia.ca http://gabriellapointe.ca 2013/7/12 Andrew Lindsay <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All, > > I was wondering if people could recommend far-red fluorophores for > immunofluorescence that are excitable with a 633nm laser, I have tried > Alexa633 with little success. Have other people had more luck with > secondary > antibodies conjugated to other fluorophores such as Cy5, DyLight649, > Alexa647.....? > > Thanks in advance, > Andrew > |
Cedric Espenel |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Andrew, You should try Atto647N. Best Cedric -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andrew Lindsay Sent: vendredi 12 juillet 2013 10:30 To: [hidden email] Subject: Fluorophores for 633nm laser ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I was wondering if people could recommend far-red fluorophores for immunofluorescence that are excitable with a 633nm laser, I have tried Alexa633 with little success. Have other people had more luck with secondary antibodies conjugated to other fluorophores such as Cy5, DyLight649, Alexa647.....? Thanks in advance, Andrew |
Rosemary.White |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The Cy5 and Cy5.5 dyes are great and emission can be separated on spectral confocals if both used together. cheers, Rosemary Dr Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 E [hidden email] On 13/07/13 12:29 AM, "Andrew Lindsay" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi All, > >I was wondering if people could recommend far-red fluorophores for >immunofluorescence that are excitable with a 633nm laser, I have tried >Alexa633 with little success. Have other people had more luck with >secondary >antibodies conjugated to other fluorophores such as Cy5, DyLight649, >Alexa647.....? > >Thanks in advance, >Andrew |
Franziska Curdt |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Alexa should work, what filters do you use? Atto 633 and 647 are more photostable, as well as Abberior STAR 635. I cannot recommend Cy5. You find a list of photostable dyes here. http://nanobiophotonics.mpibpc.mpg.de/old/dyes/ Best regards, F.Curdt ________________________________________ Von: Confocal Microscopy List [[hidden email]] im Auftrag von Andrew Lindsay [[hidden email]] Gesendet: Freitag, 12. Juli 2013 16:29 An: [hidden email] Betreff: Fluorophores for 633nm laser ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I was wondering if people could recommend far-red fluorophores for immunofluorescence that are excitable with a 633nm laser, I have tried Alexa633 with little success. Have other people had more luck with secondary antibodies conjugated to other fluorophores such as Cy5, DyLight649, Alexa647.....? Thanks in advance, Andrew |
Steffen Dietzel |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Cy5 worked great for us in the past. But so did Alexa633. Both mounted in Vectashield. As was pointed out, the human eye is a miserable detector in that range, so expect to see nothing down the eye piece. Steffen On 12.07.2013 16:29, Andrew Lindsay wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All, > > I was wondering if people could recommend far-red fluorophores for > immunofluorescence that are excitable with a 633nm laser, I have tried > Alexa633 with little success. Have other people had more luck with secondary > antibodies conjugated to other fluorophores such as Cy5, DyLight649, > Alexa647.....? > > Thanks in advance, > Andrew > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
George McNamara |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear listserv, Lee et al 2013 reports that heavy water (D2O) doubes the quantum yield for some fluorophores: * Brighter dyes in heavy water:* A simple and cost-effective method increases the brightness of a whole class of commonly used red-emitting fluorophores, including ATTO655, ATTO680, and ATTO700. Replacing water (H_2 O) by heavy water (D_2 O) in the imaging buffer doubles the fluorescence quantum yield of these dyes and significantly improves the localization precision in super-resolution imaging. I looked up some of the papers cited in their ref 4 (main text) and found that Lucifer Yellow (always reminds me of the Rolling Stones song, Sympathy for the Devil) is another that does doubles. In the (free to access) supplemental file http://onlinelibrary.wiley.com/doi/10.1002/anie.201302341/suppinfo Their Table S4 shows a small increase for Alexa Fluor 647 (0.44 to 0.40) as well as ATTO647N (0.51 to 0.66). Note that their precision localization imaging experiments were done with ascorbic acid to induce dark state(s). Other additives may be preferable for imaging for maximum