To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello, Does anybody know of microscopy techniques for 3D imaging of paper or other cellulose based fabrics? We have been trying with confocal microscopy and did imaging of the surface of the outer layer of cellulose fibers, but were unable to go through the fibers and image the underlying layers. Best regards, Dario -- Dario Donnarumma PhD student personal blog <http://dariodonnarumma.blogspot.it/> tel: +39 081 7682261 fax: +39 081 2391800 *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di Ingegneria Chimica, * *dei Materiali e della Produzione Industriale (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. Tecchio 80 Napoli* *I-80125 ITALIA* |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dario- There is a long history of imaging paper by scanning electron microscopy. Again one is limited to viewing the surface, but with stereo microscopy, one can see the surface in 3D. Carol Heckman Bowling Green State University ________________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Dar Do <[hidden email]> Sent: Tuesday, May 6, 2014 9:58 AM To: [hidden email] Subject: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello, Does anybody know of microscopy techniques for 3D imaging of paper or other cellulose based fabrics? We have been trying with confocal microscopy and did imaging of the surface of the outer layer of cellulose fibers, but were unable to go through the fibers and image the underlying layers. Best regards, Dario -- Dario Donnarumma PhD student personal blog <http://dariodonnarumma.blogspot.it/> tel: +39 081 7682261 fax: +39 081 2391800 *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di Ingegneria Chimica, * *dei Materiali e della Produzione Industriale (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. Tecchio 80 Napoli* *I-80125 ITALIA* |
|
|
In reply to this post by Dar Do
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A good method might be to use a Xray CT scanner. Most manufacturers include software for analysis and volume computations.. On Tue, May 6, 2014 at 9:58 AM, Dar Do <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or other > cellulose based fabrics? We have been trying with confocal microscopy and > did imaging of the surface of the outer layer of cellulose fibers, but were > unable to go through the fibers and image the underlying layers. > Best regards, > Dario > -- > Dario Donnarumma > PhD student > personal blog <http://dariodonnarumma.blogspot.it/> > tel: +39 081 7682261 > fax: +39 081 2391800 > *Chemical Engineering @ the Micro-Scale (ChemEMS)* > *Dipartimento di Ingegneria Chimica, * > *dei Materiali e della Produzione Industriale (DICMAPI)* > *Università degli Studi di Napoli Federico II* > *Piazzale V. Tecchio 80 Napoli* > *I-80125 ITALIA* > |
|
Oshel, Philip Eugene |
|
|
In reply to this post by Dar Do
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Scanning electron microscopy works very well for imaging papers and other such cellulosic products and their coatings. Also, the paper (etc.) can be "fractured" in liquid nitrogen to produce high quality edges to image the inner layers of the paper (etc.), and the coating layers. Cutting tends to smear the surface over the cut edge - this is more of an issue with coated papers. Phil > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or other > cellulose based fabrics? We have been trying with confocal microscopy and > did imaging of the surface of the outer layer of cellulose fibers, but were > unable to go through the fibers and image the underlying layers. > Best regards, > Dario -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
|
|
In reply to this post by Dar Do
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Heyo, On thing you could try would be to soak the paper in fluorescein, wash out the excess, and then mount the paper in water or glycerol (which is closer to the actual refractive index of cellulose - 1.469). Multi-photon will allow you look even further into the sample, if available. As others suggested, a normally more expensive option, but one that would allow you to see through the full depth of the sample with equal resolution, would be to use a microCT. Cheers, Ben Smith Benjamin E. Smith, Ph.D. Samuel Roberts Noble Microscopy Laboratory Research Scientist II University of Oklahoma Norman, OK 73019 E-mail: [hidden email] Voice 405-325-4391 FAX 405-325-7619 http://www.microscopy.ou.edu/ ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Dar Do [[hidden email]] Sent: Tuesday, May 06, 2014 8:58 AM To: [hidden email] Subject: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello, Does anybody know of microscopy techniques for 3D imaging of paper or other cellulose based fabrics? We have been trying with confocal microscopy and did imaging of the surface of the outer layer of cellulose fibers, but were unable to go through the fibers and image the underlying layers. Best regards, Dario -- Dario Donnarumma PhD student personal blog <http://dariodonnarumma.blogspot.it/> tel: +39 081 7682261 fax: +39 081 2391800 *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di Ingegneria Chimica, * *dei Materiali e della Produzione Industriale (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. Tecchio 80 Napoli* *I-80125 ITALIA* |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Second harmonic microscopy does an excellent job for this sort of thing if you happen to have access to a 2-photon microscope: http://www.opticsinfobase.org/ol/abstract.cfm?uri=ol-28-22-2207 Craig Brideau On Tue, May 6, 2014 at 9:50 AM, Smith, Benjamin E. <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Heyo, > On thing you could try would be to soak the paper in fluorescein, wash > out the excess, and then mount the paper in water or glycerol (which is > closer to the actual refractive index of cellulose - 1.469). Multi-photon > will allow you look even further into the sample, if available. > > As others suggested, a normally more expensive option, but one that > would allow you to see through the full depth of the sample with equal > resolution, would be to use a microCT. > > Cheers, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory > Research Scientist II > University of Oklahoma > Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Dar Do [[hidden email]] > Sent: Tuesday, May 06, 2014 8:58 AM > To: [hidden email] > Subject: IMAGING OF CELLULOSE > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or other > cellulose based fabrics? We have been trying with confocal microscopy and > did imaging of the surface of the outer layer of cellulose fibers, but were > unable to go through the fibers and image the underlying layers. > Best regards, > Dario > -- > Dario Donnarumma > PhD student > personal blog <http://dariodonnarumma.blogspot.it/> > tel: +39 081 7682261 > fax: +39 081 2391800 > *Chemical Engineering @ the Micro-Scale (ChemEMS)* > *Dipartimento di Ingegneria Chimica, * > *dei Materiali e della Produzione Industriale (DICMAPI)* > *Università degli Studi di Napoli Federico II* > *Piazzale V. Tecchio 80 Napoli* > *I-80125 ITALIA* > |
|
|
Yes I think second harmonic generation is useful in imaging cellulose in paper. In fact I had some experience in imaging paper ( cigarette paper) using two-photon fluorescence, second harmonic generation and coherent anti-stokes Raman. Based on the material you may be able to see strong 2-photon signal and second-harmonic signal.
-----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: May-06-14 11:16 AM To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Second harmonic microscopy does an excellent job for this sort of thing if you happen to have access to a 2-photon microscope: http://www.opticsinfobase.org/ol/abstract.cfm?uri=ol-28-22-2207 Craig Brideau On Tue, May 6, 2014 at 9:50 AM, Smith, Benjamin E. <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Heyo, > On thing you could try would be to soak the paper in fluorescein, > wash out the excess, and then mount the paper in water or glycerol > (which is closer to the actual refractive index of cellulose - 1.469). > Multi-photon will allow you look even further into the sample, if available. > > As others suggested, a normally more expensive option, but one > that would allow you to see through the full depth of the sample with > equal resolution, would be to use a microCT. > > Cheers, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory Research Scientist II > University of Oklahoma Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Dar Do [[hidden email]] > Sent: Tuesday, May 06, 2014 8:58 AM > To: [hidden email] > Subject: IMAGING OF CELLULOSE > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or > other cellulose based fabrics? We have been trying with confocal > microscopy and did imaging of the surface of the outer layer of > cellulose fibers, but were unable to go through the fibers and image the underlying layers. > Best regards, > Dario > -- > Dario Donnarumma > PhD student > personal blog <http://dariodonnarumma.blogspot.it/> > tel: +39 081 7682261 > fax: +39 081 2391800 > *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di > Ingegneria Chimica, * *dei Materiali e della Produzione Industriale > (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. > Tecchio 80 Napoli* > *I-80125 ITALIA* > |
|
|
In reply to this post by Oshel, Philip Eugene
For confocal microscopy use congo red, pontamine fast scarlet or calcofluor to stain cellulose. Alternatively fluorescently labelled cellulose binding domains are available or you can just fluorescently label some cellulase enzyme. However visualising beyond the surface layer will be difficult. To see the cross-section you can embed the paper in resin and cut cross-sections and stain them with toluidine blue for brightfield microscopy. You can also polish the cross-section of embedded paper, stain with iodine vapour and image with SEM in backscattered mode (also useful for visualising clay fillers or coatings. There is an extensive literature on this topic mainly in specialist pulp & paper journals.
