Fwd: objective lens laser damage

classic Classic list List threaded Threaded
♦
♦
1 message Options
Martin Wessendorf-2 Martin Wessendorf-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Fwd: objective lens laser damage

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I was asked to post this note on behalf of Dr. Alexander Asanov--

Martin Wessendorf



-------- Forwarded Message --------
Subject: objective lens laser damage
Date: Fri, 27 May 2016 20:19:50 -0400
From: Alexander Asanov <[hidden email]>



Dear Sripad:

Inclusions and other inhomogeneities, such as bubbles and irregularities
of refractive index are inevitable imperfections for all optical
materials.  See for example Corning specifications for optical quality
silica
https://www.corning.com/media/worldwide/csm/documents/5bf092438c5546dfa9b08e423348317b.pdf.
Note that different grades contain different concentrations of inclusions.

We work with similar lasers and incidentally have similar problems with
lenses, couplers, and even recently lost a fiber cable, likely due to
its solarization; the fiber successfully served for several months
before. More often these problems happen at interfaces. Dirt, as Dr.
Wessendorf mentioned, is often an initiator. As soon as small damage
occurs, the problem typically escalates.  Lens glue appears to be the
second most frequent source. 405 nm is certainly more damaging to the
glue than 647 of the same power.

If you focus a 1W laser into a 10-micron spot, you will get  ~1 Mega
Watt/cm^2 intensity (10^6 W/cm2), which is huge, laser-cutting
intensity. Focusing 60-100 mW into  a 30-micron spot also produces an
intensity, which may be sufficient to burn the lens glue.

Best regards,

Alexander N. Asanov, Ph.D.
President,  TIRF Labs
106 Grendon Place
Cary, NC 27519
[hidden email] <mailto:[hidden email]>

www.tirf-labs.com <http://www.tirf-labs.com/> www.TIRFmicroscopy.com
<http://www.tirfmicroscopy.com/>

*Subject:*

       

Re: objective lens laser damage

*From:*

       

S Ram <[hidden email]>

*Reply-To:*

       

Confocal Microscopy List <[hidden email]>

*Date:*

       

Fri, 27 May 2016 14:31:22 -0700

*Content-Type:*

       

text/plain

*****

To join, leave or search the confocal microscopy listserv, go to:

*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

Post images on *http://www.imgur.com* <http://www.imgur.com/> and
include the link in your posting.

*****

The bubbles is something that I observed years ago (postdoc days) even with

dry lenses (20x PApo; Zeiss) that has never been illuminated with a laser.

At that time, we suspected that someone accidentally used oil on the dry

lens and forgot to wipe it off and was never noticed for a very long time.

It is possible that it could well be that something internal was damaged.

Sripad

On Fri, May 27, 2016 at 10:41 AM, John Oreopoulos <

[hidden email]> wrote:

> *****

> To join, leave or search the  confocal microscopy listserv, go to:

>*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> Post images on*http://www.imgur.com* <http://www.imgur.com/> and include the link in
your posting.

> *****

>

> Do you really need these high laser  powers for your application?

>

> John Oreopoulos

>

> > On May 27, 2016, at 11:42 AM,  Glen MacDonald <[hidden email]>

> wrote:

> >

> > *****

> > To join, leave or search the  confocal microscopy listserv, go to:

> >*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> > Post images on*http://www.imgur.com* <http://www.imgur.com/> and include the link in your

> posting.

> > *****

> >

> > Possibly the damage is to the  optical cement between the lens elements.

> I found epoxy to be very absorbtive  and could be damaged during 2-photon

> imaging.

> >

> > Regards,

> > Glen MacDonald

> > Digital Microscopy Center

> > Box 357923

> > University of Washington

> > Seattle, WA 98195-7923  USA

> > (206) 616-4156

> > [hidden email]

> >

> >

> >

> >

> >

> >

> >

> >

> >> On May 27, 2016, at 8:03  AM, Nuno Moreno <[hidden email]>

> wrote:

> >>

> >> *****

> >> To join, leave or search  the confocal microscopy listserv, go to:

> >>*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> >> Post images on*http://www.imgur.com* <http://www.imgur.com/> and include the link in your

> posting.