brightness. I looked at Sigma-Aldrich.com for pricing - heavy water is about $1/gram ~$1/mL. Since a typical coverglass-slide preparation is 10 uL, or a typical imaging dish's "imaging volume" is about 200 uL (glassbottomdishes 20 mm diameter opening), price is trivial. Enjoy, George On 7/15/2013 7:37 AM, Steffen Dietzel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Cy5 worked great for us in the past. But so did Alexa633. Both mounted > in Vectashield. As was pointed out, the human eye is a miserable > detector in that range, so expect to see nothing down the eye piece. > > Steffen > > On 12.07.2013 16:29, Andrew Lindsay wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi All, >> >> I was wondering if people could recommend far-red fluorophores for >> immunofluorescence that are excitable with a 633nm laser, I have tried >> Alexa633 with little success. Have other people had more luck with >> secondary >> antibodies conjugated to other fluorophores such as Cy5, DyLight649, >> Alexa647.....? >> >> Thanks in advance, >> Andrew >> > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 http://works.bepress.com/gmcnamara/26/ |
JOEL B. SHEFFIELD |
In reply to this post by Steffen Dietzel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I've been following this discussion, and wondering if it would help for you to specify the detection system that you will be using for the far red signal. As suggested by several people, the problems that you are having might very well be due to that component of your optical system. Joel On Mon, Jul 15, 2013 at 8:37 AM, Steffen Dietzel <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/**wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > ***** > > Cy5 worked great for us in the past. But so did Alexa633. Both mounted in > Vectashield. As was pointed out, the human eye is a miserable detector in > that range, so expect to see nothing down the eye piece. > > Steffen > > On 12.07.2013 16:29, Andrew Lindsay wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/**wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> >> ***** >> >> Hi All, >> >> I was wondering if people could recommend far-red fluorophores for >> immunofluorescence that are excitable with a 633nm laser, I have tried >> Alexa633 with little success. Have other people had more luck with >> secondary >> antibodies conjugated to other fluorophores such as Cy5, DyLight649, >> Alexa647.....? >> >> Thanks in advance, >> Andrew >> >> > > -- > ------------------------------**------------------------------ > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Walter-Brendel-Zentrum für experimentelle Medizin (WBex) > Head of light microscopy > > Mail room: > Marchioninistr. 15, D-81377 München > > Building location: > Marchioninistr. 27, München-Großhadern > -- Joel B. Sheffield, Ph.D Department of Biology Temple University Philadelphia, PA 19122 Voice: 215 204 8839 e-mail: [hidden email] URL: http://astro.temple.edu/~jbs |
Andrew Lindsay |
In reply to this post by Andrew Lindsay
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Thanks for all your replies and based on the advice I am leaning towards Alexa647 or DyLight649. To clarify I will be using the Meta detector on a Zeiss LSM510 to detect the far-red signal, and to further clarify I intend to use the secondary antibody in combination with GFP and Cy3 to do three-color immunofluorescence on mammalian cell lines. Andrew |
Vladimir Zhukarev |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Andrew, You might find it really hard to use NIR emitting fluorophores no matter how 'bright' they are, specifically because your detector is of a LASM510 type. Segmented detectors, much more that regular PMT'es, are known to be not optimal for imaging in this spectrum range (check with Hamamatsu - source of all PMT'es for confocal microscopy). By the way this why the 'spectral scale on your machine ends at 710 nm as opposed to 800 nm with other spectral confocals. But you probably noticed this already. Best Vladimir Vladimir Zhukarev, Ph.D Advanced Microscopy Consulting Life Science and Bio-Medical Applications 'From Image-to Discovery!' [hidden email] 610.350.1510 www.from-image-to-discovery.com --------- Original Message --------- Subject: Re: Fluorophores for 633nm laser From: "Andrew Lindsay" <[hidden email]> Date: 7/16/13 1:10 am To: [hidden email] ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Thanks for all your replies and based on the advice I am leaning towards Alexa647 or DyLight649. To clarify I will be using the Meta detector on a Zeiss LSM510 to detect the far-red signal, and to further clarify I intend to use the secondary antibody in combination with GFP and Cy3 to do three-color immunofluorescence on mammalian cell lines. Andrew |
