Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

I am in the market for an EMCCD for widefield fluorescence
time-lapse imaging of genetically-labeled cultured cells. I
am acquiring images frequently (every 10 minutes) for
several days. The cells seem to be very sensitive to
phototoxicity and photobleaching (GFP is the fluorophore),
so I am hoping that replacing my Roper CoolSnapHQ CCD with
an EMCCD will allow me to use much shorter exposure times.

I realize that this is a confocal list and my application is
widefield but since EMCCDs are also used with spinning disk
confocal, I thought I'd ask. Apologies if this question has
been hashed out before... I searched the archives without
finding definitive recommendations. (Maybe that's asking for
too much!) Any help on the following would be much appreciated:

1) What EMCCDs have people found to be the most *sensitive*
(nice picture in low-light) and the most *reliable* (driver
doesn't crash, etc.)? I'd especially be interested in any
experiences or opinions with Roper/Photometrics's CascadeII
or QuantEM and the Andor iXon cameras.

2) How have people been testing cameras? I've set up demos
of several cameras so I'd like to do some standard tests.
I've seen some references on the web for calculating dark
noise, read noise, and gain, such as:
http://www.qsimaging.com/ccd_noise_measure.html
http://www.mirametrics.com/tech_note_ccdgain.htm
http://www.photomet.com/library/library_encyclopedia/library_enc_gain.php
but I'd love to hear what other tests people think are
relevant for these cameras. Obviously, I'll try imaging my
samples (using fiber-coupled LED as a constant illumination
source)... what else should I do? Fluorescent beads instead
of my cells perhaps?

3) For coupling the camera to my Olympus IX71, is there
something like the Zeiss Optovar (which, as I understand it,
has multiple selectable magnifications of the intermediate
lens) made by a third party? This would be helpful for
changing between achieving the full NA of the objective and
getting less resolution but a bigger field of view for
different experiments.

Thanks for your help. Best,

- Neville

---
Neville Sanjana
Dept. of Brain and Cognitive Science
Massachusetts Institute of Technology
Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Neville,
              While an EMCCD camera will decrease your exposure time, you may find that it will not really help solve your phototoxicity problem. The time your cells are exposed to fluorecent light will not neccesarily be dependant on the exposure time of your camera, but on how fast your entire sytem (microscope, shutters, filters, etc.) shuffles everything around.
 
What software etc are you using to capture your images?
 
We have a wide filed live cell rig here that uses an Orca ER camera on an IX81 microscope and it is all driven by metamorph. Even when the camera is capturing at 100ms exposures, the samples are exposed to flourecent light for about 2 seconds. To get rid of these sort of delays you need to move to using filter wheels, or for even shorter exposuers something like olympus' CellR system
 
 
Cheers
 
 
Cam
 
 
Cameron Nowell B. Sc (Hons)

 

Microscopy Imaging and Research Core Facility

Peter MacCallum Cancer Centre

7 St Andrews Place

East Melbourne, Victoria 3002

AUSTRALIA

 

Phone: +61396561242

Mobile: +61422882700

Fax: +61396561411

 

________________________________

From: Confocal Microscopy List on behalf of Neville Sanjana
Sent: Wed 28/11/2007 7:57 AM
To: [hidden email]
Subject: Help with EMCCD evaluation



Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

I am in the market for an EMCCD for widefield fluorescence
time-lapse imaging of genetically-labeled cultured cells. I
am acquiring images frequently (every 10 minutes) for
several days. The cells seem to be very sensitive to
phototoxicity and photobleaching (GFP is the fluorophore),
so I am hoping that replacing my Roper CoolSnapHQ CCD with
an EMCCD will allow me to use much shorter exposure times.

I realize that this is a confocal list and my application is
widefield but since EMCCDs are also used with spinning disk
confocal, I thought I'd ask. Apologies if this question has
been hashed out before... I searched the archives without
finding definitive recommendations. (Maybe that's asking for
too much!) Any help on the following would be much appreciated:

1) What EMCCDs have people found to be the most *sensitive*
(nice picture in low-light) and the most *reliable* (driver
doesn't crash, etc.)? I'd especially be interested in any
experiences or opinions with Roper/Photometrics's CascadeII
or QuantEM and the Andor iXon cameras.

