To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? Thanks much in advance, Lu |
|
Marc Reinig |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Coherent light will produce speckled images. You need the coherence length to be considerably less than the exposure time. We decorrelate our laser with a spinning disk. That results in both spatial and temporal decorrelation. Marco Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > > I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). > > > But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? > > > Thanks much in advance, > > Lu |
|
Zdenek Svindrych-2 |
|
|
In reply to this post by Lu Yan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, the quick answer is: yes, you can do SIM with LEDs (we used to do that in Guy Hagen's lab, there are other papers out there as well), but the contrast is lower and you have to use lower spatial frequencies of the patterns... There are a number of ways to create the pattern in the sample space with lasers. But with LEDs the (probably) only way is to image an amplitude grating (or a pattern on an amplitude spatial light modulator) into the sample space. Then the contrast of the pattern is given by the Opical Transfer Function (OTF) or better the Modulation Transfer Function (MTF) of the objective lens. That means the contrast of the high-frequency patterns (at the resolution limit of the objective) is essentially zero. At lower frequencies you can get some useful contrast, but the superresolution improvement will be less than factor of 2. Also, this incoherent pattern has different z-profile than that of either 2D or 3D laser-based SIM. You can increase the spatial coherence of the LED source with iris stops cleverly position in the optical setup, but that will cut your illumination intensity dramatically (and I'm not sure how much contrast improvement you get). So, you can use LEDs for SIM, but to get the most of it, the image processing involved will be different than what has been done traditionally. You can think of it as improved deconvolution, where the missing cone problem is eliminated thanks to the patterned illumination. Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 ---------- Původní e-mail ---------- Od: Yan, Lu <[hidden email]> Komu: [hidden email] Datum: 27. 7. 2017 13:32:37 Předmět: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? Thanks much in advance, Lu " |
|
Marc Reinig |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** One further thought on the SIM illumination. You should have the illumination polarization lined up with your pattern for optimum contrast. We use a laser and a polarization rotator for that reason. Plus we often image at 488 nm and LED's are hard to come by that are bright enough once you've filtered out the unwanted polarizations. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 11:54 AM, <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Lu, > the quick answer is: yes, you can do SIM with LEDs (we used to do that in > Guy Hagen's lab, there are other papers out there as well), but the contrast > is lower and you have to use lower spatial frequencies of the patterns... > There are a number of ways to create the pattern in the sample space with > lasers. But with LEDs the (probably) only way is to image an amplitude > grating (or a pattern on an amplitude spatial light modulator) into the > sample space. Then the contrast of the pattern is given by the Opical > Transfer Function (OTF) or better the Modulation Transfer Function (MTF) of > the objective lens. That means the contrast of the high-frequency patterns > (at the resolution limit of the objective) is essentially zero. At lower > frequencies you can get some useful contrast, but the superresolution > improvement will be less than factor of 2. Also, this incoherent pattern has > different z-profile than that of either 2D or 3D laser-based SIM. > > You can increase the spatial coherence of the LED source with iris stops > cleverly position in the optical setup, but that will cut your illumination > intensity dramatically (and I'm not sure how much contrast improvement you > get). > So, you can use LEDs for SIM, but to get the most of it, the image > processing involved will be different than what has been done traditionally. > You can think of it as improved deconvolution, where the missing cone > problem is eliminated thanks to the patterned illumination. > > Best, zdenek > -- > Zdenek Svindrych, Ph.D. > W.M. Keck Center for Cellular Imaging (PLSB 003) > Department of Biology,University of Virginia > 409 McCormick Rd, Charlottesville, VA-22904 > http://www.kcci.virginia.edu/ > tel: 434-982-4869 > > ---------- Původní e-mail ---------- > Od: Yan, Lu <[hidden email]> > Komu: [hidden email] > Datum: 27. 7. 2017 13:32:37 > Předmět: Illumination source coherence requirement for SIM > "***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > > I have no SIM experience so far. But I had always thought that in order to > generate the grating/grid/structured pattern on the sample plane by > interfering two beams would require coherent light sources like a laser with > reasonable coherent length (or temporal coherence). > > > But then I realized there are also SIM using low-coherence light source like > diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes > used. So I got a little confused. What is the fundamental difference bewteen > these implementation? Is it just structured pattern contrast difference? > What is the coherence requirement for SIM or similar tech. or is there any? > > > Thanks much in advance, > > Lu > " |
