Stanislav Vitha |
|
|
Hello,
Some FRAP protocols recommend using formaldehyde-fixed samples as controls to check for reversible photobleaching of the fluorescent protein, reasoning that there is no diffusion in fixed samples, thus the recovery of fluorescence in the bleached area is attributed to reversible photobleaching. From that follows that if no FRAP is observed in the fixed samples, reversible photobleaching is negligible, and FRAP in the live sample represents protein turnover (diffusion). Is this reasoning correct? I wonder if it possible that fixation (crosslinking) alters the photochemical properties of the fluorescent protein, eliminating thus reversible photobleaching even if the protein is normally capable of it. Thanks for your input. Stanislav Vitha Microscopy and Imaging Center Texas A&M University |
|
Mark Cannell |
|
|
A very valid concern. If seems to me that if photobleaching were due to
oxidation, some reversal might be due to the reducing agents in the cell that will be be destroyed by fixation. Cheers Stanislav Vitha wrote: > Hello, > > Some FRAP protocols recommend using formaldehyde-fixed samples as controls > to check for reversible photobleaching of the fluorescent protein, reasoning > that there is no diffusion in fixed samples, thus the recovery of > fluorescence in the bleached area is attributed to reversible > photobleaching. From that follows that if no FRAP is observed in the fixed > samples, reversible photobleaching is negligible, and FRAP in the live > sample represents protein turnover (diffusion). > > Is this reasoning correct? I wonder if it possible that fixation > (crosslinking) alters the photochemical properties of the fluorescent > protein, eliminating thus reversible photobleaching even if the protein is > normally capable of it. > > Thanks for your input. > > Stanislav Vitha > Microscopy and Imaging Center > Texas A&M University > > |
|
Arne Seitz |
|
|
In reply to this post by Stanislav Vitha
Hi,
what you call reversible photobleaching is most likely accumulation of non-fluorescent triplet states. They have a much longer lifetime than the excited singlet states. Fixation is definitively changing the properties of GFP. Organic solvent even seems to destroy GFP completely. http://www.bio.davidson.edu/courses/Molbio/restricted/02GFPwow/GFPwowpg1.html How the lifetime of triplet states is influenced by fixation is to my knowledge not investigates in detail but I would assume that they are definitely influenced by their chemical environment and thereby also by fixation. Thus doing FRAP experiments on fixed samples is maybe not the best control. IMHO it is much better to prolong the number of postbleach frames until you have a equilibrium between electronic ground state, excited singlet state and triplet state. Depending on your imaging conditions this might take 20-30 frames. Please keep in mind that the proportion of triplet states will depend on the imaging conditions (pixel dwell time, laser intensity etc.) so that it might also be worth playing around with these parameters in order to have less triplet states. Cheers Arne -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Stanislav Vitha Sent: jeudi 2 septembre 2010 00:09 To: [hidden email] Subject: Formaldehyde-fixed samples as FRAP controls for reversible GFP, YFP photobleaching? Hello, Some FRAP protocols recommend using formaldehyde-fixed samples as controls to check for reversible photobleaching of the fluorescent protein, reasoning that there is no diffusion in fixed samples, thus the recovery of fluorescence in the bleached area is attributed to reversible photobleaching. From that follows that if no FRAP is observed in the fixed samples, reversible photobleaching is negligible, and FRAP in the live sample represents protein turnover (diffusion). Is this reasoning correct? I wonder if it possible that fixation (crosslinking) alters the photochemical properties of the fluorescent protein, eliminating thus reversible photobleaching even if the protein is normally capable of it. Thanks for your input. Stanislav Vitha Microscopy and Imaging Center Texas A&M University |
|
ChrisWilms |
|
|
In reply to this post by Stanislav Vitha
Reversible "photobleaching" of XFPs has been described previously and
it appears to be a speciality of fluorescent proteins. I'm not sure it's simply a matter of triplet state accumulation. No matter what it is, I do agree that fixation will interfere with this process. My advice would be to move to an XFP that is less prone to darkstates. Best, Christian On/off blinking and switching behaviour of single molecules of green fluorescent protein Dickson, R. M. and Cubitt, A. B. and Tsien, R. Y. and Moerner, W. E. Nature 388 (1997), p355--358 Imaging individual green fluorescent proteins Pierce, D. W. and Hom-Booher, N. and Vale, R. D. Nature 388 (1997), p338--338 Fluorescence correlation spectroscopy reveals fast optical excitation- driven intramolecular dynamics of yellow fluorescent proteins Schwille, P. and Kummer, S. and Heikal, A. A. and Moerner, W. E. and Webb, W. W. PNAS 97 (2000), p151-156 > what you call reversible photobleaching is most likely accumulation > of non-fluorescent triplet states. They have a much longer lifetime > than the excited singlet states. |
|
Gert van Cappellen-2 |
|
|
In reply to this post by Stanislav Vitha
When you start to excite GFP's (with a high laser power) you will drive
a part of the molecules in the off-state. The molecules will remain in this state for a couple of seconds. This effect can be measured both by molecules that are mainly bound, for instance H2b-GFP as in fixed samples. Off course fixation might influence the properties of the molecules in the off state. Good FRAP analysis software takes this off state into account. Hope this helps a bit, Gert van Cappellen Stanislav Vitha schreef: > Hello, > > Some FRAP protocols recommend using formaldehyde-fixed samples as controls > to check for reversible photobleaching of the fluorescent protein, reasoning > that there is no diffusion in fixed samples, thus the recovery of > fluorescence in the bleached area is attributed to reversible > photobleaching. From that follows that if no FRAP is observed in the fixed > samples, reversible photobleaching is negligible, and FRAP in the live > sample represents protein turnover (diffusion). > > Is this reasoning correct? I wonder if it possible that fixation > (crosslinking) alters the photochemical properties of the fluorescent > protein, eliminating thus reversible photobleaching even if the protein is > normally capable of it. > > Thanks for your input. > > Stanislav Vitha > Microscopy and Imaging Center > Texas A&M University > > > |