Earl_Sanford |
I am sorry if this topic has been discussed recently but I checked the archives
and did not find anything too recent. I am currently trying to determine which software package to acquire in order to make volumetric measurements from z-stacks collected on a confocal microscope. I want a package that is rather intuitive and straight forward to obtain size information (length, width, volume, etc.). Unfortunately I am not familiar with many of the packages available. I am currently using using ImageJ but am looking to find something that is a little more refined. However since the commercial packages are not inexpensive, I need to make sure I get the best package available to make the above measurements. Please let discuss the stregths and weakness of the software packages you are familiar with. Thanks for your help Earl |
Armstrong, Louise |
*COMMERCIAL RESPONSE*
Hi Earl There are a number of commercial software packages for 3D analysis of confocal images and your choice is likely to depend on what it is you want to do. Both Volocity from PerkinElmer and Imaris from Bitplane are strong in 3D visualization and analysis and both have slightly different areas of strength depending upon the types of measurements you want to make. If you would like to contact me off list then I can arrange for one of our imaging specialists to contact you to discuss your application. Having worked with Volocity and the team since before it was launched 10 years ago, I know they will give you an honest opinion on where our strengths and weaknesses lie. If you want to try it for yourself then we recently introduced a free demonstration version of Volocity which you can download from our website. Accompanying it there is some demo data and web-based tutorials and notes to guide you through how to recreate the steps such as selecting 3D objects and measuring volumes. I hope this will let you try it out on your own images. You can always ask for a specialist to contact you at any point. www.cellularimaging.com/products/volocity/demo Best wishes Louise Louise Armstrong-Denby PhD | Global Product Manager, Bio-discovery PerkinElmer | For the Better [hidden email] Phone: +44 24 7669 2229 | Fax: +44 24 7669 0091 | Viscount Centre II, University of Warwick Science Park, Millburn Hill Road, Coventry CV4 7HS, UK www.perkinelmer.com Please consider the environment before printing this e-mail. This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc. If you are not the intended recipient of this message, please inform the sender by replying to this email or sending a message to the sender and destroy the message and any attachments. Thank you. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Earl Sanford Sent: 09 September 2010 20:23 To: [hidden email] Subject: Help in Selecting 3D Measurement Software I am sorry if this topic has been discussed recently but I checked the archives and did not find anything too recent. I am currently trying to determine which software package to acquire in order to make volumetric measurements from z-stacks collected on a confocal microscope. I want a package that is rather intuitive and straight forward to obtain size information (length, width, volume, etc.). Unfortunately I am not familiar with many of the packages available. I am currently using using ImageJ but am looking to find something that is a little more refined. However since the commercial packages are not inexpensive, I need to make sure I get the best package available to make the above measurements. Please let discuss the stregths and weakness of the software packages you are familiar with. Thanks for your help Earl |
Michael Weber-4 |
Louise, good to hear that there is finally a Volocity demo!
Earl, you can also ask your local Andor/Bitplane representative for a demo license of Imaris. A hands-on demo is always the best choise, especially if you already have samples to analyze. Both Imaris and Volocity can perform volume measurements, although the approach is a bit different. The price model might also be of interest for you, as it is different as well. Michael > *COMMERCIAL RESPONSE* > > Hi Earl > > There are a number of commercial software packages for 3D analysis of > confocal images and your choice is likely to depend on what it is you > want to do. > > Both Volocity from PerkinElmer and Imaris from Bitplane are strong in 3D > visualization and analysis and both have slightly different areas of > strength depending upon the types of measurements you want to make. > > If you would like to contact me off list then I can arrange for one of > our imaging specialists to contact you to discuss your application. > Having worked with Volocity and the team since before it was launched 10 > years ago, I know they will give you an honest opinion on where our > strengths and weaknesses lie. > > If you want to try it for yourself then we recently introduced a free > demonstration version of Volocity which you can download from our > website. Accompanying it there is some demo data and web-based > tutorials and notes to guide you through how to recreate the steps such > as selecting 3D objects and measuring volumes. I hope this will let you > try it out on your own images. You can always ask for a specialist to > contact you at any point. > > www.cellularimaging.com/products/volocity/demo > > Best wishes > > Louise > > > Louise Armstrong-Denby PhD | Global Product Manager, Bio-discovery > PerkinElmer | For the Better > [hidden email] > Phone: +44 24 7669 2229 | Fax: +44 24 7669 0091 | > Viscount Centre II, University of Warwick Science Park, Millburn Hill > Road, Coventry CV4 7HS, UK > www.perkinelmer.com > Please consider the environment before printing this e-mail. > This e-mail message and any attachments are confidential and proprietary > to PerkinElmer, Inc. If you are not the intended recipient of this > message, please inform the sender by replying to this email or sending a > message to the sender and destroy the message and any attachments. > Thank you. > > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Earl Sanford > Sent: 09 September 2010 20:23 > To: [hidden email] > Subject: Help in Selecting 3D Measurement Software > > I am sorry if this topic has been discussed recently but I checked the > archives > and did not find anything too recent. > > I am currently trying to determine which software package to acquire in > order > to make volumetric measurements from z-stacks collected on a confocal > microscope. I want a package that is rather intuitive and straight > forward to > obtain size information (length, width, volume, etc.). Unfortunately I > am not > familiar with many of the packages available. I am currently using using > ImageJ > but am looking to find something that is a little more refined. However > since > the commercial packages are not inexpensive, I need to make sure I get > the > best package available to make the above measurements. > > Please let discuss the stregths and weakness of the software packages > you > are familiar with. > > Thanks for your help > Earl |
Vitaly Boyko |
Earl, I guess there should be much cheaper and more intuitive options than Volocity, Metamorph or Imaris...