Dr Lloyd Donaldson Project Leader – Microscopy & Wood Identification Senior Scientist – Plant Cell Walls & Biomaterials Scion – Forests, Products, Innovation 49 Sala Street, Rotorua 3010 New Zealand Ph 07 343 5581 www.scionresearch.com -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philip Oshel Sent: Wednesday, 7 May 2014 2:48 a.m. To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Scanning electron microscopy works very well for imaging papers and other such cellulosic products and their coatings. Also, the paper (etc.) can be "fractured" in liquid nitrogen to produce high quality edges to image the inner layers of the paper (etc.), and the coating layers. Cutting tends to smear the surface over the cut edge - this is more of an issue with coated papers. Phil > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or > other cellulose based fabrics? We have been trying with confocal > microscopy and did imaging of the surface of the outer layer of > cellulose fibers, but were unable to go through the fibers and image the underlying layers. > Best regards, > Dario -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 This e-mail and any attachments may contain information which is confidential or subject to copyright. If you receive this e-mail in error, please delete it. Scion does not accept responsibility for anything in this e-mail which is not provided in the course of Scion’s usual business or for any computer virus, data corruption, interference or delay arising from this e-mail. |
|
|
In reply to this post by Ko, Alex
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I actually use paper as a crude test target for 2-p systems sometimes. The white bleach in the paper fluoresces nicely, and I can create specific emission wavelengths by tuning the paper to make sure that each channel is detecting properly, at least for blue and green wavelengths. Craig On Tue, May 6, 2014 at 1:27 PM, Ko, Alex <[hidden email]> wrote: > Yes I think second harmonic generation is useful in imaging cellulose in > paper. In fact I had some experience in imaging paper ( cigarette paper) > using two-photon fluorescence, second harmonic generation and coherent > anti-stokes Raman. Based on the material you may be able to see strong > 2-photon signal and second-harmonic signal. > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Craig Brideau > Sent: May-06-14 11:16 AM > To: [hidden email] > Subject: Re: IMAGING OF CELLULOSE > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Second harmonic microscopy does an excellent job for this sort of thing if > you happen to have access to a 2-photon microscope: > http://www.opticsinfobase.org/ol/abstract.cfm?uri=ol-28-22-2207 > > Craig Brideau > > > On Tue, May 6, 2014 at 9:50 AM, Smith, Benjamin E. <[hidden email] > >wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Heyo, > > On thing you could try would be to soak the paper in fluorescein, > > wash out the excess, and then mount the paper in water or glycerol > > (which is closer to the actual refractive index of cellulose - 1.469). > > Multi-photon will allow you look even further into the sample, if > available. > > > > As others suggested, a normally more expensive option, but one > > that would allow you to see through the full depth of the sample with > > equal resolution, would be to use a microCT. > > > > Cheers, > > Ben Smith > > > > Benjamin E. Smith, Ph.D. > > Samuel Roberts Noble Microscopy Laboratory Research Scientist II > > University of Oklahoma Norman, OK 73019 > > E-mail: [hidden email] > > Voice 405-325-4391 > > FAX 405-325-7619 > > http://www.microscopy.ou.edu/ > > ________________________________________ > > From: Confocal Microscopy List [[hidden email]] on > > behalf of Dar Do [[hidden email]] > > Sent: Tuesday, May 06, 2014 8:58 AM > > To: [hidden email] > > Subject: IMAGING OF CELLULOSE > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hello, > > Does anybody know of microscopy techniques for 3D imaging of paper or > > other cellulose based fabrics? We have been trying with confocal > > microscopy and did imaging of the surface of the outer layer of > > cellulose fibers, but were unable to go through the fibers and image the > underlying layers. > > Best regards, > > Dario > > -- > > Dario Donnarumma > > PhD student > > personal blog <http://dariodonnarumma.blogspot.it/> > > tel: +39 081 7682261 > > fax: +39 081 2391800 > > *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di > > Ingegneria Chimica, * *dei Materiali e della Produzione Industriale > > (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. > > Tecchio 80 Napoli* > > *I-80125 ITALIA* > > > |
|
|
In reply to this post by Lloyd Donaldson