|
|
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi zdenek, Thanks for your reply. Now I understand that using the LED sources the grid pattern has to be re-imaged onto the sample, and the contrast depends on the MTF of the lens. But for interference-based pattern generation, i.e. using a grating to essentially split a beam (into +1 and -1 orders) and then bring them together at the sample plane to get the interference pattern, this configuration will need coherent source, i.e. coherent length larger than the beam path difference, right? The resultant fringe/pattern spacing depends on the angle between two beams and of course the wavelength. However, in this configuration the sample plane is also the image plane of the grating, so how is this interference not same as re-imaging the grating pattern? This is what I am confused about. Thanks, Lu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of [hidden email] <[hidden email]> Sent: Thursday, July 27, 2017 2:54:55 PM To: [hidden email] Subject: Re: Illumination source coherence requirement for SIM ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, the quick answer is: yes, you can do SIM with LEDs (we used to do that in Guy Hagen's lab, there are other papers out there as well), but the contrast is lower and you have to use lower spatial frequencies of the patterns... There are a number of ways to create the pattern in the sample space with lasers. But with LEDs the (probably) only way is to image an amplitude grating (or a pattern on an amplitude spatial light modulator) into the sample space. Then the contrast of the pattern is given by the Opical Transfer Function (OTF) or better the Modulation Transfer Function (MTF) of the objective lens. That means the contrast of the high-frequency patterns (at the resolution limit of the objective) is essentially zero. At lower frequencies you can get some useful contrast, but the superresolution improvement will be less than factor of 2. Also, this incoherent pattern has different z-profile than that of either 2D or 3D laser-based SIM. You can increase the spatial coherence of the LED source with iris stops cleverly position in the optical setup, but that will cut your illumination intensity dramatically (and I'm not sure how much contrast improvement you get). So, you can use LEDs for SIM, but to get the most of it, the image processing involved will be different than what has been done traditionally. You can think of it as improved deconvolution, where the missing cone problem is eliminated thanks to the patterned illumination. Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 ---------- Původní e-mail ---------- Od: Yan, Lu <[hidden email]> Komu: [hidden email] Datum: 27. 7. 2017 13:32:37 Předmět: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? Thanks much in advance, Lu " |
|
|
In reply to this post by Marc Reinig
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Marco, Agreed. But isn't the two beam interference depending on the temporal coherence instead of spatial? Please correct me if I am wrong. Thanks, Lu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Marc Reinig <[hidden email]> Sent: Thursday, July 27, 2017 1:59:34 PM To: [hidden email] Subject: Re: Illumination source coherence requirement for SIM ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Coherent light will produce speckled images. You need the coherence length to be considerably less than the exposure time. We decorrelate our laser with a spinning disk. That results in both spatial and temporal decorrelation. Marco Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > > I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). > > > But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? > > > Thanks much in advance, > > Lu |
|
Zdenek Svindrych-2 |
|
|
In reply to this post by Lu Yan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Exactly! The two optical arrangements are the same, the only difference is the illumination source and possibly the grating period. Imagine this: you start with laser-based 2- or 3- beam SIM, that is, you have two or three small spot in the BFP of the objective lens. Now, if you increase the spatial (or angular) extent of the light source, that is, increase etendue and decrease spatial coherence, the spots in the BFP start to grow. E.g. 50 um / 0.22 NA fiber is still a good source for SIM, the 'spots' in the BFP are bigger, but still distinctly separated (depends on the size of the illuminated field of view, too), pattern contrast is high and you can reduce the annoying speckle... But if you increase the source etendue further, the 'spots' start to overlap and the contrast suffers. With a decent power LED (1 by 1 mm chip, Lambertian emission) you always overfill the BFP of a high magnification objective, so you are working in the incoherent regime, governed by the MTF. The temporal (in-)coherence and spectral bandwidth play minor role, imho. If you like fourier optics, the pattern intensity in focus should be the modulus of the FFT of the electric field distribution in the BFP (in the coherent case, ignoring vectorial and polarization effects). From that calculation it appears that the contrast decreases as the spots in the BFP grow, even if the field in the BFP is otherwise perfectly temporally coherent... best, zdenek ---------- Původní e-mail ---------- Od: Yan, Lu <[hidden email]> Komu: [hidden email] Datum: 27. 