> >> *****

> >>

> >> Dear Craig

> >>

> >> I sent already a message  directly to Joshua with a similar statement.

> The problem is however when the  burning in not at the first lens but

> somewhere inside. In this case this  means that the objective was not

> assembled in a proper clean  environment and has dust or other particles

> inside. Actually here at the  institute we were surprised on how often this

> happens from 2 different  manufacturers. Maybe we were just unlucky. If that

> is the case I would ask for a  replacement free of charge. I must confess

> however that in Joshua case it does  not look like but the picture is not

> clear enough. I would advice you to  bring your objective to a stereoscope,

> start to focus on the first back  optical element and check whether this is

> superficial or not.

> >>

> >> Kind regards and good luck

> >> Nuno Moreno

> >>

> >>> On 27 May 2016, at  15:07, Craig Brideau <[hidden email]>

> wrote:

> >>>

> >>> *****

> >>> To join, leave or  search the confocal microscopy listserv, go to:

> >>>*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> >>> Post images on*http://www.imgur.com* <http://www.imgur.com/> and include the link in your

> posting.

> >>> *****

> >>>

> >>> I had a glance at your  picture. The black spots look like laser burns,

> but

> >>> there is also a clear  bubble or edge that looks suspiciously like oil

> also

> >>> visible in the picture.  It's possible you had some oil contamination

> and

> >>> then it burned onto the  glass. Most optics will handle a surprising

> amount

> >>> of laser power as long  as they are extremely clean. The instant there

> is

> >>> any dust/grease the  laser will burn the surface as the contamination

> >>> increases point  absorption at the interface.

> >>>

> >>> Craig

> >>>

> >>>> On Fri, May 27,  2016 at 7:28 AM, Martin Wessendorf <[hidden email]>

> wrote:

> >>>>

> >>>> *****

> >>>> To join, leave or  search the confocal microscopy listserv, go to:

> >>>>*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> >>>> Post images on*http://www.imgur.com* <http://www.imgur.com/> and include the link in your

> posting.

> >>>> *****

> >>>>

> >>>> Dear Dr. Vaughan--

> >>>>

> >>>> Could it be dirt  and oil on the lens, rather than being burned and/or

> >>>> unglued?

> >>>>

> >>>> Martin Wessendorf

> >>>>

> >>>>

> >>>>

> >>>>

> >>>>> On 5/27/2016  2:58 AM, Joshua Vaughan wrote:

> >>>>>

> >>>>> *****

> >>>>> To join, leave  or search the confocal microscopy listserv, go to:

> >>>>>*http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*

> >>>>> Post images on  *http://www.imgur.com* <http://www.imgur.com/> and include the link in
your

> posting.

> >>>>> *****

> >>>>>

> >>>>> We damaged an  objective lens, a Nikon CFI Plan Apo Lambda 100X Oil

> 1.45 NA

> >>>>> lens on our  homebuilt TIRF/STORM microscope. Has anybody seen this

> sort of

> >>>>> thing before?  I'm interested to hear, if so. We will repair or

> replace it,

> >>>>> and if  replacing I wonder what to get. More details are below.

> >>>>>

> >>>>> Here is a link  to a pic of part of the back focal plane. It has some

> dark

> >>>>> spots that look  'burned' and some stuff that looks like 'bubbles'.

> >>>>>

> >>>>>*http://i63.tinypic.com/mmemvl.jpg*

> >>>>>

> >>>>>

> >>>>> Repairing this  lens will cost >50% of a new lens so I want to decide

> >>>>> whether

> >>>>> to repair or to  just get something else. I'd prefer something with a

> >>>>> higher

> >>>>> damage  threshold but I was unsuccessful in getting data on lens

> damage

> >>>>> thresholds.

> >>>>>

> >>>>> If replacing,  what to get? I noticed three lenses online and I

> wonder what

> >>>>> is best for  STORM and whether all have a similar damage threshold or

> not.