Cheers,
Vitaly
From: Michael Weber <[hidden email]> To: [hidden email] Sent: Fri, September 10, 2010 7:36:06 AM Subject: Re: Help in Selecting 3D Measurement Software Louise, good to hear that there is finally a Volocity demo! Earl, you can also ask your local Andor/Bitplane representative for a demo license of Imaris. A hands-on demo is always the best choise, especially if you already have samples to analyze. Both Imaris and Volocity can perform volume measurements, although the approach is a bit different. The price model might also be of interest for you, as it is different as well. Michael > *COMMERCIAL RESPONSE* > > Hi Earl > > There are a number of commercial software packages for 3D analysis of > confocal images and your choice is likely to depend on what it is you > want to do. > > Both Volocity from PerkinElmer and Imaris from Bitplane are strong in 3D > visualization and analysis and both have slightly different areas of > strength depending upon the types of measurements you want to make. > > If you would like to contact me off list then I can arrange for one of > our imaging specialists to contact you to discuss your application. > Having worked with Volocity and the team since before it was launched 10 > years ago, I know they will give you an honest opinion on where our > strengths and weaknesses lie. > > If you want to try it for yourself then we recently introduced a free > demonstration version of Volocity which you can download from our > website. Accompanying it there is some demo data and web-based > tutorials and notes to guide you through how to recreate the steps such > as selecting 3D objects and measuring volumes. I hope this will let you > try it out on your own images. You can always ask for a specialist to > contact you at any point. > > www.cellularimaging.com/products/volocity/demo > > Best wishes > > Louise > > > Louise Armstrong-Denby PhD | Global Product Manager, Bio-discovery > PerkinElmer | For the Better > [hidden email] > Phone: +44 24 7669 2229 | Fax: +44 24 7669 0091 | > Viscount Centre II, University of Warwick Science Park, Millburn Hill > Road, Coventry CV4 7HS, UK > www.perkinelmer.com > Please consider the environment before printing this e-mail. > This e-mail message and any attachments are confidential and proprietary > to PerkinElmer, Inc. If you are not the intended recipient of this > message, please inform the sender by replying to this email or sending a > message to the sender and destroy the message and any attachments. > Thank you. > > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Earl Sanford > Sent: 09 September 2010 20:23 > To: [hidden email] > Subject: Help in Selecting 3D Measurement Software > > I am sorry if this topic has been discussed recently but I checked the > archives > and did not find anything too recent. > > I am currently trying to determine which software package to acquire in > order > to make volumetric measurements from z-stacks collected on a confocal > microscope. I want a package that is rather intuitive and straight > forward to > obtain size information (length, width, volume, etc.). Unfortunately I > am not > familiar with many of the packages available. I am currently using using > ImageJ > but am looking to find something that is a little more refined. However > since > the commercial packages are not inexpensive, I need to make sure I get > the > best package available to make the above measurements. > > Please let discuss the stregths and weakness of the software packages > you > are familiar with. > > Thanks for your help > Earl |
Chris Guerin |
In reply to this post by Earl_Sanford
Hello Earl:
As someone who runs a core facility I can comment that my users find Volocity to be very powerful, intuitive and easy to use. I also agree with the comment made before though that depending on what you want to do specifically another program might be better at certain tasks. The other thing I can comment on is the support offered by Perkin Elmer after you've bought the product. Volocity support is simply outstanding. It is also great that it's multi platform (PC and Mac) and that it can run under 64 bit operating systems and if you're dealing with large data sets this is a great advantage. We frequently use it on gigabyte data sets which choke other programs and it performs very well. Finally I also agree with the others who have suggested "try before you buy". it's the only way to know if what others find easy and powerful will work as well for you. Good luck, Chris Christopher Guérin, Ph.D. Leader, Microscopy Core V.I.B., D.M.B.R., Univ. Gent Fiers-Schell-Van Montagu' building Technologiepark 927, B-9052 Ghent (Zwijnaarde) - Belgium tel : +32-9-33 13 611 web:http://www.dmbr.ugent.be/ext/public/microscopy/index.html |
Neeraj Gohad-2 |
In reply to this post by Michael Weber-4
Hi Earl,