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes, exactly as Lloyd says below. There are two other things to overcome here - light scattering with depth and light absorption with depth, and I think these are touched on in some of the literature Lloyd refers to. The inability to penetrate deeply - a depth/scattering problem - can be ameliorated by matching the refractive index of the immersion medium with that of the fabric. This might need to be determined empirically, but will be around 1.5. Glycerol, at RI 1.47, is close. However, there is also a tendency to overstain cellulose - it looks great when it's really bright, but if it's very thoroughly stained, the fibres will absorb more of the laser light, leaving less to penetrate into deeper layers. You may find that with lighter staining, you can resolve some of the deeper layers of the material. If your material has been dyed with something very absorbant, this will be difficult.... cheers, Rosemary White Dr Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 E [hidden email] On 7/05/14 5:55 AM, "Lloyd Donaldson" <[hidden email]> wrote: >For confocal microscopy use congo red, pontamine fast scarlet or >calcofluor to stain cellulose. Alternatively fluorescently labelled >cellulose binding domains are available or you can just fluorescently >label some cellulase enzyme. However visualising beyond the surface layer >will be difficult. To see the cross-section you can embed the paper in >resin and cut cross-sections and stain them with toluidine blue for >brightfield microscopy. You can also polish the cross-section of embedded >paper, stain with iodine vapour and image with SEM in backscattered mode >(also useful for visualising clay fillers or coatings. There is an >extensive literature on this topic mainly in specialist pulp & paper >journals. > > >Dr Lloyd Donaldson >Project Leader Microscopy & Wood Identification >Senior Scientist Plant Cell Walls & Biomaterials >Scion Forests, Products, Innovation >49 Sala Street, Rotorua 3010 >New Zealand >Ph 07 343 5581 >www.scionresearch.com > > > >-----Original Message----- >From: Confocal Microscopy List [mailto:[hidden email]] >On Behalf Of Philip Oshel >Sent: Wednesday, 7 May 2014 2:48 a.m. >To: [hidden email] >Subject: Re: IMAGING OF CELLULOSE > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Scanning electron microscopy works very well for imaging papers and other >such cellulosic products and their coatings. >Also, the paper (etc.) can be "fractured" in liquid nitrogen to produce >high quality edges to image the inner layers of the paper (etc.), and the >coating layers. Cutting tends to smear the surface over the cut edge >- this is more of an issue with coated papers. > >Phil > >> Hello, >> Does anybody know of microscopy techniques for 3D imaging of paper or >> other cellulose based fabrics? We have been trying with confocal >> microscopy and did imaging of the surface of the outer layer of >> cellulose fibers, but were unable to go through the fibers and image >>the underlying layers. >> Best regards, >> Dario > >-- >Philip Oshel >Microscopy Facility Supervisor >Biology Department >024C Brooks Hall >Central Michigan University >Mt. Pleasant, MI 48859 >(989) 774-3576 > > > >This e-mail and any attachments may contain information which is >confidential or subject to copyright. If you receive this e-mail in >error, please delete it. >Scion does not accept responsibility for anything in this e-mail which is >not provided in the course of Scion¹s usual business or for any computer >virus, data corruption, interference or delay arising from this e-mail. |
|
|
In reply to this post by Dar Do
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have recently done some confocal imaging of cotton fibers. For deeper penetration, less scattering in single photon fluorescence imaging, rather than Calcofluor White I suggest staining with orange or red-fluorescing dyes - some have been already mentioned, like Congo Red. The fluorescent version of Periodic Acid-Schiff staining works great (use propidium iodide as pseudo-Schiff reagent; Truernit, E., H. Bauby, et al. (2008). "High- resolution whole-mount imaging of three-dimensional tissue organization and gene expression enables the study of Phloem development and structure in Arabidopsis." Plant Cell 20(6): 1494-1503). As an alternative to fluorescence staining, I have also used vapor fixation with acrolein, which tends to induce very strong autofluorescence, at least in cotton fibers. This is useful if you do not want to expose the fibers to water and induce swelling. For imaging with oil immersion objectives I used 2,2-thioidiethanol as a mounting medium with good success. Staudt, T., M. C. Lang, et al. (2007). "2,2'-thiodiethanol: a new water soluble mounting medium for high resolution optical microscopy." Microscopy Research and Technique 70(1): 1-9. Stan Vitha Microscopy and Imaging Center Texas A&M University |
|
|
In reply to this post by Ko, Alex