7. 2017 20:26:46 Předmět: Re: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi zdenek, Thanks for your reply. Now I understand that using the LED sources the grid pattern has to be re- imaged onto the sample, and the contrast depends on the MTF of the lens. But for interference-based pattern generation, i.e. using a grating to essentially split a beam (into +1 and -1 orders) and then bring them together at the sample plane to get the interference pattern, this configuration will need coherent source, i.e. coherent length larger than the beam path difference, right? The resultant fringe/pattern spacing depends on the angle between two beams and of course the wavelength. However, in this configuration the sample plane is also the image plane of the grating, so how is this interference not same as re-imaging the grating pattern? This is what I am confused about. Thanks, Lu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of [hidden email] <[hidden email]> Sent: Thursday, July 27, 2017 2:54:55 PM To: [hidden email] Subject: Re: Illumination source coherence requirement for SIM ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, the quick answer is: yes, you can do SIM with LEDs (we used to do that in Guy Hagen's lab, there are other papers out there as well), but the contrast is lower and you have to use lower spatial frequencies of the patterns... There are a number of ways to create the pattern in the sample space with lasers. But with LEDs the (probably) only way is to image an amplitude grating (or a pattern on an amplitude spatial light modulator) into the sample space. Then the contrast of the pattern is given by the Opical Transfer Function (OTF) or better the Modulation Transfer Function (MTF) of the objective lens. That means the contrast of the high-frequency patterns (at the resolution limit of the objective) is essentially zero. At lower frequencies you can get some useful contrast, but the superresolution improvement will be less than factor of 2. Also, this incoherent pattern has different z-profile than that of either 2D or 3D laser-based SIM. You can increase the spatial coherence of the LED source with iris stops cleverly position in the optical setup, but that will cut your illumination intensity dramatically (and I'm not sure how much contrast improvement you get). So, you can use LEDs for SIM, but to get the most of it, the image processing involved will be different than what has been done traditionally. You can think of it as improved deconvolution, where the missing cone problem is eliminated thanks to the patterned illumination. Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 ---------- Původní e-mail ---------- Od: Yan, Lu <[hidden email]> Komu: [hidden email] Datum: 27. 7. 2017 13:32:37 Předmět: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? Thanks much in advance, Lu " " |
|
Marc Reinig |
|
|
In reply to this post by Lu Yan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, If you were able to only temporally decorrelate the light over time, the relative phases of the light interacting at different places would stay the same. The speckles, the places where they constructively and destructively interfere, would stay the same because the path lengths would stay constant. When you spatially decorrelate over time, the light at different places is changing phase relative to each other over time and you do not get a stable speckle field because the places of constructive and destructive interference are changing. You need them both. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 5:27 PM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Marco, > > > Agreed. But isn't the two beam interference depending on the temporal coherence instead of spatial? Please correct me if I am wrong. > > > Thanks, > > Lu > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on behalf of Marc Reinig <[hidden email]> > Sent: Thursday, July 27, 2017 1:59:34 PM > To: [hidden email] > Subject: Re: Illumination source coherence requirement for SIM > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Coherent light will produce speckled images. You need the coherence > length to be considerably less than the exposure time. We decorrelate > our laser with a spinning disk. That results in both spatial and > temporal decorrelation. > > Marco > > Marco > > Marc R. Reinig > W. M. Keck Center for Adaptive Optical Microscopy > University of California Santa Cruz > > > > On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi all, >> >> >> I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). >> >> >> But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? >> >> >> Thanks much in advance, >> >> Lu |
|