Vitaly Boyko |
We are looking for a highly sensitive EM-CCD camera but with
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, We are looking for a highly sensitive EM-CCD camera but with small pixels (8 um) and the back-thinned one. I have already received endless suggestions around "scientific" CMOS cameras, but I am a bit stubborn, as you have probably already realized. And/or, is there anything cooler then CMOS? I like Canon 1D X and Nikon D800, and the latter seems to be a proper place for CMOS... Cheers, Vitaly |
Philippe clemenceau |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Commercial response : Hi Vitaly, I don't think it exists unfortunately, and you might have to settle on either small pixels or back thinned sensitivity if you stay with EMCCD. I sell the FALCON EMCCD camera from Raptor Photonics. It is not back thinned, it has 8 x 8 micron pixels, QE goes up to 65 % and it is a 1K x 1K sensor. Sensitivity allows good performance for most applications and it is a very cost effective camera. If that is of interest to you, we can talk about it off the list ! Regards, Philippe Clémenceau, Division Manager, MS in Optical Science Imagine Optic Inc./Axiom Optics Ph:+1 (617) 401 2198 Cell: + 1 (310) 597 1347 1 Broadway, 14th floor Cambridge, MA 02142 www.axiomoptics.com Metrology, Adaptive Optics, Scientific Imaging, Lasers -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vitaly Boyko Sent: Tuesday, July 16, 2013 11:04 AM To: [hidden email] Subject: EM-CCD with 8 um/pixel back-thinned We are looking for a highly sensitive EM-CCD camera but with ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, We are looking for a highly sensitive EM-CCD camera but with small pixels (8 um) and the back-thinned one. I have already received endless suggestions around "scientific" CMOS cameras, but I am a bit stubborn, as you have probably already realized. And/or, is there anything cooler then CMOS? I like Canon 1D X and Nikon D800, and the latter seems to be a proper place for CMOS... Cheers, Vitaly |
James Pawley |
In reply to this post by Vitaly Boyko
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >We are looking for a highly sensitive EM-CCD camera but with >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear All, We are looking for a highly sensitive >EM-CCD camera but with small pixels (8 um) and >the back-thinned one. I have already received >endless suggestions around "scientific" CMOS >cameras, but I am a bit stubborn, as you have >probably already realized. And/or, is there >anything cooler then CMOS? I like Canon 1D X >and Nikon D800, and the latter seems to be a >proper place for CMOS... Cheers, Vitaly Dear Vitaliy, I will be watching answers to your request with some interest because I don't know of such a chip (yet!). I think that we need to keep a bit of geometry in mind. Let's assume that the Si has been thinned to 8µm. Photoelectrons created near the so-called "back" (but actually now the front) surface will see the attraction from the nearest, positive charge-transfer electrode, about 8µm away. However, if they are created in the border region where two electrodes are about equidistant, there will be uncertainty as to which one to go to. This uncertainty will increase as the pixel dimensions are reduced. The result is to reduce the "resolution" of the sensor below that which one would assume from the pixel count. (If you look carefully in the specs of small-dimension CCDs, you will see that the red resolution is lower than that for blue.) In other words, just making smaller pixels may not give you better resolution, just more measurements to make in the same time (and hence, higher read bandwidth and higher read noise). This is specially true for red dyes where otherwise, the BI-CCDs have a particular advantage over PMTs etc. Of course, this resolution loss also happens to some extent in front-illuminated chips, but as most photons are now absorbed pretty near the charge transfer electrodes, it is less of a problem. Of course, you could thin more, but this would reduce the QE in the red as red photons are absorbed less strongly. The drop-off of BI-CCD response in the red is largely due to this fact. There may be a lot of other engineering considerations but whatever they are, it seems that BI CCDs usually have large pixels. Hope I am wrong, Jim Pawley -- James and Christine Pawley, 5446 Burley Place (PO Box 2348), Sechelt, BC, Canada, V0N3A0, Phone 604-885-0840, email <[hidden email]> NEW! NEW! AND DIFFERENT Cell (when I remember to turn it on!) 1-604-989-6146 |