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Cellulose is a beta 1-4 linked glucan. This gives it a degree of symmetry since each glucose molecule is reversed relative to its neighbour. That in turn make it a relatively poor SH emitter. Nevertheless it can be images, especially if nothing else is there. Starch, an alpha 1-4 linked glucan has all the glucose molecules in the same orientation and is therefore non-symmetric and a much stronger emitter. So imaging cellulose when starch is present will be difficult. Also, in plant material, 3-photon fluorescence from lignin may be a confounding factor. Guy ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Ko, Alex [[hidden email]] Sent: 07 May 2014 05:27 To: [hidden email] Subject: Re: IMAGING OF CELLULOSE Yes I think second harmonic generation is useful in imaging cellulose in paper. In fact I had some experience in imaging paper ( cigarette paper) using two-photon fluorescence, second harmonic generation and coherent anti-stokes Raman. Based on the material you may be able to see strong 2-photon signal and second-harmonic signal. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: May-06-14 11:16 AM To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Second harmonic microscopy does an excellent job for this sort of thing if you happen to have access to a 2-photon microscope: http://www.opticsinfobase.org/ol/abstract.cfm?uri=ol-28-22-2207 Craig Brideau On Tue, May 6, 2014 at 9:50 AM, Smith, Benjamin E. <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Heyo, > On thing you could try would be to soak the paper in fluorescein, > wash out the excess, and then mount the paper in water or glycerol > (which is closer to the actual refractive index of cellulose - 1.469). > Multi-photon will allow you look even further into the sample, if available. > > As others suggested, a normally more expensive option, but one > that would allow you to see through the full depth of the sample with > equal resolution, would be to use a microCT. > > Cheers, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory Research Scientist II > University of Oklahoma Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Dar Do [[hidden email]] > Sent: Tuesday, May 06, 2014 8:58 AM > To: [hidden email] > Subject: IMAGING OF CELLULOSE > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello, > Does anybody know of microscopy techniques for 3D imaging of paper or > other cellulose based fabrics? We have been trying with confocal > microscopy and did imaging of the surface of the outer layer of > cellulose fibers, but were unable to go through the fibers and image the underlying layers. > Best regards, > Dario > -- > Dario Donnarumma > PhD student > personal blog <http://dariodonnarumma.blogspot.it/> > tel: +39 081 7682261 > fax: +39 081 2391800 > *Chemical Engineering @ the Micro-Scale (ChemEMS)* *Dipartimento di > Ingegneria Chimica, * *dei Materiali e della Produzione Industriale > (DICMAPI)* *Università degli Studi di Napoli Federico II* *Piazzale V. > Tecchio 80 Napoli* > *I-80125 ITALIA* > |
|
|
In reply to this post by Stanislav Vitha-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I totally endorse TDE as an embedding medium but I am puzzled by acrolein inducing fluorescence in cellulose. Can someone explain the chemistry? I thought acroleon reacted with proteins. Guy ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Stanislav Vitha [[hidden email]] Sent: 08 May 2014 00:04 To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have recently done some confocal imaging of cotton fibers. For deeper penetration, less scattering in single photon fluorescence imaging, rather than Calcofluor White I suggest staining with orange or red-fluorescing dyes - some have been already mentioned, like Congo Red. The fluorescent version of Periodic Acid-Schiff staining works great (use propidium iodide as pseudo-Schiff reagent; Truernit, E., H. Bauby, et al. (2008). "High- resolution whole-mount imaging of three-dimensional tissue organization and gene expression enables the study of Phloem development and structure in Arabidopsis." Plant Cell 20(6): 1494-1503). As an alternative to fluorescence staining, I have also used vapor fixation with acrolein, which tends to induce very strong autofluorescence, at least in cotton fibers. This is useful if you do not want to expose the fibers to water and induce swelling. For imaging with oil immersion objectives I used 2,2-thioidiethanol as a mounting medium with good success. Staudt, T., M. C. Lang, et al. (2007). "2,2'-thiodiethanol: a new water soluble mounting medium for high resolution optical microscopy." Microscopy Research and Technique 70(1): 1-9. Stan Vitha Microscopy and Imaging Center Texas A&M University |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Cotton does contain small amounts of protein. It is unlikely that the cellulose would fluoresce. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Thursday, 8 May 2014 3:54 a.m. To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I totally endorse TDE as an embedding medium but I am puzzled by acrolein inducing fluorescence in cellulose. Can someone explain the chemistry? I thought acroleon reacted with proteins. Guy ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Stanislav Vitha [[hidden email]] Sent: 08 May 2014 00:04 To: [hidden email] Subject: Re: IMAGING OF CELLULOSE ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have recently done some confocal imaging of cotton fibers. For deeper penetration, less scattering in single photon fluorescence imaging, rather than Calcofluor White I suggest staining with orange or red-fluorescing dyes - some have been already mentioned, like Congo Red. The fluorescent version of Periodic Acid-Schiff staining works great (use propidium iodide as pseudo-Schiff reagent; Truernit, E., H. Bauby, et al. (2008). "High- resolution whole-mount imaging of three-dimensional tissue organization and gene expression enables the study of Phloem development and structure in Arabidopsis." Plant Cell 20(6): 1494-1503). As an alternative to fluorescence staining, I have also used vapor fixation with acrolein, which tends to induce very strong autofluorescence, at least in cotton fibers. This is useful if you do not want to expose the fibers to water and induce swelling. For imaging with oil immersion objectives I used 2,2-thioidiethanol as a mounting medium with good success. Staudt, T., M. C. Lang, et al. (2007). "2,2'-thiodiethanol: a new water soluble mounting medium for high resolution optical microscopy." Microscopy Research and Technique 70(1): 1-9. Stan Vitha Microscopy and Imaging Center Texas A&M University This e-mail and any attachments may contain information which is confidential or subject to copyright. If you receive this e-mail in error, please delete it. Scion does not accept responsibility for anything in this e-mail which is not provided in the course of Scion's usual business or for any computer virus, data corruption, interference or delay arising from this e-mail. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Acridine orange - the "RNA stain" also stains cell walls very nicely and uniformly, especially cotton cell walls. It overcomes the irritating effect of laser polarisation seen when using calcofluor white. Propidium iodide will stain the pectic components of cotton fibres, but you also see the effect of laser polarisation there too. On 8/05/14 6:00 AM, "Lloyd Donaldson" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Cotton does contain small amounts of protein. It is unlikely that the >cellulose would fluoresce. > >-----Original Message----- >From: Confocal Microscopy List [mailto:[hidden email]] >On Behalf Of Guy Cox >Sent: Thursday, 8 May 2014 3:54 a.m. >To: [hidden email] >Subject: Re: IMAGING OF CELLULOSE > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >I totally endorse TDE as an embedding medium but I am puzzled by acrolein >inducing fluorescence in cellulose. Can someone explain the chemistry? >I thought acroleon reacted with proteins. > > Guy >________________________________________ >From: Confocal Microscopy List [[hidden email]] on >behalf of Stanislav Vitha [[hidden email]] >Sent: 08 May 2014 00:04 >To: [hidden email] >Subject: Re: IMAGING OF CELLULOSE > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >I have recently done some confocal imaging of cotton fibers. For deeper >penetration, less scattering in single photon fluorescence imaging, >rather than Calcofluor White I suggest staining with orange or >red-fluorescing dyes - some have been already mentioned, like Congo Red. >The fluorescent version of Periodic Acid-Schiff staining works great (use >propidium iodide as pseudo-Schiff reagent; Truernit, E., H. Bauby, et al. >(2008). "High- resolution whole-mount imaging of three-dimensional tissue >organization and gene expression enables the study of Phloem development >and structure in Arabidopsis." Plant Cell 20(6): 1494-1503). >As an alternative to fluorescence staining, I have also used vapor >fixation with acrolein, which tends to induce very strong >autofluorescence, at least in cotton fibers. This is useful if you do >not want to expose the fibers to water and induce swelling. >For imaging with oil immersion objectives I used 2,2-thioidiethanol as a >mounting medium with good success. >Staudt, T., M. C. Lang, et al. (2007). "2,2'-thiodiethanol: a new water >soluble mounting medium for high resolution optical microscopy." >Microscopy Research and Technique 70(1): 1-9. > >Stan Vitha >Microscopy and Imaging Center >Texas A&M University > > > >This e-mail and any attachments may contain information which is >confidential or subject to copyright. If you receive this e-mail in >error, please delete it. >Scion does not accept responsibility for anything in this e-mail which is >not provided in the course of Scion's usual business or for any computer >virus, data corruption, interference or delay arising from this e-mail. |