2) How have people been testing cameras? I've set up demos
of several cameras so I'd like to do some standard tests.
I've seen some references on the web for calculating dark
noise, read noise, and gain, such as:
http://www.qsimaging.com/ccd_noise_measure.html
http://www.mirametrics.com/tech_note_ccdgain.htm
http://www.photomet.com/library/library_encyclopedia/library_enc_gain.php
but I'd love to hear what other tests people think are
relevant for these cameras. Obviously, I'll try imaging my
samples (using fiber-coupled LED as a constant illumination
source)... what else should I do? Fluorescent beads instead
of my cells perhaps?

3) For coupling the camera to my Olympus IX71, is there
something like the Zeiss Optovar (which, as I understand it,
has multiple selectable magnifications of the intermediate
lens) made by a third party? This would be helpful for
changing between achieving the full NA of the objective and
getting less resolution but a bigger field of view for
different experiments.

Thanks for your help. Best,

- Neville

---
Neville Sanjana
Dept. of Brain and Cognitive Science
Massachusetts Institute of Technology




This email (including any attachments) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation**Commercial Response**

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

The Andor iQ system employs what we call "active blanking."

The EMCCD and our AOTF communicate directly with each other so that the
AOTF only passes excitation light while the camera is exposing.  This
great reduces effects of phototoxicity due to exposure of cells to
excitation light between captures.

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell (774) 291 - 9992
Office (860) 290 - 9211 x219
Fax (860) 290 - 9566
Web: www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, November 28, 2007 2:51 AM
To: [hidden email]
Subject: Re: Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Neville,
              While an EMCCD camera will decrease your exposure time,
you may find that it will not really help solve your phototoxicity
problem. The time your cells are exposed to fluorecent light will not
neccesarily be dependant on the exposure time of your camera, but on how
fast your entire sytem (microscope, shutters, filters, etc.) shuffles
everything around.
 
What software etc are you using to capture your images?
 
We have a wide filed live cell rig here that uses an Orca ER camera on
an IX81 microscope and it is all driven by metamorph. Even when the
camera is capturing at 100ms exposures, the samples are exposed to
flourecent light for about 2 seconds. To get rid of these sort of delays
you need to move to using filter wheels, or for even shorter exposuers
something like olympus' CellR system
 
 
Cheers
 
 
Cam
 
 
Cameron Nowell B. Sc (Hons)

 

Microscopy Imaging and Research Core Facility

Peter MacCallum Cancer Centre

7 St Andrews Place

East Melbourne, Victoria 3002

AUSTRALIA

 

Phone: +61396561242

Mobile: +61422882700

Fax: +61396561411

 

________________________________

From: Confocal Microscopy List on behalf of Neville Sanjana
Sent: Wed 28/11/2007 7:57 AM
To: [hidden email]
Subject: Help with EMCCD evaluation



Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

I am in the market for an EMCCD for widefield fluorescence
time-lapse imaging of genetically-labeled cultured cells. I
am acquiring images frequently (every 10 minutes) for
several days. The cells seem to be very sensitive to
phototoxicity and photobleaching (GFP is the fluorophore),
so I am hoping that replacing my Roper CoolSnapHQ CCD with
an EMCCD will allow me to use much shorter exposure times.

I realize that this is a confocal list and my application is
widefield but since EMCCDs are also used with spinning disk
confocal, I thought I'd ask. Apologies if this question has
been hashed out before... I searched the archives without
finding definitive recommendations. (Maybe that's asking for
too much!) Any help on the following would be much appreciated:

1) What EMCCDs have people found to be the most *sensitive*
(nice picture in low-light) and the most *reliable* (driver
doesn't crash, etc.)? I'd especially be interested in any
experiences or opinions with Roper/Photometrics's CascadeII
or QuantEM and the Andor iXon cameras.

2) How have people been testing cameras? I've set up demos
of several cameras so I'd like to do some standard tests.
I've seen some references on the web for calculating dark
noise, read noise, and gain, such as:
http://www.qsimaging.com/ccd_noise_measure.html
http://www.mirametrics.com/tech_note_ccdgain.htm
http://www.photomet.com/library/library_encyclopedia/library_enc_gain.ph
p
but I'd love to hear what other tests people think are
relevant for these cameras. Obviously, I'll try imaging my
samples (using fiber-coupled LED as a constant illumination
source)... what else should I do? Fluorescent beads instead
of my cells perhaps?