We were in the same boat couple of months ago. We evaluated Bitplane's Imaris and Visage Imaging's amira. Both the companies offered us demo licenses and gave us webminars to get us started. People from both these companies were very helpful and knowledgeable. At the end of our evaluation we found Imaris to be more intuitive and suited for our application. Imaris has many modules and the price goes up as you add modules, we have the confocal standard and statistics package that helps us do all sorts of measurements on confocal datasets, the statistics package allows you to color code objects using features like area, volume etc. We also like the key frame animation feature where you can create fly through movies with your datasets, it's a very effective way of communicating 3D data. As always, no commercial interest, Best, Neeraj. Neeraj V. Gohad, Ph.D. Postdoctoral Fellow Okeanos Research Group Department of Biological Sciences 132 Long Hall Clemson University Clemson,SC-29634 Phone: 864-656-3597 Fax: 864-656-0435 Website: http://www.clemson.edu/okeanos -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michael Weber Sent: Friday, September 10, 2010 7:36 AM To: [hidden email] Subject: Re: Help in Selecting 3D Measurement Software Louise, good to hear that there is finally a Volocity demo! Earl, you can also ask your local Andor/Bitplane representative for a demo license of Imaris. A hands-on demo is always the best choise, especially if you already have samples to analyze. Both Imaris and Volocity can perform volume measurements, although the approach is a bit different. The price model might also be of interest for you, as it is different as well. Michael > *COMMERCIAL RESPONSE* > > Hi Earl > > There are a number of commercial software packages for 3D analysis of > confocal images and your choice is likely to depend on what it is you > want to do. > > Both Volocity from PerkinElmer and Imaris from Bitplane are strong in 3D > visualization and analysis and both have slightly different areas of > strength depending upon the types of measurements you want to make. > > If you would like to contact me off list then I can arrange for one of > our imaging specialists to contact you to discuss your application. > Having worked with Volocity and the team since before it was launched 10 > years ago, I know they will give you an honest opinion on where our > strengths and weaknesses lie. > > If you want to try it for yourself then we recently introduced a free > demonstration version of Volocity which you can download from our > website. Accompanying it there is some demo data and web-based > tutorials and notes to guide you through how to recreate the steps such > as selecting 3D objects and measuring volumes. I hope this will let you > try it out on your own images. You can always ask for a specialist to > contact you at any point. > > www.cellularimaging.com/products/volocity/demo > > Best wishes > > Louise > > > Louise Armstrong-Denby PhD | Global Product Manager, Bio-discovery > PerkinElmer | For the Better > [hidden email] > Phone: +44 24 7669 2229 | Fax: +44 24 7669 0091 | > Viscount Centre II, University of Warwick Science Park, Millburn Hill > Road, Coventry CV4 7HS, UK > www.perkinelmer.com > Please consider the environment before printing this e-mail. > This e-mail message and any attachments are confidential and proprietary > to PerkinElmer, Inc. If you are not the intended recipient of this > message, please inform the sender by replying to this email or sending a > message to the sender and destroy the message and any attachments. > Thank you. > > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Earl Sanford > Sent: 09 September 2010 20:23 > To: [hidden email] > Subject: Help in Selecting 3D Measurement Software > > I am sorry if this topic has been discussed recently but I checked the > archives > and did not find anything too recent. > > I am currently trying to determine which software package to acquire in > order > to make volumetric measurements from z-stacks collected on a confocal > microscope. I want a package that is rather intuitive and straight > forward to > obtain size information (length, width, volume, etc.). Unfortunately I > am not > familiar with many of the packages available. I am currently using using > ImageJ > but am looking to find something that is a little more refined. However > since > the commercial packages are not inexpensive, I need to make sure I get > the > best package available to make the above measurements. > > Please let discuss the stregths and weakness of the software packages > you > are familiar with. > > Thanks for your help > Earl |
Arvonn Tully |
In reply to this post by Earl_Sanford
*COMMERCIAL RESPONSE* Dear Earl, I think that you will find that there are three main commercial packages, Amira (Avizo? ), Imaris and Volocity. Each package should be able provide those results, but due to the different workflows, you will find that certain software is easier than others to use.
