How to upgrade an old fluorescence microscope to LED fluorescence microscope?

classic Classic list List threaded Threaded
♦
♦
5 messages Options
Manja Schubert-2 Manja Schubert-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

How to upgrade an old fluorescence microscope to LED fluorescence microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi,

The Hg lamp transformer on our Leitz DMRBE microscope (manufactured in the early 90's) died recently. Since, getting replacement parts for this "old, but so nicely functioning" microscope becomes more and more difficult we were wondering whether it is possible to upgrade this system to a LED fluorescence microscope.
The optics are in good shape, all of the dimensional stages worked just fine but the UV light source causes headache. The microscope is not used for live cell imaging but for screening tissue samples for autoimmune reactions to certain neuronal proteins.

It would be great if  I could get advice on what is needed to turn that microscope into a LED fluorescence microscope.
Any advice, shared experience would be very helpful.


Manja


--
Dr. Manja Schubert
University of Bergen
Department of Clinical Medicine
Haukeland University Hospital
Section Neurology
Paraneoplastic Syndromes group
5021 Bergen
Norway
Avi Jacob Avi Jacob
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to upgrade an old fluorescence microscope to LED fluorescence microscope?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Manja,

One of our previous workers, Sergei, did something similar with our FV1000.
I never used a DMRBE, but it is probably a similar idea.

See the image a linked here :  http://imgur.com/5zGFdT9
"1" is the coupler from the LED light source (it came as part of the
package - the light guide goes into the back side and is locked in by the
silver knob seen in the picture).
This coupler fits into "3", which leads to the microscope via another fiber
(or directly into the microscope body).

The old part ( http://imgur.com/Wyrztf9 ) used to connect to "3".
But as the new coupler is slightly smaller than it, we had to have a ring
made "2",  that allowed the new coupler to remain centered and be locked
in.
(Hint: The ring itself has a tiny screw in it to lock the ring itself to
the coupler, then the coupler+ring is locked in by the externally
accessible screw.)

This won't work of course, if your original is already narrower than the
new part, something I doubt.

I hope this helps,

Avi





On Thu, Nov 27, 2014 at 12:32 PM, Manja Schubert <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
>
> The Hg lamp transformer on our Leitz DMRBE microscope (manufactured in the
> early 90's) died recently. Since, getting replacement parts for this "old,
> but so nicely functioning" microscope becomes more and more difficult we
> were wondering whether it is possible to upgrade this system to a LED
> fluorescence microscope.
> The optics are in good shape, all of the dimensional stages worked just
> fine but the UV light source causes headache. The microscope is not used
> for live cell imaging but for screening tissue samples for autoimmune
> reactions to certain neuronal proteins.
>
> It would be great if  I could get advice on what is needed to turn that
> microscope into a LED fluorescence microscope.
> Any advice, shared experience would be very helpful.
>
>
> Manja
>
>
> --
> Dr. Manja Schubert
> University of Bergen
> Department of Clinical Medicine
> Haukeland University Hospital
> Section Neurology
> Paraneoplastic Syndromes group
> 5021 Bergen
> Norway
>
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to upgrade an old fluorescence microscope to LED fluorescence microscope?

In reply to this post by Manja Schubert-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Manja,

consider ThorLabs LED4 (your choice of wavelengths), controller, liquid
light guide and LLG adapter.

http://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=3836&pn=LED4D249

Your microscope probably has old, degraded, Leica branded, filters. The
current generation hard coated filter sets should perform much better.
Simplest is to purchase one Quad cube and purchase ThorLabs LEDs that
match the exciter peaks of -- for example -- the Semrock
LED-DA/FI/TR/Cy5-A-000

http://www.semrock.com/SetDetails.aspx?id=3302

which is what I have on my lab's DMI6000 microscope

You can use Semrock SearchLight or Chroma's equivalent spectral viewer,
to evaluate additional options.


George
p.s. for more details on my lab's microscope, see my recent message,
044830
<http://lists.umn.edu/cgi-bin/wa?A2=ind1411&L=CONFOCALMICROSCOPY&P=R15874&X=7EE65D5189E5098F4E&Y=geomcnamara%40earthlink.net>
        2014-11-22 09:41 133 Re: Spectra data for newest FPs



On 11/27/2014 4:32 AM, Manja Schubert wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
>
> The Hg lamp transformer on our Leitz DMRBE microscope (manufactured in the early 90's) died recently. Since, getting replacement parts for this "old, but so nicely functioning" microscope becomes more and more difficult we were wondering whether it is possible to upgrade this system to a LED fluorescence microscope.
> The optics are in good shape, all of the dimensional stages worked just fine but the UV light source causes headache. The microscope is not used for live cell imaging but for screening tissue samples for autoimmune reactions to certain neuronal proteins.
>
> It would be great if  I could get advice on what is needed to turn that microscope into a LED fluorescence microscope.
> Any advice, shared experience would be very helpful.
>
>
> Manja
>
>
>    


--



George McNamara, Ph.D.
Single Cells Analyst
L.J.N. Cooper Lab
University of Texas M.D. Anderson Cancer Center
Houston, TX 77054
Tattletales http://works.bepress.com/gmcnamara/42
jerie jerie
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to upgrade an old fluorescence microscope to LED fluorescence microscope?

In reply to this post by Manja Schubert-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Maria,

you might want to check CoolLED (no commercial interest). We have
successfully used their products to upgrade several systems from Hg to
LED and are very happy with their solutions.

They have a choice of adaptors,
http://www.coolled.com/product-detail/adaptors-2/
Leica fittings for CoolLED pE-series products

It may be better though to contact them for advice on your system.

Hope that helps.
kind regards, Jens

On 11/27/14, Manja Schubert <[hidden email]> wrote:

> Hi,
>
> The Hg lamp transformer on our Leitz DMRBE microscope (manufactured in the
> early 90's) died recently...
> ...It would be great if  I could get advice on what is needed to turn that
> microscope into a LED fluorescence microscope.
> Any advice, shared experience would be very helpful.
>
>
> Manja
>
>
> --
> Dr. Manja Schubert
> University of Bergen
> Department of Clinical Medicine
> Haukeland University Hospital
> Section Neurology
> Paraneoplastic Syndromes group
> 5021 Bergen
> Norway
>


--
Visiting Scientist @ Center for Technological Development in Health (CDTS),
Oswaldo Cruz Foundation (Fiocruz), Ministry of Health, Rio de Janeiro,
Brazil.
http://br.linkedin.com/pub/jens-rietdorf/6/4a3/189/
Skype jens.rietdorf
Gerard Whoriskey Gerard Whoriskey
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: How to upgrade an old fluorescence microscope to LED fluorescence microscope?

In reply to this post by Manja Schubert-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Manja,

I first of all must declare a commercial bias.

There should be no problem putting an LED excitation system onto your Leica scope.

One thing to be aware of is that most LED systems do not sufficiently cover the full
spectrum like the bulbs did. This should not be a problem if you can match up the LED peaks
with the excitation filter passbands that are in your microscope. If you are not sure just go
for an LED system like our CoolLED pE-4000 that has 16 peaks from 365-770nm.

Best regards,

Gerry