3) For coupling the camera to my Olympus IX71, is there
something like the Zeiss Optovar (which, as I understand it,
has multiple selectable magnifications of the intermediate
lens) made by a third party? This would be helpful for
changing between achieving the full NA of the objective and
getting less resolution but a bigger field of view for
different experiments.

Thanks for your help. Best,

- Neville

---
Neville Sanjana
Dept. of Brain and Cognitive Science
Massachusetts Institute of Technology




This email (including any attachments) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.  
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi!

If it is for cell imaging you might consider a BT camera.

Forcing the previous point, which I thing is very important:

I'm having the same problem with an ORCA AG with ImagePro Plus.  What I
have now is a small program called by macros that controls an external
shutter via TTL, which makes a delay till it opens the shutter. I found
out that the driver takes almost 500ms to get the camera taking a
picture which is unacceptable! However this solution is far from being
elegant and sometimes the pseudo synchronization fails! Is this a
Hamamatsu problem? With a Hamamatsu BT1024 which has a QE>90% it
happends the same thing with metamorph.

Regarding to the previous commercial response, does anyone have a
cheaper and more flexible solution?

Regards,
NM





Nowell, Cameron wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Neville,
>               While an EMCCD camera will decrease your exposure time, you may find that it will not really help solve your phototoxicity problem. The time your cells are exposed to fluorecent light will not neccesarily be dependant on the exposure time of your camera, but on how fast your entire sytem (microscope, shutters, filters, etc.) shuffles everything around.
>  
> What software etc are you using to capture your images?
>  
> We have a wide filed live cell rig here that uses an Orca ER camera on an IX81 microscope and it is all driven by metamorph. Even when the camera is capturing at 100ms exposures, the samples are exposed to flourecent light for about 2 seconds. To get rid of these sort of delays you need to move to using filter wheels, or for even shorter exposuers something like olympus' CellR system
>  
>  
> Cheers
>  
>  
> Cam
>  
>  
> Cameron Nowell B. Sc (Hons)
>
>  
>
> Microscopy Imaging and Research Core Facility
>
> Peter MacCallum Cancer Centre
>
> 7 St Andrews Place
>
> East Melbourne, Victoria 3002
>
> AUSTRALIA
>
>  
>
> Phone: +61396561242
>
> Mobile: +61422882700
>
> Fax: +61396561411
>
>  
>
> ________________________________
>
> From: Confocal Microscopy List on behalf of Neville Sanjana
> Sent: Wed 28/11/2007 7:57 AM
> To: [hidden email]
> Subject: Help with EMCCD evaluation
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
> I am in the market for an EMCCD for widefield fluorescence
> time-lapse imaging of genetically-labeled cultured cells. I
> am acquiring images frequently (every 10 minutes) for
> several days. The cells seem to be very sensitive to
> phototoxicity and photobleaching (GFP is the fluorophore),
> so I am hoping that replacing my Roper CoolSnapHQ CCD with
> an EMCCD will allow me to use much shorter exposure times.
>
> I realize that this is a confocal list and my application is
> widefield but since EMCCDs are also used with spinning disk
> confocal, I thought I'd ask. Apologies if this question has
> been hashed out before... I searched the archives without
> finding definitive recommendations. (Maybe that's asking for
> too much!) Any help on the following would be much appreciated:
>
> 1) What EMCCDs have people found to be the most *sensitive*
> (nice picture in low-light) and the most *reliable* (driver
> doesn't crash, etc.)? I'd especially be interested in any
> experiences or opinions with Roper/Photometrics's CascadeII
> or QuantEM and the Andor iXon cameras.
>
> 2) How have people been testing cameras? I've set up demos
> of several cameras so I'd like to do some standard tests.
> I've seen some references on the web for calculating dark
> noise, read noise, and gain, such as:
> http://www.qsimaging.com/ccd_noise_measure.html
> http://www.mirametrics.com/tech_note_ccdgain.htm
> http://www.photomet.com/library/library_encyclopedia/library_enc_gain.php
> but I'd love to hear what other tests people think are
> relevant for these cameras. Obviously, I'll try imaging my
> samples (using fiber-coupled LED as a constant illumination
> source)... what else should I do? Fluorescent beads instead
> of my cells perhaps?
>
> 3) For coupling the camera to my Olympus IX71, is there
> something like the Zeiss Optovar (which, as I understand it,
> has multiple selectable magnifications of the intermediate
> lens) made by a third party? This would be helpful for
> changing between achieving the full NA of the objective and
> getting less resolution but a bigger field of view for
> different experiments.
>
> Thanks for your help. Best,
>
> - Neville
>
> ---
> Neville Sanjana
> Dept. of Brain and Cognitive Science
> Massachusetts Institute of Technology
>
>
>
>
> This email (including any attachments) may contain
> confidential and/or legally privileged information and is
> intended only to be read or used by the addressee.  If you
> are not the intended addressee, any use, distribution,
> disclosure or copying of this email is strictly
> prohibited.  
> Confidentiality and legal privilege attached to this email
> (including any attachments) are not waived or lost by
> reason of its mistaken delivery to you.
> If you have received this email in error, please delete it
> and notify us immediately by telephone or email.  Peter
> MacCallum Cancer Centre provides no guarantee that this
> transmission is free of virus or that it has not been
> intercepted or altered and will not be liable for any delay
> in its receipt.
>