> >>>>>

> >>>>> 1. CFI Plan Apo  Lambda 100X Oil 1.45 NA (current lens)

> >>>>> 2. CFI Plan Apo  Lambda DM 100X Oil 1.45 NA

> >>>>> 3. CFI Apo SR  TIRF 100X Oil 1.49 NA

> >>>>>

> >>>>>

> >>>>> I think the  Lambda lenses are supposed to be lower background,

> though I've

> >>>>> never tested  whether that is important or not. They are also Plan

> lenses,

> >>>>> but I'm not  sure how important that is for the <40 micron regions we

> >>>>> typically  image. The SR lens has a correction collar which might be

> useful

> >>>>> (?) when  imaging a bit off the glass. The online information says

> the SR

> >>>>> lens has a PSF  optimized for STORM which sounds good but I'm not

> sure what

> >>>>> it actually  means. I don't have any damage threshold information

> about

> >>>>> these

> >>>>> lenses. Oh, we  are pretty sure the damage was caused when we sent

> ~120 mW

> >>>>> at

> >>>>> 405 nm was sent  to the lens in EPI and later in TIRF. We plan to

> >>>>> substantially  attenuate the laser in the future, which didn't need

> to be

> >>>>> that big  anyway, but I am not sure how much to attenuate by.

> >>>>>

> >>>>>

> >>>>> Joshua C.  Vaughan

> >>>>> Assistant  Professor

> >>>>> Department of  Chemistry

> >>>>> Box 351700

> >>>>> University of  Washington

> >>>>> Seattle, WA  98195

> >>>>

> >>>> --

> >>>> Martin Wessendorf,  Ph.D.                   office: (612) 626-0145

> >>>> Assoc Prof, Dept  Neuroscience                 lab: (612) 624-2991

> >>>> University of  Minnesota             Preferred FAX: (612) 624-8118

> >>>> 6-145 Jackson Hall,  321 Church St. SE    Dept Fax: (612) 626-5009

> >>>> Minneapolis, MN55455                    e-mail: [hidden email]

> >>>>

Best regards,

Alexander N. Asanov, Ph.D.
President,  TIRF Labs
106 Grendon Place
Cary, NC 27519
Tel: 919-463-9545  Mobile:  919-903-4792

[hidden email] <mailto:[hidden email]>

www.tirf-labs.com <http://www.tirf-labs.com/> www.TIRFmicroscopy.com
<http://www.tirfmicroscopy.com/>

*From:*Martin Wessendorf [mailto:[hidden email]]
*Sent:* Monday, May 23, 2016 1:16 PM
*To:* Alexander Asanov
*Subject:* Re: Policy for commercial postings

Dear Dr. Asanov--

My apologies for taking so long to get back to you. Typically, people
use the list if they have a question or if they can provide the answer
to another person's question. (It's sometimes used for announcements as
well.)  If you have a question about TIRFM, feel free to send to me and
if it's appropriate I'll post it for you.

Again, you may want to look back through the Archives to get a sense of
how the List is used.

Sincerely,

Martin Wessendorf



On 5/10/2016 10:44 PM, Alexander Asanov wrote:

    Dear Dr. Wessendorf,

    I would like to initiate a discussion about TIRFM geometries. The
    Confocal community will benefit of the knowledge that will be shared
    in this discussion. Would you approve such discussion?  Without any
    intent to sell anything?

    Did you know that approximately 70% of your academic subscribers are
    owners, partial owners, or have financial interest in commercial
    entities operating in the field of microscopy? Do you have a
    mechanism to detect such hidden interests?

    Sincerely,

    Alexander N. Asanov, Ph.D.
    President,  TIRF Labs
    106 Grendon Place
    Cary, NC 27519
    Tel: 919-463-9545  Mobile:  919-903-4792

    [hidden email] <mailto:[hidden email]>

    www.tirf-labs.com <http://www.tirf-labs.com> www.TIRFmicroscopy.com
    <http://www.tirfmicroscopy.com/>


    *From:*Martin Wessendorf [mailto:[hidden email]]
    *Sent:* Tuesday, May 10, 2016 10:59 PM
    *To:* Alexander Asanov
    *Subject:* Re: Policy for commercial postings

    Dear Dr. Asanov--

    I think we've said about all we can at this point.  Please study the
    Policy and the Archives to get a feel for how other representatives
    of the private sector act.  Again, for now, if you want to post
    something, send it to me and I'll post it on your behalf.

