Hi all,

I’ve been using the “split surfaces” tool in Arivis Vision4D and it usually
works, but with some datasets the surface cutting does not occur where I
draw the line. Sometimes there is no cutting at all and sometimes it occurs
far from the manually drawn cut line, and I can’t figure out the cause or a
pattern of when/why this happens. Has anyone experienced the same and found
a solution to this issue? I’m using Vision4D 2.12.3 x64 on Win10.

Thanks,
Esteban
Arvonn Tully Arvonn Tully
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Issue with Splitting surfaces in Arivis Vision4D

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

**
Dear Esteban,

There are a couple  possible causes/reasons -
The splitting tool in Vision4d will split the surfaces along Voxel
boundaries. In larger structures, this tends to work as expected, but it
can produce unexpected results in smaller structures - in some cases  those
structures will not exist.

 I'm not specifically aware of a bug, but it could an issue related to the
now very old  version 2.12.3 (released in 10-Oct-2016) .  Please send me
your dongle number directly, so I can verify upgrade eligibility, to see
what other version you could run. We have had numerous releases since then
(current release version is 3.1.3), and function has definitely been
improved since 2.12.3.

My office contact is below,
Best Regards
Arvonn


----
*Arvonn Tully*
Application Engineer
*arivis AG - Office US*
1875 Connecticut Ave NW
10th Floor
Washington, DC 20009
USA
Phone: +1 (800) 377-6962-702
E-mail : [hidden email]
Web : http://www.arivis.com


On Fri, Nov 29, 2019 at 7:55 PM G. Esteban Fernandez <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I’ve been using the “split surfaces” tool in Arivis Vision4D and it usually
> works, but with some datasets the surface cutting does not occur where I
> draw the line. Sometimes there is no cutting at all and sometimes it occurs
> far from the manually drawn cut line, and I can’t figure out the cause or a
> pattern of when/why this happens. Has anyone experienced the same and found
> a solution to this issue? I’m using Vision4D 2.12.3 x64 on Win10.
>
> Thanks,
> Esteban
>
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Issue with Splitting surfaces in Arivis Vision4D

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Arvonn,

Thanks for your response.  We pinned the splitting problem down to
images that were cropped and resampled within Vision4D (prior to
surface creation).  Therefore we reasoned that exporting to TIFF would
abandon any bugs/glitches in the SIS file resulting from the
resampling.  Indeed exporting the resampled data to TIFF and importing
back into Vision4D fixed the surface splitting problem!  Surfaces are
split as easily and nicely in our resampled images as in
full-resolution images. I'll contact you offline fo upgrade options as
you suggested.

Regards,
Esteban

On Mon, Dec 2, 2019 at 8:03 AM Arvonn <[hidden email]> wrote:

>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> **
> Dear Esteban,
>
> There are a couple  possible causes/reasons -
> The splitting tool in Vision4d will split the surfaces along Voxel
> boundaries. In larger structures, this tends to work as expected, but it
> can produce unexpected results in smaller structures - in some cases  those
> structures will not exist.
>
>  I'm not specifically aware of a bug, but it could an issue related to the
> now very old  version 2.12.3 (released in 10-Oct-2016) .  Please send me
> your dongle number directly, so I can verify upgrade eligibility, to see
> what other version you could run. We have had numerous releases since then
> (current release version is 3.1.3), and function has definitely been
> improved since 2.12.3.
>
> My office contact is below,
> Best Regards
> Arvonn
>
>
> ----
> *Arvonn Tully*
> Application Engineer
> *arivis AG - Office US*
> 1875 Connecticut Ave NW
> 10th Floor
> Washington, DC 20009
> USA
> Phone: +1 (800) 377-6962-702
> E-mail : [hidden email]
> Web : http://www.arivis.com
>
>
> On Fri, Nov 29, 2019 at 7:55 PM G. Esteban Fernandez <
> [hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your posting.
> > *****
> >
> > Hi all,
> >
> > I’ve been using the “split surfaces” tool in Arivis Vision4D and it usually
> > works, but with some datasets the surface cutting does not occur where I
> > draw the line. Sometimes there is no cutting at all and sometimes it occurs
> > far from the manually drawn cut line, and I can’t figure out the cause or a
> > pattern of when/why this happens. Has anyone experienced the same and found
> > a solution to this issue? I’m using Vision4D 2.12.3 x64 on Win10.
> >
> > Thanks,
> > Esteban
> >