--
Nuno Moreno
Cell Imaging Unit
Instituto Gulbenkian de Ciência
http://uic.igc.gulbekian.pt
http://www.igc.gulbekian.pt
phone +351 214464606
fax   +351 214407970
Aryeh Weiss Aryeh Weiss
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Nuno Moreno wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi!
>
> If it is for cell imaging you might consider a BT camera.
>
> Forcing the previous point, which I thing is very important:
>
> I'm having the same problem with an ORCA AG with ImagePro Plus.  What I
> have now is a small program called by macros that controls an external
> shutter via TTL, which makes a delay till it opens the shutter. I found
> out that the driver takes almost 500ms to get the camera taking a
> picture which is unacceptable! However this solution is far from being
> elegant and sometimes the pseudo synchronization fails! Is this a
> Hamamatsu problem? With a Hamamatsu BT1024 which has a QE>90% it
> happends the same thing with metamorph.
>
> Regarding to the previous commercial response, does anyone have a
> cheaper and more flexible solution?
>

The way to do this is by directly controlling the shutter from the chip-exposure
line of the camera. This bypasses the software completely. Most high end cameras
  have such a control line, although it goes by various names. I know that the
PCO sensicam series, Roper Coolsnap series, Orca AG (probably the other Orcas
also),  QImaging, all have such a line. I use it with a Uniblitz shutter. You
still have to worry about the the 5-10ms shutter opening and closing time. In a
camera like the PCO (again, probably in others)  the camera can be set up to
start integrating after a delay time set by the user. This increases your sample
exposure by 10-20ms, but that is really not bad compared to the 1-2 seconds that
some software controlled systems waste before closing the shutter.

Of course, if you connect the control line directly to a system like CoolLed or
an AOTF, you can get very fast switching.

You may need to build a little bit of digital logic. For example, when doing a
live preview for focusing, we want the shutter to remain open continuously. But
when using the chip expose line, the shutter will cycle rapidly as the camera
switches between integration and image transfer. Therefore, we set it up so that
in live preview mode the Uniblitz system was commanded to stay open independent
of the hardware control line. I also added an external switch that could force
the input to the Uniblitz control to be on (ie, I OR'ed the camera control line
with a manually switched voltage). This was a matter of a few NAND gates on two
74LSxx series integrated circuits (your electronics shop will certainly have
these).

I know that this solution is used in other labs, so maybe I am missing something
in this thread about the difficulty that is being discussed.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050
Neville Sanjana Neville Sanjana
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi confocal list,

First, thanks to everyone for their wonderful suggestions.
Both commercial entities and scientists had lots of nice advice.

As for shuttering, I'm currently doing exactly what Aryeh
mentions: Using the direct chip-exposure hardware trigger
from the camera (CoolSnapHQ) to control a Uniblitz shutter
so that no software is invovled. (We use some OR and AND
gates on a breadboard as Aryeh described to control when the
shutter is responsive to the exposure signal of the
CoolSnap.) The CoolSnap can also be told to use a pre-set
delay between shutter opening and exposure, which we set at
5msec (the Uniblitz is rated for 3msec). Since we use a blue
LED as our excitation source, I could just turn that on and
off directly too (we were using a Hg lamp before). I think
the take-home message is that it is generally better to have
the camera directly control the shutter/LED rather than
worry about software synchronization, which will never be as
precise.

