Login  Register

Fura2 imaging with Olympus TIRF system

classic Classic list List threaded Threaded
2 messages Options Options
Embed post
Permalink
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Fura2 imaging with Olympus TIRF system

Not a confocal topic but maybe someone knows:
I want to image fura2 with 340/380 nm excitation using our Olympus TIRF
system (cellR system, IX81) in widefield mode. The problem is that using the
Hg/Xe burner in the MT20 lamp, i have 20 times more power at 380 than 340
out of the objective, but a more reasonable 3 times out of the fibre. I have
a UApo/340 objective on loan, which should be suitable for 340 nm. The local
Olympus representative mentioned that the mirror in the L-shaped illuminator
can cut off some of the UV light. My question: has anyone experienced such a
problem and is it possible to change the mirror in the illuminator to a UV
reflecting one? The lens in the TIRF part of the illuminator is suitable for
405 nm, what is the transmission at 340 nm? Any suggestions welcome.
Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Fura2 imaging with Olympus TIRF system

If you are going to use an optical fibre, you should also check what wavelengths
can be passed by the fibre. I know that 405 nanometers is on the edge of what my
fibre optic can pass.
Cheers,
Lisa


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute
44 Binney St.; JF 215
Boston, MA 02115
Office phone: 617-582-8824
Fax: 617-582-8750
Office location: Jimmy Fund Bldg. 220B

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Andreas Bruckbauer
Sent: Friday, June 19, 2009 6:54 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] Fura2 imaging with Olympus TIRF system

Not a confocal topic but maybe someone knows:
I want to image fura2 with 340/380 nm excitation using our Olympus TIRF
system (cellR system, IX81) in widefield mode. The problem is that using the
Hg/Xe burner in the MT20 lamp, i have 20 times more power at 380 than 340
out of the objective, but a more reasonable 3 times out of the fibre. I have
a UApo/340 objective on loan, which should be suitable for 340 nm. The local
Olympus representative mentioned that the mirror in the L-shaped illuminator
can cut off some of the UV light. My question: has anyone experienced such a
problem and is it possible to change the mirror in the illuminator to a UV
reflecting one? The lens in the TIRF part of the illuminator is suitable for
405 nm, what is the transmission at 340 nm? Any suggestions welcome.


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.