Zdenek Svindrych-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Let me just add, that it depends a lot on the way you create your pattern. If you use fancy beamsplitters to get your two or three spots in the objective BFP (the "GE OMX SIM" arrangement, for example), then the temporal coherence (and, conversely, the spectral width) is of utmost importance. In this case the spatial frequency of the pattern depends linearly on wavelength, so if you want a good pattern contrast across, lets say, 1000 lines in your field of view, you need linewidth much smaller than 1/1000 of the central wavelength. Easy with lasers, impossible with LEDs. On the other hand, in the more traditional setup (e.g. Nikon N-SIM), where you are imaging the pattern into the sample (as Lu and Marco mentioned) the frequency of the pattern is given by the grating (or SLM patter) period, so the contrast is high even with broadband source (note that for the shorter- wavelength components the pattern frequency is less than the maximum possible, and the long-wavelength components may miss the BFP aperture completely, leading to zero contrast). Or, from another point of view: the path difference between the two (or three) beams in the BFP is always zero in this case. Also, the +/-1 "spots" in the BFP are no longer spots, but little patches of rainbow... Hope it's not too confusing. Disclaimer: no commercial interest. Disclaimer: I have not confirmed this experimentally. I still may be wrong. Has anyone tried "broadband" SIM, e.g. with superluminescent diodes (not really useful for fluorescence excitation due to red-NIR wavelengths), or supercontinuum lasers? Best, zdenek ---------- Původní e-mail ---------- Od: Marc Reinig <[hidden email]> Komu: [hidden email] Datum: 28. 7. 2017 14:28:27 Předmět: Re: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, If you were able to only temporally decorrelate the light over time, the relative phases of the light interacting at different places would stay the same. The speckles, the places where they constructively and destructively interfere, would stay the same because the path lengths would stay constant. When you spatially decorrelate over time, the light at different places is changing phase relative to each other over time and you do not get a stable speckle field because the places of constructive and destructive interference are changing. You need them both. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 5:27 PM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Marco, > > > Agreed. But isn't the two beam interference depending on the temporal > > > Thanks, > > Lu > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on behalf of Marc Reinig <[hidden email]> > Sent: Thursday, July 27, 2017 1:59:34 PM > To: [hidden email] > Subject: Re: Illumination source coherence requirement for SIM > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Coherent light will produce speckled images. You need the coherence > length to be considerably less than the exposure time. We decorrelate > our laser with a spinning disk. That results in both spatial and > temporal decorrelation. > > Marco > > Marco > > Marc R. Reinig > W. M. Keck Center for Adaptive Optical Microscopy > University of California Santa Cruz > > > > On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi all, >> >> >> I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). >> >> >> But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? >> >> >> Thanks much in advance, >> >> Lu " |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** zdenek, Marco, Thanks for your input. I found a pretty neat picture of explaining spatial and temporal coherence here https://www.rp-photonics.com/coherence.html (no commercial interest 😊). I think the rotating diffuser, or some vibrating diffuser, or a 'fiber shaker' will mainly break down the spatial coherence, but at the same time degrade the temporal coherence a bit. Back to the SIM topic, supposed that you have very nice narrow linewidth single frequency laser to generate very fine interference pattern at the sample plane. However, as you record the image, this grating pattern will be, as zdenek mentioned, practical limited by the MTF of the lens. So does this mean there should be a limit to the grating pattern on sample? Or what is an optimal pattern dimensions at the sample plane? Thanks, Lu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of [hidden email] <[hidden email]> Sent: Friday, July 28, 2017 3:34:12 PM To: [hidden email] Subject: Re: Illumination source coherence requirement for SIM ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Let me just add, that it depends a lot on the way you create your pattern. If you use fancy beamsplitters to get your two or three spots in the objective BFP (the "GE OMX SIM" arrangement, for example), then the temporal coherence (and, conversely, the spectral width) is of utmost importance. In this case the spatial frequency of the pattern depends linearly on wavelength, so if you want a good pattern contrast across, lets say, 1000 lines in your field of view, you need linewidth much smaller than 1/1000 of the central wavelength. Easy with lasers, impossible with LEDs. On the other hand, in the more traditional setup (e.g. Nikon N-SIM), where you are imaging the pattern into the sample (as Lu and Marco mentioned) the frequency of the pattern is given by the grating (or SLM patter) period, so the contrast is high even with broadband source (note that for the shorter- wavelength