Since Cameron asked, I'm using custom Matlab software to
control the stage, camera, shutters, filters, etc. The PVCAM
C interface makes it easy to directly control the camera. I
am confident that the software is taking exposures
accurately. In particular, I'm taking about 200msec
exposures with 3x3 binning (blue LED, 60x/0.7NA objective,
CoolSnapHQ) every 10 minutes and the cells stay viable but
the signal is weak making it difficult for me to trace fine
structures in my cultured neurons. I'm working on changing
to a better objective and, as you know, a more sensitive CCD.


Still, I would welcome any further ideas on experiences with
different EMCCDs (not from commercial vendors) and ideas on
how to test EMCCDs. Right now, for demoing, I'm hoping to
take some dark frames at various exposure times (dark noise)
and some pictures of fluorescent beads with a set exposure
time (overall QE test). I could also test the read noise
with some bias images, but I'm hoping that these EMCCDs will
have no issue there.

Thanks again for all the help... I've certainly learned a
lot from your replies! Best,

- Neville

Aryeh Weiss wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Nuno Moreno wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi!
>>
>> If it is for cell imaging you might consider a BT camera.
>>
>> Forcing the previous point, which I thing is very important:
>>
>> I'm having the same problem with an ORCA AG with ImagePro Plus.  What
>> I have now is a small program called by macros that controls an
>> external shutter via TTL, which makes a delay till it opens the
>> shutter. I found out that the driver takes almost 500ms to get the
>> camera taking a picture which is unacceptable! However this solution
>> is far from being elegant and sometimes the pseudo synchronization
>> fails! Is this a Hamamatsu problem? With a Hamamatsu BT1024 which has
>> a QE>90% it happends the same thing with metamorph.
>>
>> Regarding to the previous commercial response, does anyone have a
>> cheaper and more flexible solution?
>>
>
> The way to do this is by directly controlling the shutter from the
> chip-exposure line of the camera. This bypasses the software completely.
> Most high end cameras  have such a control line, although it goes by
> various names. I know that the PCO sensicam series, Roper Coolsnap
> series, Orca AG (probably the other Orcas also),  QImaging, all have
> such a line. I use it with a Uniblitz shutter. You still have to worry
> about the the 5-10ms shutter opening and closing time. In a camera like
> the PCO (again, probably in others)  the camera can be set up to start
> integrating after a delay time set by the user. This increases your
> sample exposure by 10-20ms, but that is really not bad compared to the
> 1-2 seconds that some software controlled systems waste before closing
> the shutter.
>
> Of course, if you connect the control line directly to a system like
> CoolLed or an AOTF, you can get very fast switching.
>
> You may need to build a little bit of digital logic. For example, when
> doing a live preview for focusing, we want the shutter to remain open
> continuously. But when using the chip expose line, the shutter will
> cycle rapidly as the camera switches between integration and image
> transfer. Therefore, we set it up so that in live preview mode the
> Uniblitz system was commanded to stay open independent of the hardware
> control line. I also added an external switch that could force the input
> to the Uniblitz control to be on (ie, I OR'ed the camera control line
> with a manually switched voltage). This was a matter of a few NAND gates
> on two 74LSxx series integrated circuits (your electronics shop will
> certainly have these).
>
> I know that this solution is used in other labs, so maybe I am missing
> something in this thread about the difficulty that is being discussed.
>
> --aryeh
Chris Tully Chris Tully
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

In reply to this post by Nuno Moreno
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Nuno,

The drivers for all Hamamatsu cameras are written by Hamamatsu.  I would suggest that you contact them for help with your issue.  In my experience as a user of Image-Pro Plus, and as an employee of Media Cybernetics and as an employee of several of their dealers, Hamamatsu is very good about supporting their cameras on any platform and will gladly help you get it running as smoothly as possible.

Chris Tully

On Nov 28, 2007 8:59 AM, Nuno Moreno <[hidden email]> wrote:
Hi!

If it is for cell imaging you might consider a BT camera.

Forcing the previous point, which I thing is very important:

I'm having the same problem with an ORCA AG with ImagePro Plus.  What I
have now is a small program called by macros that controls an external
shutter via TTL, which makes a delay till it opens the shutter. I found
out that the driver takes almost 500ms to get the camera taking a
picture which is unacceptable! However this solution is far from being
elegant and sometimes the pseudo synchronization fails! Is this a
Hamamatsu problem? With a Hamamatsu BT1024 which has a QE>90% it
happends the same thing with metamorph.

