It's tricky but we've done it using the Photonic Instruments Micropoint FRAP laser and collecting widefield images with a Nikon TE2000 (100x 1.4 NA). You definitely need a diffraction limited spot.
Jennifer
On 9/13/07, Elaine Kunze <[hidden email]> wrote:
Search the CONFOCAL archive at <a onclick="return top.js.OpenExtLink(window,event,this)" href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank">
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Has anyone tried FRAP in bacteia? I have a user who wants to FRAP GFP. We have an Olympus 1000 and I'm not sue we can get the spot size small enough. Hints/Comments/Discouraging words?