    Sincerely,

    Martin Wessendorf


    On 5/10/2016 3:53 PM, Alexander Asanov wrote:

        Dear Dr. Wessendorf,

        Be sure that my post is about science and technology, rather
        than commerce.  It appears that you have incorrect perception
        that everything which originates in industrial settings has
        commercial intent.

        Confocal community would benefit tremendously of comparison
        lgTIRF and pTIRF with objective-type TIRF. In fact, it is quite
        obvious that it was terrible idea to use low light emission
        channel to pipe high-intensity excitation light. Several
        brilliant  papers have been published in scientific journals
        about poor quality of objective-type TIRF.  However,
        objective-TIRF is a die-hard misconception, which has received
        millions of commercial dollars to deform the perception of TIRF
        users. Read TIRF posts of your subscribers.

        Affiliation of annoyed subscriber does not matter. He or she can
        have indirect interests and motivations at subconscience.

        While super-resolution and expansion microscopy and single
        molecule detection become indispensable tools, the significance
        of TIRF increases.

        It appears that you failed to distinguish between science and
        commerce. I feel strongly that it is too difficult if not
        impossible for an industrial subscriber to contribute knowledge
        value to the confocal forum under your moderation.

        I agree with you only to your point that rules established by a
        moderator should be complied with. No questions. I did not
        intend to go over your rules. I reiterate my apologies.

        However, I firmly disagree with your attempt to downsize the
        significance of my TIRF posting to commercial intent. It appears
        that regardless of the content and knowledge value, all posting
        originated from industry will be considered as commercial. It is
        sad method of perception.

        All 2,000+ of confocal subscribers would benefit tremendously of
        the knowledge about the advantages and disadvantages of  lgTIRF
        and pTIRF with objective TIRF.

        I wish I would be wrong, because you state in your rules that
        you appreciate the contribution from industrial players. Could
        you point out to a single knowledge posting at your list from an
        industrial player?

        Sincerely,

        Alexander N. Asanov, Ph.D.
        President,  TIRF Labs
        106 Grendon Place
        Cary, NC 27519
        Tel: 919-463-9545 *www*.tirf-labs.com
        <http://www.tirf-labs.com/>www.TIRFmicroscopy.com
        <http://www.TIRFmicroscopy.com>
        Mobile:  919-903-4792 [hidden email]
        <mailto:[hidden email]>

        *From:*Martin Wessendorf [mailto:[hidden email]]
        *Sent:* Tuesday, May 10, 2016 12:04 PM
        *To:* Alexander Asanov <[hidden email]>
        <mailto:[hidden email]>
        *Subject:* Policy for commercial postings

        Dear Dr. Asanov--

        I apologize for again bringing up an uncomfortable subject
        again, but there were a couple things that I thought worth
        mentioning:

        1)  I think your idea of reformatting the Policy was excellent
        and I used it when I reposted the Policy today.

        2)  The rules of conduct on the List are important to some of
        its members.  Minutes after re-posting the Policy, I received
        the following note.  I've removed any identifiers but it came
        from the marketing director of company involved in optics and
        microscopy.  --People who make a point of playing by the rules
        are probably the most adamant about others playing by the rules.

        Sincerely,

        Martin Wessendorf
        -----------------------------------

        *Subject: *

       

        Thank you

        *Date: *

       

        Tue, 10 May 2016 16:39:05 +0100

        *To: *

       

        Martin Wessendorf <[hidden email]> <mailto:[hidden email]>

        I appreciate the post as I was quite annoyed with tirf-labs.

        Thank you for all you do!


        ----------------------------------



        --

        Martin Wessendorf, Ph.D.                   office: (612) 626-0145

        Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991

        University of Minnesota             Preferred FAX: (612) 624-8118

        6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009

        Minneapolis, MN  55455                    e-mail:[hidden email] <mailto:[hidden email]>

         

    <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=emailclient&utm_term=icon>

   

    Virus-free. www.avast.com <http://www.avast.com>



--

Martin Wessendorf, Ph.D.                   office: (612) 626-0145

Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991

University of Minnesota             Preferred FAX: (612) 624-8118

6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009

Minneapolis, MN  55455                    e-mail:[hidden email] <mailto:[hidden email]>