Regarding to the previous commercial response, does anyone have a
cheaper and more flexible solution?

Regards,
NM





Nowell, Cameron wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Neville,
>               While an EMCCD camera will decrease your exposure time, you may find that it will not really help solve your phototoxicity problem. The time your cells are exposed to fluorecent light will not neccesarily be dependant on the exposure time of your camera, but on how fast your entire sytem (microscope, shutters, filters, etc.) shuffles everything around.
>
> What software etc are you using to capture your images?
>
> We have a wide filed live cell rig here that uses an Orca ER camera on an IX81 microscope and it is all driven by metamorph. Even when the camera is capturing at 100ms exposures, the samples are exposed to flourecent light for about 2 seconds. To get rid of these sort of delays you need to move to using filter wheels, or for even shorter exposuers something like olympus' CellR system
>
>
> Cheers
>
>
> Cam
>
>
> Cameron Nowell B. Sc (Hons)
>
>
>
> Microscopy Imaging and Research Core Facility
>
> Peter MacCallum Cancer Centre
>
> 7 St Andrews Place
>
> East Melbourne, Victoria 3002
>
> AUSTRALIA
>
>
>
> Phone: +61396561242
>
> Mobile: +61422882700
>
> Fax: +61396561411
>
>
>
> ________________________________
>
> From: Confocal Microscopy List on behalf of Neville Sanjana
> Sent: Wed 28/11/2007 7:57 AM
> To: [hidden email]
> Subject: Help with EMCCD evaluation
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
> I am in the market for an EMCCD for widefield fluorescence
> time-lapse imaging of genetically-labeled cultured cells. I
> am acquiring images frequently (every 10 minutes) for
> several days. The cells seem to be very sensitive to
> phototoxicity and photobleaching (GFP is the fluorophore),
> so I am hoping that replacing my Roper CoolSnapHQ CCD with
> an EMCCD will allow me to use much shorter exposure times.
>
> I realize that this is a confocal list and my application is
> widefield but since EMCCDs are also used with spinning disk
> confocal, I thought I'd ask. Apologies if this question has
> been hashed out before... I searched the archives without
> finding definitive recommendations. (Maybe that's asking for
> too much!) Any help on the following would be much appreciated:
>
> 1) What EMCCDs have people found to be the most *sensitive*

> (nice picture in low-light) and the most *reliable* (driver
> doesn't crash, etc.)? I'd especially be interested in any
> experiences or opinions with Roper/Photometrics's CascadeII
> or QuantEM and the Andor iXon cameras.
>
> 2) How have people been testing cameras? I've set up demos
> of several cameras so I'd like to do some standard tests.
> I've seen some references on the web for calculating dark
> noise, read noise, and gain, such as:
> http://www.qsimaging.com/ccd_noise_measure.html
> http://www.mirametrics.com/tech_note_ccdgain.htm
> http://www.photomet.com/library/library_encyclopedia/library_enc_gain.php
> but I'd love to hear what other tests people think are
> relevant for these cameras. Obviously, I'll try imaging my
> samples (using fiber-coupled LED as a constant illumination
> source)... what else should I do? Fluorescent beads instead
> of my cells perhaps?
>
> 3) For coupling the camera to my Olympus IX71, is there
> something like the Zeiss Optovar (which, as I understand it,
> has multiple selectable magnifications of the intermediate
> lens) made by a third party? This would be helpful for
> changing between achieving the full NA of the objective and
> getting less resolution but a bigger field of view for
> different experiments.
>
> Thanks for your help. Best,
>
> - Neville
>
> ---
> Neville Sanjana
> Dept. of Brain and Cognitive Science
> Massachusetts Institute of Technology
>
>
>
>
> This email (including any attachments) may contain
> confidential and/or legally privileged information and is
> intended only to be read or used by the addressee.  If you
> are not the intended addressee, any use, distribution,
> disclosure or copying of this email is strictly
> prohibited.
> Confidentiality and legal privilege attached to this email
> (including any attachments) are not waived or lost by
> reason of its mistaken delivery to you.
> If you have received this email in error, please delete it
> and notify us immediately by telephone or email.  Peter
> MacCallum Cancer Centre provides no guarantee that this
> transmission is free of virus or that it has not been
> intercepted or altered and will not be liable for any delay
> in its receipt.
>

