Matthew Pearson-3 |
|
|
Hi everyone,
I was wondering whether anyone has had any experience of imaging macular pigments in live cells using either light/confocal microscopy. I have a colleague who wants to image live ARPE-19 cells in culture which they have added macular pigment crystals to, to see if the cells take them up. The macular pigment apparently absorbs light at 465nm but they do not emit any. ARPE-19 cells are also naturally heavily pigmented with melanin and lipofusin. So i am wondering A. whether the macular pigment will be visible either by light or confocal microscopy and B. We basically want to see where this pigment goes in the cell or if indeed it enters the cell at all. I would appreciate any advice people have on this subject. I have only seen papers of macular pigments being imaged in vivo using a laser scanning ophthalmoscope... Thanks for your help! Matt Pearson. |
|
Chris Tully |
|
|
Matthew,
You have an interesting problem here. To start with since you are looking for an absorber, not a fluorochrome, you MUST use transmitted light! Next, since you have many other potential absorbers in the light path (e.g. melanin and lipofusin) You will likely need to take a background shot of the cells prior to adding the macular pigment as a reference for comparison to the post treatment images. This then implies that you will need to grow the cells on a substrate that can be re-registered so that you can image the same set of cells before and after treatment. Or, depending on the time scale of the treatment and the available equipment (heating stages, etc.) it may be useful to due the experiment on the microscope stage. Now, for the filters... You will likely need to use a very narrow band filter to insure that you are only looking at 465nm light. I would start but putting this filter in the light path before the sample. You may find that you need to use a very bright light source (HG, Xe or metal halide) to get much light through such a setup. The pre-treatment image should reveal any "native" absorbers in the 465nm band. Hopefully they are not too common. Analysis: Since you are looking for changes in light that is absorbed, I would approach this by inverting both images, and then subtracting the pre treatment image from the post-treatment image (or using the pre-treatment image as the background for a background subtraction operation). Reinvert the result and the macular pigment particles should show up as dark spots in the image. You may also find it useful to take a full spectrum brightfield image for the cells (pre-treatment) to make localization of the macular pigments relative to the cells easier. --Chris Chris Tully Microscopy and Image Analysis Expert [hidden email] 240-888-1021 http://www.linkedin.com/in/christully On Fri, Feb 13, 2009 at 6:58 AM, Matthew Pearson <[hidden email]> wrote: > Hi everyone, > > I was wondering whether anyone has had any experience of imaging macular > pigments in live cells using either light/confocal microscopy. > > I have a colleague who wants to image live ARPE-19 cells in culture which > they have added macular pigment crystals to, to see if the cells take them > up. The macular pigment apparently absorbs light at 465nm but they do not > emit any. ARPE-19 cells are also naturally heavily pigmented with melanin > and lipofusin. So i am wondering A. whether the macular pigment will be > visible either by light or confocal microscopy and B. We basically want to > see where this pigment goes in the cell or if indeed it enters the cell at > all. > > I would appreciate any advice people have on this subject. I have only seen > papers of macular pigments being imaged in vivo using a laser scanning > ophthalmoscope... > > Thanks for your help! > > Matt Pearson. > > > -- > Imaging Technician > University College London > EC1V 9EL > 020 7608 6857 > [hidden email] > > |
| Free forum by Nabble | Edit this page |