Depending on the complexity of the image, you will find each software package has very different segmentation (feature extraction) tools for analyzing the objects of interest. I urge you to pay attention to these tools, which are critical for accurately defining your objects, and subsequently measuring those objects.
If you would like to try using the Imaris software, please either give us a call at: 1-888-332-4879 or visit our demo download webpage:
Someone from Bitplane will get in touch with you shortly thereafter, and they will provide you with a personalized demonstration of the software. Please note that the Imaris demonstration version software will work with any of supported Image formats, and Imaris is not limited to demo data sets (although we provide a few to get you started).
We strongly encourage you to evaluate all of the products, side by side if possible, to see what works best for your needs. Best Regards,
Arvonn Tully Technical and Application Support
Western United States [hidden email] Support Telephone:+1-888-332-4879 ext 12 fax: 866-691-9112 Bitplane, Inc. http://www.bitplane.com On Thu, Sep 9, 2010 at 12:23 PM, Earl Sanford <[hidden email]> wrote: I am sorry if this topic has been discussed recently but I checked the archives |
Bob Nienhuis |
In reply to this post by Earl_Sanford
How do these packages do volumetric measurement differently than stereological
approaches? I understand that several groups are now applying stereological methods
to confocal image stacks.
Bob Nienhuis
VA / UCLA Neurobiological Research
North Hills, CA
|
Mika Hukkanen |
We have done some volumetric measurements on podosomes and
invadopodia in epithelial-mesenchymal transition; some points to keep in mind –
deconvolution to take into account for the psf, and standard curves generated
of beads of known sizes (different sizes related to the volume of interests).
The packages that have been mentioned most likely do the job ok, however, if
you go for a commercial one the license maintenance fees per annum can be quite
high. Initial purchase price may not seem that high but are you willing, for
the years to come, to pay some 1000;- eur per each module just for bug fixes
and upgrades? Mika Hukkanen Biomedicum Imaging Unit Helsinki, Finland |
In reply to this post by Bob Nienhuis
Stereology essentially uses a random approach – you count
how your sampling pattern intersects with the structure you are
measuring. It can therefore get you measurements of (eg) organelle
volume or membrane area from random slices. Maybe you can use a full
stack of optical sections but there is no need for it. What you end up
with is an estimate based on sampling statistics. What these programs (and others) do is apply standard
segmentation methods in three dimensions, so you get an exact measure of what
is in the sampled volume (or at least, as accurate as your original
segmentation, and that’s another can of worms).
Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351
7682
Mobile 0413 281 861 ______________________________________________ From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of Bob Nienhuis How do these packages do volumetric measurement
differently than stereological approaches? I understand that several groups are now
applying stereological methods to confocal image stacks. Bob Nienhuis VA / UCLA Neurobiological Research North Hills, CA
|
Vincent Schoonderwoert |
In reply to this post by Earl_Sanford
*commercial response*
Dear all, We would like to respond to the initial request that Earl Sanford posted, concerning an interest in comparing inexpensive 3D visualization and analysis software packages. As was also mentioned in previous comments, we very much encourage testing the available software. This is why we would like to bring our Huygens software under attention, since it was not specifically mentioned yet. Huygens is known for its high quality deconvolution but offers also advanced 3D visualization and analysis tools. Huygens can be downloaded from http://www.svi.nl/download using the 'request license on-line' button. Best, Vincent Schoonderwoert Scientific Volume Imaging bv Imaging Specialist/Account Manager [hidden email] Laapersveld 63 1213 VB Hilversum, The Netherlands Tel: + 31 35 646 8216 Fax: + 31 35 683 7971 www.svi.nl FAQ: http://support.svi.nl Earl Sanford wrote: I am sorry if this topic has been discussed recently but I checked the archives and did not find anything too recent. I am currently trying to determine which software package to acquire in order to make volumetric measurements from z-stacks collected on a confocal microscope. I want a package that is rather intuitive and straight forward to obtain size information (length, width, volume, etc.). Unfortunately I am not familiar with many of the packages available. I am currently using using ImageJ but am looking to find something that is a little more refined. However since the commercial packages are not inexpensive, I need to make sure I get the best package available to make the above measurements. Please let discuss the stregths and weakness of the software packages you are familiar with. Thanks for your help Earl To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy |
Free forum by Nabble | Edit this page |