components the pattern frequency is less than the maximum possible, and the long-wavelength components may miss the BFP aperture completely, leading to zero contrast). Or, from another point of view: the path difference between the two (or three) beams in the BFP is always zero in this case. Also, the +/-1 "spots" in the BFP are no longer spots, but little patches of rainbow... Hope it's not too confusing. Disclaimer: no commercial interest. Disclaimer: I have not confirmed this experimentally. I still may be wrong. Has anyone tried "broadband" SIM, e.g. with superluminescent diodes (not really useful for fluorescence excitation due to red-NIR wavelengths), or supercontinuum lasers? Best, zdenek ---------- Původní e-mail ---------- Od: Marc Reinig <[hidden email]> Komu: [hidden email] Datum: 28. 7. 2017 14:28:27 Předmět: Re: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, If you were able to only temporally decorrelate the light over time, the relative phases of the light interacting at different places would stay the same. The speckles, the places where they constructively and destructively interfere, would stay the same because the path lengths would stay constant. When you spatially decorrelate over time, the light at different places is changing phase relative to each other over time and you do not get a stable speckle field because the places of constructive and destructive interference are changing. You need them both. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 5:27 PM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Marco, > > > Agreed. But isn't the two beam interference depending on the temporal > > > Thanks, > > Lu > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on behalf of Marc Reinig <[hidden email]> > Sent: Thursday, July 27, 2017 1:59:34 PM > To: [hidden email] > Subject: Re: Illumination source coherence requirement for SIM > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Coherent light will produce speckled images. You need the coherence > length to be considerably less than the exposure time. We decorrelate > our laser with a spinning disk. That results in both spatial and > temporal decorrelation. > > Marco > > Marco > > Marc R. Reinig > W. M. Keck Center for Adaptive Optical Microscopy > University of California Santa Cruz > > > > On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi all, >> >> >> I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). >> >> >> But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? >> >> >> Thanks much in advance, >> >> Lu " |
|
Marc Reinig |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If the spatial cutoff frequency of your lens is F, you cannot project a pattern at a spatial frequency, F, and a pattern with a spatial frequency of F will not appear on the object plane. Generally, a pattern with a spatial frequency of 0.8 to 0.9 is the best you can hope for. This will give you a 1.8 to 1.9 theoretical increase in resolution. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Fri, Jul 28, 2017 at 2:51 PM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > zdenek, Marco, > > Thanks for your input. I found a pretty neat picture of explaining spatial and temporal coherence here https://www.rp-photonics.com/coherence.html (no commercial interest 😊). I think the rotating diffuser, or some vibrating diffuser, or a 'fiber shaker' will mainly break down the spatial coherence, but at the same time degrade the temporal coherence a bit. > > Back to the SIM topic, supposed that you have very nice narrow linewidth single frequency laser to generate very fine interference pattern at the sample plane. However, as you record the image, this grating pattern will be, as zdenek mentioned, practical limited by the MTF of the lens. So does this mean there should be a limit to the grating pattern on sample? Or what is an optimal pattern dimensions at the sample plane? > > Thanks, > Lu > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on behalf of [hidden email] <[hidden email]> > Sent: Friday, July 28, 2017 3:34:12 PM > To: [hidden email] > Subject: Re: Illumination source coherence requirement for SIM > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Let me just add, that it depends a lot on the way you create your pattern. > If you use fancy beamsplitters to get your two or three spots in the > objective BFP (the "GE OMX SIM" arrangement, for example), then the temporal > coherence (and, conversely, the spectral width) is of utmost importance. In > this case the spatial frequency of the pattern depends linearly on > wavelength, so if you want a good pattern contrast across, lets say, 1000 > lines in your field of view, you need linewidth much smaller than 1/1000 of > the central wavelength. Easy with lasers, impossible with LEDs. > > On the other hand, in the more traditional setup (e.g. Nikon N-SIM), where > you are imaging the pattern into the sample (as Lu and Marco mentioned) the > frequency of the pattern is given by the grating (or SLM patter) period, so > the contrast is high even with broadband source (note that for the shorter- > wavelength components the pattern frequency is less than the maximum > possible, and the long-wavelength components may miss the BFP aperture > completely, leading to zero contrast). Or, from another point of view: the > path difference between the two (or three) beams in the BFP is always zero > in this case. Also, the +/-1 "spots" in the BFP are no longer spots, but > little patches of rainbow... > > Hope it's not too confusing. > > Disclaimer: no commercial interest. > > Disclaimer: I have not confirmed this experimentally. I still may be wrong. > Has anyone tried "broadband" SIM, e.g. with superluminescent diodes (not > really useful for fluorescence excitation due to red-NIR wavelengths), or > supercontinuum lasers? > Best, zdenek > > > > > ---------- Původní e-mail ---------- > Od: Marc Reinig <[hidden email]> > Komu: [hidden email] > Datum: 28. 