How to install a DPSS 561 on a Leica SP2

Hello List,

I am planning to install a DPSS 561 laser on a Leica SP2-AOBS, in place of the exhausted HeNe 543.

Beside the Leica upgrade kit, I am also considering to directly buy the laser myself and have Leica do just the alignment and AOTF/AOBS software reprogramming.  My concern is then how to bring the DPPS beam into the lasers-merge module inside the SP2 laser box.
I will need either a laser holder to place the DPSS into the laser box in free-space, or a fiber-coupler to the lasers-merge module so that I can keep the DPSS outside.

Has anyone done this already and can give suggestions?

I am also very interested in feedback from DPSS561 users!! I want to do FRAP and long-term time lapse, is a 25mW a reasonable size to go for (also considering the power losses in case of fiber-coupling the laser) ?

Thanks!

Tommaso







-- 

Tommaso Mello, PhD.
Gastroenterology Unit
Dept. of Clinical Pathophysiology 
University of Florence, Italy
Phone:  +39.055.4271-294
FAX:    +39.055.4271-297
Rietdorf, Jens Rietdorf, Jens
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to install a DPSS 561 on a Leica SP2

Hello Tommaso,

 

25mW is enough power on a single point scanner, the los by fiber coupling is rarely more than 50%. The HeNe 543 was typically only 1mW and even that was sufficient for bleaching at least with high NA lenses.

 

Cheers, j

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Tommaso Mello
Sent: Monday, August 30, 2010 13:13 PM
To: [hidden email]
Subject: How to install a DPSS 561 on a Leica SP2

 

Hello List,

I am planning to install a DPSS 561 laser on a Leica SP2-AOBS, in place of the exhausted HeNe 543.

Beside the Leica upgrade kit, I am also considering to directly buy the laser myself and have Leica do just the alignment and AOTF/AOBS software reprogramming.  My concern is then how to bring the DPPS beam into the lasers-merge module inside the SP2 laser box.
I will need either a laser holder to place the DPSS into the laser box in free-space, or a fiber-coupler to the lasers-merge module so that I can keep the DPSS outside.

Has anyone done this already and can give suggestions?

I am also very interested in feedback from DPSS561 users!! I want to do FRAP and long-term time lapse, is a 25mW a reasonable size to go for (also considering the power losses in case of fiber-coupling the laser) ?

Thanks!

Tommaso






 
-- 
 
Tommaso Mello, PhD.
Gastroenterology Unit
Dept. of Clinical Pathophysiology 
University of Florence, Italy
Phone:  +39.055.4271-294
FAX:    +39.055.4271-297

 

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to install a DPSS 561 on a Leica SP2

I second this.  If you have power to spare fiber coupling will be simplest.

Craig


On Mon, Aug 30, 2010 at 8:25 AM, Rietdorf, Jens <[hidden email]> wrote:

Hello Tommaso,

 

25mW is enough power on a single point scanner, the los by fiber coupling is rarely more than 50%. The HeNe 543 was typically only 1mW and even that was sufficient for bleaching at least with high NA lenses.

 

Cheers, j

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Tommaso Mello
Sent: Monday, August 30, 2010 13:13 PM
To: [hidden email]
Subject: How to install a DPSS 561 on a Leica SP2

 

Hello List,

I am planning to install a DPSS 561 laser on a Leica SP2-AOBS, in place of the exhausted HeNe 543.

Beside the Leica upgrade kit, I am also considering to directly buy the laser myself and have Leica do just the alignment and AOTF/AOBS software reprogramming.  My concern is then how to bring the DPPS beam into the lasers-merge module inside the SP2 laser box.
I will need either a laser holder to place the DPSS into the laser box in free-space, or a fiber-coupler to the lasers-merge module so that I can keep the DPSS outside.

Has anyone done this already and can give suggestions?

I am also very interested in feedback from DPSS561 users!! I want to do FRAP and long-term time lapse, is a 25mW a reasonable size to go for (also considering the power losses in case of fiber-coupling the laser) ?

Thanks!

Tommaso






 
-- 
 
Tommaso Mello, PhD.
Gastroenterology Unit
Dept. of Clinical Pathophysiology 
University of Florence, Italy
Phone:  +39.055.4271-294
FAX:    +39.055.4271-297

 


Tommaso Mello Tommaso Mello
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to install a DPSS 561 on a Leica SP2

Jens and Craig,

thank you for your comments! I will inquire Leica if they can produce an adapter to fiber-couple the laser merge module. Hopefully they won't charge the whole "upgrade kit" for that!

Cheers,

Tommaso



On lun, 2010-08-30 at 11:24 -0600, Craig Brideau wrote:
I second this.  If you have power to spare fiber coupling will be simplest.

Craig


On Mon, Aug 30, 2010 at 8:25 AM, Rietdorf, Jens <[hidden email]> wrote:
Hello Tommaso,

 

25mW is enough power on a single point scanner, the los by fiber coupling is rarely more than 50%. The HeNe 543 was typically only 1mW and even that was sufficient for bleaching at least with high NA lenses.

 

Cheers, j

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Tommaso Mello
Sent: Monday, August 30, 2010 13:13 PM
To: [hidden email]
Subject: How to install a DPSS 561 on a Leica SP2



 

Hello List,

I am planning to install a DPSS 561 laser on a Leica SP2-AOBS, in place of the exhausted HeNe 543.

Beside the Leica upgrade kit, I am also considering to directly buy the laser myself and have Leica do just the alignment and AOTF/AOBS software reprogramming.  My concern is then how to bring the DPPS beam into the lasers-merge module inside the SP2 laser box.
I will need either a laser holder to place the DPSS into the laser box in free-space, or a fiber-coupler to the lasers-merge module so that I can keep the DPSS outside.

Has anyone done this already and can give suggestions?

I am also very interested in feedback from DPSS561 users!! I want to do FRAP and long-term time lapse, is a 25mW a reasonable size to go for (also considering the power losses in case of fiber-coupling the laser) ?

Thanks!

Tommaso








 
-- 
 
Tommaso Mello, PhD.
Gastroenterology Unit
Dept. of Clinical Pathophysiology 
University of Florence, Italy
Phone:  +39.055.4271-294
FAX:    +39.055.4271-297