--
Nuno Moreno
Cell Imaging Unit
Instituto Gulbenkian de Ciência
http://uic.igc.gulbekian.pt
http://www.igc.gulbekian.pt
phone +351 214464606
fax   +351 214407970

Ian Dobbie-2 Ian Dobbie-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

In reply to this post by Neville Sanjana
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Neville Sanjana <[hidden email]> writes:

> is taking exposures accurately. In particular, I'm taking about
> 200msec exposures with 3x3 binning (blue LED, 60x/0.7NA objective,

Totally unrelated to your EMCCD question but a cheaper probably more
effective solution may be a 60x 1.4NA oil immersion lens. This will
collect 4 times as much light and hence much brighter pictures (as
well as the improved resolution). Of course your particular
application may make this impossible.

Ian
Neville Sanjana Neville Sanjana
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

That's a great idea but there are a few particulars of my
application (automated long-term time lapse) that make that
difficult. First, my autofocus algorithms (second
derivative/Laplacian filter) seem to break at higher NAs due
to the smaller depth of field (and, hence, lower tolerance
for small errors in Z). Also, for long-term timelapse with
stage tiling, after a while the oil can get bubbles in it,
which ruins the image.

I am however thinking along the lines you suggested and
trying a 40x/NA0.9 air objective to see if that could work.

- Neville

Ian Dobbie wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Neville Sanjana <[hidden email]> writes:
>
>
>>is taking exposures accurately. In particular, I'm taking about
>>200msec exposures with 3x3 binning (blue LED, 60x/0.7NA objective,
>
>
> Totally unrelated to your EMCCD question but a cheaper probably more
> effective solution may be a 60x 1.4NA oil immersion lens. This will
> collect 4 times as much light and hence much brighter pictures (as
> well as the improved resolution). Of course your particular
> application may make this impossible.
>
> Ian
Chris Tully Chris Tully
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Help with EMCCD evaluation

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Neville,

First let me say that I work for Vashaw Scientific (a Leica Microsystems dealer) and have worked for other Microscope dealers and Media Cybernetics (makers of Image-Pro Plus).  I have also been a user of microscopes and imaging systems prior to my tenure on the commercial side.

With that said I would like to suggest trying less illumination (lower voltage to the LED) and longer exposures, since phototoxicity seems to be an issue in your experiments.  I do have experience with two customers who had similar problems.  One used a Photometics Sensys 1400 (the best camera at the time) with a 16 bit digitizer in maximum gain mode and still only acquired for long enough to get ~200 gray levels from the camera in the brightest pixels.  The second put three neutral density filters (totaling 1/512 of the Hg bulb's illumination reaching the sample).  With the neutral density filters in place, his cells which bleached within seconds at full 100W Hg intensity survived 24 hours of illumination in good health.  We did have to extend his exposures to 1 minute to get reasonable signal levels and perform a dark frame subtraction.  But with those two steps he got beautiful, detailed images.

Although I cannot guarantee that my suggestion will work for you I think that it may be worth playing with different illumination intensities and exposure times.

Chris Tully
Vashaw Scientific Inc.
[hidden email]


On Dec 12, 2007 5:17 PM, Neville Sanjana <[hidden email]> wrote:
That's a great idea but there are a few particulars of my
application (automated long-term time lapse) that make that
difficult. First, my autofocus algorithms (second
derivative/Laplacian filter) seem to break at higher NAs due
to the smaller depth of field (and, hence, lower tolerance
for small errors in Z). Also, for long-term timelapse with
stage tiling, after a while the oil can get bubbles in it,
which ruins the image.

I am however thinking along the lines you suggested and
trying a 40x/NA0.9 air objective to see if that could work.

- Neville

Ian Dobbie wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Neville Sanjana <[hidden email]> writes:
>
>
>>is taking exposures accurately. In particular, I'm taking about
>>200msec exposures with 3x3 binning (blue LED, 60x/0.7NA objective,
>
>
> Totally unrelated to your EMCCD question but a cheaper probably more
> effective solution may be a 60x 1.4NA oil immersion lens. This will
> collect 4 times as much light and hence much brighter pictures (as
> well as the improved resolution). Of course your particular
> application may make this impossible.
>
> Ian