7. 2017 14:28:27 > Předmět: Re: Illumination source coherence requirement for SIM > "***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Lu, > > If you were able to only temporally decorrelate the light over time, > the relative phases of the light interacting at different places would > stay the same. The speckles, the places where they constructively and > destructively interfere, would stay the same because the path lengths > would stay constant. > > When you spatially decorrelate over time, the light at different > places is changing phase relative to each other over time and you do > not get a stable speckle field because the places of constructive and > destructive interference are changing. > > You need them both. > > Marco > > Marc R. Reinig > W. M. Keck Center for Adaptive Optical Microscopy > University of California Santa Cruz > > > > On Thu, Jul 27, 2017 at 5:27 PM, Yan, Lu <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi Marco, >> >> >> Agreed. But isn't the two beam interference depending on the temporal > coherence instead of spatial? Please correct me if I am wrong. >> >> >> Thanks, >> >> Lu >> >> ________________________________ >> From: Confocal Microscopy List <[hidden email]> on > behalf of Marc Reinig <[hidden email]> >> Sent: Thursday, July 27, 2017 1:59:34 PM >> To: [hidden email] >> Subject: Re: Illumination source coherence requirement for SIM >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Coherent light will produce speckled images. You need the coherence >> length to be considerably less than the exposure time. We decorrelate >> our laser with a spinning disk. That results in both spatial and >> temporal decorrelation. >> >> Marco >> >> Marco >> >> Marc R. Reinig >> W. M. Keck Center for Adaptive Optical Microscopy >> University of California Santa Cruz >> >> >> >> On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. > >>> ***** >>> >>> Hi all, >>> >>> >>> I have no SIM experience so far. But I had always thought that in order > to generate the grating/grid/structured pattern on the sample plane by > interfering two beams would require coherent light sources like a laser with > reasonable coherent length (or temporal coherence). >>> >>> >>> But then I realized there are also SIM using low-coherence light source > like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also > sometimes used. So I got a little confused. What is the fundamental > difference bewteen these implementation? Is it just structured pattern > contrast difference? What is the coherence requirement for SIM or similar > tech. or is there any? >>> >>> >>> Thanks much in advance, >>> >>> Lu > " |
|
Zdenek Svindrych-2 |
|
|
In reply to this post by Lu Yan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, to clarify, the point I was trying to make is that with LED illumination the contrast of the high-frequency pattern will be vanishingly small (due to nearly zero MTF near the highest spatial frequency). With laser-based illumination you can get very close to the ultimate spatial frequency limit with almost perfect contrast. The standard MTF does not apply here (tried to explain this one of the previous posts). Now, if you place a thin fluorescent layer into the focus, you will not be able to see that pattern on your camera! Here, again, you are limited by the MTF. But the pattern is there, and that's what's important. The SIM algorithm can extract the superresolution information from the variations of image intensity as the pattern wipes across your sample (of course, there is no superresolution information coming from an infinite fluorescent layer, but if you put there subresolution fluorescent beads instead, their intensities will vary periodically with pattern phase). Note, the above holds for 2D SIM (+/-1st order) only, if you introduce the 0 th order beam, another pattern with half the spatial frequency will appear, that frequency is low enough so you can actually see the pattern on your camera (with the thin fluorescent layer sample in place). So generally you want to use as high a spatial frequency as possible, so that the +/-1st order just passes through the BFP aperture of the objective lens. You can get very close to the NA limit of the lens (with a factor of 0.9 as mentioned by Marco, be also cautious about total internal reflection!). The ultimate (shortest) pattern period in the sample plane is lambda/(2*NA). It follows quite naturally from the fact, that a standing intensity pattern of two counterpropagating beams has a period of lambda/2 (the factor of 1/2 comes from the important fact that intensity = amplitude^2). Best, zdenek ---------- Původní e-mail ---------- Od: Yan, Lu <[hidden email]> Komu: [hidden email] Datum: 28. 