To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Listers, I am trying to figure out the confocal microscope scanning speed that can be achieved based on the specs of a Galvo XY scanner. What specs of a Galvo XY scanner (e.g. small angle step response, the travelling angle) should I look for determining the scanning speed? It seems the Galvo mirror size would affect the speed a lot. Any help would be greatly appreciated. Thanks in advance. Best regards, Sheng |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** It depends on the resonant frequency of the galvo modified by the weight/size of the mirror on the end of it, tweaked by how wide an angle you need it to scan. Basically you have to give the scan angle and mirror size you need to the manufacturer and they can come back to you with the resulting maximum speed. Or just look up pre-built galvos with mirrors attached and find one that suits your application. Craig On Tue, Feb 12, 2013 at 1:39 PM, yuansheng sun <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Listers, > > I am trying to figure out the confocal microscope scanning speed that can > be achieved based on the specs of a Galvo XY scanner. What specs of a > Galvo XY scanner (e.g. small angle step response, the travelling angle) > should I look for determining the scanning speed? It seems the Galvo > mirror size would affect the speed a lot. Any help would be greatly > appreciated. Thanks in advance. > > Best regards, > Sheng > |
In reply to this post by yuansheng sun
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Sheng, There's a little bit of information on this topic in chapter 9 of the 3rd edition of Jim Pawley's Hanbook of Biological Confocal Microscopy: The Intermediate Optical System of Laser-Scanning Confocal Microscopes Ernst H.K. Stelzer Some considerations about the required size of the mirrors and their relation to the scan speed are mentioned. Maybe this will help you. Cheers, John Oreopoulos Staff Scientist Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca On 2013-02-12, at 3:39 PM, yuansheng sun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Listers, > > I am trying to figure out the confocal microscope scanning speed that can > be achieved based on the specs of a Galvo XY scanner. What specs of a > Galvo XY scanner (e.g. small angle step response, the travelling angle) > should I look for determining the scanning speed? It seems the Galvo > mirror size would affect the speed a lot. Any help would be greatly > appreciated. Thanks in advance. > > Best regards, > Sheng |
In reply to this post by yuansheng sun
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Leica SP5 resonant scanner: 8,000 lines per second Leica SP8 resonant scanner: 12,500 lines per second (you should look up exact value at leica) Ignoring turnaround time and flyback ... For SP5, 8,000 lines/sec is 125 microeconds per line. If I choose 1000 pixels per line, this is a dwell time of approximately 125 nanoseconds per pixel. You would need to know the PMT performance and zoom to know how this translates to real performance (XY resolution). For commercial confocals (at least those designed in the past 10 years), I expect the manufacturer has figured out the right size for the mirror. On the SP5's I manage, Jonathan Boyd of leica applications (USA) showed me that SP5 resonant scanner mode produced a brighter signal per pixel total dwell time than standard scan mode. That is 8000 lines/sec * 100 sum (RS mode) was brighter than (now standard mode) 800 lines/sec * 10 sum was brighter than 80 lines/second. Explanation was almost certainly that slow scan drives the fluorophores into the triplet state, where they stay (or reach equilibrium) as long as the laser is on the spot. This should vary with O2, other triplet quenchers, fluorophore, laser power. George On 2/12/2013 3:39 PM, yuansheng sun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Listers, > > I am trying to figure out the confocal microscope scanning speed that can > be achieved based on the specs of a Galvo XY scanner. What specs of a > Galvo XY scanner (e.g. small angle step response, the travelling angle) > should I look for determining the scanning speed? It seems the Galvo > mirror size would affect the speed a lot. Any help would be greatly > appreciated. Thanks in advance. > > Best regards, > Sheng > > |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You don't need to be driving your fluorochrome into a triplet state to see this effect - you just have to be saturating the S1 state - ie exciting a significant proportion of the fluorochromes within the spot, so that a further photon stands a lower chance of hitting an unexcited molecule. It's still a really bad thing to do. You should always check that halving the laser power halves the signal - if not you are killing your specimen. I'll bet that if the laser power had been set so that there was no saturation at the slow scan speed, and kept the same for the fast scan, you would not have seen this effect. However, 'brightness' is really a meaningless measure here, because the resonant scan has to be adjusted for the non-constant dwell time across the sample so you are always seeing an adjusted level. Signal / noise ratio is the only meaningful criterion. Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara Sent: Thursday, 14 February 2013 12:22 PM To: [hidden email] Subject: Re: Galvo XY Scanning Speed ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Leica SP5 resonant scanner: 8,000 lines per second Leica SP8 resonant scanner: 12,500 lines per second (you should look up exact value at leica) Ignoring turnaround time and flyback ... For SP5, 8,000 lines/sec is 125 microeconds per line. If I choose 1000 pixels per line, this is a dwell time of approximately 125 nanoseconds per pixel. You would need to know the PMT performance and zoom to know how this translates to real performance (XY resolution). For commercial confocals (at least those designed in the past 10 years), I expect the manufacturer has figured out the right size for the mirror. On the SP5's I manage, Jonathan Boyd of leica applications (USA) showed me that SP5 resonant scanner mode produced a brighter signal per pixel total dwell time than standard scan mode. That is 8000 lines/sec * 100 sum (RS mode) was brighter than (now standard mode) 800 lines/sec * 10 sum was brighter than 80 lines/second. Explanation was almost certainly that slow scan drives the fluorophores into the triplet state, where they stay (or reach equilibrium) as long as the laser is on the spot. This should vary with O2, other triplet quenchers, fluorophore, laser power. George On 2/12/2013 3:39 PM, yuansheng sun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Listers, > > I am trying to figure out the confocal microscope scanning speed that can > be achieved based on the specs of a Galvo XY scanner. What specs of a > Galvo XY scanner (e.g. small angle step response, the travelling angle) > should I look for determining the scanning speed? It seems the Galvo > mirror size would affect the speed a lot. Any help would be greatly > appreciated. Thanks in advance. > > Best regards, > Sheng > > |