7. 2017 17:53:33 Předmět: Re: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** zdenek, Marco, Thanks for your input. I found a pretty neat picture of explaining spatial and temporal coherence here https://www.rp-photonics.com/coherence.html (no commercial interest 😊). I think the rotating diffuser, or some vibrating diffuser, or a 'fiber shaker' will mainly break down the spatial coherence, but at the same time degrade the temporal coherence a bit. Back to the SIM topic, supposed that you have very nice narrow linewidth single frequency laser to generate very fine interference pattern at the sample plane. However, as you record the image, this grating pattern will be, as zdenek mentioned, practical limited by the MTF of the lens. So does this mean there should be a limit to the grating pattern on sample? Or what is an optimal pattern dimensions at the sample plane? Thanks, Lu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of [hidden email] <[hidden email]> Sent: Friday, July 28, 2017 3:34:12 PM To: [hidden email] Subject: Re: Illumination source coherence requirement for SIM ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Let me just add, that it depends a lot on the way you create your pattern. If you use fancy beamsplitters to get your two or three spots in the objective BFP (the "GE OMX SIM" arrangement, for example), then the temporal coherence (and, conversely, the spectral width) is of utmost importance. In this case the spatial frequency of the pattern depends linearly on wavelength, so if you want a good pattern contrast across, lets say, 1000 lines in your field of view, you need linewidth much smaller than 1/1000 of the central wavelength. Easy with lasers, impossible with LEDs. On the other hand, in the more traditional setup (e.g. Nikon N-SIM), where you are imaging the pattern into the sample (as Lu and Marco mentioned) the frequency of the pattern is given by the grating (or SLM patter) period, so the contrast is high even with broadband source (note that for the shorter- wavelength components the pattern frequency is less than the maximum possible, and the long-wavelength components may miss the BFP aperture completely, leading to zero contrast). Or, from another point of view: the path difference between the two (or three) beams in the BFP is always zero in this case. Also, the +/-1 "spots" in the BFP are no longer spots, but little patches of rainbow... Hope it's not too confusing. Disclaimer: no commercial interest. Disclaimer: I have not confirmed this experimentally. I still may be wrong. Has anyone tried "broadband" SIM, e.g. with superluminescent diodes (not really useful for fluorescence excitation due to red-NIR wavelengths), or supercontinuum lasers? Best, zdenek ---------- Původní e-mail ---------- Od: Marc Reinig <[hidden email]> Komu: [hidden email] Datum: 28. 7. 2017 14:28:27 Předmět: Re: Illumination source coherence requirement for SIM "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Lu, If you were able to only temporally decorrelate the light over time, the relative phases of the light interacting at different places would stay the same. The speckles, the places where they constructively and destructively interfere, would stay the same because the path lengths would stay constant. When you spatially decorrelate over time, the light at different places is changing phase relative to each other over time and you do not get a stable speckle field because the places of constructive and destructive interference are changing. You need them both. Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Thu, Jul 27, 2017 at 5:27 PM, Yan, Lu <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Marco, > > > Agreed. But isn't the two beam interference depending on the temporal > > > Thanks, > > Lu > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on behalf of Marc Reinig <[hidden email]> > Sent: Thursday, July 27, 2017 1:59:34 PM > To: [hidden email] > Subject: Re: Illumination source coherence requirement for SIM > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Coherent light will produce speckled images. You need the coherence > length to be considerably less than the exposure time. We decorrelate > our laser with a spinning disk. That results in both spatial and > temporal decorrelation. > > Marco > > Marco > > Marc R. Reinig > W. M. Keck Center for Adaptive Optical Microscopy > University of California Santa Cruz > > > > On Thu, Jul 27, 2017 at 10:22 AM, Yan, Lu <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi all, >> >> >> I have no SIM experience so far. But I had always thought that in order to generate the grating/grid/structured pattern on the sample plane by interfering two beams would require coherent light sources like a laser with reasonable coherent length (or temporal coherence). >> >> >> But then I realized there are also SIM using low-coherence light source like diode lasers, LED or lamp to do SIM, and a Ronchi grating is also sometimes used. So I got a little confused. What is the fundamental difference bewteen these implementation? Is it just structured pattern contrast difference? What is the coherence requirement for SIM or similar tech. or is there any? >> >> >> Thanks much in advance, >> >> Lu " " |